2508
|
Tue Jan 12 09:37:05 2010 |
Alberto | Update | ABSL | IFO available |
I finished measuring the AbsL for this morning. The IFO is again available.
Please don't mess up with the interferometer though. I'll be back in a couple of ours |
2513
|
Wed Jan 13 12:03:00 2010 |
Alberto | Update | ABSL | Measurement now running. Please be careful |
At the moment I'm running a measurement on the PRC and I'm planning to leave it going for the time we'll be at the 40m meeting.
Please be careful in the lab. Thank you. |
2521
|
Mon Jan 18 18:34:01 2010 |
Alberto | Update | ABSL | Measurement in progress |
I started a long measurement of the PRC's transmissivity. I'm leaving the lab and I'm going to be back at about 8 tonight. Please do not disturb the interferometer. it is important that the MC and the PRC stay locked all the time. |
2522
|
Mon Jan 18 20:58:40 2010 |
Alberto | Update | ABSL | Measurement in progress |
Quote: |
I started a long measurement of the PRC's transmissivity. I'm leaving the lab and I'm going to be back at about 8 tonight. Please do not disturb the interferometer. it is important that the MC and the PRC stay locked all the time.
|
That measurement is finished. I'm now going to start another one that will take another hour or so. I'm leaving it running for the night. If you want to work on the IFO, it should be definitely done by 11pm. |
2531
|
Tue Jan 19 12:54:39 2010 |
Alberto | Update | ABSL | Measurement in progress |
A measurement will be running for the next hour. Please be careful. |
2532
|
Tue Jan 19 16:21:18 2010 |
Alberto | Update | ABSL | Watchdogs not working and then fixed |
This afternoon the watchdogs stopped working: they didn't trip when the suspension positions crossed the threshold values.
I rebooted c1susaux (aka c1dscl1epics0 in the 1Y5 rack), which is the computer that runs the watchdog processes.
The reboot fixed the problem. |
2534
|
Wed Jan 20 20:11:56 2010 |
Alberto | Update | ABSL | Some preliminary results from measuring PRC's transmissivity for an amplitude modulated beam |
Here I'm posting a plot showing a set of measurements that I made in the past few days to determine the absolute length of the PRC cavity.
As in my other AbsL measurements, I inject an auxiliary laser beam into the cavity and look at the transmission. In the PRC case, the beam is injected through the dark port and I look at a pick-off of the REFL beam. The aux laser is phase locked to the PSL beam and I control the differential frequency between the two. The PSL and the aux beam interfere and beat at their differential frequency.
The attached plot shows the transmitted power as a function of the beat frequency.
Fitting the data with the model would let me determine the cavity length.
By now I can estimate the length of the PRC at about 2.257m, but it's still a rather approximate value.
I can't provide accurate error bars yet. I need to optimize the measurements to get a more precise value.
I will go more through the details of the measurement technique and of the fitting function as soon as I have more definitive results.
The data points shown here were taken at different times and not always in optimal alignment condition of the PRC.
To get a good fit of the data I should have fewer frequency segments, taken in a shorter period of time, and in which the power circulating inside of the cavity (ie SPOB) fluctuates as little as possible.
For what regards the time needed for a measurements, I already significantly sped up the measurements (i.e. optimizing the scanning and acquisition GPIB scripts, and fixing a couple of problems with the PDH box used in the PLL), and finally now I can scan several tens of MHz in few minutes.
About the frequency segments, so far they have been determined by two factors
1) Tthe frequency generator in the PLL: the Marconi works as a continuous wave generator only in limited ranges. Switching from one to another brakes the wave in a way that causes the PLL to lose lock.
2) Getting below 18 MHz a series of other beats appear on the PLL photodiode and make the PLL lose lock.
For the first problem, I'm thinking of using two Marconis and to mix their signals. I would keep one at 300MHz and I would scan the other from 300MHz to 500MHz. In fat, in that frequency range the Marconi has not discontinuity.
To try to avoid the other beats at low frequency, I'm not entirely sure about what to do yet.
To be continued... |
Attachment 1: 2010-01-19_PRCtransmissivityVsModel.png
|
|
2536
|
Thu Jan 21 10:31:13 2010 |
Koji | Update | ABSL | Some preliminary results from measuring PRC's transmissivity for an amplitude modulated beam |
Nice and interesting plot.
I suppose slight decrease of the Schnupp asymmetry (in your model) adjusts the discrepancy in the high freq region.
At the same time, it will make the resonance narrower. So you need to put some loss at the recombination (=on the BS)?
...of course these depends on the flatness of the calibration. |
2541
|
Fri Jan 22 02:54:06 2010 |
Alberto | Update | ABSL | Overnight measurement |
I'm leaving a measurement running overnight. It should be done in about one hour.
