  1245   Thu Jan 22 12:08:59 2009 peteUpdateoplevsoplev calibration
Following the procedure described in Royal Reinecke's 2006 SURF report, I've calibrated the ETMY yaw oplev DOF. The idea is to sweep the mirror tilt, measuring the transmitted cavity power and the oplev error signal. The cavity power can be related to the mirror tilt in radians following D. Anderson APPLIED OPTICS, Vol. 23, No. 17, 1984.

I've made a simple matlab script which spits out the final number; it calls Royal's perl script to do the sweep. I get 420 microrad/ct for ETMY yaw. In 2006 Royal got 250 microrad/ct. Could something have changed this much, or is one of us wrong? I'll double check my procedure and do the other arm cavity oplevs, and describe it in detail when I have more confidence in it.

Kakeru and I plan to extend this to handle the PRM, SRM, and BS. One script to rule them all.
  1247   Thu Jan 22 23:36:50 2009 peteHowTooplevsarm cavity oplev calibration
calibrated the y-arm oplevs. the procedure is contained in a matlab script. the whereabouts of this script will be revealed in a future log entry.

ITMYpit 140 microrad/ct
ITMYyaw 98 microrad/ct
ETMYpit 400 microrad/ct
ETMYyaw 440 microrad/ct (previous measurement gave 420 microrad/ct)

procedure:

1) Start with a single arm aligned and locked. Dither the mirror tilt in a DOF. Measure arm cavity power and oplev error signal. See the first attached plot.

2) Fit the plot to a gaussian and determine mu and sigma.

3) For a spherical ETM optic, the power in the cavity P(a), as a function of translational beam axis displacement a=R*sin(theta), is proportional to exp[-a^2/(2*x^2)] where x is the waist size (D. Anderson APPLIED OPTICS, Vol. 23, No. 17, 1984). The power as a function of mirror tilt in cts, P(tilt) is proportional to exp[-(tilt-mu)^2 /(2*sigma^2)]. So if R is the mirror radius then theta = arcsin(a/R) = arcsin[(1/R)*(tilt-mu)*x/sigma)].

4) Fit theta versus mirror tilt to get the calibration. See the second attached plot.

5) For a flat ITM optic, mirror tilt causes an angular displacement of the beam. The math for this case is given in Anderson.
Attachment 1: ETMYpitpower.png
ETMYpitpower.png
Attachment 2: ETMYpitcal.png
ETMYpitcal.png
  1251   Fri Jan 23 16:33:27 2009 peteUpdateoplevsx-arm oplev calibrations
ITMXpit 71 microrad/ct
ITMXyaw 77 microrad/ct
ETMXpit 430 microrad/ct
ETMXyaw 430 microrad/ct

As with y-arm, my ITM measurements agree with Kakeru and Royal, but my ETM measurements are not quite a factor of 2 higher. Kakeru and I are investigatin.
  1327   Thu Feb 19 23:50:31 2009 peteUpdateLockingaligned pd's on AP table

Yoichi, Peter

While continuing our efforts to lock, we noticed the procedure failed at a point it had gotten past last night:  turning on the bounce/roll filters in MICH, PRC, and SRC.  We checked the MICH transfer function and noticed that the unity gain point was ~10 Hz, well below the bounce modes.   We tried increasing the gain but found saturation, and Rob suggested that there could be misalignment on the AP table, which Steve worked on today.  We went out and found two of the PDs (ASDD133 and AS166) to be badly misaligned probably due to a bumped optic upstream.  We re-aligned.

 

 

  1335   Tue Feb 24 18:42:15 2009 peteUpdateLockingmc board repair
Peter, Yoichi
Last night:


Quote:
However, when we measured the MCL loop gain with several different AO path gains, the loop shape did not change at all. This led us to suspect the AO path may not be connected. The cabling from the common mode board to the MC board seemed ok. We tested the signal flow in the MC board using a signal generator and an oscilloscope. Then we found that a signal injected to the IN2 (AO path) does not reach to the TP1A (right after the boost stages), though the signal is visible in the OUT2 (monitor BNC right after the initial amplifier (B-amp) for the AO path). The signal from IN1 (MC REFL) can be observed at TP1A. This means something is broken between the B-amp and the sum-amp in the AO path. We will check the MC board tomorrow.


