  15443   Tue Jun 30 22:00:04 2020 gautamUpdateElectronicsGlitchy POX resurfaces

This problem reared its ugly head again. I am inclined to believe the problem is electronic and not on the light, since the POY channels seem immune to this issue (see Attachment #1). I will investigate in the daytime tomorrow. Note that while the POX photodiode head has ~twice the transimpedance than POY (per measurement), the POY signal gets amplified by a ZHL-500-HLN amplifier before heading to the demod electronics (nominal gain is 19dB = x9). There is also some imbalance in the light level at the photodiodes I guess, because overall, the PDH fringe is ~twice as large for the Y arm as the X arm. Basically, the y-axes of the attached plot cannot be directly compared between POX and POY.

Mostly this is an annoyance - right now, the POX signal is only used for locking and dither aligning the X arm cavity, and so once that is done, the locking can proceed (as long as the other channels, e.g. REFL11, aren't glitching as well...)

Attachment 1: glitchyPOX.jpg
glitchyPOX.jpg
  15445   Wed Jul 1 12:50:40 2020 gautamUpdatePEMMC1 accelerometers plugged in

I re-connected the 3 accelerometers located near the MC1/MC3 chamber. It was a bit tedious to get the cabling sorted - I estimate the cable is ~80m long, and the excess length had to be wound around a spool (see Attachment #1), which wasn't really a 1 person job. It's neat-ish for now, but I'm not entirely satisfied. I think we should get shorter cables (~20m), and also mount the pre-amp/power units in a rack instead of leaving it on the floor. The pre-amp settings are x100 for all three channels. The MC2 channels are powered, but are unconnected to the seismometers - it was too tedious to unroll the other spool yesterday. Apart from this, the cable for the "Z" channel had to be re-seated in the strain relief clamp.

I did not enable any of the CDS filters that convert the raw signal into physical units, so for now, these channels are just recording raw counts.

Update 7pm: the spectra in the current config are here - not sure what to make of the MC2_Z channel appearing to show lower noise?

Update July 13 2020 430pm: This afternoon, I hooked up the MC2 accelerometer channels too...

Attachment 1: IMG_8617.JPG
IMG_8617.JPG
Attachment 2: IMG_8616.JPG
IMG_8616.JPG
  15447   Wed Jul 1 18:16:09 2020 gautamUpdateComputersrossa re-re-revival

In an effort to make a second usable workstation, I did the following (remotely) on rossa today (not necessarily in this order, I wasn't maintaining a live log so I forgot):

  1. Fixed /etc/resolv.conf, so that the other martian machines can be found.
  2. Copied over .bashrc file, and the appropriate lines from /etc/fstab from pianosa to rossa.
  3. Ran sudo apt install nfs-common. Then ran sudo mount -a to get /cvs/cds mounted.
  4. Made symlinks for /users and /opt/rtcds , and /ligo. All of these are used by various environment-setting scripts and I chose to preserve the structure, though why we need so many symlinks, I don't know...
  5. Set up the shell variable $NDSSERVER using export NDSSERVER=fb:8088. I'm not sure how, but I believe DTT, awggui etc use this on startup to get the channel list (any
  6. Followed instructions from Erik von Reis at LHO to install the cds workstation packages and dependencies. Worked like a charm 🎃
  7. As a test, I plotted the accelerometer spectra in DTT, see Attachment #1. I also launched foton from inside awggui, and confirmed that the sample rate is inherited and I could designate a filter. But I haven't yet run the noise injection to test it, I'll do that the next time I'm in the lab.
  8. Also checked that medm, StripTool and ndscope, and anaconda python all seem to work 👍🏾.

So, in summary, rossa is now all set up for use during lock acquisition. However, until this machine has undergone a few months of testing, we should freeze the pianosa config and not mess with it.

Note that this version of the "crtools" is rather new. Please, use them and if there is an issue, report the errors! I am going to occassionally try lock acquisition using rossa. 

