  17272   Wed Nov 16 12:53:36 2022 ranaUpdateASCIMC WFS ongoing

In the middle of aportioning gains and signs in the IMC WFS screen, so beware. More updates soon.

  17288   Fri Nov 18 23:21:54 2022 ranaUpdateASCIMC WFS ongoing

On Wednesday, I did some rework of the MC WFS gains. I think it should still work as before as long as the overall input gain is set to 0.1 (not 1.0 as the button on the screen sets it to).

  1. The MC_TRANS P/TY signals were very small because they are normalized by the SUM. I added a '+80 dB' gain filter to the MC2_TRANS_PIT and MC2_TRANS_YAW filter banks which increase the signal gain before the digital signals are sent from the MC2 model to the MC_WFS control screen's Input Matrix. Now if you plot the MC_TRANS and WFS signals on dataviewer, the time series all have roughly the same magnitude.
  2. I put a "-80 dB" gain button into the MC2_TRANS servo filter banks. This should make it have the same overall gain as before, since the (sensor to servo) Input Matrix is diagonal.
  3. The servo gains (WFS1_PIT, WFS2_YAW, etc.) had some negative signs. To make all the servo gains positive, I moved those signs into the Output Matrix.
  4. The Output Matrix had some values with 4-5 significant digits. I think its not necessary to have more than 2 places after the decimal point since out measurements are not that accurate, so I rounded them off. We can/should change that screen to reduce the PREC field on the matrix element display.
  5. Now, if the overall INPUT_GAIN slider is increased beyond 0.1, there is some pitch oscillation. I think that is happening because the Output Matrix is not that great. In principle, if we have diagonalized the system, putting offsets into the various loops' error points won't make offsets in the other loops, but this is not the case. The pitch loops have a lot of cross coupling (my guess is that the off-diagonal elements are of order 0.1); the yaw loops are several times better. I suggest someone redo the Output Matrix diagonalization and then use the error point offset method to check that they are diagonal.

We mainly want these loops to work well at DC, so it is perhaps better if we can measure the matrix at DC. Its less automatic than at 13 Hz, but I think it could be done with a script and some iterative matrix inversion:

  1. IMC locked, IMC ASC loops all open (by setting the overall input gain slider to zero)
  2. apply an offset in the WFS1_P basis (turn off the integrators in all the servo loops, and apply a ~400 count offset in the error point)
  3. tweak the WFS1_P output matrix until the WFS2_P and MC2_TRANS_P signals go to zero.
  4. repeat for all 6 loops.

I haven't tried this procedure before, but I think it should work. You can use something like "cdsutils servo" to slowly adjust the Output Matrix values.

 

  17311   Thu Nov 24 15:37:45 2022 AnchalUpdateASCIMC WFS output matrix diagonalization effort

I tried following the steps and the method I was using converged to same output matrix upto 2 decimal points but there is still left over cross coupling as you can see in Attachment 1. With the new output matrix, WFS loop can be turned on with full overall gain of 1.


Changes:

  • I switched off +20dB FM2 on C1IOO-WFS1_PIT and increased gain C1:IOO-WFS1_PIT_GAIN from 0.1 to 1 to be uniform with other filters.
  • Output matrix change:
    • Old matrix:
      -2.   4.8 -7.3
       3.6  3.5 -2. 
       2.   1.  -6.8
    • New Matrix:
      3.44  4.22 -7.29
      0.75  0.92 -1.59
      3.41  4.16 -7.21
  • I think the main change that allowed the WFS loop to become stable was the 0,0 element sign change.

Method:

  • I made overall gain C1:IOO-WFS_GAIN 0
  • Switched of (0:0.8) FM3 on PIT filter modules (IOO-WFS1_PIT, IOO-WFS2_PIT, IOO-MC2_TRANS_PIT)
  • Changed ramp time to 2 seconds on all these modules
  • Used offset of 10000 for WFS2 and MC2_TRANS, and 30000 for WFS1 (for some reason, response to WFS1 step was much lower than others)
  • Measured the following sensor channels
    • C1:IOO-WFS1_I_PIT_OUT
    • C1:IOO-WFS2_I_PIT_OUT
    • C1:IOO-MC_TRANS_PIT_OUT
  • First I took 30s average of these channels, then applied the offsets in the three modules one by one and recorded steps in each sensor.
  • Measured step from reference value taken before, and normalized each step to the DOF that was actually stepped to get a matrix.
  • Inverted this matrix and multiplied with existing output matrix. Made sure column norm1 is same as before and column signs are same as before.
  • Repeated a few times.

Note: The standard deviation on the averages was very high even after averaging for 30s. This data should be averaged after low passing high frequencies but I couldn't find the filter module medm screens for these signals, so I just proceeded with simple averaging of full rate signal using cdsultis avg command.


Fri Nov 25 12:46:31 2022

The WFS loop are unstable again. This could be due to the matrix balancing done while vacuum was disrupted. The above matrix does not work anymore.

  14092   Fri Jul 20 22:51:28 2018 KojiUpdateIOOIMC WFS path alignment

IMC WFS tuning

- IMC was aligned manually to have maximum output and also spot at the center of the end QPD.
- The IMC WFS spots were aligned to be the center of the WFS QPDs.
- With the good alignment, WFS RF offset and MC2 QPD offsets were tuned via the scripts.

  10646   Tue Oct 28 14:07:28 2014 KojiUpdateIOOIMC WFS sensing matrix measurement

Last night the sensing matrix for IMC WFS&QPD were measured.

C1:IOO-MC(1, 2, 3)_(ASCPIT, ASCYAW)_EXC were excited at 5.01Hz with 100 count
The output of the WFS1/WFS2/QPD were measured. They all looked well responding
i.e. Pitch motion shows pitch error signals, Yaw motion shows yaw error signals.

The below is the transfer function from each suspension to the error signals

MC1P      MC2P     MC3P
-3.16e-4  1.14e-2  4.62e-3 -> WFS1P
 5.43e-3  8.22e-3 -2.79e-3 -> WFS2P
-4.03e-5 -3.98e-5 -3.94e-5 -> QPDP

MC1Y      MC2Y     MC3Y
-6.17e-4  6.03e-4  1.45e-4 -> WFS1Y
-2.43e-4  4.57e-3 -2.16e-3 -> WFS2Y
 7.08e-7  2.40e-6  1.32e-6 -> QPDY

Taking the inverse of these matrices, the scale was adjusted so that the dc response.

  10647   Tue Oct 28 15:27:25 2014 ericqUpdateIOOIMC WFS sensing matrix measurement

 I took some spectra of the error signals and MC2 Trans RIN with the loops off (blue) and on (red) during the current conditions of daytime seismic noise.