Today we examined the MC board. With the extension board in place everything seemed fine. Without the extension board we could replicate the problem. Jiggling the IN2 jack caused the injected signal observed at TP1A to come and go. These jacks are unfortunately mounted directly on the board. We traced the problem to a resistor in this path (R30) which looked fishy. We soldered on a new 2K resistor with OWC and it fixed the problem.
  1435   Fri Mar 27 02:40:06 2009 peteSummaryIOOMC glitch investigation

Yoichi, Pete

The MC loses lock due to glitches in the MC1 coils. 
We do not know which coil for sure, and we do not know if it is a problem going into the board, or a problem on the board. 
We suspect either the UL or LR coil bias circuits (Pete would bet on UL).  If you look at the bottom 4 plots in the attached file, you can see a relatively large 3 minute dip in the UL OSEM output, with a corresponding bump in the LR (and smaller dips in the other diagonal).  
These bumps do not show up in the VMONS which is why we are suspicious of the bias.
To test we are monitoring 4 points in test channels, for UL and UR, both going into the bias driver circuit, and coming out of the current buffer before going into the coils. 
 

We ran cable from the suspension rack to the IOO rack to record the signals with DAQ channels.

The test channels:

UL coil      C1:IOO-MC_DRUM1  (Caryn was using, we will replace when we are done)

UL input   C1:IOO-MC_TMP1 (Caryn was using, we will replace when we are done)

LR coil      C1:PEM-OSA_SPTEMP

LR input   C1:PEM-OSA_APTEMP

We will leave these overnight; we intend to remove them tomorrow or Monday.

We closed the PSL shutter and killed the MC autolocker.

Attachment 1: MC1_Drift.png
MC1_Drift.png
  1463   Thu Apr 9 12:23:49 2009 peteUpdateLockingtuning ETM common mode

Pete, Yoichi

Last night, we put the IFO in FP Michelson configuration.  We took transfer functions of CARM and DARM, first using CM excitations directly on the ETMs, and then using modulations of the laser frequency via MC excitation.  We found that there was basically no coupling into DARM using the MC excitation, but that there was coherence in DARM using the ETM excitation.  Therefore, I tuned the ETM common mode in the output matrix.  I did this by taking transfer functions of PD1_Q with PD2_I (see attached plot).  I changed the  drdown_bang script to set C1:LSC-BTMTRX_14 0.98 and C1:LSC-BTMTRX_24 1.02.

Attachment 1: FPMI-DARM-CARM-ETM-fineScan.pdf
FPMI-DARM-CARM-ETM-fineScan.pdf
  1489   Thu Apr 16 16:26:57 2009 peteUpdateLockingWed. night locking
yoichi, pete

We installed the watchLockLoss script in scripts/AutoDTT/.  This script monitors arm power and uses command line
DTT to save 5 s snapshot of the interferometer when it senses loss of lock.  We ran it on linux and it seemed to
save an xml file about half the time; we'll try it on solaris.  

I managed to get up to arm power of about 20 a couple of times.  IFO lost lock a couple of times after turning
off moving zero.  MC2 would often get tripped by lock loss and need resetting.  Maybe we will try to stiffen the
op levs.
  1525   Tue Apr 28 01:48:45 2009 peteConfigurationVACVC1 open

At about 1am or so Yoichi and I opened VC1.  CC1 had fallen to about 5e-5 torr.

  1557   Thu May 7 18:12:12 2009 peteUpdateLockingarm power curve

I've plotted TRX, TRY, PD12I and PD11Q.  Arm powers after locking increase for a few tens of minutes, peak out, and then decrease before lock is lost.

 

 

Attachment 1: 2009_may_7_powers.jpg
2009_may_7_powers.jpg
  1558   Thu May 7 23:21:04 2009 peteUpdateLockingarm power curve

Quote:

I've plotted TRX, TRY, PD12I and PD11Q.  Arm powers after locking increase for a few tens of minutes, peak out, and then decrease before lock is lost.

 

 

 I should have mentioned that the AS port camera image seems to get progressively uglier over the course of these locks.  Maybe we can use the JoeCam to make a movie of it. 

  1560   Fri May 8 02:08:59 2009 peteUpdateLockinglock stretches

locks last for about an hour.  this was true last night as well (see "arm power curve" entries).   the second lock shown here evolves differently for unknown reasons.  the jumps in the arm powers of the first lock are due to turning on DC readout.  length-to-angle needs tuning.

 

 

Attachment 1: powers_oplev.pdf
powers_oplev.pdf
  1565   Fri May 8 15:40:44 2009 peteUpdateLockingprogressively weaker locks

the align script was run after the third lock here.  it would have been interesting to see the arm powers in a 4th lock 

Attachment 1: powers_3lock.pdf
powers_3lock.pdf
  1578   Tue May 12 17:26:56 2009 peteUpdateoplevsetmy oplev quad was bad

Pete, Rob

After looking at some oplev noise spectra in DTT, we discovered that the ETMY quad (serial number 115)  was noisy.  Particularly, in the XX_OUT and XX_IN1 channels, quadrants 2 (by a bit more than an order of magnitude over the ETMX ref) and 4 (by a bit less than an order of mag).  We went out and looked at the signals coming out of the oplev interface board; again, channels 2 and 4 were noise compared to 1 and 3 by about these same amounts.  I popped in the ETMX quad and everything looked fine.  I put the ETMX quad back at ETMX, and popped in Steve's scatterometer quad (serial number 121 or possibly 151, it's not terribly legible), and it looks fine.  We zeroed via the offsets in the control room, and I went out and centered both the ETMX and ETMY quads. 

