  1637   Mon Jun 1 14:33:42 2009 robConfigurationComputer Scripts / Programsop540m Monitor added to web status

I added op540m's display 0 (the northern-most monitor in the control room) to the MEDM screens webpage: https://nodus.ligo.caltech.edu:30889/medm/screenshot.html

 

Now we can see the StripTool displays that are usually parked on that screen.

 

 

  1638   Mon Jun 1 14:49:07 2009 robSummaryPSLpsl thoughts

Some thoughts on what happened with the MOPA cooling. 

Some unknown thing happened to precipitate the initial needle valve jiggle, which unleashed a torrent of flow through the NPRO.  This flow was made possible by the fact that the cooling lines are labeled confusingly, and so flow was going backwards through the needle valve, which was thus powerless to restrict it.  The NPRO got extremely cold, and most of the chiller's cooling power was being used to unnecessarily cool the NPRO.  So, the PA was not getting cooled enough.  At this, point, reversing the flow probably would have solved everything.  Instead, we turned off the chiller and thus discovered the flaky start-motor capacitor. 

Now we have much more information, flow meters in the NPRO and main cooling lines, a brand-new, functioning needle valve, a better understanding of the chiller/MOPA settings necessary for operation, and the knowledge of what happens when you install a needle valve backwards.

 

  1642   Tue Jun 2 23:12:08 2009 robConfigurationComputersntp on op440m

I restarted ntpd on op440m to solve a "synchronization error" that we were having in DTT.  I also edited the config file (/etc/inet/ntp.conf) to remove the lines referring to rana as an ntp server; now only nodus is listed.

To do this:

log in as root

/usr/local/bin/ntpd -c /etc/inet/ntp.conf

  1650   Thu Jun 4 01:32:20 2009 robConfigurationLSCAS port mode scan

Here is a set of mode scans of the AS port, using the OMC as a mode scanner.  The plot overlays various configurations of the IFO.  

To remove PZT nonlinearity, each scan was individually flattened in fsr-space by polynomial (3rd order) fitting to some known peak locations (the carrier and RF sidebands).
 

Attachment 1: modes.png
modes.png
  1656   Fri Jun 5 13:51:49 2009 robUpdateLockingtdsavg failure in cm_step script

Quote:

the command

tdsavg 5 C1:LSC-PD4_DC_IN1

was causing grievous woe in the cm_step script.  It turned out to fail intermittently at the command line, as did other LSC channels.  (But non-LSC channels seem to be OK.)  So we power cycled c1lsc (we couldn't ssh).

Then we noticed that computers were out of sync again (several timing fields said 16383 in the C0DAQ_RFMNETWORK screen).  We restarted c1iscey, c1iscex, c1lsc, c1susvme1, and c1susvme2.  The timing fields went back to 0.  But the tdsavg command still  intermittently said "ERROR: LDAQ - SendRequest - bad NDS status: 13".

The channel C1:LSC-SRM_OUT16 seems to work with tdsavg every time.

Let us know if you know how to fix this. 

 

 Did you try restarting the framebuilder?

 

What you type is in bold:

op440m> telnet fb40m 8087

daqd> shutdown

 

  1663   Tue Jun 9 23:25:24 2009 robUpdateLocking?

Quote:

Quote:

Lock acquisition is proceeding smoothly for the most part, but there is a very consistent failure point near the end of the cm_step script.

Near the end of the procedure, while in RF common mode, the sensing for the MCL path of the common mode servo is transitioned from a REFL 166I signal which comes into the LSC whitening board from the demodulator, to another copy of the signal which has passed through the common mode board, and is coming out of the Length output of the common mode board.  We do this because the signal which comes through the CM board sees the switchable low-frequency boost filter, and so both paths of the CM servo (AO and MCL) can get that filter switched on at the same time.

The problem is occurring after this transition, which works reliably.  However, when the script tries to remove the final CARM offset, and bring the offset to zero, lock is abruptly lost.  DARM, CM, and the crossover all look stable, and no excess noise appears while looking at the DARM, CARM, MCF spectra.  But lock is always lost right about the same offset. 

Saturation somewhere?

 I've seen this before. At that time, the problem was gone spontaneously the next day.

You could stop just before the offset reaches zero and then try to slowly reduce the offset manually to see where is the threshold.

 

 

Well, it hasn't gone away yet.  It happened Sat, Mon, and Tues afternoon, as well as Friday.  The threshold varies slightly, but is always around ~200-300 cnts.   I've tried reducing the offset with the signal coming from the CM board and the signal not going through the CM board, I've also tried jumping the signal to zero (rather than a gradual reduction). 

Tonight we'll measure the MC length and set the modulation frequencies, and maybe try some MZ tweaking to do RFAMMon minimization.    

