6342
|
Wed Feb 29 20:27:00 2012 |
Jenne | Update | Green Locking | X green beat - found it! |
Found it!
The actual temperature of the Xend laser is 0.02 C higher than anticipated based on the formula in elog 3759. Both the PSL and the Xend laser are at their nominal diode currents (2.100 A for the PSL, 2.003 A for Xend), so the curves should be used as they are. The PSL temp (when the slow servo offset is ~0) is 31.71 C. Using curve 2 from elog3759, the Xend laser should be 37.78, which I found was +10 counts on the Xgreen slow servo offset.
Right now the Xend laser is at 37.80 C, and the beat is around 30 MHz. This is +80 counts on the Xgreen slow servo. +60 counts gave me ~80 MHz. When (a few minutes ago) the MC unlocked and relocked, it came back to a slightly different place, so the temp of the Xend laser had to go up a few 10's of counts to get the same beat freq. Right now the PSL slow servo offset is 0.076 V.
The HP8591E is set with ResBW=100kHz, Ref Level= -39dBm (so I'm not attenuating my input signal!). The largest peak I see for the beatnote is -66dBm. The nose floor around the peak is -83dBm. Trace (trace button!) A is set to MaxHoldA, and Trace B is set to ClearWriteB, so B is giving me the actual current spectrum, while A is remembering the peak value measured, so it's easier to see if I went past the peak, and just didn't see it on the analyzer.
Also, I went back and realigned the beams earlier, to ensure that there was good overlap both near the BS which combines the PSLgreen and Xgreen beams, and at the PD. The overlap I had been looking at was okay, but not stellar. Now it's way better, which made the peak easier to see. Also, also, the waveplate after the doubling oven on the PSL table is still rotated so that I get max power on the Xgreen side of things, and not much at all on the Ygreen side. I'll need to rebalance the powers, probably after we make sure we are seeing the beatnote with the BeatBox.
Next Steps:
Lay a cable from the BBPD to the BeatBox in 1X2, make the BeatBox do its thing.
Use the dichroic locking to do a sweep of the Xarm. |
6343
|
Thu Mar 1 00:05:23 2012 |
Den | Update | PEM | seis box noise |
I've moved GUR1 seismometer from MC2 to the working tables in order not to disturb the MC while working with the seismometer box. The new place for the GUR1 for a few days is near the printer, cables and blue boxes. I've cleaned all mess and wires from the floor, so that seismometer now looks like that

I've connected 2 inputs of the N/S 1 circuit of the seismometer box with a 50 Ohm resistor and measured the noise at the output. The comparison with the seismic signal is

The noise increased at 0.5 Hz and is pretty big. This might explain the loose of coherence at low frequencies. |
6344
|
Thu Mar 1 09:26:50 2012 |
steve | Update | SUS | SOS baffle plates are ready |
Green welding glass 7" x 9" shade #14 with 40 mm hole and mounting fixtures are ready to reduce scatter light on SOS
PEEK 450CA shims and U-shaped clips will keep these plates damped.
|
Attachment 1: 03011201.PDF
|
|
6345
|
Thu Mar 1 21:48:34 2012 |
Den | Update | PEM | seis box noise |
Quote: |
The noise increased at 0.5 Hz and is pretty big. This might explain the loose of coherence at low frequencies.
|
This is because spectrum analyzer did not plot the real noise spectrum at the first few points at low frequencies. I've remeasured the noise at 1mHz - 3Hz at "output -" (TP9) and compared it to the seismometer signal

The noise seems to be much less then the signal. I've measured the noise several times and once I got a huge amount of noise

I made another measurement in some time and got the low noise again. A circuit might have a bad contact somewhere.
The plan is to change AD620 adjustable resistor (R2) from 5.49kOhm to 500Ohm to increase the gain from 20 up to 200. |
6346
|
Fri Mar 2 11:05:28 2012 |
Den | Update | PEM | seis box gain |
I've replaced R2 resistor that adjusts the gain of the AD620 amplifier. Previous value 5491Ohm, new value 464Ohm, so the gain should increase up to ~200-250. Only at the N/S 1 circuit!
LISO simulation of the circuit transfer function and noise are


LISO predicts gain ~45-46 dB = 200 and noise at the level of 10uV at 1Hz. The transfer function and noise measured are


The noise measured is 5 times higher then predicted by LISO. Though I described AD620 as an ordinary amplifier with 49.9kOhm resistor connecting output and inverted input. I specified the noise spectrum 10 nV and 1/f corner frequency 30 Hz. In the AD620 datasheet noise spectrum is 10 - 100 nV depending on the gain. However, the gain is 200 and noise spectrum should be 10 nV. May be in reality it is not the case. It also possible that the noise model used by LISO is not valid for AD620 as it is not an ordinary operational amplifier. |
6347
|
Fri Mar 2 16:05:52 2012 |
Den | Update | SAFETY | laser safety |
Today I've attended the laser safety seminar. |
6349
|
Fri Mar 2 18:55:06 2012 |
Den | Update | PEM | seis box |
I've put the seismometer box back to the 1x1, Guralp is back under MC2. When the seismometer is not plugged in, the noise is