Attached is a plot.  The reference curves are with the faulty quad (115).  The others are with the 121.

 

Attachment 1: bad_oplev_quad.pdf
bad_oplev_quad.pdf
  1580   Wed May 13 03:05:13 2009 peteUpdateoplevsetmy oplev quad was bad

Quote:

Pete, Rob

After looking at some oplev noise spectra in DTT, we discovered that the ETMY quad (serial number 115)  was noisy.  Particularly, in the XX_OUT and XX_IN1 channels, quadrants 2 (by a bit more than an order of magnitude over the ETMX ref) and 4 (by a bit less than an order of mag).  We went out and looked at the signals coming out of the oplev interface board; again, channels 2 and 4 were noise compared to 1 and 3 by about these same amounts.  I popped in the ETMX quad and everything looked fine.  I put the ETMX quad back at ETMX, and popped in Steve's scatterometer quad (serial number 121 or possibly 151, it's not terribly legible), and it looks fine.  We zeroed via the offsets in the control room, and I went out and centered both the ETMX and ETMY quads. 

Attached is a plot.  The reference curves are with the faulty quad (115).  The others are with the 121.

 

 I adjusted the ETMY quad gains up by a factor of 10 so that the SUM is similar to what it was before.

  1585   Thu May 14 02:36:05 2009 peteUpdateLockingunstable IFO

It seems that the MC3 problem is intermittent (one-day trend attached).  I tried to take advantage of a "clean MC3" night, but the watch script would usually fail at the transition to DC CARM and DARM.  It got past this twice and then failed later, during powering up.   I need to check the handoff.

 

Attachment 1: mc3.jpg
mc3.jpg
  1587   Thu May 14 16:07:20 2009 peteSummarySUSChannel Hopping: That ancient enemy (MC problems)

Quote:

Quote:
The MC side problem could also be the side tramp unit problem. Set the tramp to 0 and see if that helps.


This started around April 23, around the time that TP1 failed and we switched to the cryopump, and also when there was a mag 4 earthquake in LA. My money's on the EQ. But I don't know how.


I wonder if this is still a problem. It has been quiet for a day now. I've attached a day-long trend. Let's see what happens.
Attachment 1: mc3_5days.jpg
mc3_5days.jpg
  1588   Fri May 15 00:02:34 2009 peteUpdateSUSETMX coils look OK

I checked the four rear coils on ETMX by exciting XXCOIL_EXC channel in DTT with amplitude 1000@ 500 Hz and observing the oplev PERROR and YERROR channels.  Each coil showed a clear signal in PERROR, about 2e-6 cts.  Anyway, the coils passed this test.

 

  1610   Wed May 20 01:41:19 2009 peteUpdateVACcryopump probably not it

I found some neat signal analysis software for my mac (http://www.faberacoustical.com/products/), and took a spectrum of the ambient noise coming from the cryopump.  The two main noise peaks from that bad boy were nowhere near 3.7 kHz.

  1616   Thu May 21 18:05:03 2009 peteUpdateSUSETMX coils look OK

Quote:

I checked the four rear coils on ETMX by exciting XXCOIL_EXC channel in DTT with amplitude 1000@ 500 Hz and observing the oplev PERROR and YERROR channels.  Each coil showed a clear signal in PERROR, about 2e-6 cts.  Anyway, the coils passed this test.

 

 I also made xfer fctns of the 4 piston coils on ETMY and ETMX with OL_PIT.  (I looked at all 4 even though the attached plot only shows three.)  So it looks ike the coils are OK.

Attachment 1: etmx_etmy_coils.pdf
etmx_etmy_coils.pdf
  1620   Fri May 22 01:27:14 2009 peteUpdateSUS200 days of MC3 side

Looks like something went nuts in late April.  We have yet to try a hard reboot.

Attachment 1: mc3_side_200days.png
mc3_side_200days.png
  1641   Tue Jun 2 02:28:58 2009 peteUpdateLockingDD handoff work

alberto, pete

 

We worked on tuning the DD handoff tonight.  We checked the DD PD alignments and they looked fine.  First I tuned the 3 demod phases to minimize offsets.  Then I noticed that the post-handoff MICH xfer function needed an increase in gain to look like the pre-handoff xfer function (which has a UGF of about 25 Hz).  I increased the MICH PD9_Q gain from 2 to 7 in the input matrix.   But, the handoff to PRC still failed, so tomorrow we will try to find out why.

In the plot, ref0 is before MICH handoff, and ref1 is after MICH handoff.  There is also a PRC trace (before PRC handooff).