45.png

 

  10648   Tue Oct 28 20:47:08 2014 diegoUpdateIOOIMC WFS sensing matrix measurement

Today I started looking into the WFS problem and improvement, after being briefed by Koji and Nicholas. I started taking some measurements of open loop transfer functions for both PIT and YAW for WFS1, WFS2 and MC2_TRANS. For both WFS1 and 2 there is a peak in close proximity of the region with gain>1, and the phase margin is not very high. Tomorrow I will make measurements of the local damping open loop transfer functions, then we'll think how to improve the sensors' behaviour.

  10653   Thu Oct 30 02:12:59 2014 diegoUpdateIOOIMC WFS sensing matrix measurement

[Diego,Koji]

Today we took some measurements of transfer functions and power spectra of suspensions of the MC* mirrors (open loop), for all the DOFs (PIT, POS, SIDE, YAW); the purpose is to evaluate the Q factor of the resonances and then improve the local damping system.

  17255   Thu Nov 10 20:46:32 2022 ranaUpdateASCIMC WFS servo diagnosis

To check out the bandwidths and cross-coupling in the WFS loops, I made a script (attached) to step the offsets around, sleeping between steps. Its also in the scripts/MC/WFS/ dir.

You can see from the steps that there is some serious cross coupling from WFS1-PIT to MC_TRANS PIT. This cross-coupling is not a disaster because we run the MC2 centering loop with such a low gain. This gain hirearchy means that you can effectively consider the IMC with the WFS loops closed to be an "open loop" plant that the MC TRANS loop is trying to control.

I've started another run at 4:40 UTC since my previous one only paused for 30 seconds after turning each offset OFF/ON. This is clearly not long enough to grab the MC_TRANS loop; although you can tell sort of how slow it is from the slope of the error signal after the step is applied.

To make the plot, I used diaggui in the time series mode, with a 3 Hz BW. I applied a 4th order Butterworth filter at 0.3 Hz to low pass the data using the foton string in the time series tool.

  15165   Tue Jan 28 16:01:17 2020 gautamUpdateIOOIMC WFS servos stable again

With all of the shaking (man-made and divine), it was a hard to debug this problem. Summary of fixes:

  1. The beam was misaligned on the WFS 1 and 2 heads, as well as the MC2 trans QPD. I re-aligned the former with the IMC unlocked, the latter (see Attachment) with the IMC locked (but the MC2 spot centering loops disabled).
  2. I reset the WFS DC and RF offsets, as well as the QPD offsets (once I had hand-aligned the IMC mirrors to obtain good transmission).

At least the DC indicators are telling me that the IMC locking is back to a somewhat stable state. I have not yet checked the frequency noise / RIN.

  15170   Tue Jan 28 20:51:37 2020 YehonathanUpdateIOOIMC WFS servos stable again

I resume my IMC ringdown activities now that the IMC is aligned again.

To avoid any accidental misalignments Gautam turned off all the inputs to the WFS servo.

I set up a PD and a lens as in attachment 1 (following Gautam's setup).

I connect the REFL, TRANS and INPut PDs to the oscilloscope.

I connect a Siglent function generator to the AOM driver. I try to shut off the light to the IMC using 1V DC waveform and pressing the output button manually. However, it produced heavily distorted step function in the PMC trans PD.

I use a square wave with a frequency of 20mHz instead with an amplitude of 0.5V offset of 0.25V and dutycycle of 1% so there will be minimal wasted time in the off state. I get nice ringdowns (attachment 2) - forgot to take pictures. The autolocker slightly misaligns the M2 every time it is acting, so I manually align it everytime the IMC gets unlocked.

Data analysis will come later.

I remove the PD and lens and reenable the WFS servo inputs. The IMC locks easily. The WFS outputs are very different than 0 now though.

  12680   Wed Dec 21 21:03:06 2016 KojiSummaryIOOIMC WFS tuning

- Updated the circuit diagrams:

IMC WFS Demodulator Board, Rev. 40m https://dcc.ligo.org/LIGO-D1600503

IMC WFS Whitening Board, Rev. 40m https://dcc.ligo.org/LIGO-D1600504

- Measured the noise levels of the whitening board, demodboard, and nominal free running WFS signals.

- IMC WFS demod phases for 8ch adjusted

Injected an IMC PDH error point offset (@1kHz, 10mV, 10dB gain) and adjusted the phase to have no signal in the Q phase signals.

- The WFS2 PITCH/YAW matrix was fixed

It was found that the WFS heads were rotated by 45 deg (->OK) in CW and CCW for WFS1 and 2, respectively (oh!), while the input matrices were identical! This made the pitch and yaw swapped for WFS2. (See attachment)

- Measured the TFs MC1/2/3 P/Y actuation to the error signals

  12682   Thu Dec 22 18:39:09 2016 KojiSummaryIOOIMC WFS tuning

Noise analysis of the WFS error signals.

Attachment 1: All error signals compared with the noise contribution measured with the RF inputs or the whitening inputs terminated.

Attachment 2: Same plot for all the 16 channels. The first plot (WFS1 I1) shows the comparison of the current noise contributions and the original noise level measured with the RF terminated with the gain adjusted along with the circuit modification for the fair comparison. This plot is telling us that the electronics noise was really close to the error signal.

I wonder if we have the calibration of the IMC suspensions somewhere so that I can convert these plots in to rad/sqrtHz...?

  12683   Fri Dec 23 20:53:44 2016 KojiSummaryIOOIMC WFS tuning

WFS1 / WFS2 demod phases and WFS signal matrix

  12684   Fri Dec 23 21:05:56 2016 KojiSummaryIOOIMC WFS tuning

Signal transfer function measurements

C1:SUS-MC*_ASCPIT_EXC channels were excited for swept sine measurements.

The TFs to WFS1-I1~4, Q1~4, WFS1/2_PIT/YAW, MC2TRANS_PIT/YAW signals were recorded.

The MC1 and MC3 actuation seems to have ~30Hz elliptic LPF somewhere in the electronics chain.
This effect was compensated by subtracting the approximated time delay of 0.022sec.

The TFs were devided by freq^2 to make the response flat and averaged between 7Hz to 15Hz.
The results have been summarized in Attachment 3&4.

Attachment 4 has the signal sensing matrix. Note that this matrix was measured with the input gain of 0.1.

