  4120   Thu Jan 6 00:06:01 2011 JenneUpdateIOOMagical absorbing PZT mirror

[Kiwamu, Jenne]

We have a measely 465mW going into the MC.  We lose a boatload of power on the PZT mirror that is part of the last zigzag for steering into the MC.  Just before this mirror, we measure 1.21W .  Just after this mirror, we measure ~475mW.  Then a teeny bit gets picked off for the PSL POS/ANG.  But we're losing a factor of 3 on this one mirror.  Need to fix!!!!!!!!!

  4121   Thu Jan 6 10:47:11 2011 KojiUpdateelogELOG fixed (re: restarted)

Fixed the 40m elog crashing with the threaded display.

This morning I found that Google bot crashed the elog again. I started the investigation and found the threaded mode is fine if we use the recent 10 entries.

I gradually copied the old entries to a temporary elog and found that a deleted elog entry on August 6 had a corrupted remnant in the elog file. This kept crashed the threaded mode.

Once this entry has been eliminated again, the threaded mode got functional.

I hope this eliminates those frequent elog crashing.

Quote:

Google bot  crashed the elog again. Then, I found that Google bot (and I) can crash elogd by trying to show the threaded view.
There looks similar issue reported to the elog forum, the author did not think this is a true bag.

Note: This happens only for the 40m elog. The other elogs (ATF/PSL/TCS/SUS/Cryo) are OK for the threaded view.

Quote:

Did it again. It seemed that Google bot came to the elog and tried to obtain "http://nodus.ligo.caltech.edu:8080/robots.txt". That was the last of the log.
Bot came from the AJW's homepage. Also Google FeedFecther came to the elog.

 

 

  4122   Fri Jan 7 00:14:36 2011 kiwamuUpdateIOOEOM triple resonant box is working

 I checked the triple resonant box for the broadband EOM this afternoon, and found it was healthy.

So I installed it again together with a power combiner and succeeded in locking the MC.

Since the box has a non-50 Ohm input impedance at 29.5 MHz, so it maybe needed to adjust the phase of the LO for the demodulation of the PDH signal.

A good thing is that now we are able to impose the other sidebands (i.e. 11 MHz and 55MHz) via the power combiner.

  4123   Fri Jan 7 00:49:16 2011 JenneUpdateGreen LockingRecovered Xarm Green Lock, Preparation for Beat Note Measurement

[Kiwamu, Jenne]

We went this evening in search of a beat note signal between the Xarm transmitted green light and the PSL doubled green light. 

First, we removed our new ETMX camera (we left the mount so it should be easy to put back) from the other day.   We left the test masses exactly where they had been while flashing for IR, so even though we can no longer confirm, we expect that the IR beam axis hasn't changed.  We used the steering mirror on the end table to align the green beam to the cavity.  We turned on the loop to lock the end laser to the cavity, and achieved green lock of the arm.

Then we went to the PSL table to overlap the arm transmitted light with the PSL doubled light.  We made a few changes to the optics that take the arm transmitted light over to the PD.  We found that the arm transmitted light was very high, so we changed from having one steering mirror to having 3 (for table space / geometry reasons we needed 3, not just 2) in order to lower the beam axis.  We also found that the spot size of the arm transmitted beam was ~2x too small, so we changed the mode matching telescope from a 4x reducer to a 2x reducer by changing the 2nd lens from f=50mm to f=100mm (the first lens is f=200mm).  We made the arm transmitted beam and the PSL green beam overlap, but we saw no peak on the spectrum analyzer. 

We checked the temperature of the PSL and end lasers, and determined that we needed to adjust the set temp of the end laser.  However, we still didn't find any beat note.

We then tweaked the temperature of the doubling oven at the end station, to maximize the power transmitted, since Kiwamu said that that had worked in the past.  Alas, no success tonight.

We're stopping for the evening, with the success of reacquiring green lock of the Xarm.

  4124   Fri Jan 7 12:01:39 2011 OsamuUpdateComputersc1lsc running

I got a new adapter board for expansion chassis from CDS and exchanged the existing adapter board which was laid on the floor around ETMX to new one.

Then I connected the chassis to c1lcs,  c1lsc seems to be running now. I will return the old board to CDS since Rolf says he wants to return it to manufacture.

 I found an interface box from ADC to D-SUB37pin and a cable to connect them.

I needed to make cables to connect the interface box to existing LSC whitening filters that has a 37 pin female D-SUB connector on one end and a 40pin female flat connector on the other end. We should use shielded cables for them, but unfortunately CDS did not have right one. Temporarily I made one cable for 1-8ch using a ribbon twist cable like Joe did.

I found a saturation at ch5 of ADC0 on c1lsc. I did not check carefully but it seemed to come from the LSC whitening board. Input of ch5 of the whitening board was not terminated and had a huge output voltage, but also ch6 was not terminated and had no big output. I guess something wrong on the LSC whitening board. Needs to be checked. Anyway I unplugged the small ribbon cable between the whitening board and the next LSC AA filter board.

Finally I realized that fiber connection of RFM did not exist. What I saw was the fiber cable of Dolphin. We need a RFM PCIe interface board, and a long fiber cable between c1lsc and RFM hub.

  4125   Fri Jan 7 15:17:37 2011 steveUpdateCamerasITMX video monitor has tower wide view

The Watec 902, 1/2" CCD camera got new Tamron lens: 1/2" 10-40mm F/1.4  manual iris. IR corrective lens. It is designed to have the same focal point in the IR as

in the visible light range. However, as the depth of field in the IR range is very narrow, focus adjustment should be done carefully in the IR.

Now you can see the sus tower that will make alignment easier.

 

  4126   Sat Jan 8 21:12:12 2011 ranaUpdateCDSMegatron is back

 I started reverting Megatron into a standard Ubuntu workstation after Joe/Osamu's attempt to steal it for their real time mumbo jumbo.

First, I installed a hard drive that was sitting around on top of it. That whole area is still a mess; I'm not surprised that we have so many CDS problems in such a chaotic state. There's another drive sitting around there called 'RT Linux' which I didn't use yet.

Second, I removed the ethernet cables and installed a monitor/keyboard/mouse on it.

Then I popped in the Ubuntu 10.04 LTS DVD, wiped the existing CentOS install and started the standard graphical installation of Ubuntu.

megatron.jpg

Megatron's specs attached: 

Attachment 2: sysinfo.text
  4127   Sat Jan 8 23:18:03 2011 ranaUpdateComputersnetwork table

There's, as usual, a disconnect between the real martian host table and the one displayed on the Wiki. Basically, the point is: DO NOT TRUST THE WIKI, because people who update the actual host table only randomly update the wiki table.

The only thing worth trusting is the file

/var/named/chroot/var/named/martian.zone
on linux 1.

You will need to have 'root' or 'sudo' to get into that directory to read the file.

  4129   Mon Jan 10 16:39:36 2011 josephb, alex, rolfUpdateCDSNew Time server for frame builder and 1PPS

Alex and Rolf came over today with a Tempus LX  GPS network timing server.  This has an IRIG-B output and a 1PPS output.  It can also be setup to act as an NTP server (although we did not set that up).