 

 

Attachment 1: mich_dd.pdf
mich_dd.pdf mich_dd.pdf
  1643   Tue Jun 2 23:53:12 2009 peteDAQComputersreset c1susvme1

rob, alberto, rana, pete

we reset this computer, which was out of sync (16384 in the FE_SYNC field instead of 0)

  1645   Wed Jun 3 03:22:16 2009 peteUpdateLockingDD handoff

Rana, Alberto, Pete

We have the DD handoff nominally working.  Sometimes, increasing the SRC gain at the end makes MICH get unstable.  This could be due to a non-diagonal term in the matrix, or possibly because the DRM locks in a funky mode sometimes. 

To get the DD handoff working, first we tuned demod phases in order to zero the offsets in the PD signals handed-off-to.  Based on transer function measurements, I set the PRC PD6_I element to 0.1, and set the PD8_I signal to 0, since it didn't seem to be contributing much.  We also commented out the MICH gain increase at the end of the DD_handoff script.

It could still be more stable, but it seems to work most of the time.

 

 

  1652   Thu Jun 4 16:54:19 2009 peteUpdateLockingdaytime DD handoff

I played with the DD handoff during the day.  The DRM dark port was flickering like a candle flame in Dracula's castle.  The demod offsets for the handoff signals looked fine.  After MICH handoff, the MICH_CTRL started to get unstable at some low frequency, maybe 3 Hz (I didn't measure).  So I increased the MICH gain from 0.1 to 0.17 and it settled down.  PRC and SRC went fine.  Then the DD_handoff script raised the MICH gain to 0.7, and an instability started to grow in MICH_CTRL (at some higher frequency).  I decreased the MICH gain from 0.7 to 0.5, and it settled down and stayed stable.

  1653   Thu Jun 4 23:39:23 2009 peteUpdatePEM5 days, 20 days of accelerometers

Looks like yesterday was particularly noisy.  It's unclear to me why diurnal variation much more visible in MC1_Y, and why the floor wanders.

 

The first plot shows 5 days.  The second plot shows 20 days.

Attachment 1: acc_5day.png
acc_5day.png
Attachment 2: acc_20days.png
acc_20days.png
  1658   Fri Jun 5 17:22:55 2009 peteUpdateLockingdaytime locking

After fixing the tp problem, I tried locking again.  Grabbing and DD handoff, no problem.  Died earlier than last night, handing off CARM to REFL_DC, around arm power of 4 or so.  Seems to happen after turning off the moving zero, Rob says it might be touchy in daytime.

  1679   Tue Jun 16 16:10:01 2009 peteUpdateLockinginput matrix experiments

Last night Rob ran senseDRM and loadDRMImatrixData and came up with the following for the input matrix:

tdswrite C1:LSC-ITMTRX_b2 0.065778 \
C1:LSC-ITMTRX_d2 2.2709 \
C1:LSC-ITMTRX_f2 2.9361 \
C1:LSC-ITMTRX_122 0.42826 \
C1:LSC-ITMTRX_b3 -0.064839 \
C1:LSC-ITMTRX_d3 -0.016913 \
C1:LSC-ITMTRX_f3 -0.021576 \
C1:LSC-ITMTRX_123 -0.0025243 \
C1:LSC-ITMTRX_b5 0.3719 \
C1:LSC-ITMTRX_d5 1.3109 \
C1:LSC-ITMTRX_f5 -0.16412 \
C1:LSC-ITMTRX_125 0.39574 \
C1:LSC-ITMTRX_33 0 \
C1:LSC-ITMTRX_42 0 \
C1:LSC-ITMTRX_155 0

Today, I reran these and got the following, and DD_handoff remained happy:

tdswrite C1:LSC-ITMTRX_b2 -0.10329 \
C1:LSC-ITMTRX_d2 2.0344 \
C1:LSC-ITMTRX_f2 3.2804 \
C1:LSC-ITMTRX_122 0.22516 \
C1:LSC-ITMTRX_b3 -0.076292 \
C1:LSC-ITMTRX_d3 -0.014603 \
C1:LSC-ITMTRX_f3 -0.12101 \
C1:LSC-ITMTRX_123 0.0054128 \
C1:LSC-ITMTRX_b5 0.33521 \
C1:LSC-ITMTRX_d5 1.1425 \
C1:LSC-ITMTRX_f5 -0.32759 \
C1:LSC-ITMTRX_125 0.25877 \
C1:LSC-ITMTRX_33 0 \
C1:LSC-ITMTRX_42 0 \
C1:LSC-ITMTRX_155 0

I wanted to remeasure with the canonical output matrix (-0.7 from MICH to PRM and 0.7 from MICH to SRM), but the DRM freaked out when MICH to PRM went below -0.3.