Now, I'm going to collect some data from GUR 1 and MC_F and see if the problem with adaptive filter (increasing errror while decreasing mu) will be gone.
|
6350
|
Mon Mar 5 03:22:54 2012 |
kiwamu | Update | PSL | PMC realigned |
I realigned the steering mirrors for the PMC. The trans value went up from 0.79 to 0.83.
The misalignment was largely in the pitch direction. |
6351
|
Mon Mar 5 03:50:49 2012 |
kiwamu | Update | IOO | PZT1 PITCH railing |
PZT1 started railing in the pitch direction and because of this TRY doesn't go more than 0.7. I will leave it as it is for tonight.
Tomorrow I will shift the alignment of the MC to make the PZT1 happier.
Quote from #6300 |
PZT1, the one with Koji's custom mid-HV driver (#5447), is getting degraded.
|
|
6352
|
Mon Mar 5 05:39:36 2012 |
kiwamu | Update | LSC | REFL OSA installed |
The OSA for the REFL beam is now fully functional.
The only thing we need is a long BNC cable going from the AP table to the control room so that we can monitor the OSA signal with an oscilloscope.
The attached picture shows how they look like on the AP table. Both AS and REFL OSAs are sitting on the corner region.
Quote from #6340 |
I placed the OSA (Optical Spectrum Analyzer) on the AP table and this OSA will monitor the REFL beam.
Tomorrow I will do fine alignment of the OSA.
|
|
Attachment 1: APtable.png
|
|
6354
|
Mon Mar 5 13:11:06 2012 |
kiwamu | Update | LSC | installed a long BNC cable for REFL OSA |
A long BNC cable was installed and now the REFL OSA signal is happily shown on an oscilloscope in the control room.
Quote from #6352 |
The OSA for the REFL beam is now fully functional.
The only thing we need is a long BNC cable going from the AP table to the control room so that we can monitor the OSA signal with an oscilloscope.
|
|
6355
|
Mon Mar 5 14:10:35 2012 |
kiwamu | Update | LSC | power budget on the AP table |
I checked the laser powers on the AP table and confirmed that their powers are low enough at all the REFL photo diodes.
When the HWP( which is for attenuating the laser power with a PBS) is at 282.9 deg all of the REFL diodes receives about 5 mW.
This will be the nominal condition.
If the HWP is rotated to a point in which the maximum laser power goes through, the diodes get about 10 mW, which is still below the power rate of 18 mW (#6339).
I used the Coherent power meter for all the measurements.
The table below summarizes the laser powers on the REFL diodes and the OSA. Also the same values were noted on the attached picture.
|
nominal power [mW]
(when HWP is at 282.9 deg)
|
expected max power [mW]
(when HWP is at a point where the max power goes through)
|
REFL11 |
5.5 |
10 |
REFL33 |
4.5 |
10 |
REFL55 |
5.3 |
10 |
REFL165 |
4.8 |
10 |
REFL OSA |
0.7 |
0.7 |
A note:
I found that the OSA for the REFL beam was receiving a unnecessary bright laser.
So I put an ND1 attenuator stacked on the existing ND2 attenuator. The laser power entering in the OSA is currently at 0.7 mW. |
Attachment 1: power_budget.png
|
|
6356
|
Mon Mar 5 15:15:15 2012 |
Den | Update | PEM | RCG |
[Alex / Den]
I've encountered a problem that C1:PEM-SEIS_GUR1_X_IN1 is saved in the int format. It turned out that inside the code the signal is also in the int format. It is not just a saving error. It should not be so as ADC works at 64k and the model runs at 2k.
Why? There is a bug somewhere in the generation of the code. c1pem.c looks suspicious to Alex because there is a mismatch in the ADC numbers with the simulink model.
Solution: upgrade to 2.4 version - most probably it was fixed there. If not, Alex will handle this problem. |
6357
|
Mon Mar 5 17:07:58 2012 |
kiwamu | Update | IOO | realigned MC |
I have slightly shifted the MC beam pointing to relax the PZT1 PITCH. As a result the TRY value went to 0.97 in a first lock trial.
However another issue arose:
The polarity for controlling the PZT1 PITCH seems to have flipped for some reason.
Since it is still sort of controllable, I am leaving it as it is.
If I remember correctly, sliding the PZT1 pitch value to the positive side brought the beam spot upward in the AS CCD. But now it moves in the opposite way.
Also the ASS feedback looks tending to push the PZT1 pitch to the wrong direction.
I am not 100 % sure if the polarity really flipped, but this is my current conclusion.
(MC pointing)
- Locked the Y arm and aligned ITMY and ETMY with the ASS servos such that the beam spot on each test mass is well centered on the test mass.
- With this process the eigen axis of the Y arm cavity is well prepared.
- Checked the beam positions of the prompt reflection light and cavity leakage field in the AS CCD.
- It looked the incident beam needed to go upward in the CCD view.
- Offloaded the MC WFS feedback values to the MC suspension DC biases in a manual way.
- Disabled the MC WFS servos. The MC transmitted light didn't become worse, which means the suspensions were well aligned to the input beam
- Changed the DC bias in the MC2 PITCH, to bring the beam spot upward. I changed the DC bias by ~ 0.1 or in the EPICS counts.
- Aligned the zig-zag steering mirrors on the PSL table to match the incident beam to the new MC eigen beam axis.
- The transmitted DC light and reflected DC values went back to 27000 counts and 0.58 counts respectively without the WFS servos.
- Re-engaged the WFS servos.
Quote from #6351 |
PZT1 started railing in the pitch direction and because of this TRY doesn't go more than 0.7. I will leave it as it is for tonight.
Tomorrow I will shift the alignment of the MC to make the PZT1 happier.
|
|
6358
|
Mon Mar 5 18:12:00 2012 |
Keiko | Update | LSC | RAM simulation update |
I wrote an RAM simulation script ... it calculates the LSC signal offset and the operation point offset depending on the RAM modulation index.
Configuration : RAM is added on optC1, by the additional Mach-Zehnder ifo before the PRM.
 
