ID |
Date |
Author |
Type |
Category |
Subject |
3966
|
Mon Nov 22 18:39:53 2010 |
Jenne | Update | SUS | ETMU07: magnets glued to optic. ETMU05: magnets removed |
[Suresh, Jenne]
A story about minor disasters, and crises averted:
Once upon a time, in a cleanroom not so far away..... there lived an optic. To preserve anonymity, we shall call him "ETMU05". This optic had a rough day. When removing the grippers from the magnet-to-optic fixture, 4 out of 6 magnets broke off the dumbbells (the dumbbells were still securely glued to the optic...these had come out of the same batch that had problems last week, same problem). The remaining 2, LL and LR, were sadly of the same polarity. This is bad, because it means that the "humans" taking care of "ETMU05" didn't check the polarity of the face magnets properly, and ensure that they were laid out in an every-other pattern (LL and UR having the same polarity, and LR and UL having the opposite). So, the humans removed all magnets and dumbbells from ETMU05. All remaining glue was carefully scrubbed off the surfaces of ETMU05 using lens paper and acetone, and the magnets and dumbbells were sonicated in acetone, scrubbed with a lint-free wipe, sonicated again, and then scrubbed again to remove the glue. ETMU05 had a nice cleansing, and was drag wiped on both the AR and HR surfaces with acetone and iso. ETMU05 is now on vacation in a nice little foil hut.
His friend, (let's call him ETMU07) had a set of magnets (with polarities carefully confirmed) glued to him. The cleaned magnets and dumbbells removed from ETMU05 were reglued to their dumbbells, and should be dry by tomorrow.
.....And then they lived happily ever after. The End.
The revised schedule / status table:

|
3970
|
Mon Nov 22 20:31:58 2010 |
kiwamu | Update | Green Locking | searching for unknown loss in green PD path |
As I said in the past entry (see this entry), there was unknown loss of about 20dB in the beat detection path.
So I started fully characterizing the beat detection path.
Today I measured the frequency response of the wideband RFPD using the Jenne Laser.
Since all the data were taken by using a 1064nm laser, the absolute magnitudes [V/W] for 532nm are not calibrated yet.
I will calibrate the absolute values with a green laser which has a known power.

The data were taken by changing the bias voltage from -150V to 0V.
The shape of the transfer function looks quite similar to that Hartmut measured before (see the entry).
It has 100MHz bandwidth when the bias voltage is -150V, which is our normal operation point.
Theoretically the transfer function must keep flat at lower frequency down to DC.
Therefore for the calibration of this data, we can use the DC signal when a green beam with a known power is illuminating the PD.
|
3971
|
Tue Nov 23 01:27:33 2010 |
Kevin | Update | Electronics | POX Characterizations |
I measured the RF transimpedance of the POX photodiode by measuring the optical transfer function with the AM laser and by measuring the shot noise with a light bulb. The plots of these measurements are at http://lhocds.ligo-wa.caltech.edu:8000/40m/Electronics/POX.
I measured the noise of the photodiode at 11 MHz for different light intensities using an Agilent 4395a. The noise of a 50 ohm resistor as measured by this spectrum analyzer is 10.6 nV/rtHz. I fit this noise data to the shot noise formula to find the RF transimpedance at 11 MHz to be (2.42 ± 0.08) kΩ. The RF transimpedance at 11 MHz as measured by the transfer function is 6.4 kΩ. |
3972
|
Tue Nov 23 01:45:47 2010 |
kiwamu | Update | Green Locking | noise curve of wideband RFPD |
Quote: #3970 |
So I started fully characterizing the beat detection path.
|
As a part of the characterization works, I measured the spectra of the RFPD noise as well.
The noise is totally dominated by that of the RFPD (i.e. not by an RF amplifier).
I am going to check the noise curve by comparing with a LISO model (or a simple analytical model) in order to make sure the noise is reasonable.
(results)

The red curve represents the dark noise of the RFPD, which is amplified by a low noise amp, ZFL-1000LN.
The blue curve is a noise of only ZFL-1000LN with a 50 Ohm terminator at its input.
The last curve is noise of the network analyzer AG4395A itself.
It is clear that the noise is dominated by that of RFPD. It has a broad hill around 100MHz and a spike at 16MHz.
(notes)
Gain of ZFL-1000LN = 25.5 dB (measured)
Applied voltage to ZFL-1000LN = +15.0 V
Bias voltage on PD = -150 V |
3973
|
Tue Nov 23 10:48:31 2010 |
Koji | Update | IOO | the plan of the day |
[Kiwamu/Koji]
- The tanks are open
Plan
[done] - Remove the PZT cable currently underlying between BS and ITMY chambers
[done] - Put this PZT cable between BS and IMC chambers. Connect it on the PZT on the IMC table (SM1)
[done]- Put the two OSEM cables between BS and ITMY chambers. Connect this cable to SRM.
The connector for this cable at the BS side is coming from Bob's place on Wednesday. We left it disconnected for now.