  1769   Tue Jul 21 17:01:18 2009 peteDAQDAQtemp channel PEM-PETER_FE

I added a temporary channel, to input 9 on the PEM ADCU.    Beware the 30, 31, and 32 inputs.  I tried 32 and it only gave noise.

 

 

  1781   Wed Jul 22 20:11:26 2009 peteUpdateComputersRCG front end

I compiled and ran a simple (i.e. empty) front end controller on scipe12 at wilson house.  I hooked a signal into the ADC and watched it in the auto-generated medm screens. 

There were a couple of gotchas:

1. Add an entry SYS to the file /etc/rc.local, to the /etc/setup_shmem.rtl line, where the system file is SYS.mdl.

2. If necessary, do a BURT restore.  Or in the case of a mockup set the BURT Restore bit (in SYS_GDS_TP.adl) to 1.

 

  1805   Wed Jul 29 12:14:40 2009 peteUpdateComputersRCG work

Koji, Pete 

Yesterday, Jay brought over the IO box for megatron, and got it working.  We plan to firewall megatron this afternoon, with the help of Jay and Alex, so we can set up GDS there and play without worrying about breaking things.  In the meantime, we went to Wilson House to get some breakout boards so we can take transfer functions with the 785, for an ETMX controller.  We put in a sine wave, and all looks good on the auto-generated epics screens, with an "empty" system (no filters on). Next we'll load in filters and take transfer functions.

Unfortunately we promised to return the breakout boards by 1pm today.  This is because, according to denizens of Wilson House, Osamu "borrowed" all their breakout boards and these were the last two!  If we can't locate Osamu's cache, they expect to have more in a day or two.

Here is the transfer function of the through filter working at 16KHz sampling. It looks fine except for the fact that the dc gain is ~0.8. Koji is going to characterize the digital down sampling filter in order to try to compare with the generated code and the filter coefficients.


Attachment 1: TF090729_1.png
TF090729_1.png
Attachment 2: TF090729_1.png
TF090729_1.png
  1819   Mon Aug 3 13:47:42 2009 peteUpdateComputersRCG work

Alex has firewalled megatron.  We have started a framebuilder there and added testpoints.  Now it is possible to take transfer functions with the shared memory MDC+MDP sandbox system.  I have also copied filters into MDC (the controller) and made a really ugly medm master screen for the system, which I will show to no one.

  1826   Tue Aug 4 13:40:17 2009 peteUpdateComputersRCG work - rate

Koji, Pete

 

Yesterday we found that the channel C1:MDP-POS_EXC looked distorted and had what appeared to be doubled frequency componenets, in the dataviewer.  This was because the dcu_rate in the file /caltech/target/fb/daqdrc was set to 16K while the adl file was set to 32K.  When daqdrc was corrected it was fixed.  I am going to recompile and run all these models at 16K.  Once the 40 m moves over to the new front end system, we may find it advantageous to take advantage of the faster speeds, but maybe it's a good idea to get everything working at 16K first.

  1829   Tue Aug 4 17:51:25 2009 peteUpdateComputersRCG work

Koji, Peter

 

We put a simple pendulum into the MDP model, and everything communicates.  We're still having some kind of TP or daq problem, so we're still in debugging mode.  We went back to 32K in the .adl's, and when driving MDP,  the MDC-ETMX_POS_OUT is nasty, it follows the sine wave envelope but goes to zero 16 times per second.

 

The breakout boards have arrived.  The plan is to fix this daq problem, then demonstrate the model MDC/MDP system.  Then we'll switch to the "external" system (called SAM) and match control TF to the model.  Then we'd like to hook up ETMX, and run the system isolated from the rest of the IFO.  Finally we'd like to tie it into the IFO using reflective memory.

  1839   Wed Aug 5 17:41:54 2009 peteUpdateComputersRCG work - daq fixed

The daq on megatron was nuts.  Alex and I discovered that there was no gds installation for site_letter=C (i.e. Caltech) so the default M was being used (for MIT).  Apparently we are the first Caltech installation.  We added the appropriate line to the RCG Makefile and recompiled and reinstalled (at 16K).  Now DV looks good on MDP and MDC, and I made a transfer function that replicates  bounce-roll filter.  So DTT works too.

  1856   Fri Aug 7 16:00:17 2009 peteUpdateComputersRCG work. MDC MDP open loop transfer function

Today I was able to make low frequency transfer function with DTT on megatron.  There seems to have been a timing problem, perhaps Alex fixed it or it is intermittent.

I have attached the open loop transfer function for the un-optimized system, which is at least stable to step impulses with the current filters and gains.  The next step is to optimize, transfer this knowledge to the ADC/DAC version, and hook it up to isolated ETMX.