This was placed at waist height in the 1X7 rack.  We took the cable running to the presumably roof mounted antenna from the VME timing board and connected it to this new timing server.  We also moved the source of the 1PPS signal going to the master timer sequencer (big blue box in 1X7 with fibers going to all the front ends) to this new time server.  This system is currently working, although it took about 5 minutes to actually acquire a timing signal from the GPS satellites.  Alex says this system should be more stable, with no time jumps. 

I asked Rolf about the new timing system for the front ends, he had no idea when that hardware would be available to the 40m.

Currently, all the front ends and the frame builder agree on the time.  Front ends are running so the 1 PPS signal appears to be working as well.

  4130   Mon Jan 10 16:47:08 2011 josephb, alex, rolfUpdateCDSFixed c1lsc dolphin reflected memory

While Alex and Rolf were visiting, I pointed out that the Dolphin card was not sending any data, not even a time stamp, from the c1lsc machine.

After some poking around, we realized the IOP (input/output processor) was coming up before the Dolphin driver had even finished loading. 

We uncommented the line

#/etc/dolphin_wait

in the /diskless/root/etc/rc.local file on the frame builder.  This waits until the dolphin module is fully loaded, so it can hand off a correct pointer to the memory location that the Dolphin card reads and writes to.  Previously, the IOP had been receiving a bad pointer since the Dolphin driver had not finished loading.

So now the c1lsc machine can communicate with c1sus via Dolphin and from there the rest of the network via the traditional Ge Fanuc RFM.

  4131   Tue Jan 11 01:05:29 2011 kiwamuUpdateLSCobtained Xarm PDH signal

[Jenne, Zach, Kiwamu]

 

 We made some efforts to lock the X arm cavity with the infrared beam.

We eventually obtained the PDH signal from a photo diode at AS port, we are still in the mid way of the lock.

The PDH signal now is going into c1lsc's ADC.

We have to make sure which digital channel corresponds to our PDH signal,

 

(what we did)

- split the LO signal, which comes from a Marconi, just before the EOM into two path.

     One is going to the EOM and the other is going to the AP table for the demodulation, The driving frequency we are using is 11MHz.

- put RF amplifiers to make the RF signal bigger. The raw signal was small, it was about ~-50 dBm in the spectrum analyzer.

   So we connected two ZLN amplifiers. Now the RF signal is at about 0 dBm

- connected the LO and RF signals to a mixer.  Additionally we put a 9.5-11.5 MHz bandpass filter at the LO path since there was some amounts of 29.5MHz due to the RF reflection at the EOM resonant box.

     After a low pass filter by SR560 the signal shows typical PDH behaviors.

- strung a BNC cable which connects the demodulated signal and c1lsc.

    In order to connect the signal to the current working ADC, we disconnected AS166_I from a whitening board and plug our cable on it.

- tried checking the digital signal but we somehow couldn't configure DAQ setting, So actually we couldn't make sure which channel corresponds to our signal.

 

  4133   Tue Jan 11 11:39:45 2011 steveUpdateVACRGA is turned on

Joe updated the rga procedure in the wiki and we turned on the filament. It will take a few days to reach thermal equilibrium.

TP2 dry fore pump was replaced at 910mTorr

Attachment 1: pd70d22.jpg
pd70d22.jpg
  4134   Tue Jan 11 13:32:52 2011 josephb, kiwamuUpdateCDSUpdated some DAQ channel names

[Joe, Kiwamu]

We modified the activateDAQ.py script which lives in /opt/rtcds/caltech/c1/chans/daq/ and updates the C1SUS.ini, C1MCS.ini, C1RMS.ini, C1SCX.ini and C1SCY.ini files.  These files contain the DAQ channels for all the optics.

It has been modified so that channels like C1:SUS-ITMX_ULSEN_OUT_DAQ become C1:SUS-ITMX_SENSOR_UL.  Similarly the oplev signals go from C1:SUS-ITMX_OLPIT_OUT to C1:SUS-ITMX_OPLEV_PERROR.

After some debugging, we ran the script successfully and checked the output was correct.  We then restarted the frame builder (telnet fb 8088 and then shutdown) and also hit the DAQ reload button for all the front ends.

I tested in dataviewer that I could go back several years as well as going back just 1 hour in the history and see data for C1:SUS-ITMX_SENSOR_LL as well as C1:SUS-ITMX_OPLEV_YERROR.  I also tested realtime is also working for these channels.

 

The contents of the script are below.

 

inputfiles=["C1SUS.ini","C1RMS.ini","C1MCS.ini","C1SCX.ini","C1SCY.ini"]
prefix="[C1:SUS-"
optics=["BS_","ITMX_","ITMY_","PRM_","SRM_","MC1_","MC1_","MC2_","MC3_","ETMX_"]
#channels=["SUSPOS_IN1","SUSPIT_IN1","SUSYAW_IN1","SUSSIDE_IN1","ULSEN_OUT","URSEN_OUT","LRSEN_OUT","LLSEN_OUT","SDSEN_OUT","OL_SUM_IN1","OLPIT_IN1","OLYAW_IN1"]
channels_dict = {'SUSPOS_IN1':'SUSPOS_IN1_DAQ',
'SUSPIT_IN1':'SUSPIT_IN1_DAQ',
'SUSYAW_IN1':'SUSYAW_IN1_DAQ',
'SUSSIDE_IN1':'SUSSIDE_IN1_DAQ',
'ULSEN_OUT':'SENSOR_UL',
'URSEN_OUT':'SENSOR_UR',
'LRSEN_OUT':'SENSOR_LR',
'LLSEN_OUT':'SENSOR_LL',
'SDSEN_OUT':'SENSOR_SIDE',
'OLPIT_OUT':'OPLEV_PERROR',
'OLYAW_OUT':'OPLEV_YERROR',
'OL_SUM_OUT':'OPLEV_SUM'}

suffix="_DAQ]\n"

## set datarate
datarate=2048

## read the ini files
for inputfile in inputfiles:
    print inputfile
    outputfile=inputfile
    ifile = open(inputfile,'r')
    lines = ifile.readlines()
    ifile.close()

    for k in range(len(lines)):
        for op in optics:
            for ch in channels_dict:
                if (prefix+op+ch+suffix) in lines[k]:
                    lines[k]=prefix + op + channels_dict[ch] + "]\n"
                    lines[k+1]=lines[k+1].lstrip("#").rstrip(lines[k+1].split("=")[1])+"1\n"
                    lines[k+2]=lines[k+2].lstrip("#")
                    lines[k+3]=lines[k+3].lstrip("#").rstrip(lines[k+3].split("=")[1])+str(datarate)+"\n"
                    lines[k+4]=lines[k+4].lstrip("#")
    ofile = open(outputfile,'w')
    for k in range(len(lines)):
        ofile.write(lines[k])
        #print lines[k]
    ofile.close()

  4135   Tue Jan 11 14:05:11 2011 josephbUpdateComputersMartian host table updated daily

I created two simple cron jobs, one running on linux1 and one running on nodus, to produce an updated copy of the martian host table linkable from the wiki every day.

