  7437   Tue Sep 25 14:29:07 2012 JenneUpdateGeneralPOP in-vac optics aligned, POY beam coming out of vac

Quote:

  What was the reasoning / resolution of the POP forward/backward beam? Are we going to have the right beam for DRMI locking?

From Koji's email to me:

"With the backward beam you can see the returning beam even when the PRM is misaligned. That's the only difference. Once the PRM is aligned both beams have the same information."

So, we should be fine.

  7438   Tue Sep 25 17:21:01 2012 SteveUpdateVACpumpdown completed

 IFO P1= 1mTorr,   CC1 = 5e-4 Torr.  The IFO is ready for work. Be free to open the shutter and turn on HV

Jenne will double check the MC -REFL path for 1W power.  There is a manual block on the PSL table.

Atm1,  Pirani gauge at 9 h 10 min

Atm2, Pump configuration of pd#73 at 9h 35 min

 

Attachment 1: pd73.png
pd73.png
Attachment 2: pd73b.png
pd73b.png
  7439   Tue Sep 25 22:40:55 2012 JenneUpdateIOOIPANG ND filter installed

[Jenne, Evan Hall]

Both IPPOS and IPANG beams are (after turning on the input and output PZTs) hitting their QPDs.  However IPANG was saturating.  We went down to take a look, and we had ~2.8mW incident on the QPD.  There was an ND filter sitting unmounted, next to the diode, and an empty fork directly in front of the diode.  Since IPPOS also has an ND filter in front, we stuck this ND filter back in.  Now we are no longer saturating.

We're not hitting (yet) the center of these 2 PDs, but we're at least hitting the diodes, so it shouldn't be too hard to steer the input PZTs.

Whomever took away this ND filter without elogging it was BAD!!!  (Jamie, when we first found IPANG coming out of the vacuum during this vent, we moved some of the mirrors on the out-of-vac table in the IPANG path.  Was the ND filter removed at that time?  Or has it been out for much longer, and we never noticed because IPANG wasn't coming nicely out of the vacuum / was clipping on the oplev lens?)

  7440   Wed Sep 26 01:10:34 2012 JenneUpdateIOOPZT2 not working?!?! MC back to normal

[Jenne, Evan, Den]

MC REFL beam is back on the PD, and the mode cleaner locks.  It looks like we're a little high on the MC Refl camera, but the MC spots were measured, and don't look like they changed from Friday (or maybe Monday?), the last time they were measured. We took this to be acceptable MC alignment, and did not touch the PSL table's pointing.

The laser power reduction optics were removed, and placed out of the way on the PSL table (where do they belong?).  PSL-POS and PSL-ANG aren't quite perfectly centered, but a beam dump had been in the way of that path, so I don't know if they drifted bad, or if it was a sudden thing.  The beam is still hitting the QPDs though.  After removing the beam power reducing optics, we recentered the MC REFL beam on the REFL PD, still not touching any PSL alignment.  MC mirrors were aligned (Den did this work while I showed Evan around, so I don't know by how much), and MC Trans was maximized (really MC Refl was minimized, making sure that the unlocked MC Refl was the usual 4.something units on the EPICS readback.

We turned on the PZT high voltage supplies for the output steering PZTs and for the input steering PZTs.  We left the OMC locking PZT supplies off, since we're still not using the OMC.  Sadly, the beam coming out of the AS port looks clipped somewhere.  [SELF: attach the videocapture shot when you get to work tomorrow] We tried moving PZT2's sliders, but nothing happened!!! I can move BS and the ITMs to get the beam mostly unclipped, but I really need to be able to move the PZTs, or at least one of them.  IPPOS and IPANG beams are hitting the QPDs (although they're not centered perfectly), so the PZTs came back mostly to the same positions, but not exactly.  Evan and I sat next to the input steering PZT controllers in 1Y3, and moved the sliders around.  For most of the range, nothing changes on the LCD screen for either PZT2 pitch or yaw.  Yaw can make 2 small steps near the far negative side of the slider, but nothing happens for most of the slider.  Pitch really doesn't do anything for any part of the slider.  We ensured that the LED labeled "CL ON" was not illuminated, next to the button labeled "closed loop", for all 4 controllers (PZT1 and 2, pitch and yaw).  Sad!!   I don't know if the LCD screen on the front panel of the PZT controllers is a readback of signal supplied to the PZTs, or of the strain gauges.  I really hope it's the controller that's not working, rather than the PZTs themselves.  The PZTs were fine before we vented, and Koji and I did our centering of the PZT range check during the vent, so they were fine then.  What happened???  All PZT high voltage supplies were off during the pump-down.  I turned them off yesterday, and Evan and I turned them back on tonight around 9:30pm or 10pm.  What else could make them bad?