Attachment 1: tf_au_natural.pdf
tf_au_natural.pdf tf_au_natural.pdf
  1879   Mon Aug 10 17:36:32 2009 peteUpdateComputersRCG work. PIT, YAW, POS in MDP/MDC system

I've added the PIT and YAW dofs to the MDC and MDP systems.  The pendula frequencies in MDP are 0.8, 0.5, 0.6 Hz for POS, PIT, and YAW respectively.  The three dofs are linear and uncoupled, and stable, but there is no modeled noise in the system (yet) and some gains may need bumping up in the presence of noise.  The MDC filters are identical for each dof (3:0.0 and Cheby). The PIT and YAW transfer functions look pretty much like the one Rana recently took of POS, but of course with the different pendulum frequencies.  I've attached one for YAW.

Attachment 1: mdcmdpyaw.jpg
mdcmdpyaw.jpg
  1881   Mon Aug 10 17:49:10 2009 peteUpdateComputersRCG work - plans

Pete, Koji

 

We discussed a preliminary game plan for this project.  The thing I really want to see is an ETMX RCG controller hooked into the existing frontend via reflective memory, and the 40 m behaving normally with this hybrid system, and my list is geared toward this.  I suspect the list may cause controversy.

+ copy the MDC filters into SAM, and make sure everything looks good there with DTT and SR785.

+ get interface / wiring boards from Wilson House, to go between megatron and the analog ETMX system

+ test tying the ETMX pendulum and bare-bones SAM together (use existing watchdogs, and "bare-bones" needs defining)

+ work some reflective memory magic and create the hybrid frontend

 

In parallel with the above, the following should also happen:

+ MEDM screen design

+ add non-linear bits to the ETMX MDP/MDC model system

+ make game plan for the rest of the RCG frontend

  2182   Thu Nov 5 16:30:56 2009 peteUpdateComputersmoving megatron

 Joe and I moved megatron and its associated IO chassis from 1Y3 to 1Y9, in preparations for RCG tests at ETMY.

  2198   Fri Nov 6 18:52:09 2009 peteUpdateComputersRCG ETMY plan

  Koji, Joe, and I are planning to try controlling the ETMY, on Monday or Tuesday.  Our plan is to try to do this with megatron out of the RFM loop.   The RCG system includes pos, pit, yaw, side, and oplevs.  I will use matrix elements as currently assigned (i.e. not the ideal case of +1 and -1 everywhere).   I will also match channel names to the old channels.   We could put buttons on the medm screen to control the analog DW by hand.  

 
This assumes we can get megatron happy again, after the unhappy RFM card test today.  See Joe's elog immediately before this one.
 
We first plan to test that the ETMY watchdog can disable the RCG frontend, by using a pos step (before connecting to the suspension).   Hopefully we can make it work without the RFM.  Otherwise I think we'll have to wait for a working RFM card.
 
We plan to disable the other optics.  We will disable ETMY, take down the ETMY frontend, switch the cables, and start up the new RCG system.  If output looks reasonable we will enable the ETMY via the watchdog.   Then I suppose we can put in some small steps via the RCG controller and see if it damps.  
 
Afterwards, we plan to switch everything back.
  2233   Wed Nov 11 01:33:52 2009 peteUpdateCDSRCG ETMY code update

 I've added the side coil to the model controller and plant, and the oplev quad to the model controller and plant.  After the megatron wipe, the code now lives in /home/controls/cds/advLigo/src/epics/simLink.  The files are mdc.mdl (controller) and mdp.mdl (plant).  These RCG modules go at 16K with no decimation (no_oversampling=1 in the cdsParameters block) so hopefully will work with the old (16K) timing.

I've loaded many of the filters, there are some eft to do.  These filters are simply copied from the current frontend.  

Next I will port to the SUS module (which talks to the IO chassis).  This means channel names will match with the current system, which will be important when we plug in the RFM.

  2243   Wed Nov 11 20:46:07 2009 peteUpdateCDSRCG ETMY phase I update

The .mdl code for the mdc and mdp development modules is finished.  These modules need more filters, and testing.  Probably the most interesting piece left to do is putting in the gains and filters for the oplev model in mdp.  It might be OK to simply ignore oplevs and first test damping of the real optic without them.   However, it shouldn't be hard to get decent numbers for oplevs, add them to the mdp (plant) module, and make sure the mdc/mdp pair is stable.  In mdp, the oplev path starts with the SUSPIT and SUSYAW signals. Kakeru recently completed calibration of the oplevs from oplev cts to radians:   1403  .  From this work we should find the conversion factors from PIT and YAW to oplev counts, without making any new measurements.  (The measurements wouldn't be hard either, if we can't simply pull numbers from a document.)  These factors can be added to mdp as appropriate gains.

I've also copied mdc to a new module, which I've named "sas" to address fears of channel name collisions in the short term, and replaced the cpu-to-cpu connections with ADC and DAC connections.  sas can be the guy for the phase I ETMY test.  When we're happy with mdc/mdp, we hopefully can take the mdc filter file from chans, replace all the "MDC" strings with "SAS", and use it.