  1670   Fri Jun 12 02:01:03 2009 robUpdateComputer Scripts / ProgramsDRM matrix diagonalization

I started two scripts, senseDRM and loadDRMImatrixData.m, which Peter will bang on until they're correct.  They're in the $SCRIPTS/LSC directory.  The first is a perl script which uses TDS tools to drive the DRM optics and measure the response at the double demod photo-detectors, and write these results to a series of files loadable by matlab.  The second loads the output from the first script, inverts the resulting sensing matrix to get an input matrix, and spits out a tdswrite command which can be copied and pasted into a terminal to load the new input matrix values. 

What's left is mainly in figuring out how to do the matrix inversion properly.  Right now the script does not account for the output matrix, the gains in the feedback filters at the measurement frequency, or the fact that we'll likely want the UGF of our loops to be less than the measurement frequency.  Peter's going to hash out these details.

  1675   Tue Jun 16 02:09:31 2009 robUpdateLockingDD handoff finally working

I had trouble getting the SRC handoff from SD to DD to work.  I tried different gains, flipping the PD7 & 8 demod phases by 180 degrees, and messing with the output matrix to reduce cross-couplings in the state with MICH & PRC on DD and SRC on SD.  Eventually I decided to try to make the DRM matrix diagonalization work. 

It does, mostly.  The handoff is now stable, and the loops all have UGFs around 100Hz.  So, tonight anyways, it's possible to run senseDRM and then loadDRMImatrixData.m and run the resulting tdswrite command, and have a working handoff.  I had to eliminate a few PDs (PD5 & PD10) to get it to work properly. 

It would be nice if this script would measure all the PDs and allow individual setting of loop UGFs and measurement frequencies. 

 

 

  1676   Tue Jun 16 03:43:04 2009 robUpdateLockingsame troubles

Lock is still being lost, right at the end of the process when trying to reduce the CARM offset to zero.

  1678   Tue Jun 16 14:02:16 2009 robUpdateLockinglocked

The IFO is locked, at the operating point (zero CARM offset).  The problem with reducing the residual CARM offset in the last stage appears to have been because the common mode gain was getting too high, and so the loop was going unstable at high frequencies. 

The cm_step script is currently a confusing mess, with all the debugging statements.  I'll clean it up this afternoon and check that it still works.

  1680   Tue Jun 16 17:38:35 2009 robUpdateLockingDeWhites ON

With the common mode servo bandwidth above 30kHz and the BOOST on (1), I was able to switch on the test mass dewhitening.  Finally.

  1682   Wed Jun 17 01:07:50 2009 robConfigurationComputersmatapps on /cvs/cds

I checked out a copy of matapps into /cvs/cds/caltech/apps/lscsoft so that I could find the matlab function strassign.m, which is necessary for some old mDV commands to run.  I don't know why it became necessary or why it disappeared if it did.

  1683   Wed Jun 17 01:09:47 2009 robUpdateComputers/cvs/cds 91% full

In /cvs/cds/caltech

 


1.6M    2008-8-15.pdf
2.9M    40mUpgradeOpticalLayoutPlan01.pdf
2.4M    alh
19M     apache
18G     apps
11M     archive
4.0K    authorized_keys2
8.0K    backup.notes
8.0K    backup.notes~
1.9G    build
62G     burt
47M     cds
13M     cds40m
37M     chans
70G     conlog
52K     crontab
12K     cshrc.40m
12K     cshrc.40m~
36M     diag
1.4G    dmt
8.2M    framecpp-0.2.0
1.7M    free_080730.pdf
57M     gds
9.8G    home
60K     hooks
8.0K    hosts.40m
4.0K    id_rsa
10M     iscmodeling
110M    ldg-4.7
648M    libs
4.0K    log2.txt
224K    logs
0       log.txt
238M    medm
344M    NB
148M    NB_080304
211M    NB_080307
401M    NB40
1.2G    noisebudget.071109
837M    noisebudget.bak.20060623
3.5M    oldtarget
123M    root
5.7M    savesets
208K    schematics
655M    scripts
13G     scripts_archive
1.1M    state
3.7G    svn
4.0K    svn-commit.tmp
7.3G    target
295M    target_archive
6.7M    test
72K     test.png
4.0K    tmp
8.0K    typescript
35G     users
205M    wind

  1684   Thu Jun 18 23:08:46 2009 robUpdateLockingspectrum

Here's a noise spectrum of the RSE interferometer, in anti-spring mode, with RF readout.  I'd say the calibration is "loose."

I used the Buonanno & Chen modification of the KLMTV IFO transfer functions to model the DARM opto-mechanical response.  I just guessed at the quadrature, and normalized the optical gain at the frequency of the calibration line used (927Hz, not visible on the plot).

