  4576   Wed Apr 27 21:08:08 2011 ranaUpdateLSCAS11

I worked on AS_11 today. Its ready for its noise / optical gain calibrations. I have left it on Suresh's desk.

AS11.png

This was one of the 24.5 MHz Black Box (Ben Abbott) style RFPDs rescued from LLO. The tunable inductor that was installed was too small to get the frequency down to 11 MHz and so I swapped in one of the shielded, ferrite core ones from our '7mm' CoilCraft kit. It had a range of 1.2 - 1.8 uH according to the datasheet.

I wasn't able to simulataneously get the peak at 11.06 MHz and the notch at 59.3 MHz and so I took Koji's advice and tuned the peak best. The plot above shows how the notch is slightly off. I think its not a problem; to get it better we would have to change out the inductor for the "2-omega" notch, but I was too lazy. The thinking is that its more important to have the gain be symmetric around the signal readout frequency so as to not imbalance the audio sidebands.

Since this one is going to be AS_11, we think that the 22 MHz signal will be tiny: the transmission of the 11 MHz sidebands to the dark port is small. If we later want to put in a 22 MHz notch anyway, there is space to do this via the 'active notch' pads around the MAX4107.

For the above plot, I used the Jenne laser. The DC output of the PD was ~30 mV (~0.6 mA). The RF drive to the laser was -10 dBm: no saturations. I have calibrated out the cable responses, but not using the 1811 setup, so the absolute calibration has yet to be done.

Also, it needs some new stickers. It would be handy if someone can figure out how to get some sheets of stickers that we can put into the printer. Then we can laser printer all of the data onto the stickers and stick them to the RFPD box.

  4577   Wed Apr 27 21:19:25 2011 kiwamuUpdateLSCLSC whitening for PD1-4

On the back side of 1Y2 rack I found a cable, CAB-1X2-LSC_7, which is supposed to be connected to the whitening filter was disconnected.

I plugged it back and confirmed that the whitening filter is under control of EPICS.

Now all the gain sliders seem to be working because I can change the amplitude of signals with the sliders.

 

(method)

  To check if the gain sliders are working or not, I intentionally disconnected all the inputs to the whitening filter.

Then I brought a gain slider of interest to the maximum. Due to the big gain I was easily able to see noise lying above ADC noise.

Also if the gain slider is 0 dB, which is the minimum value, the spectrum becomes just ADC noise.

In this way I checked all the gain sliders from PD1 to PD4. The picture below is just an example screenshot when I was doing this test.

Note that each filer is designed to have two poles at 150 Hz and two zeros at 15 Hz.

Screenshot-1.png

Quote from #4570

While checking whitening filters on the LSC rack, I found some epics controls for the whitening looked not working.

So I powered two crates off : the top one and the bottom one on 1Y3 rack.

These crates contain c1iscaux and c1iscaux2. Then powered them on. But it didn't solve the issue.

  4585   Fri Apr 29 03:39:49 2011 KojiSummaryLSCCavity lengths

I tried the idea that the PRC can resonate f1 and f2 at the same time if the arm gives the reflection phase to f1 and f2 with the ratio of 1 vs 5.

The details are described on wiki. The point is this removes all of the PRC/SRC/asymmetry mumbo jumbo.

The calculated cavity lengths for f_mod of 11.065399MHz are:

  • Arm Length: 37.7974 [m]

  • PRC Length: 6.7538 [m]

  • SRC Length: 5.39915 [m]

  • Asymmetry (lx-ly): 0.0342 [m]


Here is the actual values derived from the photos.

  • Arm Length: 37.54 [m] (0.26m too short)

  • PRC Length: 6.760 [m] (6mm too long)

  • SRC Length: 5.415 [m] (16mm too long)

  • Asymmetry (lx-ly): 0.0266 [m] (8mm too long)

  4586   Fri Apr 29 05:48:52 2011 kiwamuUpdateLSCY arm locked

The Y arm has been locked with the IR beam. The purpose is to use the arm as an alignment reference for the input PZTs.

Detail will be posted later. Here is a picture of ITMY suspension. You can see there is a beam spot in the middle of the test mass.

DSC_2978_ss.jpg

  4587   Fri Apr 29 12:18:48 2011 kiwamuUpdateLSCY arm locked : details

First of all, the conclusions / results from the exercise of the Y arm locking yesterday are:

   The position of the beam spots on both ETMY and ITMY are now not so bad ( ~ 5 mm off from the center).

   The input PZTs are coarsely aligned to the Y arm.

   Nevertheless IP_ANG is still too high to come out from the view port at the Y end station.

  After the alignments of PRM, SRM and Michelson, POP is still largely clipped.

 


(what I did)

  - Alignments of the Input PZTs

    First I tried letting the incident beam hit the center of ETMY by steering PZT1 and 2 as usual.
    Then I coarsely aligned the cavities to the incident beam and checked the beam flashing spots on ETMY and ITMY with the CCD monitors.
    When the spots were far from the center I went back to the alignment of PZT1 and 2 to get better beam positions. And repeated this work several times.

  - Adjustment of the demodulation phase  for the Y arm PDH.

    First I started looking at the digital signals and tried correcting the demodulation phase by the rotation matrix, but this didn't go fast because I had to do some DAQ settings, plotting and analysis.
    Instead looking at the digital signal, I observed the analog signals with an oscilloscope. I found the demodulation phase was something like 45 deg.
    Based on the analog measurement I rotated the digital matrix by 45 deg to get the I-signal maximized. Indeed this worked well. I obtained a beautiful PDH signal from the I-signal.
    Note that we are using 11 MHz mod/demod and eventually the signals come out from "REFL33" on the digital side.