Without being able to move PZT2, just using BS and / or ITMs, I was unable to completely make the beam look nice on the AS camera.  I came close, but it still seems a little bit funny, and I had to move the suspended optics quite a bit to find that place.  This is not good. 

  7441   Wed Sep 26 11:03:04 2012 SteveUpdateVACpumpdown completed

 The new cold cathode gauge CC1 is in place. We were at atmosphere for 28 days ......more later

 

Attachment 1: 40mpd73md1.png
40mpd73md1.png
Attachment 2: 28datm.png
28datm.png
  7442   Wed Sep 26 16:59:30 2012 jamieUpdateIOOIPANG ND filter installed

Quote:

[Whomever took away this ND filter without elogging it was BAD!!!  (Jamie, when we first found IPANG coming out of the vacuum during this vent, we moved some of the mirrors on the out-of-vac table in the IPANG path.  Was the ND filter removed at that time?  Or has it been out for much longer, and we never noticed because IPANG wasn't coming nicely out of the vacuum / was clipping on the oplev lens?)

I do not remember removing anything from that setup.  We just moved some mirrors and lenses around

  7443   Wed Sep 26 17:09:15 2012 DenUpdateSUSTT

  [Koji, Steve, Den]

TT alignment is fine, yaw damping is satisfactory, pitch damping is slow. We might want to add magnets to the mirror and attach blades to the frame for pitch edge current damping.

We are moving towards electronics testing.

  7444   Wed Sep 26 23:55:14 2012 JenneUpdateIOOWFS centered

Since the MC spots are good, I put the beam back on WFS 1 and WFS 2.

Also, I changed the indicators on the LockMC screen to reflect the change in elog 7289, where we added another on/off switch for the WFS so that the ASS could be on, but the WFS off.  For the last month, the WFS could be disabled, but the MC screen's indicators would suggest that we were pushing very significantly on all 3 MC mirrors.  Now the MC screen reflects reality a little better.

I also uncommented the WFS lines in the mcup script.  Den had commented them out, but didn't elog about it!  C'mon Den, please elog stuff!!!!  (He confessed out loud the other day, but it still wasn't in the elog).

I'm leaving the WFS loops disabled (even though the MC autolocker tries to turn them on, I have them manually disabled using the extra on/off switch) since they're unstable.  I'm in the process of figuring out what's wrong.  So far, the WFS improve the MC alignment for a minute or two, and then they totally misalign the MC.  This is a work in progress.

  7445   Thu Sep 27 13:05:55 2012 Eric GustafsonUpdateLSC40 meter photodiode frequency response measurement system installation

Jenne, Mike and I installed all of the post holders we could today including: REFL11, REFL33, REFL55, AS55, MCRef, POX11 and POP55.  We did not install AS110, POY or REFL165 because there are interferences that will require moving stuff around. We also did not mount POP22 because it is a peely wally ThorLabs PD that will be replaced by a strong, straight and right thinking LIGO PD in the fullness of time.  We did move it out of the way however which is no more than it deserves. Next step this afternoon Mike and I will install all of the telescopes and launching hardware.  Then with the help of Steve we will begin routing the fibers.  The splitter module will be here by next Monday, the laser by the following Friday and then we will light up the fibers. 

  7446   Thu Sep 27 15:21:04 2012 SteveUpdateSUSTT height adaptor & PTICH damping

Quote:

  [Koji, Steve, Den]

TT alignment is fine, yaw damping is satisfactory, pitch damping is slow. We might want to add magnets to the mirror and attach blades to the frame for pitch edge current damping.