Attachment 1: DARMnoise_929352240.png
DARMnoise_929352240.png
Attachment 2: DARMnoise_929352240.pdf
DARMnoise_929352240.pdf
  1685   Thu Jun 18 23:20:40 2009 robSummaryLSCI-Q ratio with RF detected DARM

This is a ratio of PD1_I to PD1_Q (so a ratio of the two quadratures of AS166), measured in an anti-spring state.  It's not flat because our set up has single sideband RF heterodyne detection, and using a single RF sideband as a local oscillator allows one to detect different quadratures by using different RF demodulation phases.   So the variation in frequency is actually a measure of how the frequency response of DARM changes when you vary the detection quadrature.  This measure is imperfect because it doesn't account for the effect of the DARM loop.

Even though you can choose your detection quadrature with this setup, you can't get squeezed quantum noise with a single RF sideband.  The quantum noise around the other (zero-amplitude) RF sideband still gets mixed in, and negates any squeezing benefits.

Attachment 1: IQratio.jpg
IQratio.jpg
  1699   Wed Jun 24 14:43:19 2009 robUpdateComputerstdsresp on linux => pzresp

 

tdsresp is broken on our linux control room machines.  I made a little perl replacement which uses the DiagTools.pm perl module, called pzresp.  It's in the $SCRIPTS/general directory, and so in the path of all the machines.  I also edited the cshrc.40m file so that on linux machines tdsresp points to this perl replacement.

I've patched DiagTools.pm to circumnavigate the tdsdmd bug described here.  I also added a function to DiagTools.pm called diagRespNoLog, which is just like diagResp but without that pesky log file.

 


Here's the output from the tdsresp binary on CentOS:
allegra:~>tdsresp 941.54 10000 100 10 C1:LSC-ITMX_EXC C1:LSC-PD1_Q C1:LSC-PD1_I
nan nan nan nan nan nan nan
nan nan nan nan nan nan nan
nan nan nan nan nan nan nan
*** glibc detected *** tdsresp: free(): invalid next size (fast): 0x089483e8 ***
======= Backtrace: =========
/lib/libc.so.6[0xa800f1]
/lib/libc.so.6(cfree+0x90)[0xa83bc0]
/usr/lib/libstdc++.so.6(_ZdlPv+0x21)[0xf7f36571]
tdsresp[0x8057fbb]
tdsresp[0x805b394]
/lib/libc.so.6(__libc_start_main+0xdc)[0xa2ce8c]
tdsresp(__gxx_personality_v0+0x169)[0x804ddd1]
======= Memory map: ========
00242000-00249000 r-xp 00000000 fd:00 15400987                           /lib/librt-2.5.so
00249000-0024a000 r--p 00006000 fd:00 15400987                           /lib/librt-2.5.so
0024a000-0024b000 rw-p 00007000 fd:00 15400987                           /lib/librt-2.5.so
009f9000-00a13000 r-xp 00000000 fd:00 15400963                           /lib/ld-2.5.so
00a13000-00a14000 r--p 00019000 fd:00 15400963                           /lib/ld-2.5.so
00a14000-00a15000 rw-p 0001a000 fd:00 15400963                           /lib/ld-2.5.so
00a17000-00b55000 r-xp 00000000 fd:00 15400974                           /lib/libc-2.5.so
00b55000-00b57000 r--p 0013e000 fd:00 15400974                           /lib/libc-2.5.so
00b57000-00b58000 rw-p 00140000 fd:00 15400974                           /lib/libc-2.5.so
00b58000-00b5b000 rw-p 00b58000 00:00 0 
00b5d000-00b70000 r-xp 00000000 fd:00 15400984                           /lib/libpthread-2.5.so
00b70000-00b71000 r--p 00012000 fd:00 15400984                           /lib/libpthread-2.5.so
00b71000-00b72000 rw-p 00013000 fd:00 15400984                           /lib/libpthread-2.5.so
00b72000-00b74000 rw-p 00b72000 00:00 0 
00b76000-00b78000 r-xp 00000000 fd:00 15400981                           /lib/libdl-2.5.so
00b78000-00b79000 r--p 00001000 fd:00 15400981                           /lib/libdl-2.5.so
00b79000-00b7a000 rw-p 00002000 fd:00 15400981                           /lib/libdl-2.5.so
00b7c000-00ba1000 r-xp 00000000 fd:00 15400975                           /lib/libm-2.5.so
00ba1000-00ba2000 r--p 00024000 fd:00 15400975                           /lib/libm-2.5.so
00ba2000-00ba3000 rw-p 00025000 fd:00 15400975                           /lib/libm-2.5.so
00bca000-00bdd000 r-xp 00000000 fd:00 15401011                           /lib/libnsl-2.5.so
00bdd000-00bde000 r--p 00012000 fd:00 15401011                           /lib/libnsl-2.5.so
00bde000-00bdf000 rw-p 00013000 fd:00 15401011                           /lib/libnsl-2.5.so
00bdf000-00be1000 rw-p 00bdf000 00:00 0 
00dca000-00dd5000 r-xp 00000000 fd:00 15400986                           /lib/libgcc_s-4.1.2-20080825.so.1
00dd5000-00dd6000 rw-p 0000a000 fd:00 15400986                           /lib/libgcc_s-4.1.2-20080825.so.1
08048000-080b7000 r-xp 00000000 00:17 6455328                            /cvs/cds/caltech/apps/linux/tds/bin/tdsresp
080b7000-080ba000 rw-p 0006e000 00:17 6455328                            /cvs/cds/caltech/apps/linux/tds/bin/tdsresp
080ba000-080bb000 rw-p 080ba000 00:00 0 
0893d000-0896b000 rw-p 0893d000 00:00 0                                  [heap]
f5e73000-f5e74000 ---p f5e73000 00:00 0 
f5e74000-f6874000 rw-p f5e74000 00:00 0 
f692d000-f6931000 r-xp 00000000 fd:00 15400995                           /lib/libnss_dns-2.5.so
f6931000-f6932000 r--p 00003000 fd:00 15400995                           /lib/libnss_dns-2.5.so
f6932000-f6933000 rw-p 00004000 fd:00 15400995                           /lib/libnss_dns-2.5.so
f6956000-f6a12000 rw-p f6a31000 00:00 0 
f6a74000-f6a7d000 r-xp 00000000 fd:00 15400997                           /lib/libnss_files-2.5.so
f6a7d000-f6a7e000 r--p 00008000 fd:00 15400997                           /lib/libnss_files-2.5.so
f6a7e000-f6a7f000 rw-p 00009000 fd:00 15400997                           /lib/libnss_files-2.5.so
f6a7f000-f6a80000 ---p f6a7f000 00:00 0 
f6a80000-f7480000 rw-p f6a80000 00:00 0 
f7480000-f7481000 ---p f7480000 00:00 0 
f7481000-f7e83000 rw-p f7481000 00:00 0 
f7e83000-f7f63000 r-xp 00000000 fd:00 6236924                            /usr/lib/libstdc++.so.6.0.8
f7f63000-f7f67000 r--p 000df000 fd:00 6236924                            /usr/libAbort