  - Activation of oplev on ITMY

    Instead using ETMY I used ITMY for the length control because somehow I felt that ETMY coils were suspicious and they looked not so nice.
    One of the reason is that ETMY's coil actuation efficiencies looked low compared to the other test masses.
    For example a gain of 700 for SUSSIDE damping is needed on ETMY to get a reasonable Q. This is about 2 - 3 times larger gain than the other test masses.
    So I started using ITMY for the locking and activated the oplev to suppress unwanted excitation due to kicks from the control signal during the locking,
    The oplev has been misaligned, so I went to the ITMY optical bench and tweaked a steering mirror to let the He-Ne beam go into the QPD.
    I set the gains +2 for PITCH and -2 for YAW.

  - PDH locking

    The locking had been quite difficult even though the cavity alignment was quite good.
    It's because the beam on AS11_RFPD was almost falling off from the photo diode. This causes a big amplitude fluctuation in the PDH signal as the beam position moves.
    After aligning the beam by steering BS I got able to lock the Y arm. The PD whitening gains are all 0 dB and the feedback gain is -2, giving us a UGF of 250 Hz.

 

(Broken or likely broken stuff)

 * IP_ANG doesn't give a signal to the digital side.

 * ETMY coils look weak and 2 - 3 times weaker than the other test masses. (or OSEM readout gain maybe lower)

 * reload button on sitemap.adl doesn't work.

 * Farfalla, a lab laptop, seems out of network.

Quote from #4586

The Y arm has been locked with the IR beam. The purpose is to use the arm as an alignment reference for the input PZTs.

Detail will be posted later. Here is a picture of ITMY suspension. You can see there is a beam spot in the middle of the test mass.

 

  4588   Fri Apr 29 13:15:04 2011 kiwamuUpdateLSCY arm locked : details

As far as I know, this button works only once after the launch of MEDM...

Quote:

 * reload button on sitemap.adl doesn't work. 

 

  4589   Fri Apr 29 13:58:36 2011 ranaUpdateLSCY arm locked : details

Quote:

 * Farfalla, a lab laptop, seems out of network.

 If you look at the real host table instead of the misleading host table in the wiki, you will see that someone has deleted Farfalla from there. She needs to be re-added.

  4595   Mon May 2 12:51:52 2011 kiwamuUpdateLSCPRMI locking : plan

Today we will try to lock the PRMI. Here is a plan for it.

(to be done in the daytime)

 - setup REFL11 RFPD

 - setup AS55 RFPD

 - install a demod board for 55 MHz

 - install a 3-way RF combiner on EOM.

 - prepare 55 MHz RF source (Marconi or RF source box ?)

 - adjustment of each demodulation phase

 - activation of PRM oplev

 

 (control topology)

- AS55_Q ==> BS

- REFL11_I ==> PRM

  4598   Mon May 2 15:53:27 2011 kiwamuUpdateLSCPRMI locking prep : demodulation boards

I swapped the name of two demodulation boards (AS55 and REFL55).

Now the REFL11 and AS55 demodulation boards are ready to go for the PRMI locking.

The physical labels, which are on the front surface of the boards, are also corrected to avoid a confusion.

Here is the latest RF status.

RF_Work_Status.png

Quote from #4574

The files are on https://nodus.ligo.caltech.edu:30889/svn/trunk/suresh/40m_RF_upgrade/.

 

  4600   Mon May 2 18:16:24 2011 kiwamuUpdateLSCPRMI locking : status update

(done)

- AS55 RFPD

  With a help from Jamie the AS55 RFPD was installed.

- 55 MHz demodulation board

  The AS55 demod board was installed on 1Y2.

- 3-way combiner

  ZFSC-3-13 has been installed. All the RF cables from the source side were connected to the combiner.

 

(next things)

 - installation of the REFL11 RFPD

 - DAQ check for AS55 and REFL11

Quote from #4595

(to be done in the daytime)

 - setup AS55 RFPD

 - install a demod board for 55 MHz

 - install a 3-way RF combiner on EOM.

 - prepare 55 MHz RF source (Marconi or RF source box ?)

 

  4604   Tue May 3 03:53:20 2011 kiwamuUpdateLSCPRMI locking : done

The PRMI has been successfully locked

PRMI_locking.png

Also changing the sign of the PRC control gave me the lock of the carrier resonant condition.

The screenshot above is the time series of the error signals when I was locking the PRMI in the sideband resonant condition (i.e. carrier is non-resonant).

Note that I used REFL11 for the PRC control and AS55 for the MICH control as planed.

Details will be posted in the morning.

Quote from #4595

Today we will try to lock the PRMI.

 

  4605   Tue May 3 04:09:53 2011 KojiUpdateLSCREFL55 PD update

[Rana, Koji]

REFL55 was modified. The noise level confirmed. The PD is now ready to be installed.

 


Kevin's measurement report told us that something was wrong with REFL55 PD. The transimpedance looked OK, but the noise level was terrible (equivalent to the shotnoise of 14mA DC current).

Rana and I looked at the circuit, and cleaned up the circuit, by removing unnecessary 11MHz notch, 1k shunt resister, and so on.

I made a quick characterization of the PD.