  2293   Wed Nov 18 16:24:25 2009 peteConfigurationSUSETMY suspension conencted to megatron ADC/DAC

/cvs/cds/caltech/target/fb/daqd -c daqdrc

This starts the FB.

Now the dataviewer and DTT work!

Quote:

0) Now the connection for the ETMY suspension was restored in a usual state. It damps well.

1) I thought it would be nice to have dataviewer and DTT working.
   So far, I could not figure out how to run daqd and tpman.
   - I tried to configure
    /cvs/cds/caltech/target/fb/daqdrc
    /cvs/cds/caltech/target/fb/master
    /cvs/cds/caltech/chans/daq/C1TST.ini
(via daqconfig)

   - I also looked at
    /cvs/cds/caltech/targetgds/param/tpchn_C1.par
   but I don't understand how it works. The entries have dcuids of 13 and 14 although C1TST has dcuid of 10.
   The file is unmodified.

   I will try it later when I got a help of the experts.

2) Anyway, I went ahead. I tried to excite suspension by putting some offset.

It seems to have no DAC output. I checked the timing signal. It seems that looks wrong clock.

   I looked at DAC output by putting 5000,10000,15000,20000,25000cnt to UL/UR/LR/LL/SD coils.
   I could not find any voltage out of the DAC in any channels.

   Then, I checked the timing signal. This clock seems to have wrong frequency.
   What we are using now is a clock with +/-4V@4MHz. (Differential)
   Maybe 4194304Hz (=2^22Hz)?

   I went to 1Y3 and checked the timing signal for 16K. This was +/-4V@16kHz. (Diffrential)

   The possible solution would be
   - bring a function generator at the end and try to input a single end 4V clock.
   - stretch a cable from 1Y3 to 1Y9. (2pin lemo)

Quote:

I have connected ETMY sus electronics to megatron ADC/DAC.
We continue this state until 15:00 of today.

 

 

  2296   Thu Nov 19 08:53:12 2009 peteFrogsEnvironmentdiesel fumes

 Instead of doing RCG stuff, I went to Millikan to work on data analysis as I couldn't stand the fumes from the construction.  (this morning, 8am) 

  1655   Fri Jun 5 02:59:03 2009 pete, albertoUpdateLockingtdsavg failure in cm_step script

the command

tdsavg 5 C1:LSC-PD4_DC_IN1

was causing grievous woe in the cm_step script.  It turned out to fail intermittently at the command line, as did other LSC channels.  (But non-LSC channels seem to be OK.)  So we power cycled c1lsc (we couldn't ssh).

Then we noticed that computers were out of sync again (several timing fields said 16383 in the C0DAQ_RFMNETWORK screen).  We restarted c1iscey, c1iscex, c1lsc, c1susvme1, and c1susvme2.  The timing fields went back to 0.  But the tdsavg command still  intermittently said "ERROR: LDAQ - SendRequest - bad NDS status: 13".

The channel C1:LSC-SRM_OUT16 seems to work with tdsavg every time.

Let us know if you know how to fix this. 

 

  1561   Fri May 8 02:39:02 2009 pete, ranaUpdateLockingcrossover

attached plot shows MC_IN1/MC_IN2.  needs work.

This is supposed to be a measurement of the relative gain of the MCL and AO paths in the CM servo. We expect there to

be a more steep slope (ideally 1/f). Somehow the magnitude is very shallow and so the crossover is not stable. Possible

causes? Saturations in the measurement, broken whitening filters, extremely bad delay in the digital system? needs work.

 

Attachment 1: crossover.pdf
crossover.pdf
Attachment 2: photo.jpg
photo.jpg
  5   Fri Oct 19 16:11:38 2007 pkpOtherOMCOMC PZT response
Sam and I locked the laser to the OMC cavity and looked at the error signal as a function of the voltage applied to the OMC PZT.
Here are two plots showing the response as a function of frequency from 1 kHz to 100 kHz and another high-res response in the region of 4.5 kHz to 10 kHz.
Attachment 1: fullspec.jpg
fullspec.jpg
Attachment 2: 4.5to10.jpg
4.5to10.jpg
Attachment 3: 4.5to10.pdf
4.5to10.pdf
Attachment 4: fullres.pdf
fullres.pdf
  8   Mon Oct 22 19:27:14 2007 pkpOtherOMCPZT calibration/ transfer function.
We measured the PZT transfer function by comparing the PZT response of the circuit with the cavity in the loop, with that of the circuit without the cavity in the loop. Basically measure the transfer function of the whole loop with the laser/PZT and Op-amps in it. Then take another measurement of the transfer function of everything else besides the PZT and from both these functions, we can calculate the PZT response.