 
  1747   Wed Jul 15 11:38:31 2009 robUpdateLockingMC_F channel dead

It's railed.  This is what halted locking progess on Monday night, as this channel is used for the offloadMCF script, which slowly feeds back a CARM signal to the ETMs to prevent the VCO from saturating.

 

Attached is a 5 day trend, which shows that the channel went dead a few days ago.  All the channels shown are being collected from the same ICS110B (I think), but only some are dead.  It looks like they went dead around the time of the "All computers down" from Sunday.

Attachment 1: mcfdead.png
mcfdead.png
  1758   Thu Jul 16 14:41:38 2009 robUpdateLockingMC_F channel dead

Quote:

It's railed.  This is what halted locking progess on Monday night, as this channel is used for the offloadMCF script, which slowly feeds back a CARM signal to the ETMs to prevent the VCO from saturating.

 

Attached is a 5 day trend, which shows that the channel went dead a few days ago.  All the channels shown are being collected from the same ICS110B (I think), but only some are dead.  It looks like they went dead around the time of the "All computers down" from Sunday.

 Attached are the channels being recorded from the ICS110B in 1Y2 (the IOO rack).  Channels 12, 13, 16, 17, 22, 24, 25 appear to have gone dead after the computer problems on Sunday.

Attachment 1: IOO_ICS_0_15.png
IOO_ICS_0_15.png
Attachment 2: IOO_ICS_15_32.png
IOO_ICS_15_32.png
  1759   Thu Jul 16 14:54:05 2009 robUpdateLockingMC_F channel dead

Quote:

Quote:

It's railed.  This is what halted locking progess on Monday night, as this channel is used for the offloadMCF script, which slowly feeds back a CARM signal to the ETMs to prevent the VCO from saturating.

 

Attached is a 5 day trend, which shows that the channel went dead a few days ago.  All the channels shown are being collected from the same ICS110B (I think), but only some are dead.  It looks like they went dead around the time of the "All computers down" from Sunday.

 Attached are the channels being recorded from the ICS110B in 1Y2 (the IOO rack).  Channels 12, 13, 16, 17, 22, 24, 25 appear to have gone dead after the computer problems on Sunday.