Both are for PRCL sweep result. Note that REFL33I is always almost zero. Next step: Check the LSC matrix with matrix at the offset operation point. |
6359
|
Mon Mar 5 20:31:33 2012 |
Koji | Update | LSC | 22/110MHz path for POP |
This a kind of self record...
We need an RF setup at POP to extract 22 and 110 MHz components separately.
I am planning to work on this in the daytime on Tuesday. |
6361
|
Tue Mar 6 00:13:20 2012 |
keiko | Update | LSC | ASI signal offset |
 
AS55Q and AS55I signals. AS55Q is around zero while AS55I has a large offset which is about the signal amplitude. It is likely because of the RAM?
keiko, kiwamu
|
6362
|
Tue Mar 6 01:35:03 2012 |
kiwamu | Update | LSC | MICH characterization |
[Keiko / Kiwamu]
Update on the MICH characterization:
- The OSAs weren't so great because the 11 MHz sidebands were covered by the carrier's tail
- It seemed that the frequency resolution depended on the mode matching. We will try improving the mode matching tomorrow.
- The noise budget looked very bad
- There were huge peaks at 1 Hz, 3 Hz, 16.5 Hz and 23 Hz. Something is crazy in the vertex suspensions.
- Keiko will post the calibrated noise budget.
- The MICH response at AS55Q was measured and we will calibrate it into watts / meter.
|
6363
|
Tue Mar 6 15:22:02 2012 |
Keiko | Update | LSC | RAM simulation update |
Quote: |
I wrote an RAM simulation script ... it calculates the LSC signal offset and the operation point offset depending on the RAM modulation index.
Configuration : RAM is added on optC1, by the additional Mach-Zehnder ifo before the PRM.
 
Both are for PRCL sweep result. Note that REFL33I is always almost zero. Next step: Check the LSC matrix with matrix at the offset operation point.
|
On the right figure, you see the non-zero operation points even when RAM mod index = 0. Apparently they come from non-zero loss of the model. (Each mirror of 50ppm loss was assumed). |
6365
|
Tue Mar 6 16:17:36 2012 |
Jenne | Update | SUS | SUS matrix diagonalization status |
Has default inmat:
PRM, ITMX
Has fancy inmat:
BS, ITMY, SRM (but side is non-fancy), ETMX, ETMY, MC1, MC2, MC3
So it's likely that the MICH problems (giganto 1Hz peak) Keiko and Kiwamu were seeing last night had to do with ITMX having the non-optimized input matrix. I'll try to figure out where the data from the last freeswing test is, and put in a fancy diagonalized matrix. |
6366
|
Tue Mar 6 22:23:04 2012 |
Koji | Update | LSC | 22/110MHz path for POP |
As par Kiwamu's request, RF filters for POP22 and POP110 were installed. They are not really nice. We need to replace it with more fancy electronics.
More to come later.
|
6367
|
Tue Mar 6 23:14:38 2012 |
rana | Update | SUS | Optical Lever had bad filters |
We found that that bounce (16.1 Hz) and roll (23.5 Hz) modes on the ITMX were much higher than on the ITMY. After some checking, it seems that the bandstop filters for the
SUSPOS, SUSPIT, SUSYAW, and SUSSIDE loops are set to the correct frequencies. However, the OLPIT and OLYAW had not been set correctly. I have copied the SUS filters into the OL filterbanks and reloaded all the filter banks. Attached are the comparison of old, bad, OL with the SUS ones.
The same cockamamie situation was there for the BS & ITMY as well. Although we still don't have the roll mode frequencies listed in the mechanical resonances wiki, I have guessed that the ITMY roll frequency is the same as the ITMX, since they have nearly the same bounce frequency. OL filters for the BS & ITMY are now at the right frequency (probably). Keiko is on top of fixing things for the other optics.
I think this whole notching adventure was in Leo's hands several months ago, but WE forgot to point him at the OLs in addition to the SUS. I blame Kiwamu 50% for not supervising him and Koji by 45% for not supervising Kiwamu. The other 5% goes to someone else. You know who you are. |
Attachment 1: Untitled.png
|
|
6368
|
Tue Mar 6 23:37:31 2012 |
keiko | Update | SUS | MICH noise budget - SUS check |
Here are the OSEM spectrum of MICH suspensions (BS, IX, IY). Bounce and Roll modes are shown on 16 and 23 Hz. The filters for them has been checked.