The scripts live in /opt/rtcds/caltech/c1/scripts/AutoUpdate/.  One is called  updateHostTable.cron and run on linux1 everyday at 4 am, and the other is called moveHostTable.cron which is run on nodus everyday at 5am.

The new link has been added to the Martian Host table wiki page  here.

 

  4136   Tue Jan 11 16:04:17 2011 josephbUpdateCDSScript to update web views of models for all installed front ends

I wrote a new script that is in /opt/rtcds/caltech/c1/scripts/AutoUpdate/ called  webview_simlink_update.m. 

This m-file when run in matlab will go to the /opt/rtcds/caltech/c1/target directory and for each c1 front end, generate the corresponding webview files for that system and place them in the AutoUpdate directory. 

Afterwards the files can be moved on Nodus to the /users/public_html/FE/ directory with:

mv /opt/rtcds/caltech/c1/scripts/AutoUpdate/*slwebview* /users/public_html/FE/

This was run today, and the files can be viewed at:

https://nodus.ligo.caltech.edu:30889/FE/

Long term, I'd like to figure out a way of automating this to produce automatically updated screens without having to run it manually.  However, simulink seems to stubbornly require an X window to work.

  4138   Tue Jan 11 18:41:43 2011 JenneUpdateIOOPut MC PZT offset onto MC board, instead of on awkward cart

[Larisa and Jenne]

We wanted to get rid of the awkward cart that was sitting behind the 1Y1 rack.  This cart was supplying a +5V offset to the PZT driver, so that we could use the MC length signal to feedback to lock the laser to the MC cavity.  Instead, we put the offset on the last op amp before the servo out on the Mc Servo Board.  Because we wanted +5V, but the board only had +5, +15, -15V as options, and we needed -5 to add just before the op amp (U40 in the schematic), because the op amp is using regular negative feedback, we made a little voltage divider between -15V and GND, to give ourselves -5V.  We used the back side of the voltage test points (where you can check to make sure that you're actually getting DC voltage on the board), and used a 511Ohm and 1.02kOhm resistor as a voltage divider. 

Then we put a 3.32kOhm resistor in ~"parallel" to R124, which is the usual resistor just before the negative input of the op amp.  Our -5V goes to our new resistor, and should, at the output, give us a +5V offset. 

Sadly, when we measure the actual output we get, it's only +2.3V.  Sadface.

We went ahead and plugged the servo out into the PZT driver anyway, since we had previously seen that the fluctuation when the mode cleaner is locked was much less than a volt, so we won't run into any problems with the PZT driver running into the lower limit (it only goes 0-10V).

Suresh has discovered that the op amp that we're looking at, U40 on the schematic, is an AD829, which has an input impedance of a measely 13kOhm.  So maybe the 3.32kOhm resistors that we are using (because that's what had already been there) are too large.  Perhaps tomorrow I'll switch all 3 resistors (R119, R124, and our new one) to something more like 1kOhm.  But right now, the MC is locked, and I'm super hungry, and it's time for some arm locking action.

I've attached the schematic.  The stuff that we fitzed with was all on page 8.

 

Attachment 1: D040180-B.pdf
D040180-B.pdf D040180-B.pdf D040180-B.pdf D040180-B.pdf D040180-B.pdf D040180-B.pdf D040180-B.pdf D040180-B.pdf
  4140   Wed Jan 12 01:38:52 2011 KojiUpdateIOOPut MC PZT offset onto MC board, instead of on awkward cart

I can not think of any reason that the input impedance of 13kOhm between the pos/neg inputs produces such a big change at the output. In any case, the differential voltage between the pos/neg inputs produces a big output. But the output was just 2V or so. This means that the neg input was actually about zero volt. This ensures the principle of the summing amplifier of this kind.

Because the input impedance of the summing node (the additional resister you put at the negative input) is not infinity, the voltage divider is not perfect and shows 10% reduction of the voltge (i.e. the output will have +4.5V offset instead of +5V). But still it is not enough to explain such a small output like 2.3V.

What I can think of is that the earlier stages somehow have the offset for some reason. Anyway, it is difficult to guess the true reason unless all of the nodes around the last stage are checked with the multimeters.

At least, we can remove the voltage divider and instead put a 10k between -15V and the neg input in order to impose +5V offset at the output. This costs 1.5mA instead of 10mA.

Quote:

[Larisa and Jenne]

We wanted to get rid of the awkward cart that was sitting behind the 1Y1 rack.  This cart was supplying a +5V offset to the PZT driver, so that we could use the MC length signal to feedback to lock the laser to the MC cavity.  Instead, we put the offset on the last op amp before the servo out on the Mc Servo Board.  Because we wanted +5V, but the board only had +5, +15, -15V as options, and we needed -5 to add just before the op amp (U40 in the schematic), because the op amp is using regular negative feedback, we made a little voltage divider between -15V and GND, to give ourselves -5V.  We used the back side of the voltage test points (where you can check to make sure that you're actually getting DC voltage on the board), and used a 511Ohm and 1.02kOhm resistor as a voltage divider. 

Then we put a 3.32kOhm resistor in ~"parallel" to R124, which is the usual resistor just before the negative input of the op amp.  Our -5V goes to our new resistor, and should, at the output, give us a +5V offset. 

Sadly, when we measure the actual output we get, it's only +2.3V.  Sadface.

We went ahead and plugged the servo out into the PZT driver anyway, since we had previously seen that the fluctuation when the mode cleaner is locked was much less than a volt, so we won't run into any problems with the PZT driver running into the lower limit (it only goes 0-10V).

Suresh has discovered that the op amp that we're looking at, U40 on the schematic, is an AD829, which has an input impedance of a measely 13kOhm.  So maybe the 3.32kOhm resistors that we are using (because that's what had already been there) are too large.  Perhaps tomorrow I'll switch all 3 resistors (R119, R124, and our new one) to something more like 1kOhm.  But right now, the MC is locked, and I'm super hungry, and it's time for some arm locking action.

I've attached the schematic.  The stuff that we fitzed with was all on page 8.

 

 

  4141   Wed Jan 12 01:39:49 2011 kiwamuUpdateLSClocked X arm

[Suresh, Kiwamu]

 We eventually succeeded in locking X arm with the infrared beam.

The PDH signal is taken at MCL's ADC instead of c1lsc's, and fedback to MC2_POS through the MCL path.

Right now the lock is not so stable for some reasons, so we need to investigate it more.

 

(what we did)

 - strung a long BNC cable to connect the demodulated signal and the ADC of c1ioo.

          We didn't touch anything on the demodulation system, so the setup for the demodulation is exactly the same as that of yesterday (see here).

 - disconnected the actual MCL cable from the ADC breakout board at 1X2 rack. And put the demodulated signal onto it.

 - checked the analog dewhitening filter state for the MC2 coil driver, found the analog filter are always off.

  So we just made simDW and invDW always on.

- changed the gain of the MCL loop to have a stable lock for the X arm.

    right now a reasonable setup in the MCL filters are:

         FM1:ON, FM10:ON, G=0.1

 - In fact the lock of the MC is not so stable compared to before, frequently an attempt of locking the X arm leads to the unlock of the MC.