Tomorrow morning, If you need to use the interferometer, and you don't want to have the auxiliary beam going onto the dark port, you can turn down the flipping mirror and close the NPRO's mechanical shutter. |
2543
|
Fri Jan 22 14:40:49 2010 |
Alberto | Update | ABSL | Overnight measurement |
Quote: |
I'm leaving a measurement running overnight. It should be done in about one hour.
Tomorrow morning, If you need to use the interferometer, and you don't want to have the auxiliary beam going onto the dark port, you can turn down the flipping mirror and close the NPRO's mechanical shutter.
|
This is what I measured last night:

This is not a fit. It's just a comparison of the model with the data. |
2545
|
Mon Jan 25 16:30:37 2010 |
Alberto | Update | ABSL | 166 MHz sideband turned off |
I turned off the modulation at 166MHZ becasue I don't need it if I'm only locking the PRC.
It was introducing extra amplitude modulations of the beam which interfered with the AbsL's PLL photodiode.
I'm going to turn it back on later on. |
2546
|
Mon Jan 25 16:46:33 2010 |
Alberto | Update | ABSL | 166 MHz sideband turned off |
Quote: |
I turned off the modulation at 166MHZ becasue I don't need it if I'm only locking the PRC.
It was introducing extra amplitude modulations of the beam which interfered with the AbsL's PLL photodiode.
I'm going to turn it back on later on.
|
I turned back on the 166MHz modulation just a bit. I set the slider at 4.156.
When it was totally off the MZ seemd quite unhappy. |
2547
|
Tue Jan 26 03:28:56 2010 |
rana | Update | ABSL | 166 MHz sideband turned off |
You can turn the 166 off if you want. MZ is unhappy after its turned off, but that's just the thermal transient from removing the RF heat. After a several minutes, the heat goes away and the MZ can be relocked.
One of these days we should evaluate the beam distortion we get in EOMs because of the RF heat induced dn/dT. Beam steering, beam size, etc. |
2550
|
Wed Jan 27 11:02:30 2010 |
Alberto | Update | ABSL | PRC Cavity Length |
I fitted the data from scanning the PRC by changing the beat frequency of the auxiliary laser beam with the PSL beam.
The data points that I've taken so far over the entire frequency range (0-300 MHz) are not continuous. For several reasons the PLL was unable to maintain lock for such a large range and I had to break it into smaller segments. The measurements to acquire them stretched over a too long period of time during which the status of the PRC changed.
Because of that, before I get a continuous set of data points (perhaps normalized by the circulating power inside of the cavity), I restricted the fit to a 55MHz range around 100MHz. I obtained the following numbers for the fit parameters:
Length PRC = 2.169 +/- 0.007 m
Schnupp Asymmetry: 0.471+/- 0.006 m
The fit is shown in the attached plot:
When I fit over the entire set of data I get this:
Length PRC = 2.224 +/- 0.005 m
Schnupp Asymmetry: 0.457+/- 0.005 m
The results are different. Evidently I have to improve the measurement. I'm working on it.
For posterity:
The function I used to fit the transmitted beat power vs. frequency is the following:
E_trans = - t_prm * r_itm * exp(1i*2*wb*l_prc/c) .* sin(wb*l_/c) ./ ( 1 + r_prm * r_itm * exp(1i*2*wb*l_prc/c) .* cos(wb*l_/c)
Where wb is the angular frequency of the beat, l_prc and l_ are the length of the PRC and the Schnupp asymmetry, respectively; r_itm, t_itm, r_prm, t_prm are reflectances and transmittances of PRM and ITM; c is the speed of light.
|
2553
|
Fri Jan 29 12:06:58 2010 |
Alberto | Update | ABSL | Measurement running today at lunch time |
I just started a measuremtn that will be running for the next hour or so. Please be careful with the interferometer. |
2554
|
Fri Jan 29 13:14:49 2010 |
Alberto | Update | ABSL | Measurement running today at lunch time |
Quote: |
I just started a measuremtn that will be running for the next hour or so. Please be careful with the interferometer.
|
Done. IFO available |
2576
|
Mon Feb 8 14:13:03 2010 |
Alberto | Update | ABSL | PLL Characterization |
Lately I've been trying to improve the PLL for the AbsL experiment so that it could handle larger frequency steps and thus speed up the cavity scan.
The maximum frequency step that the PLL could handle withouth losing lock is given by the DC gain of the PLL. This is the product of the mixer's gain factor K [rad/V ], of the laser's calibration C [Hz/V] and of the PLL filter DC gain F(0).
I measured these quantities: K=0.226 V/rad; C=8.3e6 Hz/V and F(0)=28.7dB=21.5. The max frequency step should be Delta_f_max = 6.4MHz.
Although in reality the PLL can't handle more than a 10 KHz step. There's probably some other effect that I'm not.
I'm attaching here plots of the PLL Open Loop Gain, of the PLL filter and of a spectra of the error point measured in different circumstances.
I don't have much time to explain here how I took all those measurements. After I fix the problem, I'm going to go go through those details in an elog entry.