 This has been fixed by one of the two most powerful & useful IFO debugging techniques: rebooting.  I keyed the crate in 1Y2.

  1762   Sun Jul 19 22:38:24 2009 robOmnistructureGeneralWeb screenshots aren't being updated

Quote:

Before heading back to the 40m to check on the computer situation, I thought I'd check the web screenshots page that Kakeru worked on, and it looks like none of the screens have been updated since June 1st.  I don't know what the story is on that one, or how to fix it, but it'd be handy if it were fixed.

 Apparently I broke this when I added op540m to the webstatus page.  It's fixed now.

  1764   Mon Jul 20 12:35:21 2009 robConfigurationLockingalignment biases funny

I found the alignment biases for the PRM and the SRM in a funny state.  It seemed like they had been "saved" in badly misaligned position, so the restore scripts on the IFO configure screen were not working.  I've manually put them into a better alignment.

  1777   Wed Jul 22 11:18:49 2009 robConfigurationComputerssticky sliders

Quote:

Quote:

Quote:

Quote:

All suspentions are kicked up. Sus dampings and  oplev servos turned off.

c1iscey and c1lsc are down. c1asc and c1iovme are up-down

 The computers and RFM network are up working again. A boot fest was necessary.Then I restored all the parameters with burtgooey.

The mode cleaner alignment is in a bad state. The autolocker can't get it locked. I don't know what caused it to move so far from the good state that it was till this afternoon.  I went tuning the periscope but the cavity alignment is so bad that it's taking more time than expected. I'll continue working on that tomorrow morning.

 I now suspect that after the reboot the MC mirrors didn't really go back to their original place even if the MC sliders were on the same positions as before.

 Alberto, Rob,

we diagnosed the problem. It was related with sticky sliders. After a reboot of C1:IOO the actual output of the DAC does not correspond anymore to the values read on the sliders. In order to update the actual output it is necessary to do a change of the values of the sliders, i.e. jiggling a bit with them.

 I've updated the slider twiddle script to include the MC alignment biases.  We should run this script whenever we reboot all the hardware, and add any new sticky sliders you find to the end of the script.  It's at

 

/cvs/cds/caltech/scripts/Admin/slider_twiddle

  1780   Wed Jul 22 18:04:14 2009 robOmnistructureComputersweird noise coming from Gigabit switch

in the rack next to the printer.  It sounds like a fan is hitting something.

  1791   Sat Jul 25 16:04:32 2009 robUpdatePSLAligning the beam to the Faraday

Quote:

When I turned them on, the control signal in Pitch from WFS2 started going up with no stop. It was like the integrator in the loop was fed with a DC bias. The effect of that was to misalign the MC cavity from the good state in which it was with the only length control on (that is, transmission ~2.7, reflection ~ 0.4).

I don't know why that is happening. To exclude that it was due to a computer problem I first burtrestored C1IOO to July the 18th, but since that did not help, I even restarted it. Also that didn't solve the problem.

 

 

At least one problem is the mis-centering of the resonant spot on MC2, which can be viewed with the video monitors.  It's very far from the center of the optic, which causes length-to-angle coupling that makes the mulitple servos which actuate on MC2 (MCL, WFS, local damping) fight each other and go unstable.

  1812   Thu Jul 30 03:10:18 2009 robUpdateIOOMC tweaked further

I tilted the periscope beam and aligned the MC.  Now the spot at the Faraday entrance is near the center of the aperture in up/down space.  The arm powers are only going up to ~0.8, though.  Maybe we should try a little bit of left/right. 

I looked at the IP POS spot with a viewer card, and it looked round, so no obvious egregious clipping in the Faraday.  Someone might take a picture with one of the GigE camera and get us a beam profile there.

We no longer have an MC1 and MC3 camera view.

I can see a bright scatterer that can be seen from the east viewport of the BSC, but I can't tell what it is.  It could be a ghost beam. 

It would be nice to get an image looking into the north viewport of the IOO chamber.  I can't see in there because the BS oplev table is in the way. 

  1828   Tue Aug 4 16:12:27 2009 robUpdateComputersmini boot fest

Quote:

Last night Rana noticed that the overflows on the ITM and ETM coils were a crazy huge number.  Today I rebooted c1dcuepics, c1iovme, c1sosvme, c1susvme1 and c1susvme2 (in that order).  Rob helped me burt restore losepics and iscepics, which needs to be done whenever you reboot the epics computer.

Unfortunately this didn't help the overflow problem at all.  I don't know what to do about that.

 

Just start by re-setting them to zero.  Then you have to figure out what's causing them to saturate by watching time series and looking at spectra.