  4142   Wed Jan 12 02:41:19 2011 kiwamuUpdateGeneraldaytime tasks for tomorrow

[Rana, Kiwamu]

Here is the list for the daytime tasks of tomorrow, Jan. 12th.

The daytime task is a work basically to be done or quitted before the sun goes down.

Along with the tasks, we roughly assigned the people who are responsible for it.

The tasks below are basically separated from each other, so we can work in parallel.

 

 

--- 1.  LSC analog interface check (Joe/Koji)

   * check whitening filter

   * demodulation board check

    * check ADC connections

 

--- 2.  MC2 coil Dewhitening (Joe)

    * fix binary outputs.

    * FM9 must be the trigger of the binary outputs instead of FM10.

 

--- 3. 11MHz modulation depth (Kiwamu)

   * investigate why the depth is so low

 

--- 4. PEM DAQ name issue (Jenne)

   * change the name of seismic channels properly so that we can deal with the calibrated stuff

 

--- 5. phase adjustment for MC-PDH locking (Suresh)

 

--- 6. medm screens for C1LSC ()

    * make screens

 

 

 

  4144   Wed Jan 12 17:50:21 2011 josephbUpdateCDSWorked on c1lsc, MC2 screens

[josephb, osamu, kiwamu]

We worked over by the 1Y2 rack today, trying to debug why we didn't get any signal to the c1lsc ADC.

We turned off the power to the rack several times while examining cards, including the whitening filter board, AA board, and the REFL 33 demod board.  I will note, I incorrectly turned off power in the 1Y1 rack briefly. 

We noticed a small wire on the whitening filter board on the channel 5 path.  Rana suggested this was to part of a fix for the channels 4 and 5 having too much cross talk.  A trace was cut and this jumper added to fix that particular problem.

We confirmed would could pass signals through each individual channel on the AA and whitening filter boards.  When we put them back in, we did noticed a large offset when the inputs were not terminated.  After terminating all inputs, values at the ADC were reasonable, measuring on from 0 to about -20 counts.  We applied a 1 Hz, 0.1 Vpp signal and confirmed we saw the digital controls respond back with the correct sine wave.

We examined the REFL 33 demod board and confirmed it would work for demodulating 11 MHZ, although without tuning, the I and Q phases will not be exactly 90 degrees apart.

The REFL 33  I and Q outputs have been connected to the whitening board's 1 and 2 inputs, respectively.  Once Kiwamu  adds approriate LO and PD signals to the REFL 33 demod board he should be able to see the resulting I and Q signals digitally on the PD1 I and Q channels.

 

In an unrelated fix, we examined the suspensions screens, specifically the Dewhitening lights.  Turns out the lights were still looking at SW2 bit 7 instead of SW2 bit 5.  The actual front end models were using the correct bit (21 which corresponds to the 9th filter bank), so this was purely a display issue.  Tomorrow I'll take a look at the binary outputs and see why the analog filters aren't actually changing.

 

 

 

  4145   Wed Jan 12 22:19:54 2011 KojiUpdateIOOMC flakiness solved

[Koji Suresh Kiwamu]

Suresh modified the MC board to have +5V offset at the output. (To be reported in the separated elog)

The MC lock has not been obtained at this point. An investigation revealed that there was very small (~5mVpp) PDH signal.

Kiwamu removed his triple resonant adapter and put the 50Ohm termination insted.
This restored the signal level normal although this changed the demodulation phase about 20deg.
We left the demodulation phase as it is because this is a temporary setup and the loss of the signal is not significant.

Now the MC is steadily locked with the single super boost.

  4146   Wed Jan 12 22:33:24 2011 kiwamuUpdateCDSMC2 dewhitening are healthy except for UR

I briefly checked the MC2 analog dewhitening filters.

It turned out that the switching of the dewhitening filters from epics worked correctly except for the UR path.

I couldn't get a healthy transfer function for the UR path probably because the UR monitor at the front panel on either the AI filter or the dewhitening filter maybe broken.

Need a check again.

Quote:  #4144

Tomorrow I'll take a look at the binary outputs and see why the analog filters aren't actually changing.

  4147   Wed Jan 12 22:39:16 2011 SureshUpdateIOOPut MC PZT offset onto MC board, instead of on awkward cart

Quote:

I can not think of any reason that the input impedance of 13kOhm between the pos/neg inputs produces such a big change at the output. In any case, the differential voltage between the pos/neg inputs produces a big output. But the output was just 2V or so. This means that the neg input was actually about zero volt. This ensures the principle of the summing amplifier of this kind.

Because the input impedance of the summing node (the additional resister you put at the negative input) is not infinity, the voltage divider is not perfect and shows 10% reduction of the voltge (i.e. the output will have +4.5V offset instead of +5V). But still it is not enough to explain such a small output like 2.3V.

What I can think of is that the earlier stages somehow have the offset for some reason. Anyway, it is difficult to guess the true reason unless all of the nodes around the last stage are checked with the multimeters.

At least, we can remove the voltage divider and instead put a 10k between -15V and the neg input in order to impose +5V offset at the output. This costs 1.5mA instead of 10mA.

Quote:

[Larisa and Jenne]

We wanted to get rid of the awkward cart that was sitting behind the 1Y1 rack.  This cart was supplying a +5V offset to the PZT driver, so that we could use the MC length signal to feedback to lock the laser to the MC cavity.  Instead, we put the offset on the last op amp before the servo out on the Mc Servo Board.  Because we wanted +5V, but the board only had +5, +15, -15V as options, and we needed -5 to add just before the op amp (U40 in the schematic), because the op amp is using regular negative feedback, we made a little voltage divider between -15V and GND, to give ourselves -5V.  We used the back side of the voltage test points (where you can check to make sure that you're actually getting DC voltage on the board), and used a 511Ohm and 1.02kOhm resistor as a voltage divider. 

Then we put a 3.32kOhm resistor in ~"parallel" to R124, which is the usual resistor just before the negative input of the op amp.  Our -5V goes to our new resistor, and should, at the output, give us a +5V offset. 

Sadly, when we measure the actual output we get, it's only +2.3V.  Sadface.

We went ahead and plugged the servo out into the PZT driver anyway, since we had previously seen that the fluctuation when the mode cleaner is locked was much less than a volt, so we won't run into any problems with the PZT driver running into the lower limit (it only goes 0-10V).

Suresh has discovered that the op amp that we're looking at, U40 on the schematic, is an AD829, which has an input impedance of a measely 13kOhm.  So maybe the 3.32kOhm resistors that we are using (because that's what had already been there) are too large.  Perhaps tomorrow I'll switch all 3 resistors (R119, R124, and our new one) to something more like 1kOhm.  But right now, the MC is locked, and I'm super hungry, and it's time for some arm locking action.

I've attached the schematic.  The stuff that we fitzed with was all on page 8.

 

 

 

[Koji, Suresh]

    We looked at the board and found that the resistor R119 (the feed back) is 1.65k instead of the 3.32k that was needed for unity gain.  The gain has been intentionally reduced to 0.5 so that output range would be close to the 0-10V that is required at the input range of the PZT driver which follows.   A note to this effect is already present in the D040180-B, page 8.