Does anyone have any suggestion about what, in principle, might be limiting the frequency step?
I already made sure that both cables going to the mixer (the cable with the beat signal coming from the photodiode and the cable with the LO signal coming from the Marconi) had the same length. Although ideally, for phase locking, I would still need 90 degrees of phase shift between the mixing signals, over the entire frequency range for which I do the cavity scan. By now the 90 degrees are not guaranteed.
Also, I have a boost that adds another 20 dB at DC to the PLL's filter. Although it doesn't change anything. In fact, as said above calculating the frequency step, the PLL should be able to handle 100KHz steps, as I would want the PLL to do. |
Attachment 1: 2010-02-08_Old_PDH_Box_Filter_TF_gain_knob_0_Boost_OFF.png
|
|
Attachment 2: 2010-02-08_PLL_OLG_gain_knob_0_Boost_OFF.png
|
|
Attachment 3: 2010-02-08_PLL_Noise_Budget.png
|
|
2577
|
Mon Feb 8 14:56:17 2010 |
Alberto | Update | ABSL | Suddenly a much better alignment of PRC |
I just aligned PRM and locked PRC and I noticed that SPOB is much higehr than it used to be. It's now about 1800, vs 1200 than it used to be last week.
Isn't anyone related to that? If so, may I please know how he/she did it? |
2578
|
Mon Feb 8 15:01:46 2010 |
rob | Update | ABSL | Suddenly a much better alignment of PRC |
Quote: |
I just aligned PRM and locked PRC and I noticed that SPOB is much higehr than it used to be. It's now about 1800, vs 1200 than it used to be last week.
Isn't anyone related to that? If so, may I please know how he/she did it?
|
oops, my bad. I cranked the 33MHz modulation depth and forgot to put it back. The slider should go back to around 3. |
2579
|
Mon Feb 8 15:41:51 2010 |
Alberto | Update | ABSL | Suddenly a much better alignment of PRC |
Quote: |
Quote: |
I just aligned PRM and locked PRC and I noticed that SPOB is much higehr than it used to be. It's now about 1800, vs 1200 than it used to be last week.
Isn't anyone related to that? If so, may I please know how he/she did it?
|
oops, my bad. I cranked the 33MHz modulation depth and forgot to put it back. The slider should go back to around 3.
|
I was actually hoping that the alignment got better. |
2581
|
Tue Feb 9 09:07:06 2010 |
Alberto | Update | ABSL | PLL Characterization |
Quote: |
Lately I've been trying to improve the PLL for the AbsL experiment so that it could handle larger frequency steps and thus speed up the cavity scan.
The maximum frequency step that the PLL could handle withouth losing lock is given by the DC gain of the PLL. This is the product of the mixer's gain factor K [rad/V ], of the laser's calibration C [Hz/V] and of the PLL filter DC gain F(0).
I measured these quantities: K=0.226 V/rad; C=8.3e6 Hz/V and F(0)=28.7dB=21.5. The max frequency step should be Delta_f_max = 6.4MHz.
Although in reality the PLL can't handle more than a 10 KHz step. There's probably some other effect that I'm not.
I'm attaching here plots of the PLL Open Loop Gain, of the PLL filter and of a spectra of the error point measured in different circumstances.
I don't have much time to explain here how I took all those measurements. After I fix the problem, I'm going to go go through those details in an elog entry.
Does anyone have any suggestion about what, in principle, might be limiting the frequency step?
I already made sure that both cables going to the mixer (the cable with the beat signal coming from the photodiode and the cable with the LO signal coming from the Marconi) had the same length. Although ideally, for phase locking, I would still need 90 degrees of phase shift between the mixing signals, over the entire frequency range for which I do the cavity scan. By now the 90 degrees are not guaranteed.
Also, I have a boost that adds another 20 dB at DC to the PLL's filter. Although it doesn't change anything. In fact, as said above calculating the frequency step, the PLL should be able to handle 100KHz steps, as I would want the PLL to do.
|
I might have found the problem with the PLL that was preventing me from scanning the frequencies by 100KHz steps. A dumb flimsy soldering in the circuit was making the PLL unstable.
After I fixed that problem and also after writing a cleverer data acquisition script in Python, I was able to scan continuosly the range 10-200MHz in about 20min (versus the almost 1.5-2 hrs that I could do previously). I'm attaching the results to this entry.
The 'smears' on the right side of the resonance at ~33MHz, are due to the PSL's sideband. I think I know how to fix that.
As you can see, the problem is that the model for the cavity transmission still does not match very well the data. As a result, the error on the cavity length is too big (~> 10 cm - I'd like to have 1mm).
Anyway, that was only my first attempt of scanning. I'm going to repeat the measurement today too and see if I can come out better. If not, than I have to rethink the model I've been using to fit. |
Attachment 1: 2010-02-08_PRCtransmissivity_EntireFreqRange_VsFit.png
|
|
2582
|
Tue Feb 9 10:10:58 2010 |
Alberto | Update | ABSL | back to analog |
I want to try to do the measurement with the network analyzer used as local oscillator, instead of the Marconis that I'm using now. Tha could give me better noise rejection. It would also give me information about the phase.