Input matrix for diagonalizing the actuation/sensor response

Pitch

\begin{pmatrix} -1.58983 & -0.901533 & -5592.53 \\ 0.961632 & -0.569662 & 1715.12 \\ 0.424609 & 1.60783 & -5157.38 \end{pmatrix}

e.g. To produce pure WFS1P reaction, => -1.59 MC1P + 0.962 MC2P + 0.425 MC3P

Yaw

\begin{pmatrix} 1.461 & -0.895191 & -4647.9 \\ 0.0797164 & 0.0127339 & -1684.11 \\ 0.223054 & -1.31518 & -4101.14 \end{pmatrix}

  12685   Sun Dec 25 14:39:59 2016 KojiSummaryIOOIMC WFS tuning

Now, the output matrices in the previous entry were implemented.
The WFS servo loops have been engaged for several hours.
So far the REFL and TRANS look straight. Let's see how it goes.

  12686   Mon Dec 26 12:45:31 2016 KojiSummaryIOOIMC WFS tuning

It didn't go crazy at least for the past 24hours.

  12688   Thu Dec 29 13:22:21 2016 ranaSummaryIOOIMC WFS tuning
  • For the rough calibration below 10 Hz, we can use the SUS OSEM cal: the SUSPIT and SUSYAW error signals are in units of micro-radians.
  • It seems from the noise plots that the demod board is now dominating over the whitening board noise.
  • If the RF signals at the demod input are low enough, we can consider either increasing the light power on the WFS or increasing the IMC mod. depth.
  • We should look at the out-of-lock light power on the WFS and re-examine what the 'safe' level is. We used to do this based on the dissipated electrical power (bias voltage x photocurrent).

At Hanford, there is this issue with laser jitter turning into an IMC error point noise injection. I wonder if we can try out taking the acoustic band WFS signal and adding it to the MC error point as a digital FF. We could then look at the single arm error signal to see if this makes any improvement. There might be too much digital delay in the WFS signals if the clock rate in the model is too low.

  12689   Thu Dec 29 16:52:51 2016 KojiSummaryIOOIMC WFS tuning

Koji responding to Rana

> For the rough calibration below 10 Hz, we can use the SUS OSEM cal: the SUSPIT and SUSYAW error signals are in units of micro-radians.

I can believe the calibration for the individual OSEMs. But the input matrix looked pretty random, and I was not sure how it was normalized.
If we accept errors by a factor of 2~3, I can just naively believe the calibration factors.

> If the RF signals at the demod input are low enough, we can consider either increasing the light power on the WFS or increasing the IMC mod. depth.

The demod chip has the conversion factor of about the unity. We increased the gains of the AF stages in the demod and whitening boards. However, we only have the RMS of 1~20 counts. This means that we have really small RF signals. We should check what's happening at the RF outputs of the WFS units. Do we have any attenuators in the RF chain? Can we skip them without making the WFS units unstable?

  12690   Thu Dec 29 21:35:30 2016 ranaSummaryIOOIMC WFS tuning

The WFS gains are supposedly maximized already. If we remotely try to increase the gain, the two MAX4106 chips in the RF path will oscillate with each other.

We should insert a bi-directional coupler (if we can find some LEMO to SMA converters) and find out how much actual RF is getting into the demod board.

  12669   Tue Dec 6 16:47:40 2016 KojiUpdateIOOIMC WFS whitening filter investigation

The whitening board saids it is Rev B, but the actual component values are more like Rev. C.

The input stage (AD602) has an input resistor of 909 Ohm.
This is causing a big attenuation of the signal (x1/10) because the input impedance of AD602 is not high. And this screws up the logarithm of the gain.
I don't think this is a right approach.

  12670   Tue Dec 6 17:54:08 2016 KojiUpdateIOOIMC WFS whitening filter investigation

The input resistor 909Ohm of AD602 was shorted. I've confirmed that the gain (= attenuation by voltage division) was increased by a factor of 10.
This modification was done for WFS2-I1 and WFS2-Q1. Also the thick film resistors for the WFS2-I1 channel was all replaced with thin film resistors.

Attachment 1 shows the comparison of the noise levels. The curves were all calibrated referred to the response of the original whitening filter configuration.
(i.e. measurement done after the gain change was compensated by the factor of 10.)

Now the AF chain is not limited by the noise in the whitening filter board. (Brown)
In fact, this noise level was completely identical between I1 and Q1. Therefore, I don't think we need this resistor replacement for the whitening filter board.

We can observe the improvement of the overall noise level below 10Hz. (Comparison between green and red/blue)
As the signal level goes up, the noise above 100Hz was also improved.

Now we need to take care of the n x 0.7Hz feature which is in the demod board...
 

  12671   Tue Dec 6 22:41:49 2016 KojiUpdateIOOIMC WFS whitening filter investigation

I have implemented the same modification (shorting the input resistor of AD602) to the two whitening boards.

  12676   Tue Dec 13 17:26:42 2016 KojiUpdateIOOIMC WFS whitening filter investigation

Rana pointed out that this modification (removal of 900Ohm) leave the input impedance as low as 100Ohm.
As OP284 can drive up to 10mA, the input can span only +/-1V with some nonlinearity.

Rather than reinstalling the 900Ohms, Rana will investigate the old-days fix for the whitening filter that may involve the removal of AD602s.
Until the solution is supplied, the IMC WFS project is suspended.

  12678   Thu Dec 15 03:46:19 2016 ranaUpdateIOOIMC WFS whitening filter investigation

https://dcc.ligo.org/LIGO-D1400414

As it turns out, its not so old as I thought. Jenne and I reworked these in 2014-2015. The QPD whitening is the same as the IMC WFS whitening so we can just repeat those fixes here for the IMC.

Quote:

Rana pointed out that this modification (removal of 900Ohm) leave the input impedance as low as 100Ohm.
As OP284 can drive up to 10mA, the input can span only +/-1V with some nonlinearity.

Rather than reinstalling the 900Ohms, Rana will investigate the old-days fix for the whitening filter that may involve the removal of AD602s.
Until the solution is supplied, the IMC WFS project is suspended.

 

  17063   Fri Aug 5 12:42:12 2022 KojiUpdateIOOIMC WFS: Overnight observation

The IMC lock survived overnight and the WFS servo loops kept it aligned. The IMC was unlocked in the morning.
The left 6 plots are the WFS servo outputs and the right most two plot show the transmission and reflection of the IMC.

If the WFS is making the lock unstable under high seismic conditions, please turn the loops off.

 

  17192   Sat Oct 15 17:22:56 2022 ChrisUpdateOptimal ControlIMC alignment controller testing

[Anchal, Radhika, Jamie, Chris]

We conducted a test of three alternative controllers for the IMC pitch DOFs on Friday. These were loaded into a new RTS model c1sbr, which runs on the c1ioo front end as a user-space program at 256 Hz. It communicates with the c1ioo controller via shared memory IPCs to exchange error and control signals.