keiko, kiwamu, Rana |
Attachment 1: Mar6sus1
|
Attachment 2: Mar6sus2
|
Attachment 3: Mar6sus3
|
6369
|
Wed Mar 7 04:08:48 2012 |
kiwamu | Update | SUS | BS SIDE gain was too low |
The BS SIDE damping gain seemed too low. The gain had been 5 while the rest of the suspensions had gains of 90-500.
I increased the gain and set it to be 80.
I did the "Q of 5" test by kicking the BS SIDE motion to find the right gain value.
However there was a big cross coupling, which was most likely a coupling from the SIDE actuator to the POS motion.
Due to the cross coupling, the Q of 5 test didn't really show a nice ring down time series. I just put a gain of 80 to let the Q value sort of 5.
I think we should diagonalize the out matrices for all the suspensions at some point. |
6370
|
Wed Mar 7 11:20:23 2012 |
steve | Update | SUS | oplev qpd offsets zeroed |
All oplev qpd quadrons were zeroed by offset in blocked dark condition. |
6371
|
Wed Mar 7 11:44:29 2012 |
Jenne | Update | Green Locking | Xgreen beatnote cable made, laid |
The Xgreen PD now has a cable going over to the beatbox. Once beatbox characterization is done I can re-find the beat, and we can do some stuff with the beatbox. |
6372
|
Wed Mar 7 13:30:17 2012 |
Jim | Update | PEM | added TPs and JIMS channels to PEM front-end model |
[Jim Ryan]
The PEM model has been modified now to include a block called 'JIMS' for the JIMS(Joint Information Management System) channel processing. Additionally I added test points inside the BLRMS blocks that are there. These test points are connected to the output of the sqrt function for each band. I needed this for debugging purposes and it was something Jenny had requested.
The outputs are taken out of the RMS block and muxed, then demuxed just outside the JIMS block. I was unable to get the model to work properly with the muxed channel traveling up or down levels for this. Inside the JIMS block the information goes into blocks for the corresponding seismometer channel.
For each seismometer channel the five bands are processed by comparing to a threshold value to give a boolean with 1 being good (BLRMS below threshold) and 0 being bad (BLRMS above threshold). The boolean streams are then split into a persistent stream and a non-persistent stream. The persistent stream is processed by a new library block that I created (called persist) which holds the value at 0 for a number of time steps equal to an EPICS variable setting from the time the boolean first drops to zero. The persist allows excursions shorter than the timestep of a downsampled timeseries to be seen reliably.
The EPICS variables for the thresholds are of the form (in order of increasing frequency):
C1:PEM-JIMS_GUR1X_THRES1
C1:PEM-JIMS_GUR1X_THRES2
etc.
The EPICS variables for the persist step size are of the form:
C1:PEM-JIMS_GUR1X_PERSIST
C1:PEM-JIMS_GUR1Y_PERSIST
etc.
I have set all of the persist values to 2048 (1 sec.) for now. The threshold values are currently 200,140,300,485,340 for the GUR1X bands and 170,105,185,440,430 for the GUR1Y bands.
The values were set using ezcawrite. There is no MEDM screen for this yet.
PEM model was restarted at approx. 11:30 Mar. 7 2012 PST.
|
6374
|
Wed Mar 7 15:56:36 2012 |
Koji | Update | LSC | 22/110MHz path for POP |
The RF separator installed comprises of the Minicircuits filters cascaded as in the figure below.
This has one input and 4 output ports for 11, 22, 30-60, and 110MHz signal.
As seen in this entry #6167, we have 22 and 110MHz signals together with 11, 44, 66MHz signals.
They may be demodulated via a harmonic characteristic of the mixers. (Remeber mixers are not multipliers.)

Of course the big concern is the impedance matching for those signals as usual.
The 2nd attachment shows measured impedance of the circuits with all of the ports terminated.
From the complex impedance, we can calculate the reflection coefficient. The 44 and 110MHz
components look correctly matched while the others seems largely reflected.
This certainly is not a nice situation, as the reflection can make the amplifier next to the PD unhappy
(although the reflected power is tiny in our case).
In our case more eminent problem is that the amplitude of the 22MHz signal can vary depending on the cable length by
factor of 10 in amplitude. (c.f. VSWR on the 2nd attachment.)
The transmission to each port was measured. The separation of the signals looks good. But the attenuation of the
targetted signals (i.e. insertion losses) are qulitatively consistent with the impedance. Again these losses are depend
on the cable length.
|
Attachment 2: impedance.pdf
|
|
Attachment 3: transmission.pdf
|
|
6375
|
Wed Mar 7 16:32:09 2012 |
keiko | Update | LSC | OSA |
I swap an OSA at PSL and OSA at REFL. It was because the PSL-OSA had a better resolution, so we place this better one at REFL. The ND filter (ND3) which was on the way to REFL OSA was replaced by two BSs, because it was producing dirty multiple spots after transmitting. |
6376
|
Wed Mar 7 17:39:40 2012 |
keiko | Update | LSC | MICH noise budget on 5 Mar |
This is the calibrated MICH noise budget on Mar 5. There was a sharp peak at 1Hz and a blob on 3 Hz. The demod phase was adjusted for AS55Q.