First page:

The transimpedance ws measured as a function of the frequency. The resonance was tuned at 55MHz. The notch was tuned at 110MHz in order to reject the second harmonics. The transimpedance was ~540V/A at 55MHz. (For the calibration, I believed the DC transimpedance of 50V/A and 10000V/A for the DC paths of this PD and #1611, respectively, as well as the RF impedance (700V/A0 of #1611.

Second page:

Output noise levels were measured with various amount of photocurrent using white light from a light bulb. The measurement was perforemed well above the noise level of the measurement instruments.

Third page:

The measured output noise levels were converted into the equivalent current noise on the PD. The dark noise level agrees with the shot noise level of 1.5mA (i.e. 22pA/rtHz). In deed, the noise level went up x~1.5 when the photocurrent is ~1.4mA.

Attachment 1: REFL55.pdf
REFL55.pdf REFL55.pdf REFL55.pdf
  4606   Tue May 3 05:32:04 2011 kiwamuUpdateLSCdaytime tasks

Daytime tasks :

 - PRM & BS oplev (Steve)

 - LSC binary outputs (Joe/Jamie)

 - installation of the REFL55 RFPD (Suresh/Jamie)

 - Adjustment of demodulation phases (Kiwamu)

 - Bounce-Roll filters on BS and PRM (Suresh/Joe)

 - Suspension diagnostic using the free-swinging spectra (Leo)

 - PMC alignment (Jenne/Koji)

  4607   Tue May 3 10:21:25 2011 KojiUpdateLSCdaytime tasks

I think the installation of the PD DC signals are quite important. What to do
1) Connect the DC signals to the right top whitening board (be aware that there may be the modification of the whitening circuit).
2) Reconfigure the LSC model such that the DC signal is passed to the right channels (modify the left top part of the model)

Quote:

Daytime tasks :

 - PRM & BS oplev (Steve)

 - LSC binary outputs (Joe/Jamie)

 - installation of the REFL55 RFPD (Suresh/Jamie)

 - Adjustment of demodulation phases (Kiwamu)

 - Bounce-Roll filters on BS and PRM (Suresh/Joe)

 - Suspension diagnostic using the free-swinging spectra (Leo)

 - PMC alignment (Jenne/Koji)

 

  4610   Tue May 3 11:49:03 2011 KojiUpdateLSCdaytime tasks

Done. C1:PSL-PMC_PMCTRANSPD was improved from ~0.75 to 0.87.

Quote:

- PMC alignment (Jenne/Koji)

 

  4617   Tue May 3 16:47:55 2011 kiwamuUpdateLSCPRMI locking : details

Here are some details about the PRMI locking done last night.

 

(REFL11 installation)

 REFL11 has been installed on the AP table. The RF signal from the RFPD is sent by a heliax cable which has been called ASDD133.

Before the beam goes into the RFPD a HWP and PBS are installed such that we can adjust the amount of light entering to the photo diode.

One thing I didn't like was that I had to introduce a big amount of the light into the PD to get a reasonably big RF signal.

I was trying to look for an RF signal by looking at a spectrum analyzer, then I realized that the RF signal at 11 MHz was quite tiny when the DC_MON was less than 1.5 V.

After I increased the amount of the light up to 1.9 V in DC_MON, which sounds already too much, I then got able to see the 11 MHz signal on the analyzer.

Note that I decreased the amount of the light down to 0.5 V after I finished locking the PRMI.

We should make sure what is going on with the 11 MHz modulation.

APtable.png

 

(Locking)

First I started locking the MIchelson with AS55. The demodulation phase was already somewhat optimized to the I-signal port.

So I decided not to touch the demodulation phase matrix because it may take some times.

After I eliminated electrical offsets in the digital side, I was easily able to lock the Michelson. The control sign was plus.

Then I started playing with the PRC control too. The demodulation phase in REFL11 looked nearly 45 deg although I didn't carefully measure it.

I made a 45 deg rotational matrix to maximize the I-port signal and tried to lock the PRC. Then immediately I was able to lock PRC as well as MICH.

 GAIN_MICH = 100

 GAIN_PRC = 100

Also GAIN_PRC = -100 gave a carrier resonant lock.

The control filters are the same in MICH and PRC. I used my favorite filters as usual.

 FM1 = 1000 : 10

 FM6 = 0.1 : 1

 FM7 = 1 : 50

Somehow I frequently failed to engage the boost filters (i.e. FM6 and FM7) it looks offsets in the control path kicks either BS or PRM.

Quote from #4604

The PRMI has been successfully locked

Details will be posted in the morning. 

 

  4619   Tue May 3 18:25:38 2011 kiwamuSummaryLSCPRMI locking : plan

[Rana/Jamie/Kiwamu]

 Since we've got the PRMI locked we now should be able to do more qualitative measurements.

Here is a task list that we will measure/develop in the PRMI condition.

 

 

 - Optical gain measurements

 - Characterization of control loops

 - MICH and PRC calibrations

 - Noise budget

 - Development of automatic noise budget scripts

 - Arm loss measurement 

 - Shnupp asymmetry measurement

  4623   Wed May 4 13:45:56 2011 kiwamuUpdateLSClocking last night

Last night I was trying to calibrate the MICH error signal and the actuators on BS and ITMs.

However I gave up taking the data because the MC locking was unstable. MC3 drifted a lot.

  4631   Thu May 5 00:08:59 2011 JenneUpdateLSC(Almost) New Screens for RFPDs

I modified C1LSC.mdl to use the CDSphase blocks, which automatically calculate the R and D phase rotation for us.  Now each of the RFPDs has 2 channels in place of the old IQ_MTRX channels:  C1:LSC-RFPD_PHASE_R and C1:LSC-RFPD_PHASE_D.