- Energize all of four PZTs and check the functionality. |
3974
|
Tue Nov 23 10:53:20 2010 |
josephb | Update | CDS | timing issues |
Problem:
Front ends seem to be experiencing a timing issue. I can visibly see a difference in the GPS time ticks between models running on c1ioo and c1sus.
In addition, the fb is reporting a 0x2bad to all front ends. The 0x2000 means a mismatch in config files, but the 0xbad indicates an out of sync problem between the front ends and the frame builder.
Plan:
As there are plans to work on the optic tables today and suspension damping is needed, we are holding off on working on the problem until this afternoon/evening, since suspensions are still damping. It does mean the RFM connections are not available.
At that point I'd like to do a reboot of the front ends and framebuilder and see if they come back up in sync or not. |
3975
|
Tue Nov 23 11:20:30 2010 |
josephb | Update | CDS | Cleaning up old target directory |
Winter Cleaning:
I cleaned up the /cvs/cds/caltech/target/ directory of all the random models we had built over the last year, in preparation for the move of the old /cvs/cds/caltech/target slow control machine code into the new /opt/rtcds/caltech/c1/target directories.
I basically deleted all the directories generated by the RCG code that were put there, including things like c1tst, c1tstepics, c1x00, c1x00epics, and so forth. Pre-RCG era code was left untouched. |
3977
|
Tue Nov 23 14:52:28 2010 |
Jenne | Update | PEM | PEM Model Started |
Joe showed me what was what with adding DAQ channels, and I have begun building a simulink model to acquire the PEM channels.
My models is in: /cvs/cds/rtcds/caltech/c1/core/advLigoRTS/src/epics/simLink/c1pem.mdl
Next on the to do list in this category: test which input connector goes with which channel (hopefully it's linear, exactly as one would think), and give the channels appropriate names. |
3978
|
Tue Nov 23 16:55:14 2010 |
josephb | Update | CDS | Updated apps |
Updated Apps:
I created a new setup script for the newest build of the gds tools (DTT, foton, etc), located in /opt/apps (which is a soft link from /cvs/cds/apps) called gds-env.csh.
This script is now sourced by cshrc.40m for linux 64 bit machines. In addition, the control room machines have a soft link in the /opt directory to the /cvs/cds/apps directory.
So now when you type dtt or foton, it will bring up the Centos compiled code Alex copied over from Hanford last month. |
3979
|
Tue Nov 23 18:08:28 2010 |
Jenne | Update | SUS | ETMU07: Balanced, standoff glued. ETMU05: Magnets glued to optic |
[Koji, Jenne]
ETMU07 had its wire winched to the correct height, was balanced, standoff glued. Can be ready for going into the oven tomorrow, if an oven is available. (One of Bob's ovens has a leak, so he's down an oven, which puts everything behind schedule. We may not be able to get anything into the oven until Monday).
ETMU05 had magnets glued to the optic. Hopefully tomorrow we will winch the wire and balance the optic, and glue the standoff, and be ready to go into the oven on Monday.
The spring plungers were sonicated, but have not yet been baked. I told Daphen that we'd like the optics baked first, so that we can get ETMX in the chamber ASAP, and then the spring plungers as soon as possible so that we can install ETMY and put the OSEMs in.
The updated status table:

|
3980
|
Tue Nov 23 22:01:49 2010 |
kiwamu | Update | IOO | immigration plan for in-vac PZT mirrors |
[Suresh, Kiwamu]
We found that two of three PZT mirrors are at wrong place in the chambers.
Therefore we have to move these PZT mirrors together with their connections.
Here is a diagram for the current situation and the plan.

Basically mirror (A) must be associated to the output beam coming out from the SRM, but it was incorrectly put as a part of the input optics.
Similar to that, mirror (C) must belong to the input optics, but it is incorrectly being used as a part of OMC stuff.
Therefore we have to swap the positions of mirror (A) and mirror (C) as shown in the diagram above.
In addition to the mirror immigration, we also have to move their cables as well in order to keep the right functions.
We took a look at the length of the cables outside of the chambers in order to check if they are long enough or not.
And we found that the cables from c1asc (green line in the diagram) is not long enough, so we will put an extension D-sub cable. |
3981
|
Tue Nov 23 22:45:59 2010 |
kiwamu | Update | Computers | Altium |
I installed and activated Altium, a PCB design software, on the Windows machine M2.
With Altium I am going to design the triple resonant circuit for the broadband EOM. |
3983
|
Tue Nov 23 23:52:49 2010 |
rana | Update | CDS | Updated apps |
Wow. I typed DTT on rossa and it actually worked! No complaints about testpoints, etc. I was also able to use its new 'NDS2' function to get data off of the CIT cluster (L1:DARM_ERR from February). You have to use the kinit/kdestroy stuff to use NDS2 as usual (look up NDS2 in DASWG if you don't know what I mean). |
3984
|
Wed Nov 24 17:57:24 2010 |
Jenne | Update | SUS | ETMU07: Baking. ETMU05: Needs side magnets reglued |
[Jenne, Koji]
We removed ETMU07 from the suspension tower, after confirming that the balance was still good. Bob put it in the oven to bake over the weekend. The spring plungers and our spare magnets are all in there as well.