We are moving towards electronics testing.

Atm1,  TT 1.5" high adaptor base will be back from the shop in 10 days.

Atm2,  There is no PITCH damping, YAW edie current damping works well at 0.5 mm gap

Atm3,  Adjustable Al -disc that contains a small magnet is purely designed.

We have to come up with a solution to have damping in PITCH

Attachment 1: 09271201.PDF
09271201.PDF
Attachment 2: IMG_1656.JPG
IMG_1656.JPG
Attachment 3: IMG_1654.JPG
IMG_1654.JPG
  7447   Thu Sep 27 16:26:11 2012 SteveUpdateLSCRF fibre protection in cable trays

Quote:

Jenne, Mike and I installed all of the post holders we could today including: REFL11, REFL33, REFL55, AS55, MCRef, POX11 and POP55.  We did not install AS110, POY or REFL165 because there are interferences that will require moving stuff around. We also did not mount POP22 because it is a peely wally ThorLabs PD that will be replaced by a strong, straight and right thinking LIGO PD in the fullness of time.  We did move it out of the way however which is no more than it deserves. Next step this afternoon Mike and I will install all of the telescopes and launching hardware.  Then with the help of Steve we will begin routing the fibers.  The 1x16 splitter module will be here by next Monday, the laser by the following Friday and then we will light up the fibers. 

 I'm proposing split loom tubing that would run in the cable tray  to protect the fibers  inside of it.  This tubing diameter in the cable tray can be 1.5-2"  and out of the tray 0.75"

 

  7448   Thu Sep 27 17:00:41 2012 Eric GustafsonUpdateLSC40 meter photodiode frequency response measurement system installation

Mike and I installed all of the telescopes and launching hardware for REFL11, REFL33, REFL55, AS55, MCRef, POX11 and POP55. On Monday afternoon Steve will work with us on the fiber routing.  Steve is buying some protective covers for the fibers.

  7449   Thu Sep 27 19:03:06 2012 ManasaUpdate40m UpgradingNew ETMX layout - Version 1.0

 I am working towards redesigning the endtables. I've attached the first version of the layout. As per Steve's comment I've tried to leave a 2" empty space on all sides of the table. It still has to be updated with the whole 40m layout to be more precise about the pickoff and the ingoing beam directions.

Attachment 1: ETMX_OldNew.png
ETMX_OldNew.png
  7450   Thu Sep 27 20:05:57 2012 KojiUpdate40m UpgradingNew ETMX layout - Version 1.0

1. I wonder how the mode profiling/matching was considered in the new layout.

I can see the distances between the components and lenses are largely different from the old ones.

This is OK if you plan to go through a new mode matching solution with new lenses.
But it takes
 a certain amount of time.

Note that we don't care the distance after the last lens as the Rayleigh range there is supposed
to be long enough to allow this kind of change.

2. The huge frustration of the green alignment in the old setup was caused by the 3D beam steering
at the last two 2" mirrors. i.e. the beam elevation on the table does not match with the beam elevation of the cavity.
In order to avoid this, I suggest you to use three 45 deg 2" mirrors instead of two. In fact these mirrors are supposed to be used at 45deg incidence!

3. The incident green beam and the transmitted IR beam should share a same path as they
 share a same cavity mode.
This means that you should use a harmonic separator for the transmitted light pick-off.

4. Use the harmonic separator for the fiber path too. Get the mirror spec from Jamie.

5. Since the optical window on the chamber has a wedge angle, the beam paths are not straightforward.
The cavity beams can't be moved as they are constrained by the arm cavity.
Probably there is almost no freedom to move even for the oplev beams.
It would be safe just to follow the old positions and angles on the window.
Make sure the beam on the drawing is realistic. The angles of the oplev beams in the old setup look strange.

Is there a possibility to replace the optical window so that it has an AR for 532 and 1064 at least???

6. I wonder if the rejected beam by the Farady have a realistic angle or not. Check it with the old setup.
It is definitely better to have a steering mirrror and a lens before the refl PD.

7. The IR QPD and trans PD are intended to be used for the low and high power detection.
I forgot which is which. So check the range of them and think about the power distribution.