Also I wouldn't dislike abandoning the GPIB interfaces to acquire data. |
2679
|
Thu Mar 18 10:46:51 2010 |
Koji | Update | ABSL | PLL reconstructed |
Last night (Mar 17) I checked the PLL setup as Mott have had some difficulty to get a clean lock of the PLL setting.
- I firstly found that the NPRO beam is not going through the Faraday isolator well. This was fixed by aligning the steering mirrors before the Faraday.
- The signal from the RF PD was send to the RF spectrum analyzer through a power splitter. This is a waist of the signal. It was replaced to a directional coupler.
- Tee-ing the PZT feedback to the oscilloscope was producing the noise in the laser frequency. I put the oscilloscope to the 600Ohm output of the SR560, while connectiong the PZT output to the 50Ohm output.
- In addition, 6dB+6dB attenuators have been added to the PZT feedback signal.
Now the beating signal is much cleaner and behave straight forward. I will add some numbers such as the PD DC output, RF levels, SR560 settings...
Now I am feeling that we definitely need the development of really clean PLL system as we use PLL everywhere! (i.e. wideband PD, nice electronics, summing amplifiers, stop poking SR560, customize/specialize PDH box, ...etc) |
2680
|
Thu Mar 18 12:27:56 2010 |
Alberto | Update | ABSL | PLL reconstructed |
Quote: |
Last night (Mar 17) I checked the PLL setup as Mott had some difficulty to get a clean lock of the PLL setting.
- I firstly found that the NPRO beam is not going through the Faraday isolator well. This was fixed by aligning the steering mirrors before the Faraday.
- The signal from the RF PD was send to the RF spectrum analyzer through a power splitter. This is a waist of the signal. It was replaced to a directional coupler.
- Tee-ing the PZT feedback to the oscilloscope was producing the noise in the laser frequency. I put the oscilloscope to the 600Ohm output of the SR560, while connectiong the PZT output to the 50Ohm output.
- In addition, 6dB+6dB attenuators have been added to the PZT feedback signal.
Now the beating signal is much cleaner and behave straight forward. I will add some numbers such as the PD DC output, RF levels, SR560 settings...
Now I am feeling that we definitely need the development of really clean PLL system as we use PLL everywhere! (i.e. wideband PD, nice electronics, summing amplifiers, stop poking SR560, customize/specialize PDH box, ...etc)
|
I also had noticed the progressive change of the aux NPRO alignment to the Farady.
I strongly agree about the need of a good and robust PLL.
By modifying the old PDH box (version 2008) eventually I was able to get a PLL robust enough for my purposes. At some point that wasn't good enough for me either.
I then decided to redisign it from scratch. I'm going to work on it. Also because of my other commitments, I'd need a few days/1 week for that. But I'd still like to take care of it. Is it more urgent than that? |
2681
|
Thu Mar 18 13:40:35 2010 |
Koji | Update | ABSL | PLL reconstructed |
We use the current PLL just now, but the renewal of the components are not immediate as it will take some time. Even so we need steady steps towards the better PLL. I appreciate your taking care of it.
Quote: |
Quote: |
Last night (Mar 17) I checked the PLL setup as Mott had some difficulty to get a clean lock of the PLL setting.
- I firstly found that the NPRO beam is not going through the Faraday isolator well. This was fixed by aligning the steering mirrors before the Faraday.
- The signal from the RF PD was send to the RF spectrum analyzer through a power splitter. This is a waist of the signal. It was replaced to a directional coupler.
- Tee-ing the PZT feedback to the oscilloscope was producing the noise in the laser frequency. I put the oscilloscope to the 600Ohm output of the SR560, while connectiong the PZT output to the 50Ohm output.
- In addition, 6dB+6dB attenuators have been added to the PZT feedback signal.
Now the beating signal is much cleaner and behave straight forward. I will add some numbers such as the PD DC output, RF levels, SR560 settings...
Now I am feeling that we definitely need the development of really clean PLL system as we use PLL everywhere! (i.e. wideband PD, nice electronics, summing amplifiers, stop poking SR560, customize/specialize PDH box, ...etc)
|
I also had noticed the progressive change of the aux NPRO alignment to the Farady.
I strongly agree about the need of a good and robust PLL.
By modifying the old PDH box (version 2008) eventually I was able to get a PLL robust enough for my purposes. At some point that wasn't good enough for me either.
I then decided to redisign it from scratch. I'm going to work on it. Also because of my other commitments, I'd need a few days/1 week for that. But I'd still like to take care of it. Is it more urgent than that?
|
|
2684
|
Thu Mar 18 21:42:26 2010 |
Koji | Update | ABSL | PLL reconstructed |
I checked the setup further more.