I tried to remove the grippers from ETMU05, and when I did, both side dumbbells came off of the optic. Unfortunately, I was working on getting channels into the DAQ, so I did not clean and reglue ETMU05 today. However Joe told me that we don't have any ETMY controls as yet, and we're not going to do Yarm locking (probably) in the next week or so, so this doesn't really set any schedules back.
The cleaning of ETMU05 will be tricky. Getting the residual glue off of the optic will be fine, but for the dumbbells, we'd like to clean the glue off of the end of the dumbbells using a lint free wipe soaked in acetone, but we don't want to get any acetone in the magnet-to-dumbbell joint, and we don't want to break the magnet-to-dumbbell joint. So we'll have to be very careful when doing this cleaning.
The Status Table:

|
3985
|
Wed Nov 24 18:11:26 2010 |
Jenne | Update | PEM | Some progress on getting PEM channels |
I have made a little bit of progress on the PEM channels. I have begun writing up detailed instructions in the DAQ Wiki page on how to add a channel to the new DAQ system. I have followed those instructions thus far, and can see my channels in the .ini file (and in the daqconfig gui thing), but I don't have any channels in Dataviewer or DTT.
There are some tricky "gotchas" involved in creating new models and channels. Some examples include: No use of the characters "DUMMY" in any channel name. The makefile is specifically hardcoded to fail if that string of characters is used. Also, you must have at least 2 filter banks in every model. Why? No one knows. You just do. The model won't compile unless you have 2 or more filter banks.
My efforts today included ~3 reboots of the frame builder, and ~2 reboots of c1sus. When Kiwamu and I rebooted c1sus, we burt restored to some time in the last 24 hrs. Some of the SUS filters on some of the optics were not set correctly (things like the bounce roll filter), so we turned all of them on, and reset all of the input and output matricies to be the correct combination of +1 and -1's to make Pit, Pos and Yaw. The tuning seems to happen now-a-days in the gains for each DOF, and the gains were set correctly by the burt restore for every optic except PRM. We made some educated guesses for what the gains should be based on the other optics, and PRM is damping pretty well (these guesses included reducing the SIDE gain by ~10 from the BS SIDE value, since the analog gain of the PRM SIDE sensor is much higher than others). We'll have to fine tune these gains using some Yuta-developed method soon. Or find a burt snapshot that had some non-unity values in there. |
3986
|
Thu Nov 25 02:49:39 2010 |
kiwamu | Update | CDS | installation of C1LSC: still going on |
[Joe, Kiwamu]
We tried installing C1LSC but it's not completely done yet due to the following issues.
(1) A PCIe optical fiber which is supposed to connect C1LSC and its IO chasis is broken at a high probability.
(2) Two DAC boards (blue and golden board) are missing.
We will ask the CDS people at Downs and take some more of those stuff from there.
( works we did )
- took the whole C1ASC crate out from the 1Y3 rack.
- installed an IO chasis to the place where C1ASC was.
- strung a timing optical fiber to the IO chasis.
- checked the functionality of the PICe optical fiber and found it doesn't work.
Fig.1 c1asc taken out from the rack Fig.2 IO chasis installed to the rack

Fig.3 PCIe extension fiber (red arrow for an obvious bended point)
|
3987
|
Fri Nov 26 16:37:29 2010 |
kiwamu | Update | Photos | pictures on PIcasa |
I uploaded some pictures taken in the last and this week. They are on the Picasa web albums.
in vac work [Nov. 18 2010]
in vac work [Nov 23 2010]
CDS work [Nov 24 2010]
 |
3988
|
Mon Nov 29 11:51:41 2010 |
kiwamu | Update | IOO | swaped in-vac PZT mirrors |
This morning I opened the chambers and started some in-vac works.
As explained in this entry, I swapped pzt mirror (A) and (C) successfully.
The chambers are still open, so don't be surprised.
(today's missions for IOO)
- cabling for the pzt mirrors
- energizing the pzt mirrors and slide them to their midpoint.
- locking and alignment of the MC
- realignment of the pzt mirrors and other optics.
- letting the beam go down to the arm cavity
|
3989
|
Mon Nov 29 17:45:28 2010 |
Zach | Update | elog | restarted elog |
The elog was down so I restarted it. The instructions on the wiki do not work as the process has some complicated name (i.e. it is not just 'elogd'). I used kill and the pid number.
I will get around to updating the restart script to work with 2.8.0. |
3991
|
Mon Nov 29 22:50:07 2010 |
Suresh | Update | SUS | ETMU05 Side Magnets glued back |
[Suresh, Jenne]
ETMU05 : Gluing Side magnets back on to the optic.