|
Attachment 1: Mar5-MICHbudget.png
|
|
6378
|
Wed Mar 7 19:10:06 2012 |
kiwamu | Update | LSC | REFL OSA : how the signal look like |
Just a quick report on the REFL OSA.
The attached plot below shows the raw signal from the REFL OSA which Keiko installed in this afternoon.
When the data was taken the beam on the REFL OSA was a direct reflection from PRM with the rest of the suspended mirrors misaligned.
One of the upper and lower 11 MHz sidebands is resolved (it is shown at 0.12 sec in the plot) while the other one is still covered by the carrier tail.
The 55 MHz upper and lower sidebands are well resolved (they are at 0.06 and 0.2 sec in the plot).
One of the oscilloscopes monitoring the OSA signals in the control room has a USB interface so that we can record the data into a USB flash memory and plot it like this.

Quote from #6375 |
I swap an OSA at PSL and OSA at REFL. It was because the PSL-OSA had a better resolution, so we place this better one at REFL. The ND filter (ND3) which was on the way to REFL OSA was replaced by two BSs, because it was producing dirty multiple spots after transmitting.
|
|
6379
|
Wed Mar 7 20:06:23 2012 |
Koji | Update | LSC | REFL OSA : how the signal look like |
I'm puzzled why the 11MHz peak can be such high considering 1.7~2 times smaller the modulation depth. |
6380
|
Wed Mar 7 20:53:13 2012 |
keiko | Update | LSC | MICH noise budget on 5 Mar |

This is the MICH noise budget on 6th March. 1Hz peak got a bit better as the BS sus control gain was increased.
|
6381
|
Wed Mar 7 21:13:30 2012 |
rana | Update | DAQ | NDS2 |
I noticed that NDS2 was not running on mafalda as it should be. Instead, there were a couple of zombie MEDMs using up 99% of the CPU. I killed the zombies and have run the 'build channel list' script. When it finished, I tried to restart the nds server, but got the following error in the log file. Email has been dispatched to JZ.
mafalda:logs>less nds2-mafalda-201203072111.log
Configuring from file: nds2.conf Allow list: ALL terminate called after throwing an instance of 'std::runtime_error' what(): Insufficient arguments
|
6382
|
Wed Mar 7 22:04:05 2012 |
kiwamu | Update | LSC | REFL OSA : how the signal look like |
I was also wondering about the same thing, comparing with what Mirko obtained before with the same OSA ( #5519).
Quote from #6379 |
I'm puzzled why the 11MHz peak can be such high considering 1.7~2 times smaller the modulation depth.
|
|
6384
|
Wed Mar 7 23:29:28 2012 |
keiko | Update | LSC | REFL OSA observation |
kiwamu, keiko

We measure the REFL OSA spectrum when (1) direct reflection from the PRM (2) CR lock at PRC (3) SB lock at PRC. When CR lock, both SBs are reflected from the PRC and when SB lock (ref line), some SB is sucked by PRM and looked lower than the other two lines.
|
6385
|
Thu Mar 8 00:57:48 2012 |
keiko | Update | LSC | MICH noise budget on Mar 5, Mar 6, and old |
Here is the recent two noise budgets of MICH, with the old measurement by Jenne. The most latest Mar 6 data is quite close to the old data, even better around 20-30 Hz. Probably some scattering source was improved?