    The voltage divider with 1k and 0.5k provides 4.5V (ref Koji's note above) this provides 2.25V at the output due to the gain of 0.5.  To get to the original goal of introducing a 5V offset on the output, we introduced the modification shown in the  'D040180-B with 5V offset.pdf' uploaded below.  Please check page 8, the changes are marked in red.  We checked to make sure that the output is 5V when the input is disconnected. 

D040180-B_with_5V_offset.pdf

The PCB pics at the end are also attached.  The 4.99k resistor is glued onto the PCB with epoxy and placed as close to the opamp possible.

Attachment 1: P1120508.JPG
P1120508.JPG
Attachment 2: P1120509.JPG
P1120509.JPG
  4148   Thu Jan 13 03:00:01 2011 JenneUpdateIOOWFS shenanigans

My goal this afternoon was to measure the quantum efficiency of the MC WFS.  In the process of doing this, I discovered that when I reverted a change in the MCWFS path (see elog 4107 re: this change), I had not checked the max power going to the WFS when the MC unlocks.

Current status:

MC locks (is locked now).  No light going to WFS at all (to prevent MC WFS french-fry action).  Quantum Efficiency measured.

The Full Story:

Power to WFS:

Rana asked me to check out the quantum efficiency of the WFS, so that we can consider using them for aLIGO.  This involves measuring the power incident on the PDs, and while doing so, I noticed that WFS1 had ~160mW incident and WFS2 had ~240mW incident while the mode cleaner was unlocked.  This is bad, since they should have a max of ~10mW ever.  Not that 200mW is going to destroy the PD immediately, but rather the current out, with the 100V bias that the WFS have, is a truckload of power, and the WFS were in fact getting pretty warm to the touch.  Not so good, if things start melting / failing due to extended exposure to too much heat.

The reason so much power was going to the WFS is that it looks like Yuta/Koji et. al., when trying to use the WFS as a MC1 oplev, changed out 2 of the beam splitters in the MC WFS / MC Refl path, not just one.  Or, we've just been crispy-frying our WFS for a long time.  Who knows?  If it is option A, then it wasn't elogged.  The elog 3878 re: BS changeout only mentions the change of one BS.

Since the MC Refl path has a little more than ~1W of power when the MC is unlocked, and the first BS (which was reverted in elog 4107) is a 10% reflector, so ~100mW goes to the MC Refl PD, and ~900mW goes to the MC WFS path.  In front of a Black Hole beam dump was sitting a BS1-33, so we were getting ~300mW reflected to be split between the 2 WFS, and ~600mW dumped.  The new plan is to put a W2 window in place of this BS1-33, so that we get hopefully something like 0.1% reflected toward the WFS, and everything else will be dumped.  I could not find a W2-45S (everything else is S, so this needs to be S as well).  I found a bunch of W2-0deg, and a few W2-45P.  Does anyone have a secret stash of W2-45S's???  To avoid any more excessive heat just in case, for tonight, I have just left out this mirror entirely, so the whole MC WFS beam is dumped in the Black Hole.  The WFS also have aluminum beam dumps in front of them to prevent light going in.  None of this affects the MC Refl path, so the MC can still lock nice and happily.

Quantum Efficiency Measurement:

I refer to Jamie's LHO elog for the equation governing quantum efficiency of photodiodes: LHO 2 Sept 2009

The information I gathered for each quadrant of each WFS was: [1] Power of light incident on PD (measured with the Ophir power meter), [2] Power of light reflected off the PD (since this light doesn't get absorbed, it's not part of the QE), and [3] the photo current output by the PD (To get this, I measured the voltage out of the DC path that is meant to go to EPICS, and backed out what the current is, based on the schematic, attached). 

I found a nifty 25 pin Dsub breakout board, that you can put in like a cable extension, and you can use clip doodles to look at any of the pins on the cable.  Since this was a PD activity, and I didn't want to die from the 100V bias, I covered all of the pins I wasn't going to use with electrical tape.  After turning down the 100V Kepco that supplies the WFS bias, I stuck the breakout board in the WFS.  Since I was able to measure the voltage at the output of the DC path, if you look at the schematic, I needed to divide this by 2 (to undo the 2nd op amp's gain of 2), and then convert to current using the 499 Ohm resistor, R66 in the 1st DC path.  

I did all 4 quadrants of WFS1 using a 532nm laser pointer, just to make sure that I had my measurement procedure under control, since silicon PDs are nice and sensitive to green.  I got an average QE of ~65% for green, which is not too far off the spec of 70% that Suresh found.

I then did all 8 WFS quadrants using the 1064nm CrystaLaser #2, and got an average QE of ~62% for 1064 (58% if I exclude 2 of the quadrants....see below).  Statistics, and whatever else is needed can wait for tomorrow.

Problem with 2 quadrants of WFS2?

While doing all of this, I noticed that quadrants 3 and 4 of WFS2 seem to be different than all the rest.  You can see this on the MEDM screens in that all 6 other quadrants, when there is no light, read about -0.2, whereas the 2 funny quadrants read positive values.  This might be okay, because they both respond to light, in some kind of proportion to the amount of light on them.  I ended up getting QE of ~72% for both of these quadrants, which doesn't make a whole lot of sense since the spec for green is 70%, and silicon is supposed to be less good for infrared than green.  Anyhow, we'll have to meditate on this.  We should also see if we have a trend, to check how long they have been funny.

Attachment 1: D990249-B-1_MCWFS_schematic.pdf
D990249-B-1_MCWFS_schematic.pdf
  4149   Thu Jan 13 12:56:57 2011 ranaUpdateIOOWFS shenanigans

Actually, I just found out that there are different flavors of 'YAG-444'.

There's a YAG-444AH and also a YAG-444-4AH. I'm not sure which one we have or even which is better. The diode's internal resistance is not listed.

They also say explicitly that he 'YAG Enhancement' is just using thicker Silicon. Since the absorption of 1064 nm light in Silicon is very small, most of the light just goes in and then comes back out without depositing much of the power.

Attachment 1: PerkinElmerQPDs.pdf
PerkinElmerQPDs.pdf PerkinElmerQPDs.pdf
  4150   Thu Jan 13 14:21:13 2011 josephbUpdateCDSWebview of front end model files automated

After Rana pointed me to Yoichi's MEDM snapshot script, I learned how to use Xvfb, which is what Yoichi used to write screens without a real screen.  With this I wrote a new cron script, which I added to Mafalda's cron tab to be run once a day at 6am.

The script is called webview_update.cron and is in /opt/rtcds/caltech/c1/scripts/AutoUpdate/.