The following steps taken in this process:
1) The two magnet+dumbell units which had come loose from the optic needed to be cleaned. A lint free wipe was placed on the table top and a few cc of acetone was poured on to it. The free end of the dumbbell was then scrubbed on this wipe till the surface regained its shine. The dumbell was held at its narrow part with a forceps to avoid any strain on the magnet-dumbbell joint.
2) The optic was then removed from its gluing fixture (by loosening only one of the three retaining screws) and placed in an Al ring. The glue left behind by the side magnets was scrubbed off with a optical tissue wetted with Acetone.
3) The optic was returned to the gluing fixture. The position of the optic was checked by inserting the brass portion of the gripper and making sure that the face magnets are centered in it [Jenne doubled checked to be sure we got everything right].
4) The side magnets were glued on and the optic in the fixture has been placed in the foil-house.
If all goes well we will be able to balance the ETMU05 and give it to Bob for baking.
ETMU07 : It is still in the oven and we need to ask Bob to take out. It will be available for installation in the 40m tomorrow.
|
3992
|
Tue Nov 30 04:22:52 2010 |
kiwamu | Update | IOO | swaped in-vac PZT mirrors |
As a result of the vacuum work, now the IR beam is hitting ETMX.
The spot of the transmitted beam from the cavity can be found at the end table by using an IR viewer.
Quote: #3988 |
(today's missions for IOO)
- cabling for the pzt mirrors
- energizing the pzt mirrors and slide them to their midpoint.
- locking and alignment of the MC
- realignment of the pzt mirrors and other optics.
- letting the beam go down to the arm cavity
|
|
3993
|
Tue Nov 30 11:44:36 2010 |
Zach | Update | elog | elog restarted -- SCRIPT adapted for 2.8.0 |
I have created an updated version of the "start-elog-nodus" script and put it in .../elog/elog-2.8.0. It seems to work fine. |
3994
|
Tue Nov 30 12:10:44 2010 |
josephb | Update | elog | Elog restarted again |
The elog seemed to be down at around 12:05pm. I waited a few minutes to see if the browser would connect, but it did not.
I used the existing script in /cvs/cds/caltech/elog/ (as opposed to Zach's new on in elog/elog-2.8.0/) which also seems to have worked fine. |
3995
|
Tue Nov 30 12:25:08 2010 |
josephb | Update | CDS | LSC computer to chassis cable dead |
Problem:
We seemed to have a broken fiber link for use between the LSC and its IO chassis. It is unclear to mean when this damage occurred. The cable had been sitting in a box with styrofoam padding, and the kink is in the middle of the fiber, with no other obvious damage near by. The cable however may have previously been used by the people in Downs for testing and possibly then. Or when we were stringing it, we caused a kink to happen.
Tried Solutions:
I talked to Alex yesterday, and he suggested unplugging the power on both the computer and the IO chassis completely, then plugging in the new fiber connector, as he had to do that once with a fiber connection at Hanford. We tried this this morning, however, still no joy. At this point I plan to toss the fiber as I don't know of any way to rehabilitate kinked fibers.
Note this means that I rebooted c1sus and then did a burt restore from the Nov/30/07:07 directory for c1suspeics, c1rmsepics, c1mcsepics. It looks like all the filters switched on.
Current Plan:
We do, however, have the a Dolphin fiber which originally was intended to go between the LSC and its IO chassis, before Rolf was told it doesn't work well that way. However, we were going to connect the LSC machine to the rest of the network via Dolphin.
We can put the LSC machine next to its chassis in the LSC rack, and connect the chassis to the rest of the front ends by the Dolphin fiber. In that case we just need the usual copper style cable going between the chassis and the computer.
|
3997
|
Tue Nov 30 12:45:36 2010 |
kiwamu | Update | SUS | found a loose connection : ITMX damping |
Last night I found that the response of ITMX against the angle offsets were strage.
Eventually I found a loose connection at the feedthrough connectors of ITMX chamber.
So I pushed the connector hard, and then ITMX successfully became normal.
It looked like someone had accidentally kicked the cable during some works.
This bad connection had made unacceptable offsets in the OSEM readout, but now they seem fine. |
3998
|
Tue Nov 30 14:21:21 2010 |
Zach | Update | elog | Script unnecessary |
I didn't realize that the script in .../elog was configured to start 2.8.0 already. I will revert the instructions on the wiki to point to that
one. |
3999
|
Tue Nov 30 16:02:18 2010 |
josephb | Update | CDS | status |
Issues:
1) Turns out the /opt/rtcds/caltech/c1/target/gds/param/testpoint.par file had been emptied or deleted at one point, and the only entry in it was c1pem. This had been causing us a lack of test points for the last few days. It is unclear when or how this happened. The file has been fixed to include all the front end models again. (Fixed)
2) Alex and I worked on tracking down why there's a GPS difference between the front ends and the frame builder, which is why we see a 0x4000 error on all the front end GDS screens. This involved several rebuilds of the front end codes and reboots of the machines involved. (Broken)
3) Still working on understanding why the RFM communication, which I think is related to the timing issues we're seeing. I know the data is being transferred on the card, but it seems to being rejected after being red in, suggesting a time stamp mismatch. (Broken)
4) The c1iscex binary output card still doesn't work. (Broken)
Plan:
Alex and I will be working on the above issues tomorrow morning.