The IMC maintained lock during the handoffs, and we were able to take one minute of data for each (circa GPS 1349807926, 1349808426, 1349808751; spectra attached), which we can review to assess the performance vs the baseline. (On the first trial, lock was lost at the end when the script tried to switch back to the baseline controller, because we did not take care to clear the integrators. On subsequent trials we did that part by hand.)

The method of setting up this test was convoluted, but now that we see it working, we can start putting in the merge requests to get the changes better integrated into the system. First, modifications were required to the realtime code generator, to get controllers running at the new sample rate of 256 Hz. (This was done in a separate filesystem image on fb1, /diskless/root.buster256, which is only loaded by c1ioo, so as to isolate the changes from the other front end machines.) The generated code then needed hand-edits to insert additional header files and linker options, so that the alternative controllers could be loaded from .so shared libraries. Also, the kernel parameters had to be set as described here, to allow the user-space controller to have a CPU core all to itself. Finally, isolating the core was done following the recipe in this script (skipping the parts related to docker, since we didn’t use it).

  17314   Sun Nov 27 15:30:22 2022 ChrisUpdateOptimal ControlIMC alignment controller testing

Five more mode cleaner alignment controllers were tested this morning (remotely). These were designed to run in tandem with the standard controller, instead of supplanting it. Before the test, c1ioo was burt restored back to the settings of the previous test on Oct 28, and in MC TRANS PIT/YAW filter banks the 80 dB gain filters were disengaged and outputs were enabled. Subsequently, all settings were returned to the original values. Each test consisted of five minutes with pitch alignment uncontrolled, five minutes with the standard controller only, and twenty minutes with both controllers enabled. GPS times for each phase of testing are the following:

  • musgo
    • OL start 1353602764
    • CL start 1353603074
    • policy start 1353603410
  • musgo_ghost
    • OL start 1353604697
    • CL start 1353605007
    • policy start 1353605355
  • musgo_stumble
    • OL start 1353606574
    • CL start 1353606884
    • policy start 1353607229
  • musgo_goldfish
    • OL start 1353608446
    • CL start 1353608756
    • policy start 1353609099
  • musgo_late
    • OL start 1353610321
    • CL start 1353610631
    • policy start 1353610971
  14310   Tue Nov 20 13:13:01 2018 gautamUpdateVACIMC alignment is okay

I checked the IMC alignment following the vent, for which the manual beam block placed on the PSL table was removed. The alignment is okay, after minor touchup, the MC Trans was ~1200 cts which is roughly what it was pre-vent. I've closed the PSL shutter again.

  15227   Wed Feb 26 22:05:06 2020 gautamUpdateIOOIMC checks

Today, I did the following tests (and so was touching electronics/cables at/around 1X2):

  • Measured the IMC OLTF.
  • Measured the TF from injection at IN2 to response at the IMC error point (T-eed the I out of the IMC demodulator as there is no longer a monitor point available).
  • Measured the IMC in loop error signal with 0.25 Hz resolution from DC-2kHz.
  • Confirmed that the IMC IN2 (a.k.a. AO path) gain slider performs as advertised. This is a useful test to run post Acromag switchover on Friday.

Results to follow.

After this work, I reverted the EPICS channels to the usual values. The IMC can be locked.

  15229   Wed Feb 26 23:50:51 2020 gautamUpdateIOOIMC checks

In the style of the KA characterization of the CM board, the AO path gain EPICS slider (IN2) of the IMC servo board was stepped by 1 dB through the full available range of -32 dB to +31 dB. For each value of the requested gain, I measured the TF from the injected signal (to IN2) to TP1A on the IMC servo board. I used the BNC connector for this test, whereas we use the LEMO connector for the AO path. The source was tee-d off at the SR785 side, with one leg going to IN2 of the IMC servo board, and the other going to CH1A of the SR785. TP1A of the IMC board was connected to CH2A of the SR785. 

Attachment #1 - Measured gain vs requested gain.

  • When debugging the CM board, it was this kind of test that revealed the faulty latch ICs.
  • -12 dB to -11 dB gain step looks anomalous, but overall the trend seems linear.
  • I was confused by why there should be a discontinuity at this stage of the gain stepping - seems like the scanning script I use changes the SR785 excitation amplitude at this point (from 300mV to 100mV). But why should the size of the excitation signal change the magnitude of the transfer function? Is this indicative of some loading issue?
  • There is an overall offset between the requested gain and measured gain of ~2-3 dB. This seems large.
  • There is nothing in the schematic which would have me expect this - there is a 1/2 divider at the positive input of the differential receiving stage, but this just cancels out the non-inverting gain of x2.

Attachment #2 - Frequency dependent transfer functions

  • There seem to be two families of curves - they correspond to <-12 dB and >-12 dB.
  • The feature at 90 kHz is strange - need to look at the schematic to see what this could be.

The motivation here is to try and figure out why I cannot engage the AO path smoothly in the CARM handoff part of lock acquisiton. I plan to use this information to do some loop modeling and project laser frequency noise coupling in various stages of the lock acquisition process.

  16782   Fri Apr 15 11:59:16 2022 YehonathanUpdateIOOIMC completely misaligned

Came this morning, opened the PSL and there was not even a beam on the MC REFL.

Looking at the big monitor it seems like the WFS signals went through the roof during the "auto-alignment" night session.

I restored the MC alignment from before the misalignment happen and wait for the SUS to damp. Once the RMS values went below 200 I enabled the watchdog and the coil outputs.

I opened the PSL shutter and the IMC locked immediately. I turned on the WFS servo and the MC REFL DC went down to 0.3. I run the WFS relief script.

  14653   Tue Jun 4 10:56:31 2019 gautamUpdateIOOIMC diagnostics

I briefly managed to lock the IMC today - it stayed locked for ~10 minutes. Attachment #1 shows spectra of a few error and control signals for today's lock, and from a stretch yesterday before the problems surfaced*. The 60 Hz lines are much bigger, and MC_F signals broadband excess noise above a few Hz. I suspect a problem somewhere in the electronics.

*I confess the comparison isn't entirely valid because I had to tweak the FSS FAST gain from its nominal value of 22 to 25 in order to get the PC drive RMS down to the ~1.5V level. At the nominal gain setting, with the laser frequency locked to the cavity length, the PC Drive RMS was ~4 V. Still, indicative of something being off in the electronics.

  14659   Thu Jun 6 22:11:53 2019 KojiUpdateIOOIMC diagnostics

As per Gautam's request, I looked at the IMC situation.