I have not yet compiled / rebooted / done CDS magic to actually make these installed.  So far the change is only in the simulink model.

I was going to wait until morning to compile/reboot/magic, so I can do it under Joe's supervision.

In the meantime, I also modified the RFPD screens.  They have white boxes for the _R and _D channels just now, but that's because the new model hasn't been put in.  They now look like phase rotators, instead of Koji's temporary matrix.

Still to do:  Find the EPICS database where the phase rotation calculation is done (you give it an angle, it gives you sin(angle) and cos(angle) ).  I want to put a "90-angle" in the database so that we can type in the measured relative phase between I and Q, and it will calculate how many more degrees it needs to get to 90deg. 

 

  4632   Thu May 5 04:38:20 2011 KojiSummaryLSCComparison between S3399 and FFD-100

Comparison between Hamamatsu S3399 and Perkin Elmer FFD-100

These are the candidates for the BB PD for the green beat detection as well as aLIGO BB PD for 532nm/1064nm.

FFD-100 seems the good candidate.

 

Basic difference between S3399 and FFD-100

- S3399 Si PIN diode: 3mm dia., max bias = 30V, Cd=20pF

- FFD-100 Si PIN diode: 2.5mm dia., max bias = 100V, Cd=7pF

 

The circuit at the page 1 was used for the amplifier.

- FFD-100 showed 5dB (= x1.8) larger responsivity for 1064nm compared with S3399. (Plot not shown. Confirmed on the analyzer.)

- -3dB BW: S3399 180MHz, FFD-100 250MHz for 100V_bias. For 30V bias, they are similar.

Attachment 1: PD_response.pdf
PD_response.pdf PD_response.pdf PD_response.pdf
  4636   Thu May 5 13:13:32 2011 JenneUpdateLSCNew RFPD screens are in

Joe helped me compile the lsc simulink model, and now we have R&D phase rotation.

Right now, we have to do our own math, and figure out what relative phase to put in.  Soonly, I'll figure out how to do subtraction, and we can put in the measured value.

 

More details when I'm not running around like crazy...

--------------------------------------------------------------------------------------------

Okie dokie.  Last night I had modified the c1lsc.mdl to accommodate the R & D phase rotation.  I also made pretty new screens.  This morning however, the adventures began.....

Under Joe's supervision, I ran "make c1lsc".  The error that came up was something about things not being connected.  Joe assures me that this is a temporary problem, that Rolf is already working on.  The reason is that right now the LSC model is "flat", i.e. it doesn't have a bunch of sub-boxes and sub-screens in the simulink model.  Somehow this causes badness.  Joe stuck all the guts of the LSC model into a sub-model.  He then enabled "top_names", which makes the channels use the name of the sub-model, not the sub-model AND the main model (so since the sub-model is called LSC, our channels are just C1:LSC-OTHER_STUFF, rather than C1:LSC-LSC_OTHER_STUFF).  This fixed things so that the compiling worked (when we did "make c1lsc").  The one other thing that we changed was to delete all of the little "Outs" that were attached to EPICS readouts.  These are unneccessary and don't go anywhere, and when we made the sub-model, they made a bunch of empty outputs (unconnected outs on the main simulink model).  So, after doing that, we were able to compile, and do "make install-c1lsc", and all was good in the world.  Mostly.

Joe then noticed that I was using the CDS part "cdsPhase", which only takes one phase input.  I wanted "cdswfsPhase", which actually does the R&D phase rotation that we want.  Perhaps Alex/Rolf/whoever should change the name of that CDS part.  We switched all of the cdsPhase blocks to be cdswfsPhase, and recompiled.  All was still good in the world.  Mostly.

The last thing that was funny was that when I wanted to execute the medm screens, they would still look at the old _IQ_MTRX_1_1 and _IQ_MTRX_2_1 values, rather than the newly defined _PHASE_R and _PHASE_D channels, even though while editing the medm screen, it looked like it was pointing to the right place.  Anyhow, I opened the text file version of the C1LSC_PDX.adl, and changed the channel names to the _R and _D versions by hand.  I don't know if we edit the screens and run generate_screens.py again, if we'll have to re-edit the .adl text files.

After fixing this, all really was good in the world. 

Perhaps though, this making a subsystem business broke the filters somehow?  Foton is looking at the wrong text file now?  Something?  The filters are all still there, they just got moved down a level.  Joe said that he and Rolf are on it, and he should be able to put the LSC model back to being "flat" in the next few days.

  4637   Thu May 5 13:14:15 2011 kiwamuUpdateLSCchans file for LSC

Foton doesn't correctly display the LSC filter bank file : C1LSC.txt.

Foton tells a lie that they all are empty.

 

The file itself looks fine to me i.e. I can find correct filters in text format.

Looks like someone (maybe Joe and Jenne ?) updated the file. I am not sure if this is the reason or not.

allegra:chans>ls -al | grep LSC
-rw-r--r--  1 controls controls   20659 May  5 11:46 C1LSC.txt

 

NEEDS TO BE FIXED SOON

  4638   Thu May 5 13:55:08 2011 kiwamuUpdateLSCMI locking : calibration of BS and ITMs actuators

EDIT by KI on 15/5/2011:

The calibration of MICH error signal was wrong by a factor of 2.

 

The open loop transfer functions of the Michelson locking have been measured.