|
6386
|
Thu Mar 8 04:13:12 2012 |
kiwamu | Update | LSC | update on the locking activity |
[Keiko / Kiwamu]
Some updates on the locking activity:
- Started summarizing the data of the Michelson lock in a wiki page:
- Gradually moving on to the PRMI lock
- The lock stays for reasonably a long time (~20 min or more)
- POP22/110 demod signals seemed just ADC noise.
- A first noise budget is in process
- The glitches make the noise level worse above 40 Hz or so in both the MICH and PRCL budgets.
- Sensing matrix will be measured tomorrow
- The data will be also summarized in a wiki page
|
6388
|
Thu Mar 8 23:37:03 2012 |
kiwamu | Update | IOO | dither Y arm dither script |
I disabled the feedback to the PZT1 PITCH in the Y arm dithering scripts so that it won't push the beam away from the good point.
Currently one has to do a manual alignment only for the PZT PITCH but the rest of DOFs are still able to be automatically aligned with the script.
Quote from #6357 |
The polarity for controlling the PZT1 PITCH seems to have flipped for some reason.
|
|
6390
|
Fri Mar 9 10:44:57 2012 |
steve | Update | RF System | OSA |
Optical spectrum analyzers like the Attachment made by Coherent , Meles Griot- CVI and Spectral Product are all discontinued.
The 40m have Coherent models C240 analyzer with controller C251 Their Finesse measured in 2004: sn205408 F302, sn205409 F396,
Jenne borrowed Jan's Meles Griot model 13SAE006, Peter King has the same model. FSR 300 MHZ, finnees 200 minimum |
Attachment 1: OSA.pdf
|
|
6393
|
Fri Mar 9 13:34:13 2012 |
keiko | Update | LSC | update on the locking activity |
We tried to measure the sensing matrix for MICH and PRCL last night. They look too much mixed as we expect... the matrix may be posted later. We suspect the IX and IY of the MICH excitation is not balanced very well, although Kiwamu adjusted that about two weeks ago, and it is mixing the dof. We'll try to balance it again, ans see the matrix.
Keiko, Kiwamu
Quote: |
[Keiko / Kiwamu]
Some updates on the locking activity:
- Started summarizing the data of the Michelson lock in a wiki page:
- Gradually moving on to the PRMI lock
- The lock stays for reasonably a long time (~20 min or more)
- POP22/110 demod signals seemed just ADC noise.
- A first noise budget is in process
- The glitches make the noise level worse above 40 Hz or so in both the MICH and PRCL budgets.
- Sensing matrix will be measured tomorrow
- The data will be also summarized in a wiki page
|
|
6395
|
Fri Mar 9 16:00:46 2012 |
steve | Update | Green Locking | laser emergency shut down switch replaced at the south end |
Over-sized local laser emergency switch was held by large C clamp at the south end. This was replaced by a smaller one and it is mounted with magnets.
The Innolight laser was turned off, while the interlock was wired. |
6397
|
Fri Mar 9 20:44:24 2012 |
Jim Lough | Update | CDS | DAQ restart with new ini file |
DAQ reload/restart was performed at about 1315 PST today. The previous ini file was backed up as c1pem20120309.ini in the /chans/daq/working_backups/ directory.
I set the following to record:
The two JIMS channels at 2048:
[C1:PEM-JIMS_CH1_DQ] Persistent version of JIMS channel. When bit drops to zero indicating something bad (BLRMS threshold exceeded) happens the bit stays at zero for >= the value of the persist EPICS variable.
[C1:PEM-JIMS_CH2_DQ] Non-persistent version of JIMS channel.
And all of the BLRMS channels at 256:
Names are of the form:
[C1:PEM-RMS_ACC1_F0p1_0p3_DQ]
[C1:PEM-RMS_ACC1_F0p3_1_DQ]
On monday I intend to look at the weekend seismic data to establish thresholds on the JIMS channels.
256 was the lowest rate possible according to the RCG manual. The JIMS channels are recorded at 2048 because I couldn't figure out how to disable the decimation filter. I will look into this further. |
6398
|
Sat Mar 10 02:00:03 2012 |
keiko | Update | LSC | update on the locking activity |
ITMX and ITMY balance for the MICH excitation (lockin) is adjusted again. Now it's ITMx = -0.992, ITMy = 1 for MICH (lockin output matrix values).
RA: what were the old values? Does this change make any difference for the signal mixing noticed before? |
6400
|
Mon Mar 12 01:04:18 2012 |
keiko | Update | LSC | RAM simulation update, RAM LSC matrix |
I calculated the DRMI RAM LSC matrix with RAM and the operation point offsets.
- configuration: C1 DRMI
- RAM is added by an Mach-Zehnder ifo placed before the PRM
- demodulation phases are optimised for each DoF
- the operation points offset from the PDH signals are calculated and added to the optical configuration as mirror position offsets
- Then the matrix is calculated with the offsets and the RAM
- The set of the scrips are found as RAMmatrix.m, normMAT.m, newGetMAT.m, on CVS/ifomodeling/40m/fullIFO_Optickle. They are a bit messy scripts at this moment.
Results:
(1) No RAM LSC matrix
|
PRCL |
MICH |
SRCL |
REFL11I |
1 |
-0.001806 |
-0.000147 |
AS 55Q |
0.000818 |
1 |
0.000474 |
AS 55 I |
1.064561 |
902.292816 |
1 |
(2) With 1% RAM mod index of PM (normalised by (1) )
|
PRCL |
MICH |
SRCL |
REFL11I |
1.000618 |
-0.001837 |
-0.000163 |
AS 55Q |
0.000919 |
1.000521 |
0.000495 |
AS 55 I |
1.169741 |
924.675187 |
1.018479
|
(3) With 5% RAM mod index of PM (normalised by (1) )
|
PRCL |
MICH |
SRCL |
REFL11I |
0.999986 |
-0.001812 |
-0.000150 |
AS 55Q |
0.000838 |
1.000028 |
0.000479 |
AS 55 I |
1.084598 |
906.83668 |
1.003759
|
|
6401
|
Mon Mar 12 18:57:58 2012 |
keiko | Update | LSC | RAM simulation update, RAM LSC matrix |
Quote: |
I calculated the DRMI RAM LSC matrix with RAM and the operation point offsets.
- configuration: C1 DRMI
- RAM is added by an Mach-Zehnder ifo placed before the PRM
- demodulation phases are optimised for each DoF
- the operation points offset from the PDH signals are calculated and added to the optical configuration as mirror position offsets
- Then the matrix is calculated with the offsets and the RAM
- The set of the scrips are found as RAMmatrix.m, normMAT.m, newGetMAT.m, on CVS/ifomodeling/40m/fullIFO_Optickle. They are a bit messy scripts at this moment.
Results:
(1) No RAM LSC matrix
|
PRCL |
MICH |
SRCL |
REFL11I |
1 |
-0.001806 |
-0.000147 |
AS 55Q |
0.000818 |
1 |
0.000474 |
AS 55 I |
1.064561 |
902.292816 |
1 |
(2) With 1% RAM mod index of PM (normalised by (1) )
|
PRCL |
MICH |
SRCL |
REFL11I |
1.000618 |
-0.001837 |
-0.000163 |
AS 55Q |
0.000919 |
1.000521 |
0.000495 |
AS 55 I |
1.169741 |
924.675187 |
1.018479
|
(3) With 5% RAM mod index of PM (normalised by (1) )
|
PRCL |
MICH |
SRCL |
REFL11I |
0.999986 |
-0.001812 |
-0.000150 |
AS 55Q |
0.000838 |
1.000028 |
0.000479 |
AS 55 I |
1.084598 |
906.83668 |
1.003759
|
|
Adding some more results with more realistic RAM level assumption.
(4) With 0.1% RAM mod index of PM (normalized by (1) )
|
PRCL |
MICH |
SRCL |
REFL11I |
0.99999 |
-0.001807 |
-0.000148 |
AS 55Q |
0.000822 |
1.000002 |
0.000475 |
AS 55 I |
1.068342 |
906.968167 |
1.00559
|
(5) With 0.5% RAM mod index of PM (normalized by (1) )
|
PRCL |
MICH |
SRCL |
REFL11I |
0.999978 |
-0.001810 |
-0.000149 |
AS 55Q |
0.000830 |
1.000010 |
0.000476 |
AS 55 I |
1.075926 |
904.321433 |
1.001677
|
|
6402
|
Mon Mar 12 22:14:56 2012 |
Suresh | Update | RF System | Calibration of Demod Board Efficiency. |
I have completed the calibration of the demod board efficiencies. Here is the schematic of the set-up.