Quote:

wiped and install Debian 10 on rossa today

still to be done: config it as CDS workstation

please don't try to "fix" it in the meantime

Attachment 1: MCacc.pdf
MCacc.pdf
  15452   Mon Jul 6 00:37:28 2020 gautamUpdateComputersrossa: lib symlink

This is strange - I was definitely able to launch medm when I was working on this machine remotely on Friday. But now, there does seem to be a problem with this shared library being missing.

First of all, I installed mlocate to find where the shared library files are installed. Then I made the symlink, and now sitemap seems to work again.

Weirdly, my changes to /etc/resolv.conf got overwritten somehow. Was this machine rebooted? Uptime suggests it's only been running for ~6 hours at the time of writing of this elog.

sudo apt install mlocate
sudo updatedb
sudo ln -s /usr/lib/x86_64-linux-gnu/libreadline.so.7 /usr/lib/x86_64-linux-gnu/libreadline.so.6
Quote:

when I try 'sitemap' on rossa I get:

medm: error while loading shared libraries: libreadline.so.6: cannot open shared object file: No such file or directory

  15455   Mon Jul 6 12:51:41 2020 gautamUpdateComputersrossa: resolvconf installed

Indeed, this is now fixed by following instructions from here. I rebooted rossa at ~1250 PDT and confirmed that resolv.conf didn't get overwritten. The resolv.conf file also now has the following useful lines at the head:

~>cat /etc/resolv.conf
# Dynamic resolv.conf(5) file for glibc resolver(3) generated by resolvconf(8)
#     DO NOT EDIT THIS FILE BY HAND -- YOUR CHANGES WILL BE OVERWRITTEN
Quote:

yes, I rebooted yesterday to fix the 'steaking white lines' problem in the video/display

maybe we're supposed to edit something besides resolv.conf since that gets over-written on boot for some linux OS

  15457   Mon Jul 6 17:41:19 2020 gautamUpdateLSCAn LSC puzzler

Last Tuesday evening, while attempting the PRFPMI locking, I noticed a strange feature in the LSC signals, which is shown in Attachment #1 (the PDF exported by dataviewer is 14MB so I upload the jpeg instead). As best as I can tell, the REFL33 and POP22 channels show an abrupt jump in the signal levels, while the other channels do not. POP110 shows a slight jump at around the same time, and the large excursion in AS110_Q actually occurs a few seconds later, and is probably some angular excursion of the PRC/BS. I'm struggling to interpret how this can be explained by some interferometric mechanism, but haven't come up with anything yet. The LO for the 3f error signals is the 2f field, but then why doesn't the POP110 channel show a similar jump if there is some abrupt change in the resonant condition? Is such a change even feasible from a cavity length change point of view? Or did the sideband frequency somehow abruptly jump? But if so, why is the jump much more clearly visible in one sideband than the other?

Does anyone have any ideas as to what could be going on here? This may give some clue as to what's up with the weird sensing matrices, but may also be something boring like broken electronics... 

Attachment 1: LSCsignals.jpg
LSCsignals.jpg
  15458   Tue Jul 7 14:06:10 2020 gautamUpdateASCSome more thoughts about ASC

Summary:

I want to be able to run the dither alignment servo with the PRFPMI locked - I've been thinking about what the scheme should be, and I list here some questions I had while thinking about this.

Details:

  1. ITM Oplev DC coupling
    • In the current scheme, I DC couple the ITM Oplev servos after the arms have been aligned to maximize POX/POY transmission.
    • However, looking back at data from when the CARM offset is reduced (e.g. Attachment #1), it looks like the ITMs are being torqued quite a bit during this process (ITMX PIT changes by ~20urad, ITMY YAW by ~10urad in this particular lock attempt). 
    • So the spots are not actually being centered on the test-masses? I guess the spot position on ITMX isn't actually controlled because we have only one actuator (BS) for the XARM beam axis. Is it unexpected that ITMY gets torqued so much? 
    • It is unclear what would happen if the ITM Oplev servos are not DC coupled. I wonder if I'd still be able to reach the high circulating powers and then rely solely on the TR QPDs for the arm cavity angular control.
    • Another possibility is to offload the DC part of the control signal to the optic's slow bias voltage slider, and then turn off the DC coupling. After that, the dither alignment can optimize the cavity alignment.
  2. Dither alignment at high circulating power
    • think that the system should work with the same settings as for the POX/POY locking, with the following changes:
      • Scale the overall loop gain by the arm transmission.
      • Change LSC2ASS matrix element from XARM/YARM ---> DARM.
        Does this sound right?
    • In light of the above, I was thinking that we should introduce a gain scaling field in the c1ass RTCDS model (like we have for the LSC power normalization matrix). Is it worth to go through the somewhat invasive process of model recompilation etc?
    • With the PRFPMI locked, I am wondering if I can simultaneously run the dither alignment loops for all the DoFs. Probably not, especially for MICH, since the actuator is the BS, which is also the actuator for the XARM loop?
Attachment 1: ITM_OL_DCcoupling.png
ITM_OL_DCcoupling.png
  15460   Wed Jul 8 22:50:33 2020 gautamUpdateComputersrossa: more symlinks

I wanted to try using rossa as my locking workstation today. However, a few problems became quickly evident. Basically, any of our scripts that rely on the cdsutils package (there are MANY) will not work on rossa, because of some library error. This machine is running Debian 10, while the cdsutils package is being loaded from a pre-compiled install on the shared drive, so perhaps this isn't surprising?

Digging a little more, I found that actually, a version of cdsutils that actually works with python3 is actually shipped with the standard cds-workstation meta-package. This is great news, and we should try and use this where possible I guess. Deferring further debugging for daytime work.

Anyway, I added a symlink: sudo ln -s /usr/lib/x86_64-linux-gnu/libncurses.so.6 /usr/lib/x86_64-linux-gnu/libncurses.so.5, and installed wmctrl using sudo apt install wmctrl

  15463   Thu Jul 9 16:16:20 2020 gautamUpdateComputersrossa: graphics driver issues?

I noticed these streaky lines again today (but they were not a problem last night). It is annoying if we have to reboot this machine all the time. I wonder if this has something to do with missing drivers. When I ran sudo apt update && sudo apt upgrade, I got several lines like (this isn't the whole stack trace)

W: Possible missing firmware /lib/firmware/nvidia/gp108/acr/ucode_unload.bin for module nouveau
W: Possible missing firmware /lib/firmware/nvidia/gp108/acr/ucode_load.bin for module nouveau
W: Possible missing firmware /lib/firmware/nvidia/gp108/acr/unload_bl.bin for module nouveau
W: Possible missing firmware /lib/firmware/nvidia/gp108/acr/bl.bin for module nouveau

Is this indicative of the graphics drivers being installed incorrectly? I am hesitant to mess with this because I think in the past, it was always trying to update some graphics driver that crashed the whole machine into some weird state where we have to wipe the drive and do a fresh re-install of the OS.

Should we just follow these instructions? The graphics card is apparently Quadro P400, which is one of the supported ones according to the list of supported devices.

Or just swap donatella and rossa monitors and defer the problem for later?

Quote:

yes, I rebooted yesterday to fix the 'steaking white lines' problem in the video/display

  15464   Thu Jul 9 17:12:52 2020 gautamUpdateBHDIn-air BHD

Summary:

We can probably learn something about the interferometer / top level BHD plan with an in-air BHD setup, even if the noise is bad. Here are some thoughts about how we would do it. 

LO delivery:

For this first attempt, we don't really care about the PRC filtering. So possible places to pick off an LO beam are:

LO beam pickoff options
Location Pros Cons
IP POS
  • Filtered by IMC
  • Medium level of invasiveness  
  • We lose the IP POS diagnostic, which is kind of useful nowadays given the drifty TTs.
  • Only few mW LO power available
PSL table IR beam currently going to green doubling setup
  • Least invasive w.r.t. normal IFO operation
  • Plenty of light (~100 mW) available. But how much can we safely couple into fiber?
  • Beam not filtered by IMC (although it is filtered by the PMC)
POX/POY
  • Since this beam is extracted from inside the PRC, probably enjoys the best filtering.
  • Possibly drifts a lot, so tricky to reliably couple into a fiber?
  • Maximally invasive w.r.t. regular IFO operations.