8. We should have separated CCDs for IR Trans and Green Refl.
We had a terrible ghost green beam on the IR trans CCD.
Thus, think about the amount of ghost reflection and consider filtering if necessary.

  7451   Fri Sep 28 07:49:30 2012 SteveUpdate40m UpgradingNew ETMX layout - Version 1.0

  Where is IP-ANG ? It is good practice to use two mirrors at launching and detecting the beam, so you can walk it - precisely adjust it.

The window can be replaced at  ~$1,500 ea. 10 weeks as optical quality BK7 with dual AR

  7452   Fri Sep 28 21:17:41 2012 KojiUpdateIOOMC WFS adjustment

MC WFS was fixed. Now it is running constantly with the autolocker.

Found a bug in the IOO screen: All of the 6 WFS signal indicators is liked to the same info (C1:IOO-MC1_PIT_OUTPUT).
Fix this, Jenne! Baaaaagghhhhh! 


What I did:

1. C1:IOO-MC_RFPD_DCMON indicator was saturating. "HOPR" of this entry was  set to 5 by running the following command:

ezcawrite C1:IOO-MC_RFPD_DCMON.HOPR 5

2. Scan MC2 spot position by using /opt/rtcds/caltech/c1/scripts/MC/moveMC2 scripts.
or the adjustment, C1:SUS-MC2_ASCPIT_EXC and C1:SUS-MC2_ASCYAW_EXC were excited with 300cnt at 12Hz and 10Hz, respectively.
The corresponding peaks (i.e. ANgle to length coupling) in C1:IOO-MC_F were monitored on DTT and adjusted so that the peaks are approximately nulled.

3. moveMC2 scripts are not perect to keep the maximum of the transmission. So, the alignment was adjusted with MC1 and MC3.

4. Repeated 2 and 3 until the alignment converges.

5. Once I got satisfied with the MC2 spot position, I went to the MC2 table and aligned the steering mirror before the QPD.

6. As these actions above moves the REFL beam, I went to the MC REFL path and adjusted the MC REFL PD position and the MC WFS spot positions.

7. Checked if the alignment is still good. The MC REFL is 0.50~0.51. Pretty good.

8. Run /opt/rtcds/caltech/c1/scripts/MC/WFS/WFS_FilterBank_offsets to register the current WFS offset etc.

9. At this point, MC WFS started working fine. I also confirmed the autolocker worked with this setting.

--------

Checked how the things are going in the morning. There were several unlocks. But the autolocker and WFS kept the cavity lcoked again.
Very good.

Some power fluctuation of ~1% is observed in the MC trans. I checked the PMC trans and found it is also fluctuating by 1% in a coherent way.
So I judge the WFS itself is fine. (See attached)

 

 

Attachment 1: MC_WFS_12h.png
MC_WFS_12h.png
  7453   Mon Oct 1 07:25:48 2012 SteveUpdateSUSMC2 & ETMX sus damping restored
  7454   Mon Oct 1 09:02:17 2012 SteveUpdateVACpumpdown reached vacuum normal

Quote:

 The new cold cathode gauge CC1 is in place. We were at atmosphere for 28 days ......more later

 

 cc1 = 2.3e-5 Torr at day 6 vacuum normal

Attachment 1: pd73md6.png
pd73md6.png
Attachment 2: pd73md6b.png
pd73md6b.png
  7455   Mon Oct 1 11:08:25 2012 JenneUpdateIOOMC WFS adjustment

Quote:

Found a bug in the IOO screen: All of the 6 WFS signal indicators is liked to the same info (C1:IOO-MC1_PIT_OUTPUT).
Fix this, Jenne! Baaaaagghhhhh! 

 My bad.  As it turns out, you can't copy and paste between MEDM instances.  It is now fixed.

  7456   Mon Oct 1 13:11:43 2012 JenneUpdateIOOPZT inspection elogs

I'll come back to the PZTs later, but I want to write down all the elogs I have found so far that look relevant.

http://nodus.ligo.caltech.edu:8080/40m/699

http://nodus.ligo.caltech.edu:8080/40m/5368

nodus.ligo.caltech.edu:8080/40m/5431

  7457   Mon Oct 1 16:05:01 2012 jamieUpdateCDSmx stream restart required on all front ends

For some reason the frame builder and mx stream processes on ALL front ends were down.  I restarted the frame builder and all the mx_stream processes and everything seems to be back to normal.  Unclear what caused this.  The CDS guys are aware of the issue with the mx_stream stability and are working on it.