The purpose of this excise is to calibrate the coil-magnet actuators on BS and ITMs.

The estimated actuation coefficients are :

 BS = 3.69e-08 [m/counts]
 ITMX = 8.89e-09 [m/counts]
 ITMY = 9.22e-09 [m/counts] 

I guess the accuracy is something like 5 % because the calibration of the MI optical gain relies on a peak-to-peak measurement.

A next step is to calibrate the PRM actuator and the PRC optical gain.


 

(measurement)

The Michelson was locked with different actuators in every measurement. I locked the Michelson to the dark fringe with BS, ITMX and ITMY in each time.

The measured peak-to-peak value in the error signal was 20.2 counts, corresponding to a sensor gain of 5.96e+07 [counts/m]. Note that I used AS55_Q for the locking.

After locking the MI I took the open loop transfer function by injecting broadband noise from DTT.

Then the data were fitted coarsely.  In the fitting I used the resonant frequencies that Leo reported recently (http://blue.ligo-wa.caltech.edu:8000/40m/Mechanical_Resonances).

The Q-values are assumed to be 5 because of the local dampings. As a result the fitting gives us the actuator coefficients.

Here is a plot showing the measured open loop transfer functions. The solid lines represent the fittings.

MICH_OL.png

 

(by the way)

- The delay time including ADC/DAC and RFM looks quite big. According to the fitting the delay is something like 600 usec.

This is about two times larger than the one reported before (see this entry). I will re-measure it with empty filters.

  4639   Thu May 5 14:40:14 2011 KojiUpdateLSCMI locking : calibration of BS and ITMs actuators

I've got confused

1) Are these the DC responses of the coils? If that is true, we need to specify the resonant frequency of each suspension to get the AC response.

2) Are these the AC responses well above the resonant freqs? In that case, The responses should be x.xxx / f^2 [m/counts]

Quote:

The open loop transfer functions of the Michelson locking have been measured.

The purpose of this excise is to calibrate the coil-magnet actuators on BS and ITMs.

The estimated actuation coefficients are :

 BS = 3.69e-08 [m/counts]
 ITMX = 8.89e-09 [m/counts]
 ITMY = 9.22e-09 [m/counts]

  4640   Thu May 5 14:50:30 2011 JenneUpdateLSCchans file for LSC

Quote:

Foton doesn't correctly display the LSC filter bank file : C1LSC.txt.

Foton tells a lie that they all are empty.

 

The file itself looks fine to me i.e. I can find correct filters in text format.

Looks like someone (maybe Joe and Jenne ?) updated the file. I am not sure if this is the reason or not.

allegra:chans>ls -al | grep LSC
-rw-r--r--  1 controls controls   20659 May  5 11:46 C1LSC.txt

 

NEEDS TO BE FIXED SOON

 See my updated elog 4636 for what Joe and I did this morning, and what a possible problem is (making the LSC model into a sub-model).

  4643   Thu May 5 15:28:23 2011 kiwamuUpdateLSCRe: MI locking : calibration of BS and ITMs actuators

They are the DC responses.

I put the resonant frequencies that Leo reported in the wiki to obtain the DC response.

The resonant frequencies I used are :

  f_BS = 0.957 Hz

  f_ITMX = 0.966 Hz

  f_ITMY = 0.988 Hz

Also I assumed that all the Q-values are 5 due to the damping.

Quote:

I've got confused

1) Are these the DC responses of the coils? If that is true, we need to specify the resonant frequency of each suspension to get the AC response.

2) Are these the AC responses well above the resonant freqs? In that case, The responses should be x.xxx / f^2 [m/counts]

Quote:

 BS = 3.69e-08 [m/counts]
 ITMX = 8.89e-09 [m/counts]
 ITMY = 9.22e-09 [m/counts]

 

  4645   Thu May 5 16:11:22 2011 JamieUpdateLSCchans file for LSC

Quote:

Foton doesn't correctly display the LSC filter bank file : C1LSC.txt.

 This was because of a bug in the RCG for foton filter module naming when top names is being used.  Rebuilding the LSC front-end model with top_names (which was needed to get around another bug in the RCG) broke the filter file.  I manually fixed the file, so it should work now.

  4656   Sat May 7 02:52:53 2011 kiwamuUpdateLSCTRY photo diode installed

[Suresh / Kiwamu]

 We installed the TRY photo diode (Thorlabs one) and the ETMYT CCD camera in place on the ETMY table.

Now we can see a signal on the TRY digital channel.

It will be quite useful for the Y arm locking, for instance we can do a triggered locking and the maximization of the intracavity power.

Someone has to install the EMTY trans QPD at some point.

  4664   Mon May 9 12:33:40 2011 kiwamuUpdateLSCC1:LSC-TRIG_MTRX : wrong matrix size

I found that C1:LSC-TRIG_MTRX has a wrong matrix size. It needs to be fixed.

It is designed to have a 11x8 matrix in the simlink model file, but it's been compiled as a 6x8 matrix.

I recompiled c1lsc but it didn't fix the issue. Here below is the matrix statement in the C file :c1lsc.c. Indeed it is 6x8 matrix ....