Locking path

  • Acquisition: IMC IN1 Gain +4 (nominal), Boost 0, VCO Gain (-32), FSS Common +6 (nominal), FSS FAST +20
    This is too low gain. So oscillate VCO Gain between -32 and ~0 until TEM00 lock is acquired
  • Once lock is acquired, bring the VCO gain to +11 (new nominal), and increase the FSS FAST to +23 (new nominal). Change the IMC BOOST to 3 (nominal)

Diagnosis

  • The PMC servo gain was checked. The control signal monitor for the PMC actuation was hooked up to SR785. The nominal gain was +18dB. Increasing the gain to 20dB made the servo oscillating. So the nominal gain of +18dB seems still reasonable.
  • The status of NOISE EATER was checked. Both the PMC REFL and TRANS were looked at by AG4395A. The power spectrum of them did not change much around the kHz~MHz region. It made the PSD slightly (x2~3) improved below 1kHz. I also did not recognize the relaxation oscillation peak. So I could not figure out where to see. NOISE EATSER was on and is still on.
  • IFO Modulation Freq: I took this chance to look at the IMC absolute length using the peak at 3.6MHz. The TP1A output of the IMC servo board was hooked up to AG4395A.
    The new FSR of the IMC (and thus the modulation frequency for the IFO) is 11.066275MHz (instead of the previous 11.066209MHz).
    This corresponds to 0.16mm difference in the roundtrip length.
  • (*Still working) IMC SERVO configuration:
    • FAST 25 (nominal) sometimes invoke the oscilattion. 24 has gain peaking ~30kHz. There is a big line peak at 35kHz so wanted to avoid the servo bump (PZT-EOM cross over). So decided to use 23dB. (This is not optimal for the CM servo as we need as much as bandwidth for CM servo.)
    • IMC VCO GAIN (bad name. this is actually overall output gain for IMC) was increased from the nominal 7 to 11. Increasing this above 11 makes the servo oscillating at ~200kHz.
  • (*Still working) Measured power spectrum of the error signal. Too many line peaks.
  • (*Still working) Single trigger observation: Oscilloscope monitoring started from 35kHz going up and ~20kHz oscillation +/-6V of the IMC servo output was observed. Could not capture good data for this. Try the other day.

I'll complete the entry later.

  14670   Thu Jun 13 18:01:18 2019 aaronUpdateIOOIMC diagnostics

Continuing this investigation of the IMC, today I am getting familiar with the PMC and FSS. I'd like to measure the frequency noise of the PSL referenced to the PMC.

I checked that the PSL shutter is off, so no light reaches the IMC.

I'm not really sure what I'm looking for on the FSS boxes. I found a few documents to guide:

I ran the FSS autolock script from C1PSL_FSS, nothing obvious changes when I do so. The FSS error signal (which I think is PSL-FSS_MIXERM) is flatlined, and the RC-RF_PD has no LO (PSL-FSS_LOCALC is nan).

  14675   Fri Jun 14 13:10:00 2019 aaronUpdateIOOIMC diagnostics

The circuit diagram for the PMC servo card is D980352. From this diagram, I see that I can send an excitation from the network analyzer to FP2TEST (9.09 kOhm input impedance) where it is added to the PMC error signal before going to the loop filters.

I hook up the following

  • Agilent 4395A output to SR560 (300 Hz HP, gain of 1)
  • SR560 to FP2TEST and to Agilent's channel R
  • PMC error signal IF (mixer box mounted to rack, I noticed the IF BNC->SMA were a bit loose/stressed by a hanging LP RF filter) to SR560 (also 300 Hz HP, gain of 2)
  • SR560 to Agilent channel A

I 'Enable' Test 2 on the PSL screen, so FP2TEST gets added to the error signal.

PDH signal

  • TDS 3034B with four channels
    • 1. PMC servo box external drive (split off from the function generator)
    • 2. PMC servo box output monitor (mirrors the drive, shows when drive is saturating)
    • 3. IF signal (split off after the mixer)
    • 4. PMC Trans (long BNC from the PSL table)
  • SRS DS345 function generator (into the PMC servo box' external drive)
    • 100 Hz signal (there's a 10 Hz pole on the PZT drive, so any faster than this and I can't see both sidebands without the HV output mon clipping)
    • 3.19 Vpp amplitude (smallest amplitude at 100 Hz such that both sidebands are well resolved)
    • 1.52 V offset (center the carrier's PDH error signal at pi/2 out of phase with the drive)

I was able to see the carrier and both sidebands.

I tried to grab this data from the scope via ethernet, but was unsuccessful (timeout errors, I'm using the scripts from scripts/tektronix/tek-dump, and the GPIB box that Kruthi had been using for the GigE cam; I also tried plugging in directly scope->ethernet. Never got anything but timeout errors, so maybe I'm not specifying the port correctly. Anyway the trace is frozen on the scope for later use, or I can easily repeat this now that I know how).

Spectrum

Next, I locked the PMC (Test1 is off, tune DC output adjust until I get some transmission, turn on the loop at Test1, increase the gain to before the loop goes unstable). I'm sending the following channels to SR560 (gain = 2, no filtering, high dynamic reserve, 50 Ohm outputs), and reading spectra from the Agilent 4395A:

  • R-- PMC TRANS PD
  • A-- PDH IF
  • B-- PMC PZT HV MON

The HV mon was always saturating the preamp, so I disconnected it; I added a 50 Hz (6db) high pass to the Trans PD signal, since it has a DC component.

I got to take a look at the traces on the spectrum analyzer front panel, but too tired to do the GPIB for now. There are peaks, things look reasonable.

  14677   Mon Jun 17 12:37:16 2019 aaronUpdateIOOIMC diagnostics

Grabbed the PMC data

I went to set up the spectrum analyzer measurements through GPIB, but inadvertently deleted the contents of ~/Agilent/netgpibdata/ (made a soft link in my folder, decided I wanted it gone but rm'ed instead of unlink). I copied what I think was in that folder back (from /opt/rtcds/caltech/c1/scripts/general/labutils/netgpibdata).

Again, the spectra are:

  • R-- PMC TRANS PD into SR560 with G=2 DC coupled, no filtering
  • A-- PDH IF into SR560 with G=2, DC coupled, no filtering
  • B-- PMC PZT HV MON into SR560 with G=2, AC coupled, no filtering

I recorded the three spectra with the following parameters:

  • Three separate spans:
    • 10 Hz to 150 kHz
    • 100 to 550 kHz
    • 500 kHz to 2.5 MHz
  • bwSpanRatio = 0.1 %
  • averages: 10
  • number of points: 801
  • spec type: noise (PSD units)

I then ran AGmeasure with the above parameters in the yaml, with the rest following the defaults in AgilentTemplate.yaml. I saved the data in /users/aaron/40m/data/PMC/190617/

Looks like the header contains all of the parameters, so I shouldn't have trouble distinguishing the spectra. I didn't get the instant plotting working, but the data seem to be there.