Status:
Currently, the c1ioo, c1sus and c1iscex computers are running with their front ends. They all still have 0x4000 error. However, you can still look at channels on dataviewer for example. However, there's a possibility of inconsistent timing between computer (although all models on a single computer will be in sync).
All the front ends where burt restorted to 07:07 this morning. I spot checked several optic filter banks and they look to have been turned on. |
4000
|
Tue Nov 30 18:15:52 2010 |
Jenne | Update | SUS | ETMX ready to be installed. ETMY ready for winching |
[Jenne, Kiwamu]
We put ETMX back in its tower, and confirmed its balance. It might be pointing a teensy bit upward, but it is way less than the DC pointing offset we see when we put the OSEMs in the towers (since the PDs and LEDs have some magnetic bits to them).
Discussions are ongoing as to where the ETM should sit on its table, but we'll probably toss it into the chamber later this evening.
I took ETMY out of the magnet gluing fixture, and put it in a ring, in the foil house. It is ready to have the wire winched and get balanced at our convenience.
The updated status table:

|
4001
|
Tue Nov 30 19:35:09 2010 |
Zach | Update | elog | restarted again |
Restarted the elog again, this time using .../elog/start-elog.csh, which Joe pointed out works just fine. I have amended the wiki instructions to point to this script, instead. |
4002
|
Wed Dec 1 02:39:00 2010 |
Suresh | Update | SUS | Installation of ETMU07 as ETMX |
[Kiwamu, Jenne, Koji, Suresh]
The following steps in this process were completed.
1) Secured the current ETMX (Old ETMY) with the earth quake stops.
2) Removed the OSEMs and noted the Sl no. of each and its position
3) Placed four clamps to mark the location of the current ETMX tower (Old ETMY's position on the table)
4) Moved the ETMX (Old ETMY) tower to the clean table flow bench. In the process the tower had to be tilted during removal because it was too tall to pass upright through the vacuum chamber port. It was scary but nothing went wrong.
5) Koji calculated the location of the new ETMX and told us that it should be placed on the north end of the table.
6) Moved the OSEM cables, counter balancing weights and the 'chopper' out of the way. Had to move some of the clamps securing the cables.
7) Moved the ETMU07 tower from the clean room to the ETMX table
8) Positioned the OSEMs as they were placed in the earlier tower and adjusted their position to the middle of the range of their shadow sensors. The four OSEMs on the face did not give us any trouble and were positioned as required. But the side OSEM could not be put in place. The magnet on the left side, which we are constrained to use since the tower is not designed to hold an OSEM on the right side, seems a little too low (by about a mm) and does not interrupt the light beam in the shadow sensor. The possible causes are
a) the optic is rotated. To check this we need to take the tower back to the clean room and check the location of the optic with the traveling microscope. If indeed it is rotated, this is easy to correct.
b) the magnet is not located at the correct place on the optic. This can also be checked on the clean room optical bench but the solution available immediately is to hold the OSEM askew to accommodate the magnet location. If time permits the magnet position can be corrected.
We have postponed the testing of the ETMU07 tower to 1st of Nov Dec.
|
4003
|
Wed Dec 1 12:02:49 2010 |
josephb, alex | Update | CDS | Rebuilding frame builder with latest code |
Problem:
The front ends seem to have different gps timestamps on the data than the frame builder has when receiving them.
One theory is we have fairly been doing SVN checkouts of the code for the front ends once a week or every two weeks, but the frame builder has not been rebuilt for about a month.
Current Action:
Alex is currently rebuilding the frame builder with the latest code changes.
It also suggests I should try rebuilding the frame builder on a semi-regular basis as updates come in.
|
4004
|
Wed Dec 1 13:41:21 2010 |
josephb, alex, rolf | Update | CDS | Timing is back |
Problem:
We had timing problems across the front ends.
Solution:
Noticing that the 1PPS reference was not blinking on the Master Timing Distribution box. It was supposed to be getting a signal from the c0dcu1 VME crate computer, but this was not happening.
We disconnected the timing signal going into c0dcu1, coming from c0daqctrl, and connected the 1PPS directly from c0daqctrl to the Ref In for the Master Timing distribution box (blue box with lots of fibers coming out of it in 1X5).
We now have agreement in timing between front ends.
After several reboots we now have working RFM again, along with computers who agree on the current GPS time along with the frame builder.
Status:
RFM is back and testpoints should be happy.
We still don't have a working binary output for the X end. I may need to get a replacement backplane with more than 4 slots if the 1st slot of this board has the same problem as the large boards.