  1858   Fri Aug 7 16:14:57 2009 robOmnistructureVACUPS failed

Steve, Rana, Ben, Jenne, Alberto, Rob

 

UPS in the vacuum rack failed this afternoon, cutting off power to the vacuum control system.  After plugging all the stuff that had been plugged into the UPS into the wall, everything came back up.  It appears that V1 closed appropriately, TP1 spun down gracefully on its own battery, and the pressure did not rise much above 3e-6 torr. 

 

The UPS fizzed and smelled burnt.  Rana will order a new, bigger, better, faster one.

 

  1863   Fri Aug 7 18:06:24 2009 robOmnistructureVACopening V1 when PTP1 is broken

We've had a devil of a time getting V1 to open, due to the Interlock code. 

 

The short story is that if C1:Vac-PTP1_pressure > 1.0, the interlock code won't let you push the button to open V1 (but it won't close V1). 

 

PTP1 is broken, so the interlock was frustrating us.   It's been broken for a while, but this hasn't bitten us till now.

 

We tried swapping out the controller for PTP1 with one of Bob's from the Bake lab, but it didn't work. 

 

It said "NO COMM" in the C1:Vac-PTP1_status, so I figured it wouldn't update if we just used tdswrite to change C1:Vac-PTP1_pressure to 0.0.  This actually worked, and V1 is now open.  This is a temporary fix.

  1874   Mon Aug 10 15:24:17 2009 robSummaryPSLMZ bad

I think the MZ pzt is broken/failing.  I'm not sure how else to explain this behavior.

The first bit of the time series is a triangle wave into the DC offset (output) field, over approximately the whole range (0-250V).   You can see the fringe visbility is quite small.  The triangle wave is stopped, and I then maxed out the offset slider to get to the "high" power point from the triangle wave sweep. Then for a little while with the PZT is held still, and the power still increases.  The MZ is then locked, and you can see the PZT voltage stay about the same but the power continues to rise over the next ~10 minutes or so.

Attachment 1: brokenMZpzt.png
brokenMZpzt.png
  1875   Mon Aug 10 15:56:12 2009 robSummaryPSLMZ bad redux

Quote:

I think the MZ pzt is broken/failing.  I'm not sure how else to explain this behavior.

 

The first bit of the time series is a triangle wave into the DC offset (output) field, over approximately the whole range (0-250V).   You can see the fringe visbility is quite small.  The triangle wave is stopped, and I then maxed out the offset slider to get to the "high" power point from the triangle wave sweep. Then for a little while with the PZT is held still, and the power still increases.  The MZ is then locked, and you can see the PZT voltage stay about the same but the power continues to rise over the next ~10 minutes or so.

 

 

 

This plot answers the previous question, and raises a new one--what the heck is MZTRANSPD?  I'd guess the pins are unconnected--it's just floating, and somehow picking up the MZ_PZT signal.

 

 

Attachment 1: badMZtrans.png
badMZtrans.png
  1876   Mon Aug 10 16:37:27 2009 robUpdatePSLMZ alignment touched

I aligned the MZ.  The reflection went from .86 to .374

  1878   Mon Aug 10 17:27:47 2009 robConfigurationLSCTRX, TRY gain

 

These are the settings which determine the transmon (eg, TRX) amplitude, and which are updated by the matchTransMon scripts.

For the X arm

 

op440m:AutoDither>tdsread C1:LSC-TRX_GAIN C1:LSC-LA_PARAM_FLT_01 C1:LSC-LA_PARAM_FLT_00
0.0023
0.155
119.775

 

For the Y arm

op440m:AutoDither>tdsread C1:LSC-TRY_GAIN C1:LSC-LA_PARAM_FLT_04 C1:LSC-LA_PARAM_FLT_03
0.00237196
-0.116
19.9809


  1884   Tue Aug 11 01:21:55 2009 robUpdatePSLMZ needs some attention

the servo needs some work. 

 

2 day trend

 

Attachment 1: badMZservo.png
badMZservo.png
  1889   Wed Aug 12 02:00:32 2009 robUpdateLockingreport

Spent a lot of time aligning tonight.  The BS is not staying put--sometimes after a lock loss it gets badly mis-aligned. 

DD handoff is working, after putting beam on REFL diodes and running senseDRM script.

  1919   Mon Aug 17 09:52:04 2009 robSummaryGeneralconlogger restarted

I restarted the conlogger on op340m.  This needs to be done when op340m is rebooted--it wasn't done for some reason and so we've lost several days of controls records.

  1921   Mon Aug 17 17:48:49 2009 robSummaryGeneralconlogger restarted

Quote:

I restarted the conlogger on op340m.  This needs to be done when op340m is rebooted--it wasn't done for some reason and so we've lost several days of controls records.