  7458   Mon Oct 1 17:03:01 2012 steveUpdateSUSETMY oplev relayed

Quote:

Quote:

Quote:

 

 The typical sign of a dying gas laser is that it glows for a few minutes only. The power supplies are fine.

Two new  JDS - Uniphase 1103P lasers ( NT64-104 )  arriving on Monday, May 21

 Yesterday I swapped in new He/Ne laser with output power 3.5 mW  The return spot on qpd is large ~6mm in diameter and 20,500 counts

The spot size reduction require similar layout as ETMX oplev.

 The oplev path is relayed and the spot size on the qpd is reduced. I still have to clean up and replace "Miki Mouse" lens holder.

There was no IP-ANG coming out of the chamber at this time!

 I did relayed the oplev path with new  f 500 mm lens

Attachment 1: IMG_1673oplevETMY.jpg
IMG_1673oplevETMY.jpg
  7460   Tue Oct 2 07:45:49 2012 SteveUpdateSUS MC2 & PRM sus damping restored
  7461   Tue Oct 2 09:13:16 2012 SteveUpdateVACRGA scan at day 7

Quote:

Quote:

 The new cold cathode gauge CC1 is in place. We were at atmosphere for 28 days ......more later

 

 cc1 = 2.3e-5 Torr at day 6 vacuum normal

 

Attachment 1: pd73md7.png
pd73md7.png
  7463   Tue Oct 2 15:14:54 2012 jenne, jamieUpdateIOOPZT diagnosis

pzt2 mod signals matched slider vals for both pitch and yaw

  pzt2 yaw mon output = 6
  pzt2 pitch mon output = 11.3

From the PZT connector-converter board we determined the following pin-outs:

  X=Yaw:  red=1, white=14, black=3 
  Y=Pitch:  red=2, white=15, black=16

We believe that red is signal, white/black/shield are all ground.  We also believe (although this is from the PMC PZT) that the expected capacitance of the PZTs should be in the 100's of nF range.

Here are the readings from the two PZT dsub connectors:

  pin 1:14   PZT1 = ".003" on 2uF scale
             PZT2 = ".184"
   
  pin 2:15   PZT1 = ".002" on 2uF scale
             PZT2 = ".202"

So we think this means (given this crappy capacitance meter) that PZT2 is showing roughly 200nF, which sounds ok, but that PZT1 is indeed bad.

So next we investigate the PZT2 driver.

 
  7465   Tue Oct 2 16:32:43 2012 JenneUpdateIOOPZT diagnosis

[Koji, Jenne]

Jamie and I pulled the whole PZT driver for both PZT1 and PZT2. 

Koji and I found that each HV power supply (the left-most module) has 2 fuses.  Both HV supplies (PZT1 and PZT2) have one blown fuse.  The "T2L250A" measures low resistance for both HV supplies, but the "T250mAL250V" measures Open for both HV supplies.

I have ordered 10 pieces of each kind of fuse, Next Day shipping, from DigiKey.

  7466   Wed Oct 3 07:41:23 2012 SteveUpdateSUS MC2 sus damping restored

This is the third morning in a row that the MC2 was tripped.  Would you look at it Koji?

  7467   Wed Oct 3 11:43:33 2012 SteveUpdateSAFETYsafety training

Den Martynov received 40m specific safety training.

  7468   Wed Oct 3 15:37:04 2012 SteveUpdatePEMants on the PSL table

We observed one or two ants climbing over PMC optics without booties and safety glasses.

The floor was mopped with strong Bayer Home Pest Control solution in the Vertex area.

Do not work inside the 40m lab if you are sensitive to chemicals!