I have burt restored the c1ioo, c1mcs, c1rms, c1sus, and c1scx processes, and optics look to be damped.
|
4006
|
Thu Dec 2 02:50:12 2010 |
kiwamu | Update | SUS | ETMX installed |
[Suresh, Kiwamu]
We finished the installation of ETMX into the chamber.
In order to clear the issue of the side OSEM, we put a spacer such that the OSEM can tilt itself and accommodate the magnet.
Though we still don't fully understand why the side magnet is off from the center.
Anyway we are going to proceed with this ETMX and perform the REAL green locking.
(what we did)
- took the ETM tower out from the chamber, and brought it to the clean room again.
- checked the rotation of the ETM by using a microscope. It was pretty good.
The scribe lines at the both sides are at the same height within the diameter of the scribe line.
- checked the height of the ETM by measuring the vertical distance from the table top to the scribe line. This was also quite good.
The height is correctly 5.5 inch within the diameter of the scribe line.
- checked the magnet positions compared with the OSEM holder holes.
All the face magnets are a little bit off upward (approximately by 1mm or less).
The side magnet is off toward the AR surface by ~ 1-2mm.
(yesterday we thought it was off downward, but actually the height is good.)
- raised the position of the OSEM holder bar in order to correct the miscentering of the face magnets.
Now all the face magnets are well centered.
- brought the tower back to the chamber again
- installed the OSEMs
We put a folded piece of aluminum foil in between the hole and the side OSEM as a spacer.
- leveled the table and set the OSEMs to their mid positions.
- slided the tower to place
|
4007
|
Fri Dec 3 05:21:11 2010 |
kiwamu | Update | Green Locking | locked the laser to the cavity |
I succeeded in locking the end green laser to X arm with the new ETM.
Though the lock is still not so stable compared to the previous locking with the old ETM. Also the beam centering is quite bad now.
So I will keep working on the end green lock a little bit more.
Once the lock gets improved and becomes reasonably stiff, we will move onto the corner PLL experiment.
(to do)
- beam centering on ITMX
- check the mode matching
- revise the control servo |
4008
|
Fri Dec 3 14:34:23 2010 |
rana | Update | CDS | fooling around in the FB rack |
This morning (~0100) I started to redo some of the wiring in the rack with the FB in it. This was in an effort to activate the new Megatron (Sun Fire 4600) which we got from Rolf.
Its sitting right above the Frame Builder (FB). The fibers in there are a rats nest. Someone needs to team up with Joe to neaten up the cabling in that rack - its a mini-disaster.
While fooling around in there I most probably disturbed something, leading to the FB troubles today. |
4009
|
Fri Dec 3 15:37:10 2010 |
josephb | Update | CDS | fb, front ends fixed - tested RFM between c1ioo and c1iscex |
Problem:
The front ends and fb computers were unresponsive this morning.
This was due to the fb machine having its ethernet cable plugged into the wrong input. It should be plugged into the port labeled 0.
Since all the front end machines mount their root partition from fb, this caused them to also hang.
Solution:
The cable has been relabled to "fb" on both ends, and plugged into the correct jack. All the front ends were rebooted.
Testing RFM for green locking:
I tested the RFM connection between c1ioo and c1scx. Unfortunately, on the first test, it turns out the c1ioo machine had its gps time off by 1 second compared to c1sus and c1iscex. A second reboot seems to have fixed the issue.
However, it bothers me that the code didn't come up with the correct time on the first boot.
The test was done using the c1gcv model and by modifying the c1scx model. At the moment, the MC_L channel is being passed the MC_L input of the ETMX suspension. In the final configuration, this will be a properly shaped error signal from the green locking.
The MC_L signal is currently not actually driving the optic, as the ETMX POS MATRIX currently has a 0 for the MC_L component. |
4014
|
Mon Dec 6 11:59:41 2010 |
josephb | Update | CDS | New c1lsc computer moved to lsc rack |
Computer moved:
The c1lsc computer has been moved over to the 1Y3 rack, just above the c1lsc IO chassis.
It will talking to the c1sus computer via a Dolphin PCIe reflected memory card. The cards have been installed into c1lsc and c1sus this morning.
It will talk to its IO chassis via the usual short IO chassis cable.
To Do:
The Dolphin fiber still needs to be strung between c1sus and c1lsc.
The DAQ cable between c1lsc and the DAQ router (which lets the frame builder talk directly with the front ends) also needs t to be strung.
c1lsc needs to be configured to use fb as a boot server, and the fb needs to be configured to handle the c1lsc machine. |
4015
|
Mon Dec 6 16:49:43 2010 |
josephb | Update | CDS | c1lsc halfway to working |
C1LSC Status:
The c1lsc computer is running Gentoo off of the fb server. It has been connected to the DAQ network and is handling mx_streams properly (so we're not flooding the network error messages like we used to with c1iscex). It is using the old c1lsc ip address (192.168.113.62). It can ssh'd into.