I'm still having trouble getting the data from the oscilloscope. I'm not sure why the tektronix scrips I've used before aren't working, I'm checking it now.

update: Grabbed the data, the issue was just using the wrong IP address. 

 

  14681   Tue Jun 18 20:35:07 2019 aaronUpdateIOOIMC diagnostics

I made a script (/users/aaron/40m/GPIB/tds_dump.py) that grabs data from a Tektronix scope and packages into a pickled dict with the following structure:

  1. ch1
    1. times ("ts")
    2. values ("vals")
    3. channel info ("info")
  2. ch2
    1. ""
  3. etc

I made a python notebook that does the following:

  1. Grab the data from the pickle above
  2. Fit a triangle wave to the drive signal
  3. Determine the (change in Volts) / second from the triangle wave, as well as define the times of a single sweep of the PDH error signal
  4. Trim the error signal data to contain the PDH signal from the carrier and two sidebands only (the original trace was for three periods).
  5. Fit the functional form of the PDH signal to the trimmed error signal. 
    1. The sideband frequency is fixed at 35.5 MHz, and the scaling of Volts-to-Hz is left free, so this fit gives the calibration of IF volts to Hz.
  6. Grab the spectra (already saved from the Agilent with the netgpib scripts) and apply this V-to-Hz scaling
  7. Plot the spectra

The fit in step (5) is still looking quite bad, despite the fitted values being close to the expected. Since we really just want a calibrated spectrum, I'll instead fit a line to the linear portion of the PDH error signal for the carrier and both sidebands, then determine the scaling from that.

  14683   Wed Jun 19 19:12:51 2019 aaronUpdateIOOIMC diagnostics

Here are the results from the fit. Data can be found on nodus in /users/aaron/40m/data/PMC/190617/. I've put a jupyter notebook with the analysis in /users/aaron/40m/analysis/PDH_calibrate.ipynb (might be some filename issues due to different directory structure on my laptop).

Here's a summary of the current measurement. I'll be referencing the diagram for the PMC servo card.

  1. With the PMC servo loop open, sweep the PMC PZT by sending a triangle wave in to J5 (external drive on the servo card).
    1. I used a 100Hz drive, but should use something slower so my drive isn't filtered out by the 100Hz pole on the servo card and the 10Hz pole on the PZT.
  2. Monitor the voltage at the HV drive, as well as at "mixer out" (J8 on the servo board)
    1. Note that I took this PDH error signal from FP2TEST rather than "mixer out", which means my error signal was not low pass filtered.
  3. Calibrate the HV mon in units of Hz by fitting the PDH signal. The sidebands should be 35.5Mhz away from the carrier peak.
    1. This part needs to be done differently to account for thermal locking in the PMC.
  4. You now have the PDH error signal as a function of PMC resonance in Hz, and can use that to calibrate the PDH error spectrum.
    1. The spectrum is taken when the PMC is locked, so the Hz/V scaling is the slope of the PDH error signal.

In the figures below, I obesrved that for fast (100Hz) drives, the PDH error signal had a pi/2 phase shift relative to the triangle wave, which means even though the resonance appears near the turnaround of the triangle, it is actually occuring near the center of the range.

There are several problems with this data:

  • PMC error signal spectrum is not properly calibrated, even according to the process described above
  • The drive was faster than the response of the PZT.
  • I was driving with a ~1V excitation, so I've lost a factor of 10 somewhere on the way to the external drive curve. Probably just a problem with how I've read the data dump from the scope.
  14686   Fri Jun 21 19:36:26 2019 KojiUpdateIOOIMC diagnostics

The IMC REFL error signal was measured to compare it with the other spectra (if we have).

The blue curve is the in-loop IMC error and the red is the dark noise. So they are not an apple-to-apple comparison. But the red noise is going to be suppressed by the loop, and still the red is below blue. This means that the blue curve is the measured noise rather than the readout noise.

We suspect that the current issue is the PC drive saturation (as usual). Does this indicate that the laser freq noise is actually increased?

----

Another suspect was that the degradation of the LO level. We used to have the issue of slowly dying ERA-5 (ERA-5SM indeed). The RF levels on the demod board were measured using an active probe.

The LO input: 0dBm, ERA-5 input: -2.7dBm and -2.1dBm for I and Q. I found that the outputs of the ERA-5SM were +10.5dBm and +10.6dBm.
This lead me to replace the chips but the situation was not changed. Then I realized that the LO levels should have been measured with the load replaced from the mixers to a 50Ohm load. Somehow these mixers lower the apparent LO levels. So I decided to say this is OK.

  14693   Mon Jun 24 15:49:05 2019 aaronUpdateIOOIMC diagnostics

aI went to repeat these measurements using the mixer out channel from the servo box, and with a slower sweep for the PDH calibration.

I had trouble getting the PDH signal, here are some notes:

  • I added a 50 Ohm terminator to BNC T on the mixer box. This had been terminated before I started, but I noticed no terminator today.
  • Noticed some distortion of my driving triangle wave if I measured it on ch3 and 4 of the tektronix scope, not present on ch 1/2
  • Initially wasn't finding a signal because I was opening the loop by turning off the Test 1 switch, but this meant the mixer mon on the servo box also did not receive the PDH signal. Instead, I cut the loop with the "BLANK" switch on the PMC screen, which instead blanks out the op amp between the mixer mon and the PMC drive conditioning (so the external drive still reaches the PZT).

attachment 1 is the configuration of the PMC screen when I was trying to get some PDH signal; I did move the DC output adjust to 0V, but found that this led to the output being railed; this makes sense, the op amp at U9 has a negative bias at GND.

Rana came by and gave me some tips.

  • I'd been using the wrong servo board diagram, it should be in D1400221
  • We removed the LP filter from the mixer output (before going to FP1TEST on the servo board), since the board itself already is filtering the IF.
  • We might have observed the thermal locking? See for yourself, the trans and refl signals while sweeping the PZT drive at 5 Hz and 30 Hz respectively are in attachments 2 and 3.
  • Rather than using an SR560, I should use an RF coupler between the mixer and FP1TEST to measure the error signal spectrum. I found a ZFDC-20-5-S+ (0.1-2000 MHz) and sent an SMA cable from the coupled port to channel R of the Agilent 4395.

We finally got the PDH signal again, and I recorded the PDH signal while driving with the following settings on the Siglent function generator.