 

Attachment 1: IMG_1690.JPG
IMG_1690.JPG
  7470   Wed Oct 3 16:26:58 2012 ManasaUpdatePEMants on the PSL table

I spotted around 4 within 30 minutes working at the PSL table even after the deathly spray. They seem to be running down from the cables on the oscilloscope rack to the table and the optics.

  7472   Wed Oct 3 18:45:51 2012 jamieUpdateIOOwiki page for active IO tip-tilts

I made a wiki page for the active IO tip-tilts.  I should have made this a long time ago.

  7473   Wed Oct 3 23:24:05 2012 KojiUpdateSUS MC2 sus damping restored

Quote:

This is the third morning in a row that the MC2 was tripped.  Would you look at it Koji?

This may be cause by the impact of crazy WFS signal after the lock loss.
The auto locker is not fast enough to shut the WFS down before the mirrors are kicked.

Jenne and I discussed the issue and agreed that this can be solved by implementing
the same triggering algorithm as the LSC triggers.

Give us a bit more time to work on this.

  7475   Thu Oct 4 01:06:52 2012 JenneUpdateIOOPZT diagnosis

[Koji, Jenne]

We naively hoped that just replacing the fuses would fix the problem with the PZT HV drivers.  Alas, this was not the case. 

All of our investigations (other than visual inspections) today have been of the PZT2 module.  We have not applied any electricity to any PZT1 components/modules today.

After blowing a few more fuses (not good, we know, but we really didn't know what was going on at the time and were convinced that our changes between fuse installations should prevent fuse-blowing, including removing all modules except the HV driver), we found that the YAW driver for both PZT1 and PZT2 has severe discoloration on the PCB, and several resistors and other solder joints are damaged near some high voltage regulators. Pitch on PZT1 looks a tiny bit discolored, but doesn't look totally cooked like the 2 YAW modules do.  So, at least PZT1's Yaw was cooked before we started replacing fuses, since we haven't plugged it in yet today.

We then began some more methodical checks:

We bypassed the fuses by applying 10 Vpp = ~7.2 Vrms to the input side of the big transformer on the PZT2 HV driver board.  (This usually sees the 120 Vrms from the wall AC, so we were looking at things with a factor ~16 attenuation from what they normally see.)  We then measured things on the other side of the transformer, and made sure that they made some sense (one path for 5V stuff, one path for 15V stuff, one path for 180V stuff).  One of the rectifying diode bridges (the one for HV) didn't seem to be working, and didn't seem to have all of its pins connected, as if perhaps one or more diodes inside was destroyed.

When I went home for dinner, Koji continued looking at the low voltage supply capability of the PZT2 driver.  He removed the diode bridge from the HV path, and also removed the FET that lives on the output side of the HV driver board.  He was then able to energize the HV driver and the non-burnt pitch module.  So the +\-5 V and +\-15 V paths have been confirmed okay for PZT2's driver stuff.

What I will do tomorrow (when there is someone here to rescue me if I crispy-fry myself) is solder a wire to the now open pin of the backplane connector on the HV driver board, so that we can supply an external 180V to the pitch / yaw modules (although, obviously we won't be using the burnt yaw modules as-is).  Tomorrow I'll start by applying a nice small voltage, check that things still look okay, no shorts, and then I'll slowly increase the voltage until I get to the nominal 180V.

Since the low voltage stuff on the driver board is working, once we supply an external 180V (if successful), we should be able to re-install the PZT driver and drive PZT2. 

Since both Yaw modules that we have are burnt, I am proposing that we use the PZT2 HV board (which has been checked and modified this evening) with the 2 pitch modules.  Since we are not actively utilizing the strain gauge sensors, the fact that the calibrations on these modules are not exactly the same (rather, that PZT1's pitch is not the same as PZT2's yaw) should not matter at all.  This means that we will not be able to energize PZT1 at all, but that shouldn't be a problem.  Even when PZT 2 was working, PZT1 had very, very, very limited motion through the full range of applied voltage, so having no driver connected shouldn't have an impact.