In all cases, I think the easiest option to actually route whatever beam we choose into a fiber, and then bring it over to whatever cavity we choose to use for an OMC. I'm assuming whatever phase control technique we end up using can cancel the fiber phase noise at relevant frequencies.

LO phase control

  • Stress the fiber? This will require us to purchase some custom hardware, and interface it to the CDS system.
  • PZT mirror? We should have sufficient hardware available to drive a PI style PZT mirror.

There is a question about the range, but I think these are the only two realistic options we can implement on a reasonable time scale.

OMC:

Again, there are a few options. Here are some pros and cons that come to my mind.

OMC cavity options
Option Pros Cons
Old copper OMC
  • Probably the simplest option in terms of the peripherals.
  • PZT driver recently verified to work
  • We can get the OMMT and DCPDs out as well.
  • Allows us to not compromise on the RF darm optical gain (not sure if locking will be as easy if we cut the power to the AS55 photodiode by 50-75%).
  • Requires a vent.
  • Probably not the most efficient use of the space on the AP table.
  • Filtering performance isn't quantified.
Spare PMC
  • Doesn't require a vent.
  • Compact footprint.
  • Need to build the cavity.
  • Need to check if the drive electronics from the old copper OMC can easily be interfaced with whatever PZT we use on this cavity.
  • Filtering performance kind of unknown?
Custom cavity with spare mirrors
  • Doesn't require a vent.
  • Probably no more difficult than the spare PMC option?
  • We need at least one actuatable mirror, so we'd need some PZT mounted optic + associated drive electronics.

If we can do a vent (we'd just need a single chamber open), I'd go for the option of getting the copper OMC out and using that. Attachment #1 shows the approximate sizes of the various components (OMMT, OMC cavity, DCPDs), while Attachment #2 shows a rough sketch of where things would go on the AP table, with the rectangles approximately to scale.

CDS:

I'd made a c1omc model sometime ago. Basically, I think we have sufficient ADC/DAC channels in the c1ioo machine for any of the options listed above - but using the copper OMC and associated peripherals would allow the easiest interfacing.

Criticisms/comments/thoughts please.

Attachment 1: OMCchamber.pdf
OMCchamber.pdf
Attachment 2: AP_Table_20180328.pdf
AP_Table_20180328.pdf
  15466   Fri Jul 10 01:25:28 2020 gautamUpdateLSCLocking notes

More tomorrow, but I tried the following tonight:

  1. Dither alignment for PRC / MICH seems to work when the PRFPMI is locked. Unfortunately, the correct settings for the arm cavity dither alignment loops continue to elude me.
  2. I tried some arm ASC loop characterization by stepping the error points of these loops - I saw some weird cross coupling between the loops that needs investigation.
  3. I'm unable to turn an integrator on for the "Common YAW" QPD loop - unclear why this is, but every time I attempt to engage said integrator, the lock is immediately blown. Needs some investigation of the signals.
  4. With the PRC / MICH angular DoFs aligned with the dither alignemnt, and the arm ASC loops hand tuned, I was able to get the darkest dark port I've seen. In terms of ASDC counts, it was ~ 200, which after undoing all the digital gains etc corresponds to ~100 uW of light. I think we can get a rough estimate of the contrast defect by accounting for (i) T_SRM, (ii) OMC pickoff fraction (iii) other losses between the BS dark port and the AP table (iv) 50/50 BS between AS55 and AS110 PD (the ASDC signal is derived from the former) and (v) the throughput of the 55 MHz sideband to the dark port, although there are many uncertainties. 
  15468   Fri Jul 10 15:26:28 2020 gautamUpdateLSCMC2 coils need DC balancing?

I was looking at some signals from last night, see Attachment #1.