  • 1.1 Hz triangle wave, 6Vpp, -7Vdc offset, high impedance mode

I tried getting a spectrum using the coupler, the mixer mon is seeing a DC offset though and causing the PZT to rail. Will try to understand why, but in the meantime removing the coupler (still no LP filter) lets us lock the PMC again.

RXA: Kruthi thinks all of our subsequent IMC locking problems are Aaron's fault (she was quick to give him up as soon as the thumb screws were tightened...)

  14721   Tue Jul 2 19:36:18 2019 aaronUpdateIOOIMC diagnostics

The latest in my fling with the PMC. Though PMC trans is back to nominal levels (~0.713 V), we'd still like to understand the PMC noise.

Last time, I took some spectra with the RF probe (Agilent 41800A). I had already measured the PDH error signal by sweeping the PZT at ~1 Hz. The notebook I used for analysis has been updated in /users/aaron/analysis/PDH_calibrate.ipynb. The analysis was the following:

  • fit the PDH error signal, assuming a 35.5MHz modulation frequency. Here are the (approximate) fit parameters:
    • Mapping of PZT mon voltage to Hz: 5.92 Hz/V_{PZT_mon}
    • P_carrier*P_signal: 0.193 W^2
    • HV mon voltage on resonance: 0.910 V
    • Error signal far off resonance: 0.249 V
    • Transmission: 0.00238
      • ​yikes. The nominal transmission is T=0.003. I let this parameter be free as a check, and to avoid overconstraining the data; is this consistent with measurements of the PMC optics' transmission?
    • Length: 0.0210 m
      • This is consistent with the nominal PMC length
  • Using the fit of the full PDH error signal, I am able to plot error signal vs frequency, and fit the linear portion of the carrier PDH signal. The results of this fit are:
    • -9.75e-7 V_PDH per Hz
    • 0.105 V error signal at DC
  • I then divide the power spectra by the squared slope of the linear fit above (V_PDH^2/Hz^2) to get the spectra in Hz^2/rtHz
    • I've plotted both the spectrum I took directly at the mixer I using the agilent probe, as well as the spectrum taken by sending the PMC servo card's mixer mon to an SR560 (G=2) then to the spectrum analyzer

There are a few problems remaining:

  • There should be a gain of 100 between the mixer I and the servo board's mixer out. It's not clear to me that this is reflected in the spectra. Moreover, the header files on the spectra I grabbed from the Agilent say that the R (mixer I) channel has 20dB of input attenuation, which is also not reflected. If I have swapped the two spectra and not accounted for either the gain of the servo card or the attenuation of the spectrum analyzer, these two gains would cancel, but I'm not confident that's what's going on.
  11561   Thu Sep 3 00:14:09 2015 ranaConfigurationIOOIMC fast gain change for lock acq

The IMC often was making that scratchy noise when first catching lock and sometimes breaking. Thinking of the crappy crossover sit that EQ showed in his latest plots, I decided that it didn't make sense to acquire lock with an unstable PZT/EOM crossover, so I have changed mcdown to acquire with +13 dB Fast Gain and its much fast now and no longer makes that sound.

I also changed the caput command from 'caput -l' to 'caput -c -l' to see if the async 'wait for callback' feature will insure that the commands get sent. I witnessed the mcdown not actually writing all of its commands once or twice tonight. With the MC Boost left on its never going to lock.

mcdown has been committed to SVN. Please, if you have recently edit mcup and Autolock, commit them to the SVN or else I will delete them and do an svn up.

  12751   Wed Jan 25 01:27:45 2017 gautamUpdateIMCIMC feedforward checkup

This is probably just a confirmation of something we discussed a couple of weeks back, but I wanted to get more familiar with using the multi-coherence (using EricQs nice function from the pynoisesub package) as an indicator of how much feedforward noise cancellation can be achieved. In particular, in light of our newly improved WFS demod/whitening boards, I wanted to see if there was anything to be gained by adding the WFS to our current MCL feedforward topology.

I used a 1 hour data segment - the channels I looked at were the vertex seismometer (X,Y,Z) and the pitch and yaw signals of the two WFS, and the coherence of the uncorrelated part of these multiple witnesses with MCL. I tried a few combinations to see what is the theoretical best achievable subtraction:

  1. Vertex seismometer X and Y channels - in the plot, this is "Seis only"
  2. Seis + WFS 1 P & Y
  3. Seis + WFS 2 P & Y
  4. Seis + WFS 1 & 2 P
  5. Seis + WFS 1 & 2 Y

The attached plot suggests that there is negligible benefit from adding the WFS in any combination to the MCL feedforward, at least from the point of view of theoretical achievable subtraction

I also wanted to put up a plot of the current FF filter performance, for which I collected 1 hour of data tonight with the FF on. While the feedforward does improve the MCL spectrum, I expected better performance judging by previous entries in the elog, which suggest that the FIR implementation almost saturates the achievable lower bound. The performance seems to have degraded particularly around 3Hz, despite the multi-coherence being near unity at these frequencies. Perhaps it is time to retrain the Weiner filter? I will also look into installation of the accelerometers on the MC2 chamber, which we have been wanting to do for a while now...

  11794   Sat Nov 21 00:45:30 2015 KojiUpdateIOOIMC fix

Based on the observation of the PMC error signal, I started measuring the IMC OLTF. Immediately, it was found that the overall IMC loop gain was too low.
The UGF was ~40kHz, which was really marginal. It had been >100kHz when I have adjusted it about a year ago. (Next entry for the detail)

The first obvious thing was that the SMA cables around the IMC servo have visible degradation (Attatched photos).
I jiggled the signal cable from the demodulator Q_out to the MC servo. The openloop gain seemed fluctuating (increased) based on the cabling.
I decided to repair these cables by adding solder on the shield.

Even after the repair, the open loop TF didn't show any improvement. I checked the LO level and found that it was -16.7dBm.

I traced the problem down to the frequency generator unit (T1000461). The front panel of the unit indicates the output power for the 29.5MHz output is 13dBm,
while measurement showed it was 6~8dBm (fluctuating). The T1000461 document describes that there is only a wenzel oscillator inside. Does this mean the oscillatorwas degraded??? We need to open the box.

I was not sure what was the LO level. I naively assumed the input is 0dBm. Reducing the attenuation of the dial on the AM Stabilizer unit from 12dB attn to 0dB.
This made the LO level -3.3dBm.


Later at home, I thought this nominal LO level of 0dBm could have been wrong.

The demodulator circuit (D990511) has the amplifier ERA-5 (G=~20dB) at the input. Between the input and the ERA-5 there is a pattern for an attenuator.
Assuming we have no attenuator, the ERA-5 has to spit out 20dBm. That is too much for this chip. I need to pull out the box to see how much is the nominal LO for this box using an active probe.