 

  7476   Thu Oct 4 08:39:58 2012 SteveUpdateGeneralpower outage

There had to be a power outage. Laser and air condition turned back on. The vacuum is OK

Sorensen DC power supplies were tripped, so they were reset: at AUX OMC South 18V and 28V for RF PS and at 1X1 24V

 

Power Outage confirmed:

** Notification **

 

CALIFORNIA INSTITUTE OF TECHNOLOGY

                 FACILITIES MANAGEMENT

 

**PLEASE POST**

 

 

Building:         Campus

 

Date:             Thursday October 04,2012

 

This morning at 2:17 a.m. much of the City of Pasadena including our Campus experienced a electric power sag of short duration, approximately 1/10 of a second. The cause was a fault on one of Pasadena’s 17KV circuits. Some sensitive equipment have been impacted.

                 

Contact:          Mike Anchondo x-4999

 

Attachment 1: Oct4R2012.png
Oct4R2012.png
  7477   Thu Oct 4 14:04:21 2012 jamieUpdateCDSfront ends back up

All the front end machines are back up after the outage.  It looks like none of the front end machines came back up once power was restored, and they all needed to be powered manually.  One of the things I want to do in the next CDS upgrade is put all the front end computers in one rack, so we can control their power remotely.

c1sus was the only one that had a little trouble.  It's timing was for some reason not syncing with the frame builder.  Unclear why, but after restarting the models a couple of times things came back.

There's still a little red, but it mostly has to do with the fact that c1oaf is busted and not running (it actually crashes the machine when I tried to start it, so this needs to be fixed!).

  7478   Thu Oct 4 14:08:49 2012 jamieUpdateSUSsuspensions damped

All suspension damping has been restored.

  7483   Thu Oct 4 22:16:40 2012 DenUpdatePSLPMC is locked

PZT monitor is not lying to us, I've measured it with a voltmeter. But PMC SERVO is still interesting. If I break the loop after PMCERR signal monitor (C1:PSL-PMC_BLANK=0), I can change PZT voltage from 0V (C1:PSL-PMC_RAMP = -7.3) to 132V (C1:PSL-PMC_RAMP=-10.0). If I break the loop by enabling TEST1, PZT voltage goes up to 294 V, though voltage on the TEST1 and MIXER MON is 0.

  7484   Thu Oct 4 22:27:54 2012 KojiUpdateSUSHow about the slow machines?

One terrible concern of mine is that the slow machines were rebooted at the power interruption.
Based on the elog entries, I assume they have not been burtrestored...

If this is true, they may cause some weird behaviors of the PSL/IOO electronics.

 

  7485   Thu Oct 4 22:35:16 2012 DenUpdateSUSHow about the slow machines?

Quote:


Based on the elog entries, I assume they have not been burtrestored...

 Do you know how to burtrestore or restart slow machines?

Edit by Den: I did burtrestore of c1psl.snap from 2 days ago. Still slow machines behave not normal. For example, if I sweep C1:PSL-FSS_SLOWDC, SLOW monitor value does not change.

  7487   Fri Oct 5 00:29:34 2012 DenUpdatePEMreadout box power

Guralp readout box received +13.7 /0/ -15V instead +15V because of the broken fuse. Power provided by the source is normal.

Edit by Den: I've found a similar fuse on one of the tables and borrowed it. Guralp is not working again.

  7489   Fri Oct 5 04:34:31 2012 ranaUpdateSUSHow about the slow machines?

Quote:

Quote:


Based on the elog entries, I assume they have not been burtrestored...

 Do you know how to burtrestore or restart slow machines?

Edit by Den: I did burtrestore of c1psl.snap from 2 days ago. Still slow machines behave not normal. For example, if I sweep C1:PSL-FSS_SLOWDC, SLOW monitor value does not change.

 Problems with Slow Machines?

Read ELOG

  7490   Fri Oct 5 11:11:00 2012 DenUpdatePEMreadout box power

Quote:

Guralp readout box received +13.7 /0/ -15V instead +15V because of the broken fuse. Power provided by the source is normal.

Edit by Den: I've found a similar fuse on one of the tables and borrowed it. Guralp is not working again.