The data is given below and the data-file is attached in several different formats.

|
Attachment 3: Demod_calib.txt
|
Measurements After corrections Efficiency= out/in
Demod Board mV_ampl mV_pk-pk mV_pk-pk mV_ampl mV_ampl mV_ampl Vout/Vin Vout/Vin
PD in Q out I out PD in Q out I out Q out I out
REFL33 10.6 10.0 10.0 9.4 5.0 5.0 0.53 0.53
AS11 24.0 10.0 11.0 21.3 5.0 5.5 0.23 0.26
REFL11 22.5 240.0 255.0 20.0 120.0 127.5 6.00 6.38
POX11 24.0 9.2 8.5 21.3 4.6 4.3 0.22 0.20
POY11 22.4 10.5 9.0 19.9 5.3 4.5 0.26 0.23
AS55 17.6 268.0 268.0 15.6 134.0 134.0 8.57 8.57
REFL55 19.7 15.8 15.5 17.5 7.9 7.8 0.45 0.44
POP55 18.8 278.0 274.0 16.7 139.0 137.0 8.32 8.20
REFL165 21.2 16.0 16.4 18.8 8.0 8.2 0.42 0.44
POY110 23.4 14.7 14.4 20.8 7.4 7.2 0.35 0.35
POY22 17.5 11.9 9.3 15.6 6.0 4.6 0.38 0.30
|
Attachment 4: Demod_calib.xlsx
|
6403
|
Tue Mar 13 07:04:55 2012 |
kiwamu | Update | LSC | evolution of the sensing matrix in PRMI as a function of time |
The punch line is -- the sensing matrix still looks strange in the PRMI configuration.
I have been measuring the sensing matrix of the PRMI configuration because it didn't make sense (#6283).
One strange thing I have noticed before was that all the I-phase signals showed a weird behavior -- they fluctuate too much in time series.
Tonight I measured the sensing matrix again but this time I recorded them as a function of time using the realtime LOCKINs in the LSC front end.
The attached plots are the responses (optical gains) of PRCL and MICH in watts / meter at various sensors in time series.
I will explain some more details about how I measured and calibrated the data in another elog entry.