 I added a cronjob on op340m to check every half-hour if the conlog is running, and if not, restart it. 

  1924   Tue Aug 18 15:16:15 2009 robUpdateIOOMC WFS working again

Rob, Yoichi

 

The MC WFS have apparently been bad for a few days, causing the MC alignment to drift away at DC.  We tried a few things to fix it, including jiggling some EPICS settings in the WFS head & demod screens.  This seemed to work for WFS1 but not WFS2.  Confused, we decided to go stare at the rack 1Y2.  While doing that, we noticed that the top two Sorensens in 1Y1 (these are directly below the Guralp box) were at different voltages from nominal.  The 5V had dropped to 4.2V and the 24V was at 24.6V.  We adjusted the knobs until these were set correctly.  After this, the MC WFS appear to work again.

 

When working in a rack, you must be as careful about accidentally touching things as when working on an optical table.

  1930   Wed Aug 19 23:57:35 2009 robUpdateLockingreport

 

locking work proceeding apace tonight.

diagonalized DRM with setDDphases & senseDRM

initial locks are fairly quick, aqstep script succeeds reliably.

first part of cm_step (handoff CARM-> MCL) usually works.

tuning up later parts of cm_step (presumably due to optical gain changes resulting from MOPA decline). 

got to arm powers ~60.

  1960   Fri Aug 28 13:49:07 2009 robUpdateLockingRF CARM hand off problem

Quote:
Last night, the lock script proceeded to the RF CARM hand-off about half of the time.
However, the hand off was still unsuccessful.

It failed instantly when you turn on the REFL1 input of the CM board, even
when the REFL1 input gain was very low, like -28dB.

I went to the LSC rack and checked the cabling.
The output from the PD11_I (REFL_2) demodulation board is split
into two paths. One goes directly to the ADC and the other one goes
to an SR560. This SR560 is used just as an inverter. Then
the signal goes to the REFL1 input of the CM board.

I found that the SR560 was set to the A-B mode, but B input was open.
This made the signal very noisy. So I changed it to A only mode.
There was also a 1/4 attenuator between the PD11_I output and the SR560.
I took it out and reduced the gain of SR560 from 10 to 2.
These changes allowed me to increase the REFL1 gain to -22dB or so.
But it is still not enough.

I wanted to check the CM open loop TF before the hand-off, but I could
not do that because the lock was lost instantly as soon as I enabled the
test input B of the CM board.
Something is wrong with the board ?

Using the PD11_I signal going into the ADC, I measured the transfer functions
from the CM excitation (digital one) to the REFL_DC (DC CARM signal) and PD11_I.
The TF shapes matched. So the PD11_I signal itself should be fine.

We should try:
* See if flipping the sign of PD11_I signal going to REFL1 input solve the problem.
* Try to measure the CM analog TF again.
* If the noise from the servo analyzer is a problem, try to increase the input gains
of the CM board and reduce the output gain accordingly, so that the signal flowing
inside the CM board is larger.



I'd bet it's in a really twitchy state by the time the script gets to the RF CARM handoff, as the script is not really validated up to that point. It's just the old script with a few haphazard mods, so it needs to be adjusted to accomodate the 15% power drop we've experienced since the last time it was locked.

The CM servo gain needs to be tweaked earlier in the script--you should be able to measure the AO path TF with the arm powers at 30 or so. I was able to do this with the current SR785 setup earlier this week without any trouble.

The 1/4 attenuator is there to prevent saturations on the input to the SR560 when there's still a CARM offset.

Not sure if flipping the sign of PD11 is right, but it's possible we compensated the digital gains and forgot about it. This signal is used for SRCL in the initial acquisition, so we'd have noticed a sign flip.
  1989   Thu Sep 17 14:17:04 2009 robUpdateComputersawgtpman on c1omc failing to start

[root@c1omc controls]# /opt/gds/awgtpman -2 &
[1] 16618
[root@c1omc controls]# mmapped address is 0x55577000
32 kHz system
Spawn testpoint manager
no test point service registered
Test point manager startup failed; -1

[1]+  Exit 1                  /opt/gds/awgtpman -2

 

 

 

  1990   Thu Sep 17 15:05:47 2009 robUpdateComputersawgtpman on c1omc failing to start

Quote:

[root@c1omc controls]# /opt/gds/awgtpman -2 &
[1] 16618
[root@c1omc controls]# mmapped address is 0x55577000
32 kHz system
Spawn testpoint manager
no test point service registered
Test point manager startup failed; -1

[1]+  Exit 1                  /opt/gds/awgtpman -2

 

 

 

 

 

 

This turned out to be fallout from the /cvs/cds transition.  Remounting and restarting fixed it.