#!/bin/bash
DISPLAY=:6
export DISPLAY
#Check if Xvfb server is already running
pid=`ps -eaf|grep vfb | grep $DISPLAY | awk '{print $2}'`
if [ $pid ]; then
        echo "Xvfb already running [pid=${pid}]" >/dev/null
else
# Start Xvfb
echo "Starting Xvfb on $DISPLAY"
Xvfb $DISPLAY -screen 0 1600x1200x24 >&/dev/null &
fi
pid=$!
echo $pid > /opt/rtcds/caltech/c1/scripts/AutoUpdate/Xvfb.pid
sleep 3

#Running the matlab process
/cvs/cds/caltech/apps/linux/matlab/bin/matlab -display :6 -logfile /opt/rtcds/caltech/c1/scripts/AutoUpdate/webview.log -r webview_simlink_update

  4151   Thu Jan 13 16:34:02 2011 josephbUpdateComputers32 bit matlab updated

There was a problem with running the webview report generator in matlab on Mafalada.  It complained of not having a spare report generator license to use, even though the report generator was working before and after on other machines such as Rosalba.  So I moved the old 32 bit matlab directory from /cvs/cds/caltech/apps/Linux/matlab to /cvs/cds/caltech/apps/Linux/matlab_old.  I installed the latest R2010b matlab from IMSS in /cvs/cds/caltech/apps/Linux/matlab and this seems to have made the cron job work on Mafalda now.

  4152   Thu Jan 13 16:41:07 2011 josephbUpdateCDSChannel names for LSC updated

I renamed most of the filter banks in the c1lsc model.  The input filters are now labeled based on the RF photodiode's name, plus I or Q.  The last set of filters in the OM subsystem (output matrix) have had the TO removed, and are now sensibly named ETMX, ETMY, etc.

We also removed the redundant filter banks between the LSCMTRX and the LSC_OM_MTRX.  There is now only one set, the DARM, CARM, etc ones.

The webview of the LSC model can be found here.

  4153   Fri Jan 14 01:55:26 2011 kiwamuUpdateLSCX arm locked with C1LSC digital control

 [Koji, Kiwamu]

 We succeeded in locking the X arm with the C1LSC digital control.

As we did on the day before yesterday, the feedback signal goes to MCL (#4141), but this time the signal is transfered from C1LSC through the RFM.

 


 (key points) 

- checking the state of the analog whitening filters at C1LSC rack.

   We took the transfer function of them and found that they were always on regardless of the clicking any buttons on medm.

To cancel the filter shape of the whitening, we put an unWhitening filter so that these transfer functions becomes flat in total.

The whitening filter approximately has : pole:150Hz, pole:150Hz, zero:15Hz, zero:15Hz (although these numbers came from by our eye ball fitting)   

 

 - demodulation phase adjustment

   We performed the same measurement as that of Suresh and Koji did yesterday (#4143) to adjust the phase of the PDH demodulation.

By changing the cable length we roughly adjusted the I-phase to eventually ~10 deg, which is close enough to 0 deg.

(probably some more efforts should be made as a part of daytime tasks)

Note that we are currently using the REFL33 demodulation board for this purpose (#4144). The LO power we put is about 16dBm.

The angle between I and Q at 11MHz is actually almost 90 deg.

This fact has been confirmed by putting a sinusoidal signal with a slightly different frequency (~100Hz) from that of the LO onto the RF input.

 

 - attenuation of RF signal

  Since the PDH signal taken by C1LSC's ADC had been saturated somewhat, we introduced a ND filter of 10 on the photo diode to attenuate the RF signal.

As a result the amplitude of the PDH signal on dataviewer became more reasonable. No more saturations.

 

(some notes)

 unWhitening filter           pole:15Hz. pole:15Hz, zero:150Hz, zero:150Hz

 C1LSC_MC_FM1   pole:1kHz, zero:10Hz

 Gain in digital control       G ~ -1

measured UGF  ~  200-300 Hz

 measured RFM delay ~ 125 usec 

  4154   Fri Jan 14 11:29:00 2011 kiwamuUpdateLSCexpected open loop TF of X arm locking

Here shows a plot of the expected open loop transfer function (TF) for the X arm locking.

xarm_oltf.png

I assume that the delay time of the digital system associated with the ADC/DAC and the digital filtering process is ~100 usec independently from the RFM delay according to Yuta's measurement (#3961).

Also I assume the MC2 pendulum has a pole at 1Hz with Q of ~5, and the X arm has its cavity pole at ~3kHz.

When the lock acquisition takes place, we used the red curve shown above in order to avoid a big DC feedback onto MC2.

Once the X arm became resonant at TEM00, we manually switched FM3 on, which is a boost filter containing a pole at  1Hz and a zero at 50Hz in order to suppress the residual motion below 1Hz.

The expected curve for the boosted state is drawn by the blue curve in the plot. 

With this open loop TF, the UGF can be realized only around 100-300 Hz due to the phase margin condition.

This expectation of the UGF is consistent with our measurement because we obtained the UGF around 200-300Hz.

In fact above 300Hz we observed that the control became unstable and started oscillating.

 

Quote:

 (some notes)

 unWhitening filter           pole:15Hz. pole:15Hz, zero:150Hz, zero:150Hz

 C1LSC_MC_FM1   pole:1kHz, zero:10Hz

 Gain in digital control       G ~ -1

measured UGF  ~  200-300 Hz

 measured RFM delay ~ 125 usec 

 

  4155   Fri Jan 14 12:29:57 2011 KojiUpdateLockingNext steps for the green

These are the next steps for a better operation of the arm locking. They are suitable for the day time activities

Reconfiguration of the X-End table

- End transmission power monitor (CDS model exists, need to configure the PD)

- IR steering mirror for the transmon

- Restore/align end green beam

- Relocate the end trans CCD

- Connect the video output cable for the X-end CRT monitor

LSC Whitening

- LSC Whitening binary IO connection

 


They are not urgent but also good things to do

MC servo characterization

- Error signal: frequency noise

- Loop characterization

Arm cavity characterization with cavity sweep

- Arm finesse for 1064nm and 532nm

- Arm FSR measurement with 1064 (and optionally with 532nm simultaneously)

- Arm g-factor for 1064nm and 532nm

  4156   Fri Jan 14 12:34:08 2011 KojiUpdateLSCX arm locked with C1LSC digital control

My feeling was that the saturation was caused by the LSC whitening filter which was always on.
Once the LSC whitening filter is controlled from C1LSC, we would be able to remove the attenuator.

Quote:

  - attenuation of RF signal

  Since the PDH signal taken by C1LSC's ADC had been saturated somewhat, we introduced a ND filter of 10 on the photo diode to attenuate the RF signal.

As a result the amplitude of the PDH signal on dataviewer became more reasonable. No more saturations.

 

  4157   Fri Jan 14 17:13:39 2011 josephbUpdateCamerasPylon driver for Basler Cameras installed on Megatron

After getting some help from the Basler technical support, I was directed to the following ftp link:

ftp://Pylon4Linux-ro:h50UZgkl@ftp.baslerweb.com

I went to the pylon 2.1.0 directory and downloaded the pylon-2.1.0-1748-bininst-64.tar.bz2 file.  Inside of this tar file was another one called pylon-bininst-64.tar.bz2 (along with some other sample programs). I ran tar -jxf on pylon-bininst-64.tar.bz2 and placed the results into the /opt/pylon directory.  It produced a directory of includes, libraries and binaries there.

After playing around with the make files for several sample programs they provided, I finally have been able to compile  them.  At several places I had to have the make files point to /opt/pylon/lib64 rather than /opt/pylon/lib.  I'll be testing the camera with these programs on Monday.  I'd also like to see if this particular distribution will work on Centos machines.  There's some comments in one of the INSTALL help files suggesting packages needed for an install on Fedora 9, which may mean its possible to get this version working on the Centos machines.