The calibration was done by using the error signal response to a triangular wave of volts applied to the PZT. A measurement of the slope of the error signal , which has three zero-crossings as the cavity sweeps through the sidebands, gives us the Volts/Hz response. In order to derive a frequency calibration of the x axis, we assume that the first zero crossing corresponds to the first side band (-29.5 MHz) and the third one corresponds with the other sideband (+29.5 MHz). And then by using the fact that we know the response of the cavity to a constant frequency shift, we can use the Volts/Hz measurement to calculate the Volts/nm calibration. The slope that was calculated was 3.2e-6 V/Hz and using the fact that the cavity is 1 m in length and the frequency is 1064 nm, we get a calibration of 0.9022 V/nm.

Attachment 1: calib.pdf
calib.pdf
Attachment 2: calibpzt2.pdf
calibpzt2.pdf
Attachment 3: all2.pdf
all2.pdf
Attachment 4: noPZT2.pdf
noPZT2.pdf
  82   Thu Nov 8 00:55:44 2007 pkpUpdateOMCSuspension tests
[Sam , Pinkesh]

We tried to measure the transfer functions of the 6 degrees of freedom in the OMS SUS. To our chagrin, we found that it was very hard to get the OSEMs to center and get a mean value of around 6000 counts. Somehow the left and top OSEMs were coupled and we tried to see if any of the OSEMs/suspension parts were touching each other. But there is still a significant coupling between the various OSEMs. In theory, the only OSEMS that are supposed to couple are [SIDE] , [LEFT, RIGHT] , [TOP1, TOP2 , TOP3] , since the motion along these 3 sets is orthogonal to the other sets. Thus an excitation along any one OSEM in a set should only couple with another OSEM in the same same set and not with the others. The graphs below were obtained by driving all the OSEMS one by one at 7 Hz and at 500 counts ( I still have to figure out how much that is in units of length). These graphs show that there is some sort of contact somewhere. I cant locate any physical contact at this point, although TOP2 is suspicious and we moved it a bit, but it seems to be hanging free now. This can also be caused by the stiff wire with the peek on it. This wire is very stiff and it can transmit motion from one degree of freedom to another quite easily. I also have a graph showing the transfer function of the longitudnal degree of freedom. I decided to do this first because it was simple and I had to only deal with SIDE, which seems to be decoupled from the other DOFs. This graph is similar to one Norna has for the longitudnal DOF transfer function, with the addition of a peak around 1.8 Hz. This I reckon could very be due to the wire, although it is hard to claim for certain. I am going to stop the measurement at this time and start a fresh high resolution spectrum and leave it running over night.

There is an extra peak in the high res spectrum that is disturbing.
Attachment 1: shakeleft.pdf
shakeleft.pdf
Attachment 2: shakeright.pdf
shakeright.pdf
Attachment 3: shakeside.pdf
shakeside.pdf
Attachment 4: shaketop1.pdf
shaketop1.pdf
Attachment 5: shaketop2.pdf
shaketop2.pdf
Attachment 6: shaketop3.pdf
shaketop3.pdf
Attachment 7: LongTransfer.pdf
LongTransfer.pdf LongTransfer.pdf LongTransfer.pdf
Attachment 8: Shakeleft7Nov2007_2.pdf
Shakeleft7Nov2007_2.pdf
Attachment 9: Shakeleft7Nov2007_2.png
Shakeleft7Nov2007_2.png
  87   Fri Nov 9 00:23:12 2007 pkpUpdateOMCX and Z resonances
I got a couple of resonance plots going for now. I am still having trouble getting the Y measurement going for some reason. I will investigate that tommorow. But for tonight and tommorow morning, here is some food for thought. I have attached the X and Z transfer functions below. I compared them to Norna's plots - so just writing out what I was thinking -

Keep in mind that these arent high res scans and have been inconviniently stopped at 0.5 Hz Frown.

Z case --

I see two small resonances and two large ones - the large ones are at 5.5 Hz and 0.55 Hz and the small ones at 9 Hz and 2 Hz respectively. In Norna's resonances, these features arent present. Secondly, the two large peaks in Norna's measurement are at 4.5 Hz and just above 1 Hz. Which was kind of expected, since we shortened the wires a bit, so one of the resonances moved up and I suppose that the other one moved down for the same reason.

X case --

Only one broad peak at about 3 Hz is seen here, whereas in Norna's measurement, there were two large peaks and one dip at 0.75 Hz and 2.5 Hz. I suspect that the lower peak has shifted lower than what I scanned to here and a high res scan going upto 0.2 Hz is taking place overnight. So we will have to wait and watch.

Pitch Roll and Yaw can wait for the morning.
Attachment 1: Xtransferfunc.pdf
Xtransferfunc.pdf Xtransferfunc.pdf Xtransferfunc.pdf
Attachment 2: Ztransferfunc.pdf
Ztransferfunc.pdf Ztransferfunc.pdf Ztransferfunc.pdf