- I replaced the PD from NewFocus 1GHz one to Thorlabs PDA255.
- I macthed the power level of the each beam.
Now I have significant fraction of beating (30%) and have huge amplitude (~9dBm).
The PLL can be much more stable now.
Koji |
2696
|
Mon Mar 22 22:11:26 2010 |
Mott | Update | ABSL | PLL reconstructed |
It looks like the PLL drifted alot over the weekend, and we couldn't get it back to 9 dBm. We switched back to the new focus wideband PD to make it easier to find the beat signal. I replaced all the electronics with the newly fixed UPDH box (#17) and we aligned it to the biggest beat frequency we could get, which ended up being -27 dBm with a -6.3V DC signal from the PD.
Locking was still elusive, so we calculated the loop gain and found the UGF is about 45 kHz, which is too high. We added a 20 dB attenuator to the RF input to suppress the gain and we think we may have locked at 0 gain. I am going to add another attenuator (~6 dB) so that we can tune the gain using the gain knob on the UPDH box.
Finally, attached is a picture of the cable that served as the smb - BNC adaptor for the DC output of the PD. The signal was dependent on the angle of the cable into the scope or multimeter. It has been destroyed so that it can never harm another innocent experiment again! |
Attachment 1: IMG_0150.JPG
|
|
2697
|
Mon Mar 22 23:37:32 2010 |
Mott | Update | ABSL | PLL reconstructed |
We have managed to lock the PLL to reasonable stability. The RF input is attenuated by 26 dBm and the beat signal locks very close to the carrier of the marconi (the steps on the markers of the spectrum analyzer are coarse). We can use the marconi and the local boost of the pdh box to catch the lock at 0 gain. Once the lock is on, the gain can be increased to stabilize the lock. The locked signals are shown in the first photo (the yellow is the output of the mixer and the blue is the output to the fast input of the laser. If the gain is increased too high, the error signal enters an oscillatory regime, which probably indicates we are overloading the piezo. This is shown in the second photo, the gain is being increased in time and we enter the non-constant regime around mid-way through.
Tomorrow I will use this locked system to measure the PZT response (finally!). |
Attachment 1: 2010-03-22_23.14.00.jpg
|
|
Attachment 2: 2010-03-22_23.24.50.jpg
|
|
2703
|
Tue Mar 23 18:44:46 2010 |
Mott | Update | ABSL | PLL reconstructed |
After realigning and getting the lock today, I tried to add in the SR785 to measure the transfer function. As soon as I turn on the piezo input on the PDH box, however, the lock breaks and I cannot reacquire it. We are using an SR650 to add in the signal from the network analyzer and that has worked. We also swapped the 20 dB attenuator for a box which mimics the boost functionality (-20 dB above 100 Hz, 0 dB below 6Hz). I took some spectra with the SR750, and will get some more with the network analyzer once Alberto has finished with it.
The SR750 spectra is posted below. The SR750 only goes up to 100 kHz, so I will have to use the network analyzer to get the full range. |
Attachment 1: NPRO_PLL_freqresp.png
|
|
4832
|
Fri Jun 17 16:05:07 2011 |
kiwamu | Update | ABSL | LightWave out of MOPA box |
[Suresh / Kiwamu]
We did the following things :
- Took the LightWave NPRO out from the MOPA box
- Temporarily took out the laser controller which has been connected to the Y end laser.
- Put the LightWave on AP table and plugged the laser controller and confirmed that it still emits a beam

[Things to be done]
- measure the beam profiles and power
- get a laser controller, which will be dedicated for this laser, from Peter King
[Background and Motivation]
The PRC and SRC length have to be precisely measured before the vent.
In order to measure those absolute length we are going to use the Stochino technique, which requires another laser to scan the cavity profiles.
The LightWave NPRO laser in the MOPA box was chosen for the Stochino laser because it has a large PZT range of 5 MHz/V and hence allows us to measure a wider frequency range.
The laser in the MOPA box had been connected to home-made circuits, which are not handy to play with. So we decided to use the laser with the usual laser controller.
Peter King said he has a LightWave laser controller and he can hand it to us.
Until we get the controller from him we do some preparations with temporary use of the Y end laser controller. |
4840
|
Mon Jun 20 11:38:49 2011 |
kiwamu | Update | ABSL | I-P curve of LightWave M126-1064-700 |
The I-P curve of the LightWave NPRO (M126-1064-700), which was taken out from the MOPA box, was measured. It looks healthy.
The output power can go up to about 1 W, but I guess we don't want it to run at a high power to avoid any further degradation since the laser is old.

X-axis is the current read from the display of the controller. Y-axis is the output power, directly measured by Coherent PM10.
The measurement was done by changing the current from the controller.
Quote from #4832 |
[Things to be done]
- measure the beam profiles and power
- get a laser controller, which will be dedicated for this laser, from Peter King
|
|
4841
|
Mon Jun 20 13:48:25 2011 |
Koji | Update | ABSL | I-P curve of LightWave M126-1064-700 |
Hmm. Was the current within the operating range? (i.e. Is it a 700mW laser or a 1W one?)