  1991   Fri Sep 18 14:25:00 2009 robOmnistructurePSLwater under the laser chiller

rob, koji, steve

We noticed some water (about a cup) on the floor under the NESLAB chiller today.  We put the chiller up on blocks and took off the side panel for a cursory inspection, but found no obvious leaks.  We'll keep an eye on it.

  1994   Wed Sep 23 17:32:37 2009 robAoGComputersGremlins in the RFM

A cosmic ray struck the RFM in the framebuilder this afternoon, causing hours of consternation.  The whole FE system is just now coming back up, and it appears the mode cleaner is not coming back to the same place (alignment).

 

rob, jenne

  1997   Thu Sep 24 15:45:27 2009 robUpdateIOOMC OLG

I measured the mode cleaner open loop gain with the HP3563A

The UGF is 64kHz, phase margin is 28 deg.

  2013   Mon Sep 28 17:39:34 2009 robUpdatePSLproblems

The PSL/IOO combo has not been behaving responsibly recently. 

The first attachment is a 15 day trend of the MZ REFL, ISS INMON, and MC REFL power.  These show two separate problems--recurring MZ flakiness, which may actually be a loose cable somewhere which makes the servo disengage.  Such disengagement is not as obvious with the MZ as it is with other systems, because the MZ is relatively stable on its own.  The second problem is more recent, just starting in the last few days.  The MC is drifting off the fringe, either in alignment, length, or both.  This is unacceptable.

The second attachment is a two-day trend of the MC REFL power.  Last night I carefully put the beam on the center of the MC-WFS quads.  This appears to have lessened the problem, but it has not eliminated it. 

It's probably worth trying to re-measure the MCWFS system to make sure the control matrix is not degenerate. 

Attachment 1: problems.png
problems.png
Attachment 2: problems2.png
problems2.png
  2016   Tue Sep 29 01:50:10 2009 robConfigurationLSCnew modulation frequencies

Mode cleaner length measured tonight.

 

33196198

132784792

165980990

199177188

[Tag by KA: modulation frequency, MC length]

  2019   Tue Sep 29 16:14:44 2009 robConfigurationElectronicsRob is breaking stuff....

Quote:

Koji and I were looking for an extender card to aid with MZ board testing.  Rob went off on a quest to find one.  He found 2 (in addition to the one in the drawer near the electronics bench which says "15V shorted"), and put them in some empty slots in 1X1 to test them out.  Somehow, this burned a few pins on each board (1 pin on one of them, and 3 pins on the other). We now have 0 functioning extender cards: unfortunately, both extender cards now need fixing.  The 2 slots that were used in 1X1 now have yellow electrical tape covering the connectors so that they do not get used, because the ends of the burnt-off pins may still be in there. 

In other, not-Rob's-fault news, the Martian network is down...we're going to try to reset it so that we have use of the laptops again.

 

This happened when I plugged the cards into a crate with computers, which apparently is a no-no.  The extender cards only go in VME crates full of in-house, LIGO-designed electronics.

  2024   Tue Sep 29 23:43:46 2009 robUpdateSUSITMY UL OSEM

We had a redo of elog entry 975 tonight.  The noisy OSEM was fixed by jiggling the rack end of the long cable.  Don't know exactly where--I also poked around the OSEM PD interface board.

In the attached PDF the reference trace is the noisy one.

Attachment 1: ITMYosemBAD.pdf
ITMYosemBAD.pdf
  2026   Wed Sep 30 01:04:56 2009 robUpdateComputersgrief

much grief.  somehow a burt restore of c1iscepics failed to work, and so the LSC XYCOM settings were not correct.  This meant that the LSC whitening filter states were not being correctly set and reported, making it difficult to lock for at least the last week or so.

Attachment 1: C1LSC_XYCOM.jpg
C1LSC_XYCOM.jpg
  2027   Wed Sep 30 02:01:28 2009 robUpdateLockingweek

It's been a miserable week for lock acquisition, with each night worst than the last.  The nadir was around Sunday night, when I couldn't even get a PRM to lock stably, which meant that the auto-alignment scripts could not finish successfully.  It now appears that was due to some XYCOM mis-settings. 

We've also been having problems with timing for c1susvme2.  Attached is a one-hour plot of timing data for this cpu, known as SRM.  Each spike is an instance of lateness, and a potential cause of lock loss.  This has been going on for a quite a while.

Tonight we also encountered a large peak in the frequency noise around 485 Hz.  Changing the MZ lock point (the spot in the PZT range) solved this.

 

Attachment 1: srmcpu.png
srmcpu.png
ELOG V3.1.3-