This decrease/increase of the LO level affects the WFS demod too. According to D980233-B, the input stage has the comparator chip AD96687, which can handle differential voltage of 5.5V.
Therefore the effect is minimal.

  11796   Sun Nov 22 07:09:01 2015 ranaUpdateIOOIMC fix

On the demod board there is a 10 dB attenuator (AT1), which lowers the level to -10 dBm before the ERA-5. Then it should be 10 dBm before going to the rest of the parts. But I guess the ERA-5 chips which come later on in the circuit could be decaying like the ones in the PMC LO board.

Quote:

Later at home, I thought this nominal LO level of 0dBm could have been wrong.

  11797   Sun Nov 22 12:07:09 2015 KojiUpdateIOOIMC fix

TO DO

Done (Nov 23, 2015) - Check if the attenuator is still there in the input chain

Done (Nov 23, 2015) - Check if the actual LO levels at the 17dBm mixers are reasonable.

- Check if the actual LO levels for the LSC demods are OK too

  15897   Wed Mar 10 15:35:25 2021 Paco, AnchalSummaryIMCIMC free swinging experiment set to trigger at 5:00 am

A tmux session named "MCFreeSwingTest" will run on Rossa. This session is running script scripts/SUS/freeSwingMC.py (also attached) which will trigger at 5:00 am to impart 30000 counts kick to MC1, MC2, and MC3 after shutting PSL shutter and disabling the MC autolocker. It will let them freely swing for 1050 sec and will repeat 15 times to allow some averaging. In the end, it will undo all the changes it does and switches on autolocker on IMC. The script is set to restore any changes in case it fails at any point or a Ctrl-C is detected.

  15893   Wed Mar 10 11:46:22 2021 Paco, AnchalSummaryIMCIMC free swinging prep

[Paco, Anchal]

# Initial State
- MC is locked. The PRM monitor shows some oscillations.
- POP monitor shows light flashing once in a while.
- AS monitor shows one beam along with some other flashing beam around it.
- PRM Watchdog is tripped and shutdown. Everything else is normal except for overload on SRM OpLevs.
- Donatella got a mouse promotion

# Reenabling PRM watchdog:
- The custom reEnablePRMWatchdog.py has been deleted.
- Tried enabling the coil outputs manually and switching watchdog to Normal.
- Again saw large fluctuations like yesterday.
- Probably still the same issue of how current calculated actuations to the coils is in range -600 to -900 and gives and impulse to the optics when suddenly turned on.
- Waiting for PRM to damp down a little.
- Today we plan to change the position bias on PRM C1:SUS-PRM_POS_OFFSET instead of changing biases in pitch and yaw.
- Changing C1:SUS-PRM_POS_OFFSET from 0 to +/- 100 without enabling the coils, it seems upper and lower coils are anticorrelated with just changing the position. So going back to changing pitch.
- Changing C1:SUS-PRM_PIT_OFFSET from 0 -> 780. Switched on watchdog to normal.
- PRM damped down. OpLev errors are also within range.
- Enabled both OpLevs.

# Try locking Y-Arm
- IFO>CONFIGURE>YARM>Restore YARM (POY) using Donatella. See a bunch of python error messages in the call complaining about unable to find some python 2 files. Closed it with Ctrl-C after a stuck state.
- Tried running it on Pianosa, the script ran without error but Y-Arm didn't lock.

# Try locking X-Arm
- IFO>CONFIGURE>XARM>Restore XARM (POX) on Donatella. Again a bunch of OSError messages. Donatella is not configured properly to run scripts.
- Tried running it on Piasnosa, the script ran without error but X-Arm didn't lock.
- This might mean that both arms are misaligned or the BS/PRM is misaligned.
- Moving around C1:SUS-PRM_PIT_OFFSET and C1:SUS-PRM_YAW_OFFSET in order to see if the transmitted light is misalgined. Both arms are set to acquire lock if possible. No luck.

# Hypothesis: The Arm cavity is not aligned within itself (ITM-ETM)
- Will try to lock X-Arm with green light while tuning the ETMX. Hopefully the BS and ITM are aligned so that once we align ETMX to get a green lock, the IR will also lock from the other side.
- Running IFO>CONFIGURE>XARM>Restore XARM (ALS) on Pianosa. No lock, moving forward with tunning ETMX pitch and yaw offsets. Nothing changed. Brought back to same values.

[Rana joined, Anchal moved to Rossa from Pianosa]

# Moving on to IMC suspensions characterization:
- Closed the PSL shutter, to our suprise, the MC was still locked. We thought this would take away any light from IMC but it doesn't. Maybe the IFO Overview needs to show the schematic in a way where this doesn't happen: "No light from any laser entering the MC but it still is locked with a resonating field inside."
- Shutting IMCR shutter (hoping that would unlock the IMC), still nothing happend.
- Tried shutting PSL shutter from Rossa, nothing happened to MC lock still.
- Closed shutter IOO>Lock MC> Close PSL and this unlocked the IMC. Found out that this shutter channel is C1:PSL-PSL_ShutterRqst while the one from the sitemap>Shutter>PSL changes C1:AUX-PSL_ShutterRqst. Some clarification on these medm screens would be nice.
- Disabled the MC autolocked from IOO>Lock MC screen (C1:IOO-MC_LOCK_ENABLE).
- Checked the scripts/SUS/freeswing.py to understand how kick is delivered and optic is left to swing freely.
- Next, we are looking at the C1SUS_MC1 screen to understand what channels to read during data acquisition.
- In sensor matrix, we see INMON for each sensor which is probably raw counts data from the OSEMs. Rana mentioned that OSEM data comes out in units of microns. These are C1:SUS-MC1_ULSEN_OUTPUT (and so on for UR, LL, LR, SD).

- In prep for finishing, recovered Autolocker by first opening the PSL mechanical shutter, then re-enabling the Autolocker. The IMC lock didn't immediately recover, and we saw some fuzz on the PSL-FSS_FAST trace, so we closed the shutter again, waited a minute, then re-opened it and MC caught its lock.
 

  15895   Wed Mar 10 15:00:16 2021 gautamSummaryIMCIMC free swinging prep

Did you fix this issue? It is helpful to post a screenshot of the offending MEDM screen in addition to witticisms. The elog says "sitemap>Shutter>PSL" but I can't find PSL under the dropdown for shutters from Sitemap.

# Moving on to IMC suspensions characterization:
- Closed the PSL shutter, to our suprise, the MC was still locked. We thought this would take away any light from IMC but it doesn't. Maybe the IFO Overview needs to show the schematic in a way where this doesn't happen: "No light from any laser entering the MC but it still is locked with a resonating field inside."