However, it is not talking properly to the IO chassis. The IO chassis turns on when the computer turns on, but the host interface board in the IO chassis only has 2 red lights on (as opposed to many green lights on the host interface boards in the c1sus, c1ioo, and c1iscex IO chassis). The c1lsc IO processor (called c1x04) doesn't see any ADCs, DACs, or Binary cards. The timing slave is receiving 1PPS and is locked to it, but because the chassis isn't communicating, c1x04 is running off the computer's internal clock, causing it to be several seconds off.
Need to investigate why the computer and chassis are not talking to each other.
General Status:
The c1sus and c1ioo computers are not talking properly to the frame builder. A reboot of c1iscex fixed the same problem earlier, however, as Kiwamu and Suresh are working in the vacuum, I'm leaving those computers alone for the moment, but a reboot and burt restore probably should be done later today for c1sus and c1ioo
Current CDS status:
MC damp |
dataviewer |
diaggui |
AWG |
c1ioo |
c1sus |
c1iscex |
RFM |
Dolphin RFM |
Sim.Plant |
Frame builder |
TDS |
|
|
|
|
|
|
|
|
|
|
|
|
|
4016
|
Mon Dec 6 22:18:39 2010 |
kiwamu | Update | Green Locking | aligned the beam axis |
[Suresh and Kiwamu]
We aligned the green beam to the X arm cavity more carefully.
Now the green beam is hitting the centers of ETMX, ITMX and BS.
Also we confirmed that the green beam successfully comes out from the chamber to the PSL table.
(what we did)
- opened the BS, ITMX and ETMX chambers.
- checked the positions of the beam spots on ITMX, BS and ETMX
The spot position on ETMX was fine,
But at BS and ITMX, the spots were off downward.
We decided to move the beam angle by touching a steering mirror at the end green setup.
- changed the beam axis by touching the steering mirror at the end station.
- checked the spot positions again, they all became good. It looks the errors were within ~ 1mm.
- moved the position of a TT, which is sitting behind the BS, by ~10mm, because it was almost clliping the beam.
- aligned the green optics
- got the beam coming out from the chamber.
|
4018
|
Mon Dec 6 23:33:15 2010 |
Jenne | Update | SUS | ETMU05 winched, balanced, glued!!!!!! |
[Suresh, Jenne]
We Finished!!!
ETMU05 (ETMY) had its wire winched to the correct height, was balanced, and had the standoff glued. Since it's kind of like final exam week at Caltech, Suresh had his suspension exam today, and did most of this work himself, with me hanging around and watching.
As you can see in my almost entirely green table, all that is left to do with the whole suspensions project is bake the optic (hopefully Bob has time / space this week), and then stick it in the chamber! Hooray!!! (Can you tell I'm excited to not spend too much more time in the cleanroom?)
The table:

|
4019
|
Tue Dec 7 12:12:40 2010 |
kiwamu | Update | CDS | added some more DAQ channels |
[Joe and Kiwamu]
We added some more DAQ channels on c1sus.
We wanted to try diagonalizing the input matrices of the ITMX OSEMs because the motion of ITMX looked noisier than the others
So for this purpose we tried adding DAQ channels so that we can take spectra anytime.
After some debugging, now they are happily running.
(DAQ activation code)
There is a code which activates DAQ channels written by Yuta in this October.
/cvs/cds/rtcds/caltech/c1/chans/daq/activateDAQ.py
If you just execute this code, it is supposed to activate the DAQ channels automatically by editing C1AAA.ini files.
However there were some small bugs in the code, so we fixed them.
Now the code seems fine.
(reboot fb DAQ process)
When new DAQ channels are added, one has to reboot the DAQ process running on fb.
To do this, log in to a certain port on fb,
telnet fb 8088
shutdown
Then the process will automatically recover by itself.
After doing the above reboo job, we found tpman on C1IOO got down.
We don't fully understand why only C1IOO was affected, but anyway rebooting of the c1ioo front end machine fixed the problem.
|
4020
|
Tue Dec 7 16:09:53 2010 |
josephb | Update | CDS | c1iscex status |
I swapped out the IO chassis which could only handle 3 PCIe cards with the another chassis which has space for 17, but which previously had timing issues. A new cable going between the timing slave and the rear board seems to have fixed the timing issues.
I'm hoping to get a replacement PCI extension board which can handle more than 3 cards this week from Rolf and then eventually put it in the Y-end rack. I'm also still waiting for a repaired Host interface board to come in for that as well.
At this point, RFM is working to c1iscex, but I'm still debugging the binary outputs to the analog filters. As of this time they are not working properly (turning the digital filters on and off seems to have no effect on the transfer function measured from an excitation in SUSPOS, all the way around to IN1 of the sensor inputs (but before measuring the digital fitlers). Ideally I should see a difference when I switch the digital filters on and off (since the analog ones should also switch on and off), but I do not. |
4021
|
Tue Dec 7 17:19:33 2010 |
kiwamu | Update | Computers | pyNDS available |
I moved the Pynds package from Yuta's local directory to the public place.