You can obtain the nominal operating current from the old EPICS values (or some elog entries).
Note that NPROs are designed to be healthy only at around the nominal pumping power
(i.e. thermal gradient, and thermal lensing of the crystal, etc.)
ALSO:
Be aware that this laser should be used under the old SOP. So the appropriate interlocking is mandatory.
And probably we need to modify the SOP such that it reflects the latest situation.
Quote: |
The I-P curve of the LightWave NPRO, which was taken out from the MOPA box, was measured. It looks healthy.
The output power can go up to about 1 W, but I guess we don't want it to run at a high power to avoid any further degradation since the laser is old.
|
|
4842
|
Mon Jun 20 16:44:02 2011 |
steve | Update | ABSL | I-P curve of LightWave M126-1064-700 |
Put the serial numbers into the elog. So we can identify the laser and controller in the future.
The old days the NPRO ( inside the MOPA ) was running ~1.7A 500 mW |
4855
|
Wed Jun 22 15:24:10 2011 |
kiwamu | Update | ABSL | got a laser controller for LightWave |
Peter King came over to the 40m with a laser controller and gave it to us.
We will test it out with the LightWave NPRO, which was used for MOPA. |
Attachment 1: DSC_3150.JPG
|
|
Attachment 2: DSC_3153.JPG
|
|
4872
|
Thu Jun 23 22:59:45 2011 |
kiwamu | Update | ABSL | I-P curve of LWE |
The I-P curve was measured again, but this time in a lower current range of 1.0-1.9 [A].
The plot below is the latest I-P curve.

(Decision)
Based on the measurement and some thoughts, I decided to run this laser at about 1.8 [A] which gives us a middle power of ~ 360 [mW].
In the 40m history, the laser had been driven at 2.4 [A] in years of approximately 2006-2009, so it's possible to run it at such a high power,
but on the other hand Steve suggested to run it with a smaller power such that the laser power doesn't degrade so fast.
(notes)
The laser controller handed from PK (#4855) was used in this measurement.
The nominal current was tuned to be 1.8 [A] by tuning a potentiometer on the laser head (see page.18 on the manual of LWE).
There was a huge bump around 1.4 [A] and sudden power drop at 1.48 [A] although I don't know the reason.
Quote from #4842 |
The old days the NPRO ( inside the MOPA ) was running ~1.7A 500 mW
|
|
4874
|
Fri Jun 24 00:13:24 2011 |
kiwamu | Update | ABSL | beam profile measurement of LWE |
The beam profile of the LWE (LightWave Electronics) NPRO was measured.
Mode matching telescopes will be designed and setup soon based on the result of the measurements.
Here is a plot of the measured beam profile.

(some notes)
The measurement was done by using Kevin's power attenuation technique (#3030).
An window was put just after the NPRO and the reflected beam was sampled for the measurement to avoid the beam scan saturated. |
4877
|
Fri Jun 24 07:49:23 2011 |
steve | Update | ABSL | I-P curve of LWE with serial numbers |
Quote: |
The I-P curve was measured again, but this time in a lower current range of 1.0-1.9 [A].
Quote from #4842 |
The old days the NPRO ( inside the MOPA ) was running ~1.7A 500 mW
|
|
Lightwave Laser Head M126-1064-700 sn238, mounted on full size Al base and side heat sink on
Controller 125/126 Smart Supply sn 201M |
4917
|
Thu Jun 30 03:26:40 2011 |
kiwamu | Update | ABSL | status update |
Status update of the absolute length (ABSL) measurement:
- To accommodate the ABSL stuff, the AS path was relocated on the AP table.
(In this evening Jenne was able to lock MICH with AS55, so it's working fine.)
- On the AP table all of the necessary items, including the NPRO, a Faraday, some mirrors and etc., were in place
- The mode matching was coarsely done. The Rayleigh range looked reasonably long.
- Fine alignments will be done tomorrow
- Also a picture of the setup will be uploaded in the morning. |
4923
|
Thu Jun 30 14:11:55 2011 |
kiwamu | Update | ABSL | a photo of ABSL setup on the AP table |
Here is a picture of the latest ABSL setup at the east part of the AP table.

(Some notes )
- The ABSL laser is injected from the AP port.
- A 90 % reflection BS was installed just after the NPRO, this is for sampling a 10% of the laser to the PSL table.
However, I've just realized that this is not a nice way because the 10 % beam doesn't go through the Faraday. Whoops.
- A polarzser cell at the input side of the Faraday doesn't let any beam go through it for some reasons (broken ?).
Therefore instead of having such a bad cell, a cube PBS was installed.
- A room was left on the table for the AS165 RFPD (green-dashed rectangular in the picture).