  4160   Fri Jan 14 20:39:20 2011 ranaUpdateCDSUpdated some DAQ channel names

I like this activateDAQ script, but someone (Jenne with Joe's help) still needs to add the PEM channels - we still cannot see any seismic trends.

  4161   Sun Jan 16 02:20:59 2011 SureshUpdateLockingcomparing the PSL with the X-end-NPRO through the green beat

Objective:

      We wish to study the coherence of the two NPROs i.e. PSL and the X-end-NPRO by locking both of them to the X-arm and then observing the green beat frequency fluctuations. 

What we did:

   a) locked the PSL to the X-arm as described in 4153

   b) locked the x-end-NPRO to the X-arm with a PDH lock to the reflected green from the ETMX

   c) Obtained the green beat signal with a spectrum analyser as described in 3771

Observations:

   Please see the attached screen shots from the spectrum analyser.   They are taken with different BW and sweep range settings.  They give a estimate of the width of the green beat signal and the range of the frequency fluctuations of the beat-note.

P1160510.JPGP1160511.JPGP1160515.JPG

P1160516.JPG

 

Estimates:

   a) width of the beat note is less than 6KHz if measured over time scales of a few milli seconds

   b) the frequency fluctuations of the beat note are about 100KHz over time scales longer than 100ms

Next Step:

    We wish to record the beat note frequency as a function of time in order to establish the stability over time scale of a day.

 

 

  4163   Mon Jan 17 15:31:50 2011 josephbUpdateCamerasTest the Basler acA640-100gm camera

The Basler acA640-100gm is a power over ethernet camera.  It uses a power injector to supply power over an ethernet cable to the camera.  Once I got past some initial IP difficulties, the camera worked fine out of the box.

You need to set some environment variables first, so the code knows where its libraries are located.

setenv PYLON_ROOT /opt/pylon
setenv GENICAM_ROOT_V1_1 /opt/pylon
setenv GENICAM_CACHE /cvs/cds/caltech/users/josephb/xml_cache
setenv LD_LIBRARY_PATH /opt/pylon/lib64:$LD_LIBRARY_PATH

I then run the /opt/pylon/bin/PylonViewerApp

Notes on IP:

Initially, you need to set the computer connecting to the camera to an ip in the 169.254.0.XXX range.  I used 169.254.0.1 on megatron's eth1 ethernet connection.  I also set mtu to 9000.

You can then run the IpConfigurator in /opt/pylon/bin/ to change the camera IP as needed.

Attachment 1: PylonViewer.jpg
PylonViewer.jpg
  4164   Mon Jan 17 20:13:38 2011 ranaUpdateElectronicsPOX_11 Optical TF

I used 50 mA to drive the laser diode. The light is split 50/50 between the DUT (Device Under Test) and the New Focus 1611 (1 GHz BW) diode used as the reference.

This measurement is the TF of DUT/(New Focus). The resonances are there, but clearly there's an issue with instability around 200 MHz. The setup is still powered up, so please be careful around the RFPD testing table (don't stomp around yank the cables out of the power supplies).

I looked at the RF Photodiode wiki that Alberto has started - most of the TF features are replicated there. Todo:

* Update the 'schematic' with a real schematic instead of the cartoon.
* Change the circuit to remove the resistor in the RF path.
* Add compensation to avoid the 200 MHz instability.
* Make sure to include opamp current noise in the noise model (it is the dominant noise source but has been left out in the noise estimation plot).
* Make the output into a true 50 Ohms.
Attachment 1: A.TIF
A.TIF
  4165   Tue Jan 18 10:25:23 2011 steveUpdatePEMvertex crane work this week

The crane people are here. The Vertex chambers are covered with plastic. The PSL HEPAs will be running on high during the day.

It is going to be disturbingly noisy and dirty during the day. Please try start working in the evening if it can not wait till next week.

  4166   Wed Jan 19 03:37:30 2011 SureshUpdateLockingcomparing the PSL with the X-end-NPRO through the green beat

 

 

 [Kiwamu, Suresh]

Today we attempted to convert the beat-note frequency into an analog voltage using the SR620 frequency counter.

First an observation: the stability of the green beat was seen to be much better than the 100kHz fluctuation seen yesterday. Probably because Kiwamu noticed that one of the MC mirrors had a large variance in its motion and changed the  gain and filter parameters to decrease this.  The PSL was therefore more stable and the green peak fluctuation was less than 10kHz over time scales of a few seconds. 

SR620 D/A output resolution given by the manufacturer is 5mV over the -10 to +10V range and this range corresponding to 300MHz.  We, however saw fluctuations of 100mV on the screen which looked as if they corresponded to the  least significant bit.  This would imply a resolution of 1.5MHz at this range.   Even if the manufacturer's claim was true it would lead to a resolution of 75kHz, far in excess of the required resolution a few hundred Hz.

We therefore require to set up the VCO-PLL to obtain a finer frequency resolution.

In the mean time the green beat drifted beyond the 100MHz detection band of the green-PD.  So we changed the x-end-NPRO temperature by -0.05 to bring it back into the detection band.

 

We are also considering, Rana and Koji's suggestion of using a set of 14 flip-flops to divide the ~80MHz beat frequency so that it comes down to about 4kHz.  This could then be sampled by the usual 16-bit, 64kSa/s ADC cards and brought into the digital domain where further digital processing would be needed to extract the the required frequency variations  in the 0 to 10kHz band.  Found a nice paper on this object

Attachment 1: Phase_noise_in_digital_frequency_dividers.pdf
Phase_noise_in_digital_frequency_dividers.pdf Phase_noise_in_digital_frequency_dividers.pdf Phase_noise_in_digital_frequency_dividers.pdf Phase_noise_in_digital_frequency_dividers.pdf Phase_noise_in_digital_frequency_dividers.pdf Phase_noise_in_digital_frequency_dividers.pdf Phase_noise_in_digital_frequency_dividers.pdf Phase_noise_in_digital_frequency_dividers.pdf
  4167   Wed Jan 19 04:25:54 2011 KevinUpdateElectronicsPOX Transfer Functions

I redid the optical POX transfer functions and updated the wiki at http://lhocds.ligo-wa.caltech.edu:8000/40m/Electronics/POX.

I measured each transfer function several times to calculate uncertainties for each measured point. There is one large transfer function from 1 MHz to 500 MHz showing a resonance peak at 11 MHz and notches at 22 MHz and 55 MHz. I also made more detailed measurements around each of these resonance peaks. These measurements were fit to a resonance curve to determine the resonant frequency, transimpedance at resonance, and Q for each peak. These measurements agree with the shot noise measurement for the transimpedance at 11 MHz taken earlier considering that this measurement was made at 11 MHz instead of at the resonant frequency of 11.14 MHz.

I measured these transfer functions with the Agilent 4395a using the netgpib.py script last week. I realized that when using this script to save multiple copies of the same measurement after setting up the instrument, the first and second measurements are saved but all measurements saved after are identical to the second measurement until the instrument is physically reset. This happens because the analyzer switches the trigger from continuous to hold after making a measurement using this script. Kiwamu said that the script can be modified to return the trigger to continuous after saving the data so that multiple measurements can be saved without being at the analyzer physically. I did not want to waste more time figuring out how to modify the script to do this so I used one of the netbooks and sat at the analyzer manually returning the trigger to continuous after each measurement.