// MuxMatrix:  LSC_TRIG_MTRX
for(ii=0;ii<8;ii++)
{
  lsc_trig_mtrx[ii] =
          pLocalEpics->lsc.LSC_TRIG_MTRX[ii][0] * lsc_imux_trigger[0] +
          pLocalEpics->lsc.LSC_TRIG_MTRX[ii][1] * lsc_imux_trigger[1] +
          pLocalEpics->lsc.LSC_TRIG_MTRX[ii][2] * lsc_imux_trigger[2] +
          pLocalEpics->lsc.LSC_TRIG_MTRX[ii][3] * lsc_imux_trigger[3] +
          pLocalEpics->lsc.LSC_TRIG_MTRX[ii][4] * lsc_imux_trigger[4] +
          pLocalEpics->lsc.LSC_TRIG_MTRX[ii][5] * lsc_imux_trigger[5];
 }

  4665   Mon May 9 13:14:48 2011 josephbUpdateLSCC1:LSC-TRIG_MTRX : wrong matrix size

[Joe, Kiwamu]

There is a feature/bug of the RCG code that you can only have 1 receiving tag for every sending tag.  There were 5 tags which were being received by two tags each, for two different matrices.  Only the first tag was receiving, the second was apparently ignored.

This has been fixed temporarily by putting in direct lines in place of these 5 tags.

Quote:

I found that C1:LSC-TRIG_MTRX has a wrong matrix size. It needs to be fixed.

It is designed to have a 11x8 matrix in the simlink model file, but it's been compiled as a 6x8 matrix.

 

  4671   Mon May 9 18:54:29 2011 kiwamuUpdateLSCtriggerd locking logic : screen prepared

I made a new medm screen for the triggering logics. Have fun.

I put a button on C1LSC.adl to invoke this screen.

trigger_medm.png

Quote from #4656

It will be quite useful for the Y arm locking, for instance we can do a triggered locking and the maximization of the intracavity power.

 

  4689   Wed May 11 13:54:32 2011 steveUpdateLSCLSC power supplies

Aux- rack _1Y2 is just behind 1Y3  It contains Kepco Analoge DC power supplies for +- 5, 15 and 24V

Placing these power supplies away from the LSC rack was an effort to minimize pick up from them.

 

Attachment 1: P1070679.JPG
P1070679.JPG
  4694   Wed May 11 22:52:55 2011 kiwamuUpdateLSCstatus update and plan

Rana forced me to write this entry for summary because he didn't come to the 40m meeting today.

Status update :

    Interferometer Input Beam alignment with the PZTs.

     60 % done. The rest of the 40 % is to make the procedures automated.

     The beam spots on ITMY and ETMY are centered within ~1 mm accuracy.

     PRM, BS, ITMX, & ITMY actuator calibrations and PRC/MICH error calibrations

    Ongoing: First we will do it by hand, then some scripts will be made for the calibraion and resultant noise budget.

     F2A Suspension filter calculations.

      ETMY and ITMY are done. Need volunteer for ETMX, ITMX, BS, PRM, & SRM !!

      Bounce-Roll notch filters

      Leo is working on it. 25% complete...

     DC signal from RFPDs

      The RFPDs have a local SMA DC output as well as a DC output from their PD Interface cards in the LSC rack. We have hooked up some of the PD Interface DC outputs to the LSC ADCs but not tested??

Next steps:

  Installation of a temporary (Thorlabs) DCPD on either POY to see the intra-cavity power in the PRC. It would be ridiculous to put detectors on POX or POP since they're still clipped.

  D-phase and amplitude imbalance adjustment of the demod baords.

          make a script which uses pynds.

  Alignment of the full interferomter, starting from the X arm

  Loss measurements for the arms

  Schnupp asymmetry measurement

  

  4710   Fri May 13 01:42:35 2011 kiwamuUpdateLSCready for Schnupp asymmetry measurement

[Valera / Kiwamu]

 We are able to lock each arm smoothly.  It is ready for the Schnupp asymmetry measurement.

( to be done )

 - Manual D-phase adjustment of the AS55 channel.

 - A script to adjust the D-phase.

 - Trigger logic for the boost filters.

 - Modification of some old alignment scripts to adopt them to the new LSC channels

  4717   Sat May 14 14:50:21 2011 KojiUpdateLSC1Y2 5V Blown Fuse found -> Fixed

Incidentally, a blown fuse on 5V line at 1Y2 rack was found during the intallation of Sorensens.
The fuse (5A 125V) has been replaced and fixed.

When I plugged the fuse in, I heard some sound like relays were switched. Are there any relays in the LSC rack?

It was a 9th fuse from the top as seen in the picture.

Attachment 1: 1Y2_5V_Fuse.JPG
1Y2_5V_Fuse.JPG
  4721   Sun May 15 19:10:12 2011 kiwamuUpdateLSCCalibration of actuators : BS, ITMX and ITMY

The AC response of the actuators on BS, ITMX and ITMY were re-measured by another technique.

Last time I estimated them by measuring the open-loop transfer functions, but this time the responses were measured in a more direct way.

The measured AC responses (60 Hz - 200 Hz) are :

      BS   = 1.643e-98 / f2  [m/counts] (corrected based on the plot below - Manasa)

     ITMX = 3.568e-9 / f2 [m/counts]

     ITMY = 3.542e-9 / f2 [m/counts]

Next : measurement of the PRM actuator response


(The technique) 

 This time a technique that Rana told me a week ago was used.

This technique allows us to directly measure the response of an actuator at high frequency without any loop corrections.

First of all, MICH has to be locked to keep MICH within the linear range of the error signal. So now MICH is a linear sensor to the mirror motions.

In the MICH control a steep low pass filter should be inserted in order to avoid unwanted effects from the control loop at the high frequencies.