 I've meant Guralp is NOW working again 

  7491   Fri Oct 5 14:40:55 2012 SteveUpdateVACRGA scan after power outage

Quote:

All the front end machines are back up after the outage.  It looks like none of the front end machines came back up once power was restored, and they all needed to be powered manually.  One of the things I want to do in the next CDS upgrade is put all the front end computers in one rack, so we can control their power remotely.

c1sus was the only one that had a little trouble.  It's timing was for some reason not syncing with the frame builder.  Unclear why, but after restarting the models a couple of times things came back.

There's still a little red, but it mostly has to do with the fact that c1oaf is busted and not running (it actually crashes the machine when I tried to start it, so this needs to be fixed!).

 The RGA was not effected by the short power outage.

Attachment 1: afterpoweroutaged10.png
afterpoweroutaged10.png
Attachment 2: pd73m10d.png
pd73m10d.png
  7492   Fri Oct 5 14:53:29 2012 JenneUpdatePSLTTFSS board not fully seated!

[Den, Jenne]

Den noticed that the -15V LED on the TTFSS board was not illuminated.  A further symptom of the FSS being funny was that the PC RMS Drive was constantly high (3.6 ish) and the FAST Monitor was very high, often saturated. 

We took the TTFSS board out, and put an extender card in, and it looked like all of the correct power is being supplied to the board (the +\- 24V LEDs on the extender card were illuminated).  Just to check, we put the board back in, and this time both +\- 15V LEDs came on.  So it looks like the board is fine, it just wasn't seated in there all the way.

Now the readbacks on the FSS screen look good (PC RMS Drive is less than 1, FAST Mon is 5ish), the MC is locked, and I think we're back in business. 

 

  7493   Fri Oct 5 16:21:48 2012 SteveUpdateGeneralfreshmen visiters

40 plus freshmen visited the 40m today

Attachment 1: IMG_1693.JPG
IMG_1693.JPG
Attachment 2: IMG_1694.JPG
IMG_1694.JPG
  7495   Sun Oct 7 12:11:00 2012 AidanUpdateComputersRebooted cymac0

I rebooted cymac0 a couple of times. When I first got here it was just frozen. I rebooted it and then ran a model (x1ios). The machine froze the second time I ran ./killx1ios. I've rebooted it again.

  7496   Sun Oct 7 15:05:42 2012 DenUpdateSUS MC2 sus damping restored

Quote:

This is the third morning in a row that the MC2 was tripped.

 MC2 was tripped again. I think the answer is that watchdog's critical value was too small C1:SUS-MC2_PD_MAX_VAR = 10, so seismic could trip MC2. I've changed the value to 100.

mc2.png

  7497   Sun Oct 7 23:39:10 2012 DenUpdateModern Controlstate estimation

I've applied online state estimation technique using Kalman filter to LQG controller. It helps to estimate states that we do not measure. I've considered MC2 local damping, we measure position and want to estimate velocity that we need for control. We can either differentiate the signal or apply state estimation to avoid huge noise injection at high frequencies. In state estimation we need to know noise covariance, I've assumed that LID sensor noise is 0.1 nm. Though covariance can be calculated better.

In the time-domain figure C1:SUS-MC2_SUSPOS_IN1 = MC2 postion, C1:SUS-MC2_SUSPOS_OUT = MC2 velocity obtained by differentiation, 2 other channels are estimations of position and velocity.

Attachment 1: est_time.png
est_time.png
Attachment 2: est_freq.pdf
est_freq.pdf
  7498   Mon Oct 8 09:45:28 2012 jamieUpdateComputersRebooted cymac0

Quote:

I rebooted cymac0 a couple of times. When I first got here it was just frozen. I rebooted it and then ran a model (x1ios). The machine froze the second time I ran ./killx1ios. I've rebooted it again.

For context, there's a is stand-alone cymac test system running at the 40m.  It's not hooked up to anything, except for just being on the martian network (it's not currently mounting any 40m CDS filesystems, for instance).  The machine is temporarily between the 1Y4 and 1Y5 racks.

  7499   Mon Oct 8 09:51:30 2012 ranaUpdateModern Controlstate estimation

 

 I guess that the estimated state has the same low pass filter, effectively, that we use to low pass the feedback signal in SUSPOS. I wonder if there is an advantage to the state estimation or not. Doesn't the algorithm also need to know about the expected seismic noise transmission from the ground to the optic?