Quote from #4917 |
- Also a picture of the setup will be uploaded in the morning.
|
|
4925
|
Thu Jun 30 21:00:20 2011 |
kiwamu | Update | ABSL | fine alignment done |
(Just a quick report)
The fine alignment of the ABSL laser injection was successfully done.
I was able to see the DRMI fringings at the AS camera. The ABSL beam is injected from the AS port, therefore what I saw on the camera was the reflection back from the interferometer.
(Things to be done)
- A beat-note setup on the PSL table.
- Refinement of the mode matching. The beam spot on the AS camera is a bit bigger, so I should more tightly focus the injected beam.
Quote from #4917 |
- Fine alignments will be done tomorrow
|
|
4939
|
Tue Jul 5 16:09:54 2011 |
kiwamu | Update | ABSL | some photos for ABSL setup |
Here I show two photos of the latest ABSL (ABSolute Length measurement) setup.

Figure.1 : A picture of the ABSL setup on the AP table.
The setup has been a little bit modified from the before (#4923).
As I said on the entry #4923, the way of sampling the ABSL laser wasn't so good because the beam, which didn't go through the faraday, was sampled.
In this latest configuration the laser is sampled after the faraday with a 90% beam splitter.
The transmitted light from the 90% BS (written in pink) is sent to the PSL table through the access tube which connects the AP and PSL table .

FIgure.2: A picture of the ABSL setup on the PSL table.
The 10% sampled beam ( pink beam in the picture) eventually comes to the PSL table via the access tube (the hole on the left hand side of the picture).
Then the ABSL beam goes through a mode matching telescope, which consists of a combination of a concave and a convex lens.
The PSL laser (red line in the picture) is sampled from a point after the doubling crystal.
The beam is combined at a 50 % BS, which has been setup for several purposes( see for example #3759 and #4339 ) .
A fast response PD (~1 GHz) is used for the beat-note detection. |
4940
|
Tue Jul 5 17:38:46 2011 |
kiwamu | Update | ABSL | ABSL laser frequency-locked |
In this past weekend the ABSL laser was successfully frequency-locked to the PSL laser with a frequency offset of about 100 MHz.
In the current setup a mixer-based frequency discriminator is used for detection of the beat-note frequency.
Setup for frequency locking
The diagram below shows the setup for the frequency locking.

(UGF)
According to a brief check of the loop oscillation it implies that the UGF is around 40 kHz.
Of course I will measure the open-loop over a wide frequency range at some point.
The lock was quite robust and it was able to stay locked for more than an hour as far as I observed.
(Beat-note detection and its amplification)
The fast RFPD, which is standing on the PSL table to detect the beat-note ( see the picture on #4939), showed the signal level of -16 dBm.
Then the signal goes through the RF amplifier stage to have an appropriate power level at the mixer in the frequency detection stage.
Before the signal goes into the amp. stage I put a power splitter so that I can see a signal on a spectrum analyzer.
(Frequency detection)
In the mixer-based frequency discriminator, the length difference between two coax. cable was chosen to be 1 m.
This length difference gives us a zero cross point (operation point) of ~100 MHz and linear range of ~ +/-100MHz, which can sufficiently cover the FSR of PRC: 22MHz.
(Control filter)
After the signal goes through the freq. discriminator, a low noise amplifier, ITHACO 1201 was installed as a control filter.
The reason I chose ITHACO 1201 was that it has a larger output range than that of SR560.
ITHACO 1201 can go to +/- 15V, which means the ABSL laser should be able to track the frequency by about +/- 65 MHz without a thermal actuation.
Right now a single pole was put at 1 Hz for an easy lock.
The output from 1201 goes directly to the laser PZT, whose input is on the front panel of the laser controller.
Temperature setpoints
During the work I found three temperature points, where we can observe the beat-note signal within a bandwidth of 1 GHz.
Here is a summary of the temperature set points :
PSL temperature = 31.71 deg
ABSL temperature = 44.19 deg (not good),
47.25 deg (chosen to be nominal set point),
50. 23 deg
The first one (44.19 deg) wasn't good because changing the temperature of the ABSL laser also changes the amplitude of the beat-note significantly.
In a worst case the signal completely disappeared when the ABSL laser was at a certain temperature close to 44.19 deg.
The other two temperature points seemed good. I eventually chose the second one (47.25 deg) for the nominal set point.
Note that the current readout has been 1.81 A on the laser controller. |
5079
|
Mon Aug 1 04:08:24 2011 |
kiwamu | Update | ABSL | Arm length measurement : cavity kick technique |
I made some attempts to measure the current length of the arm cavities by using the mass-kicking technique.
However unfortunately I am running out my energy to complete the measurement,
so I will finish the measurement at some time today.
I still have to set an appropriate kick amplitude. Right now I am injecting AWG into ETMY_LSC_EXC at 0.2 Hz with amplutde of 400 cnts.
I guess it needs a little bit more amplitude to get more psuedo-constant velocity.