For example I put a low pass filter composed of an elliptical filter whose cut-off frequency is at 50 Hz such that the control loop doesn't push the mirrors above the cut-off frequency.

Hence the error signal of MICH above 50 Hz directly corresponds to the motion of the mirrors including BS, ITMX and ITMY.

Taking a transfer function from an actuator to the MICH error signal directly gives the actuator response.

In my measurements MICH was locked by feeding the signal back to BS. The plot below is the expected open-loop transfer function for the MICH control.

oltf.png

You can see that the open loop TF suddenly drops above 50 Hz. The UGF was at about 20 Hz, confirmed by looking at the loop oscillation on DTT.

 

(Measurement)

 In the technique the error signal has to be calibrated to [m]. This time AS55_Q was used and calibrated based on a peak-to-peak measurement.

The peak to peak value in the MICH error signal was 8 counts, which corresponds to the sensor efficiency of 4.72e+07 [counts/m].

Then I took transfer functions from each suspension (i.e. C1:SUS-XXX_LSC_EXC) to the error signal at AS55_Q over a frequency range from 60 Hz to 200Hz.

For the transfer function measurements I ran the swept sine on DTT to get the data. Note that the PD whitening filters were on.

The plot below is the results of the measurements together with the fitting lines.

calib_actuators.png

In the fitting I excluded the data pints at 60 Hz, because their coherence was low due to the power line noise.

  4729   Tue May 17 01:05:56 2011 kiwamuUpdateLSCAsymmetry measurement prep : recentering works

I re-centered beams on several PDs and a camera including :

  AS55, ETMY_QPD, TRY and ETMYT_CCD.

 

The most important one was AS55.

When I was locking each arm I found that the error signal from AS55 was very coupled to the angular motion of the arms.

I checked the beam on the AS55 RFPD and found the beam on the edge of the photo diode. This is possibly because Valera and I had been touching the input beam alignment.

At that time the DC signal from AS55 without aligning PRM and SRM was about 5 mV.

Adjusting the beam position by a steering mirror brought the DC signal up to 20 mV.

Then the lock of each arm became more stable.

  4735   Tue May 17 22:50:00 2011 JamieConfigurationLSCNew digital lockin added to LSC model/screen

I added a lockin to the LSC model, and added it's corresponding control to the LSC screen.  Here's is what I added to the LSC model:

lsc-mdl-lockin.png

On the left are the froms from the normalized RF PD outputs. Those go into a matrix, whose single row goes into the lockin signal input. The clock output from the lockin goes into the LSC output matrix (last row).

Here is what I added to the LSC master screen:

lsc-screen-lockin.png

I also modified the lockin overview screen:

lsc-lockin-screen.png

I think this new screen looks a lot cleaner.  Maybe we could start using this one as a template for lockin screens.

  4743   Wed May 18 22:11:31 2011 ranaConfigurationLSCNew digital lockin added to LSC model/screen

Untitled.png

 This is OK....but, the input matrix should come from the same place as the regular input matrix: i.e. it should be just another row like CARM, DARM, etc. rather than have its own screen.

Also, I think a nice mod to all the matrices would be if the ORANGE triangle was only visible when there's a signal flowing through it.

  4749   Thu May 19 16:46:20 2011 kiwamuUpdateLSCc1lsc model : input channels rearanged

According to Suresh's LSC rack design I rearranged the input channels of the c1lsc model such that the analog signals and the ADC channels are nicely matched.

Also I updated the c1lsc model in the svn with a help from Joe. The picture below is a screen shot of the input channels in the model file after I edited it.

c1lsc.png

  4753   Fri May 20 03:01:17 2011 kiwamuUpdateLSClocking status

(PRMI locking)

Since REFL11 has gone I tried locking the PRMI with combination of REFL55 and AS55.

Without any pain the lock of PRMI was achieved successfully. AS55 was used to sense MICH and REFL55 was used for PRC.

 

(scripting)

Additionally I was modifying several scripts which are invoked from C1IFO_CONFIGURE.adl. Some details about the scripts will be uploaded on the wiki later.

An important thing is that now we are able to use the "restore" commands for the Y arm, X arm, Michelson and PRM locking.

The scripts will automatically acquire the lock of each DOF.  The image below is just a screen shot of the medm screen where you can call the scripts.

Screen_shot_2011-05-20_at_2.50.00.png

 

( Still to do)

   * PRM actuator response measurement

   * PRC noise budget

   * MICH-PRC actuator decoupling

  4757   Sat May 21 06:19:46 2011 kiwamuUpdateLSCDRMI trial : no luck

I will try with POY55 that Koji prepared today.

  4759   Mon May 23 00:36:51 2011 kiwamuUpdateLSCDRMI trial : sucess

Eventually the DRMI was locked.

I was struggling to find a good signal port for SRC over the weekend and finally found AS55_I worked somehow. I used :

   REFL11_I --> PRC

   AS55_Q   --> MICH

   AS55_I    --> SRC

A configuration script was prepared such that someone can try this configuration by clicking a button on the C1IFO_CONFIGURE.adl screen.

I don't think this signal extraction scheme is the best, but now we can find better signal ports by shaking each DOF and looking at each signal port.

More details will be reported in the morning.

Quote:

I will try with POY55 that Koji prepared today.

 

  4760   Mon May 23 12:27:26 2011 kiwamuUpdateLSCDRMI trial : details

(PRMI locking with slightly misaligned SRM)

 First I tried locking PRC and MICH with a little bit misaligned SRM. This condition allowed me to search for a good signal port for SRC.