|
6404
|
Tue Mar 13 13:28:31 2012 |
Ryan Fisher | Update | CDS | DAQ restart with new ini file |
Extra note: This was the ini file that was edited:
/cvs/cds/rtcds/caltech/c1/chans/daq/C1PEM.ini |
6405
|
Tue Mar 13 16:40:06 2012 |
kiwamu | Update | LSC | evolution of the sensing matrix in PRMI as a function of time: details |
Here I describe the measurement of the sensing matrix.
Motivations
There were two reasons why I have been measuring the sensing matrix :
- I wanted to know how much each element in the sensing matrix drifted as a function of time because the sensing matrix didn't agree with what Optickle predicted (#6283).
- I needed to estimate the MICH responses in the 3f demodulated signals, so that I can decide which 3f signal I should use for holding MICH.
I will report #2 later because it needs another careful noise estimation.
Measurement
In order to measure the sensing matrix, the basic steps are something like this:
- Excite one of the DOF at a certain frequency, where a notch filter is applied in the LSC servos so that the servos won't suppress the excitation signal.
- Demodulate the LSC signals (e.g. C1:LSC-REFL11_I_ERR and etc.,) by the realtime LOCKINs (#6152) at the same frequency.
- Calibrate the obtained LOCKIN outputs to watts/meter.
In the actual measurement I choose the frequency of the excitation signal to be at 283.1 Hz,
at which any of the LSC servos don't have gains of more than 1 and there were no particular structures in the spectra.
For the amplitude of the excitation, I usually choose it to be 1000 - 2000 counts.
Because all the actuators have response functions of approximately 10 -9 / f^2 meter/counts ( #5637), the actual displacement in the excited DOF should be about 10 pm level.
Therefore the excited displacements must be always in the linear ranges and also the amplitude in counts is reasonably smaller than the DAC range.
LOCKIN detection
The attached cartoon below shows how the LOCKIN system works for the MICH response measurement.
In the case of the PRCL response measurement, the setup is the same except that only PRM is shaken.
Here is some notes about the LOCKIN detection.

Calibration of the LOCKINs
The calibration of the LOCKIN detectors is easy because all the processes takes place in the digital land, where we know all the parameters.
In this phase the goal is to calibrate the signals into counts / meter.
To calibrate the LOCKIN output signals, the following equation is used :
[The obtained LOCKIN output in counts ] = H x ADOF x CLO x CEXC x 1/2 ,
where H is the response of a sensor (e.g. AS55_I, AS55_Q and so on) against a particular DOF in unit of counts / m and this the quantity which we want to measure here,
ADOF is the actuator efficiency of the DOF at the excitation frequency in unit of m/counts,
CLO is the amplitude of the local oscillator signal for demodulating the sensor signals in unit of counts,
CEXC is the amplitude of the excitation signal in unit of counts,
the last 1/2 term comes from the fact there is a low pass filter in each demodulation path.
Therefore once we measure the response of a sensor, dividing the obtained LOCKIN output by ADOF x CLO x CEXC x 1/2 gives the calibrated response in unit of counts/meter.
A DOF are well known as they have been measured several times ( #5637).
For the MICH actuator I assumed that AMICH = 2 x (ITMY response) since they are balanced through the actuation coefficients.
Note that a confirmation of this calibration has been done
when the configuration is in the simple Michelson, where we can easily estimate the response of a sensor by letting the MICH freely swing.
Calibration of the responses to watts/meter
With the calibration process described above, we obtain the sensor responses in unit of counts/m.
Then we need to do another calibration to make them into unit of W/m.
If we think about how the RFPD signal flows, we get the following gain chain.
[raw response in counts/m ] = Hopt x CADC x Ldemod x GWF x Ztrans x RPD
Hopt is the optical gain at a sensor which we want to calibrate. It is in unit of W/m.
CADC is the conversion factor of the ADCs and the value is CADC = 1638.4 counts/m because their resolution is 16 bit and the range is +/-20 V.
L demod is the conversion efficiency of the demodulation boards in unit of V/V. I used the values which Suresh measured yesterday ( #6402).
GWF is the gain of the whitening filter in unit of V/V,
Z trans is the transimpedance gain of an RFPD in unit of V/A and I used the values summarized in ( the wiki),
and RPD is the responsivity of the photo diodes and I assumed RPD = 0.75 A/W for all the RFPDs.
Therefore the calibration can be done by dividing the raw response value by the entire gain chain of CADC x Ldemod x GWF x Ztrans x RPD.
Settings and parameters
- LSC RF demodulation phases
- AS55 = 17.05 deg (minimizing the PRCL sensitivity in the Q-phase)
- REFL11 = -41.05 deg (maximizing the PRCL sensitivity in the I-phase)
- REFL33 = -25.85 deg (maximizing the PRCL sensitivity in the I-phase)
- REFL55 = 4 deg (maximizing the PRCL sensitivity in the I-phase)
- REFL165 = 39 deg (random number)
- Whitening filters
- AS55 = 30 dB
- REFL11 = 0 dB
- REFL33 = 42 dB
- REFL55 = 30 dB
- REFL165 = 45 dB
- MICH servo
- AS55_Q for the sensor
- G = -5 in the digital gain
- FM2, FM3, FM5 and FM9 actiavted
- UGF ~ 100 Hz
- Feedback to ITMs differentially
- PRCL servo
- REFL33_I for the sensor
- G = 1 in the digital gain
- FM2, FM3, FM4, FM5 and FM9 activated
- UGF ~ 100 Hz
- Feedback to PRM
Quote from #6403 |
Tonight I measured the sensing matrix again but this time I recorded them as a function of time using the realtime LOCKINs in the LSC front end.
I will explain some more details about how I measured and calibrated the data in another elog entry.
|
|