Volunteers are always welcome !
(some notes)
The procedure was well-described in entry #555 by Dr.Stochino.
Here is just an example of the time series that I took today showing how the time series looks like.

|
5095
|
Tue Aug 2 16:55:21 2011 |
rana | Update | ABSL | Arm length measurement : cavity kick technique |
Quote: |
I made some attempts to measure the current length of the arm cavities by using the mass-kicking technique.
|
Why not just scan the Green laser to measure the arm lengths instead? The FSR of the arm is ~3.75 MHz and so all you have to do is lock the arm green and then sweep the PZT until the resonance is found at 3.75 MHz.

|
5117
|
Thu Aug 4 09:42:19 2011 |
Koji | Update | ABSL | ABSL Laser shutter closed |
The shutter of the ABSL laser is closed for the vent work. |
8257
|
Fri Mar 8 12:57:57 2013 |
Annalisa | Update | ABSL | BS installed on ITMY table |
Sendhil and I installed the S polarized BS on the ITMY table to steer the NPRO beam through the AR wedge and align it to the POY beam.
We took a shutter from the BSPRM table (which was not used) and a beam dump from the AS table (which was used by the auxiliary laser already removed and installed on the ITMY).
To do: do better alignment of the NPRO beam, maybe installing some iris after the BS and before the AS wedge, phase lock the two beams. |
8258
|
Fri Mar 8 13:42:35 2013 |
Jenne | Update | ABSL | BS installed on ITMY table |
Re: POY beam reduction.
We are able to lock the Yarm with the beam / gain as it is. I had thought we might need to increase the DC gain in the whitening board by a factor of 2, but so far it's fine. |
8303
|
Mon Mar 18 12:02:12 2013 |
Annalisa | Configuration | ABSL | ABSL setup for g-factor measurement of PRC |
Motivations
The ABSL technique has been already used in the past to measure the absolute length of the interferometer's optical cavities by means of an auxiliary laser source, as described in LIGO-P1200048-v3 and in Alberto Stochino thesis work.
Using the same technique it is possible to measure the g-factor of the power recycling cavity by measuring the cavity Transverse Mode Spacing.
Plan for experimental setup
The auxiliary laser is set on the POY table and is injected through the ITMY window in way to follow the same path of the POY beam. It hits the AR wedge of ITMY and is reflected back to the BS and the PRM.
Since the main beam is P-polarized, all the optics in the central IFO are P-polarization dependent, so it is useful to P-polarize the auxiliary beam before it enters the IFO.
I made a mode matching calculation with a la mode script, in order to mode match the auxiliary beam waist to the waist of the main laser.
However, before ordering and installing steering optics and mode maching lenses, I'm waiting to know whether someone has an NPRO laser to install on the END table in place of the broken one, otherwise the one I'm using could be taken.
In this case a possibility could be to take the auxiliary beam from the end table with an optical fiber, but it means to use the auxiliary laser alternately to lock the arm or make a measurement of TMS. If so, a new calculation for the mode matching needs to be done.
Anyway, I hope that another laser will be found!
In order to phase lock the auxiliary beam with the main beam, the latter will be taken from the PSL table after the PMC through a single mode fiber, which will be brought up to the POY table. This solution results to be more reliable then taking the POY beam to phase lock the two laser, because POY is related to the locking.
The signal with the beat note between the two lasers can be detected by the transmission from PR2 (POP).
|
8361
|
Wed Mar 27 21:53:21 2013 |
Annalisa | Update | ABSL | Beat note of ATF auxiliary laser |
After measuring the beat note, the "Alberto" NPRO auxiliary laser has been moved from the PSL table to the POY table. Its beam profile is going to be measured. It's going to be used as green laser on the END table, in place of the broken one.
The auxiliary laser borrowed form ATF lab (which will be used for the ABSL measurement) has been set on the PSL table to make a measurement of the beat note between it and the main laser.
The setup is mostly the same of the previous beat note measurement . In this case, laser input power is 326 mW, so I needed to replace one of the mirrors of the steering optics with a BS 50% reflecting in order to have less than 1 mW on the PD.
Now, the total power on the PD is less than 0.5 mW.
I didn't measure the beat note yet to leave the PSL table as quite as possible for the locking procedures.
To do:
Measure the beat note, fiber coupling the NPRO laser to bring it to the POY table.
|
8369
|
Thu Mar 28 23:00:30 2013 |
Annalisa | Update | ABSL | Beat note of ATF auxiliary laser found |
The beat note for the ATF lab laser has been found.
The measurement has been carried out in the same way as described in elog 8368.
The only difference is that in this case I started from a temperature of 35.2 degC, and I reduced it until the minimum which was 30.71 degC. No beat note in this range.
Then I rised on the temperature and I found the first beat note at 41.46 degC. It has been detected at a frequency of about 120 MHz with an RF power of -53 dBm and a frequency fluctuation of about +/- 5 MHz.