In this locking, REFL11_I was used to lock PRC and AS55_Q was used for MICH. This is the same scheme as the current PRMI locking.

Since the alignment of SRM was close to the good alignment, I expected to see fringes from SRC in some signal ports (i.e. REFL55, POY55 and so on).

Sometimes a fringe of SRC disturbed AS55_Q and broke the MICH locking, so I had to carefully misalign SRM so that the SRC fringes are small enough to maintain the lock of MICH.

 

(Looking for a good signal port for SRC)

 After I locked the PRMI with slightly misaligned SRM, I started looking for a good signal port for SRC.

At the beginning I tried finding a good SRC port by shaking SRM at 100 Hz and looked at the power spectra of all the available LSC signals.

I was expecting to see a 100 Hz peak in the spectra, but this technique didn't work well because SRC wasn't within the linear range and hence didn't produce linear signals.

So I didn't see any strong signals at 100 Hz and finally gave up this technique.

Then I started looking for a PDH-like signal in time series and immediately found AS55_I showed large PDH-like signals.

So I started using the AS55_I for the SRC locking and eventually succeeded.
 

 

(Two tips for the DRMI locking)

During the locking of DRMI, I found two tips that made the locking quite smooth.

 - Triggered locking

   Since every LSC signal ports showed large signals from PRC somehow, feeding back the signals made the suspensions crazy.

   So I used triggered locking for the PRC and MICH locking to avoid unwanted kicks on BS and PRM.

   If  the DC of REFL goes above a certain level, the control of  PRC starts. Also if the DC of AS goes below a certain level the control of MICH starts.

  These triggers make the lock smoother.

 - Do not use resonant gain filters

  This is really a stupid tip. When I was trying to lock MICH, the lock became quite difficult for some reasons.

  It looked there was an oscillation at 3 Hz every time the MICH control started. It turned out that a 3 Hz resonant gain filter had been making it difficult.

  All the resonant gain filters should be off when a lock acquisition is taken place.

Quote from #4759

Eventually the DRMI was locked.

More details will be reported in the morning.

 

  4761   Mon May 23 14:28:23 2011 kiwamuConfigurationLSCPOY55 installed

Last Saturday the POY55 RFPD (see this entry) was installed on the ITMY optical bench for the trial of the DRMI locking.

Since the amount of the light coming into the diode is tiny, the DC monitor showed ~ 3 mV even when the PRC was locked to the carrier.

In order to amplify the tiny RF signal from the photo diode a ZHL amplifier was installed next to the RFPD. The RF amp is sitting on delrin posts for insulation from the table.

POY55.png

  4762   Mon May 23 18:10:41 2011 kiwamuUpdateLSCf2p filters on PRM : not good

During the DRMI trial I noticed that the f2p filters on PRM is not quite effective (i.e. pushing PRM in POS direction makes misalignments).

I checked the f2p filters in an easy way. I pushed POS at 0.01 Hz with an amplitude of 1000 counts and looked at the oplev error signals with / without the f2p filters.

The picture below is a time series of the POS excitation, the oplev's PITCH and YAW error signals.

You can see there still is a big coupling from POS to YAW after the f2p filters were enabled. (Its supposed to be like this)

I will redo the f2p measurement on PRM.

f2p_PRM.png

  4763   Mon May 23 18:16:42 2011 KojiConfigurationLSCPOY55 installed

The DC Transimpedance of POP55 was increased from 50 Ohm to 10010 Ohm. There is the offset of 46mV. This should be cancelled in the CDS.

Quote:

Last Saturday the POY55 RFPD (see this entry) was installed on the ITMY optical bench for the trial of the DRMI locking.

Since the amount of the light coming into the diode is tiny, the DC monitor showed ~ 3 mV even when the PRC was locked to the carrier.

In order to amplify the tiny RF signal from the photo diode a ZHL amplifier was installed next to the RFPD. The RF amp is sitting on delrin posts for insulation from the table.

 

  4766   Wed May 25 20:12:55 2011 JamieConfigurationLSCNew digital lockin added to LSC model/screen

Quote:

This is OK....but, the input matrix should come from the same place as the regular input matrix: i.e. it should be just another row like CARM, DARM, etc. rather than have its own screen.

 You're absolutely right.  That was a brain-fart oversight.  I fixed the model so that the input from the lockin comes from another output row in the RFPD input matrix.  I then fixed the C1LSC medm screen accordingly:

lsc-lockin-adl.png

This is obviously much simpler and more straight-forward.

A future improvement would be to modify the DCPD input matrix to be able to route those signals to the lockin as well.  This is actually currently possible since the DCPD input matrix is just a subset of the full input matrix, but it's not available via medm yet.

  4767   Thu May 26 17:10:21 2011 JamieConfigurationLSCNew digital lockin added to LSC model/screen

Quote:

A future improvement would be to modify the DCPD input matrix to be able to route those signals to the lockin as well.  This is actually currently possible since the DCPD input matrix is just a subset of the full input matrix, but it's not available via medm yet.

 I went ahead and added the lockin output to the DCPD input matrix.

 

  4768   Fri May 27 17:52:53 2011 steveUpdateLSCLSC rack cables strain relieved & labeled

LSC rack 1Y2 cables are strain relieved and labeled. Spare and/or obsolete cables are laid out on the top of the beam tube and on the outside of the rack.

The POY 110 MHZ demodboard has a very touchy position in the VME crate. Watch out for it! It has to be fixed.

ELOG V3.1.3-