  4168   Wed Jan 19 10:31:24 2011 josephbUpdateelogElog restarted again

Elog wasn't responding at around 10 am this morning.  I killed the elogd process, then used the restart script.

  4169   Wed Jan 19 10:45:00 2011 KojiUpdateElectronicsPOX Transfer Functions

TF looks fine except for the large peak at around 200MHz which has been reported by Rana. The time series and the spectrum without the light are pathetic...

I still prefer to see the fit by LISO as the pole/zero fitting of LISO as the fit result is more physically understandable.
Anyone can ask me about the instruction how to use LISO

I guess Idc of 24mA would be just a mistake. It looks like ~0.2mA from the plot that sounds normal for the transimpedance of 2kOhm.

Question: What is the HWHM of the reesonance when you have f0 and Q.

 

  4170   Wed Jan 19 17:00:23 2011 KevinUpdateElectronicsPOX Transfer Functions

 The value of I_dc was a mistake. The value should be 240 µA.

The widths of the resonance peaks are listed below the fits to each peak on the wiki.

  4172   Thu Jan 20 01:50:30 2011 KevinUpdateElectronicsPOX Transfer Functions

[Koji, Kevin]

We fit the entire POX optical transfer function from 1 MHz to 500 MHz in LISO. The fit is on the wiki at http://lhocds.ligo-wa.caltech.edu:8000/40m/Electronics/POX. Using LISO's root fitting mode, we found that the transfer function has five poles and four zeros.

I will work on making plots of the residuals. This is difficult because by default, LISO does not calculate the fitting function at the frequencies of the data points themselves and I haven't figured out how to force it to do this yet.

  4173   Thu Jan 20 04:03:02 2011 kiwamuUpdateCDSc1scy error

 I found that c1scy was not running due to a daq initialization error.

 I couldn't figure out how to fix it, so I am leaving it to Joe.


 Here is the error messages in the dmesg on c1iscey
[   39.429002] c1scy: Invalid num daq chans = 0
[   39.429002] c1scy: DAQ init failed -- exiting
 
 
Before I found this fact, I rebooted c1iscey in order to recover the synchronization with fb.
The synchronization had been lost probably because I shutdowned the daqd on fb.
  4174   Thu Jan 20 04:43:28 2011 kiwamuUpdateGreen Lockingstatus update: PLL connected to ADC

 I connected the PLL signal to the ADC on c1ioo. 

So now we are able to take the data into the digital world, and will be able to feedback signals to the suspensions.

 The output signal from the VCO box goes to a black beakout board on 1X2 rack though a BNC cable.  

Then the signal comes out from the back side of the board with DB39 style, so I put a DB39 to SCSI adapter so that we can take it to the IO chasis.

Now the SCSI is connected to ADC_1 (the second ADC card) on the IO chasis at 1X1. 

 

  Additionally I modified the green locking simulink model, C1GCV, in order to pick the right ADC channels.

A medm screen for green locking is now under the construction. I put a link on the sitemap screen, so anyone can look at the half-baked green locking screen.

Any comments and suggestions are really welcome.

  4175   Thu Jan 20 10:15:50 2011 josephbUpdateCDSc1scy error

This is caused by an insufficient number of active DAQ channels in the C1SCY.ini file located in /opt/rtcds/caltech/c1/chans/daq/.  A quick look (grep -v # C1SCY.ini) indicates there are no active channels.  Experience tells me you need at least 2 active channels.

Taking a look at the activateDAQ.py script in the daq directory, it looks like the C1SCY.ini file is included, by the loop over optics is missing ETMY.  This caused the file to improperly updated when the activateDAQ.py script was run.  I have fixed the C1SCY.ini file (ran a modified version of the activate script on just C1SCY.ini).

I have restarted the c1scy front end using the startc1scy script and is currently working.

Quote:
 Here is the error messages in the dmesg on c1iscey
[   39.429002] c1scy: Invalid num daq chans = 0
[   39.429002] c1scy: DAQ init failed -- exiting
 

 

  4176   Thu Jan 20 15:15:39 2011 kiwamuUpdateGreen Lockingstatus update: PLL connected to ADC

I realized that the black AA board I mentioned on the last entry has the same range issue as Valera reported before (see #3911)

Basically our ADC card has +/- 10V input range, but on the other hand the AA board is already limited by approximately +/- 2V.

We have to fix it.

Quote: #4174

  The output signal from the VCO box goes to a black beakout board on 1X2 rack though a BNC cable.  

Then the signal comes out from the back side of the board with DB39 style, so I put a DB39 to SCSI adapter so that we can take it to the IO chasis.

Now the SCSI is connected to ADC_1 (the second ADC card) on the IO chasis at 1X1. 

  4177   Thu Jan 20 15:39:59 2011 AidanUpdateLockingUpper limit on frequency noise of ADC

[Aidan, Kiwamu]

Kiwamu and I plugged the output from a DS3456 function generator into the ADC and started recording the data. The func. generator output a 237.8Hz, 1Vpp sine wave. We chose this value because it corresponds to the FSR of a 38.5m cavity (=3.896MHz) divided by 2^14, the frequency divider amount we intend to use.

Since 1 FSR divided down is 237.8Hz and corresponds to a length change of the cavity of 532nm/2 = 266nm, then we can, roughly, say that a frequency change of 1Hz corresponds to a length change in the cavity of approximately 1nm. The width of the 237.8Hz peak in the spectra corresponds to an upper limit on the noise floor due to digitizing the signal (this could be limited by the ADC, or the function generator, or the windowing on the FFT).

The FWHM of the peak in the spectrum was approximately 5mHz, corresponding to an uncertainty in the length of the cavity of about 6pm (we used a Hanning Window, 50% overlap and a BW of 2.92mHz, 7 averages). Regardless of what is the dominant contribution to the width of the peak, this implies that the frequency noise associated with digitizing a signal in the ADC is much smaller than we require and will not limit our performance if we choose to use a frequency divider and digital PLL with the Green Locking.

RA: Here's the previous measurement

Attachment 1: Sine-wave-width-test_of_ADC.pdf
Sine-wave-width-test_of_ADC.pdf
  4179   Thu Jan 20 18:20:55 2011 josephbUpdateCDSc1iscex computer and c1sus computer swapped

Since the 1U sized computers don't have enough slots to hold the host interface board, RFM card, and a dolphin card, we had to move the 2U computer from the end to middle to replace c1sus.

We're hoping this will reduce the time associated with reads off the RFM card compared to when its in the IO chassis.  Previous experience on c1ioo shows this change provides about a factor of 2 improvement, with 8 microseconds per read dropping to 4 microseconds per read, per this elog.

So the dolphin card was moved into the 2U chassis, as well as the RFM card.  I had to swap the PMC to PCI adapter on the RFM card since the one originally on it required an external power connection, which the computer doesn't provide.  So I swapped with one of the DAC cards in the c1sus IO chassis.

But then I forgot to hit submit on this elog entry..............