Now the package is living under :
/cvs/cds/caltech/apps/linux64/python/lib64/python2.4/site-packages
Also I added the PATH on cshrc.40m, so you don't have to setenv every time.
(This package can not run on non-64bit linux)
Typing "import nds" on python allows you to use the nds functions.
Quote from Yuta's past entry
Quote: #3722 |
3. I installed pyNDS to /cvs/cds/caltech/users/yuta/pynds
|
|
4022
|
Tue Dec 7 18:37:15 2010 |
Suresh | Update | SUS | ETMU05 ready for baking |
The ETMU05 has been removed from the suspension and put into the little foil house.
Before removing it I checked the position and pitch of the optic with reference to the table top.
The height:
Using the traveling microscope I checked the height of the scribe lines from the table top. They are at equal heights, centered on 5.5 inches, correct to about a quarter of the width of the scribe line.
The pitch
The retro-reflection of the He-Ne laser beam is correct to within one diameter of the beam at a distance of about 1.5m. This is the reflection from the rear, AR coated, surface. The reflection from the front, HR coated, surface was down by about two diameters.
Jenne has checked with Bob and agreed on a date for baking the optic.
|
4023
|
Tue Dec 7 19:34:58 2010 |
kiwamu | Update | CDS | rebooted DAQ and all the front end machines |
I found that all the front end machine showed the red light indicators of DAQ on the XXX_GDS_TP.adl screens.
Also I could not get any data from both test points and DAQ channels.
First I tried fixing by telneting and rebooting fb, but it didn't help.
So I rebooted all the front end machines, and then everything became fine.
|
4024
|
Tue Dec 7 20:38:17 2010 |
kiwamu | Update | SUS | watchdogs off at ITMX and ETMX |
I am leaving ITMX and ETMX freely swinging, so that later I can take the spectra and diagonalize the input matrices.
Please don't restore the watchdogs until tomorrow morning. |
4025
|
Wed Dec 8 12:26:56 2010 |
josephb | Update | CDS | megatron set up - as a test front end |
[josephb, Osamu]
Megatron Setup:
To show Osamu how to setup a a front end as well as provide a test computer for Osamu's use, we used the new megatron (sunfire x4600 with 16 cores and 8 gigabytes of memory) as a front end without an IO chassis.
The steps we followed are in the wiki, here.
The new megatron's IP address is 192.168.113.209. It is running the c1x99 front end code. |
4026
|
Wed Dec 8 12:47:18 2010 |
kiwamu | Update | SUS | diagonalisation of ITMX input matrix |
The input matrix of ITMX has been diagonalized.
The evaluation of this diagonalisation will be done tonight by freely swinging ITMX again.
(Somehow I couldn't get any data for ETMX from the DAQ channels. I will try it again tonight.)
(details)
For solving the matrix, I used Yuta's python code called inmartixoptimizer.py.
I took the transfer functions of UL->UR, UL->LL and UL->LR as described in this entry.
In the measurement, the frequency bin was set to 0.001 Hz and the data were 50 times averaged on dtt.
Here is the new input matrix.
[[ 0.87059649 1.14491977 1.07992057 0.90456317]
[ 0.64313916 0.55555661 -1.44997325 -1.35133098]
[ 1.13979571 -1.19186285 -0.89606597 0.77227546]]
This matrix should give a better performance than before. |
4027
|
Wed Dec 8 14:46:19 2010 |
josephb, kiwamu | Update | CDS | Why the ETMX daq channels were not recorded last night |
When adding the ETMX DAQ channels using the daqconfig gui (located in /opt/rtcds/caltech/c1/scripts/) on C1SCX.ini, we forgot to set the acquire flag to 1 from 0.
So the frame builder was receiving the data, but not recording it.
We have since then added ETMX and the C1SCX.ini file to Yuta's useful "activateDAQ.py" script in /opt/rtcds/caltech/c1/chans/daq/, so that it now sets the sensor and SUSPOS like channels to be acquired at 2k when run. You still need to restart the frame builder (telnet fb 8087 and then shutdown) for these changes to take effect.
The script now also properly handles files which already have had channels activated, but not acquired. |
4028
|
Wed Dec 8 14:51:09 2010 |
josephb | Update | CDS | c1pem now recording data |
Problem:
c1pem model was reporting all zeros for all the PEM channels.
Solution:
Two fold. On the software end, I added ADCs 0, 1, and 2 to the model. ADC 3 was already present and is the actual ADC taking in PEM information.
There was a problem noted awhile back by Alex and Rolf that there's a problem with the way the DACs and ADCs are number internally in the code. Missing ADCs or DACs prior to the one you're actually using can cause problems.
At some point that problem should be fixed by the CDS crew, but for now, always include all ADCs and DACs up to and including the highest number ADC/DAC you need to use for that model.
On the physical end, I checked the AA filter chassis and found the power was not plugged in. I plugged it in.
Status:
We now have PEM channels being recorded by the FB, which should make Jenne happier. |