  • It looks like as the DC control signal to the MC2 suspension increases, the MC transmission decreases.
    • I confirmed that the IMC REFL level doesn't correspondingly trend up, but didn't include it here for plot compactness, so I think the cavity length isn't being detuned.
    • So the problem is suggestive of some L2A coupling, and the MC2 coil actuators need to be balanced better at DC?
    • You can see from the IMC WFS control signals that the WFS servo is presumably trying to counteract this L2A action, but doesn't succeed, probably because the servo isn't tuned correctly.
    • This is a problem that is distinct from the drifting TT alignment. So it complicates the alignment situation.
    • Ideally, if the dither alignment servos could be made to work for the arm cavities when locked in the PRFPMI config, this wouldn't be so much of a problem, as the TTs would just adjust the beam pointing to match the cavity axes of the arms. But since I haven't managed to get that servo working yet...
  • But why should MC2 need such a large DC control signal ever?
    • In the PRFPMI lock, the CARM servo is supposed to match the laser frequency to the average length of the two arm cavities.
    • The MC2 suspension is used as a frequency actuator in order to realize this matching.
    • But, as you can see, the digital CARM control signal picks up a significant DC offset the deeper we go into the lock.
    • Can't we offload this DC signal to the laser crystal temperature servo? Is there going to be some weird interaction with the existing slow loop? Or is the idea itself flawed?

Attachment #2 shows some ASC metrics. My conclusion here is that running the PRCL and MICH dither alignment servos (former demodulating REFLDC and latter demodulating ASDC to get an error signal) that running the dither alignment servo and hand tuning the arm ASC loop offsets improves the mode matching to the IFO, because:

  1. The arm transmission increases.
  2. POPDC increases.
  3. ASDC decreases.

The REFLDC behavior needs a bit more interpretation I think, because if the IFO is overcoupled (as I claim it is), then better alignment would at some point actually result in REFLDC increasing. 

All the DC signals recorded by the fast system come from the backplane P2 connector of the PD interface boards. According to the schematic, these signals have a voltage gain of 2. The LSC photodiodes themselves have a nominal DC gain of 50 ohms. So, the conversion from power to digital counts is: 0.8 A/W * 50 V/A * 2 * 3276.8 cts/V * whtGain. Inverting, I get 3.8 uW/ct for a whitening gain of 1. This is power measured at the photodiode - optical losses upstream of the photodiode will have to be accounted for separately.

Assuming a modulation depth of 0.2, the 55 MHz sideband power should be ~20 mW. The Schnupp asymmetry is supposed to give us O(1) transmission of this field to the AS port. Then, the SRM will attenuate the field by a factor of 10, so we expect ~2 mW at the AS port. Let's assume 80 % throughput of this field to the AP table, and then there is a 50/50 beamsplitter dividing the light between the AS55 and AS110 photodiodes. So, we expect there to be ~700 uW of power in the TEM00 mode 55 MHz sideband field. This corresponds to 1600 cts according to the above calibration (the ASDC whitening gain is set to 18 dB). The fact that much smaller numbers were seen for ASDC indicates that (i) the schnupp asymmetry is not so perfectly tuned and the actual transmission of the sideband field to the dark port is smaller, or (ii) one or more optical splitting fractions assumed above is wrong. If the former is true, we can still probably infer the contrast defect if we can somehow get an accurate measurement of the sideband transmission to the dark port.

Attachment 1: MC2_balancing.png
MC2_balancing.png
Attachment 2: ASDC.png
ASDC.png
  15469   Sat Jul 11 00:10:22 2020 gautamUpdateComputersrossa: more developmental work

After some consultation with Erik von Reis at LHO, this workstation is progressing towards being usable for most commissioning tasks. DTT, awggui, foton, and MEDM are all now working well. The main limitation now comes from the fact that many of our python scripts are written for python2, and rossa doesn't have many dependencies installed for python2. I see no reason to build these dependencies on rossa for python2, we should not have to work with an unsupported language. But at the same time, I don't want to completely wipe all our python2 scripts, and make them python3, because this would involve a lot of tedious testing that I'm not prepared to undertake at the moment (the problem is compounded by the fact that pianosa does not have many dependencies installed for python3).

So what I have done in the interim is make python3 versions of the most important scripts I need to get the PRFPMI locking working - they are in the scripts directory and have the same names as their python2 counterparts, but have a 3 appended to their names. So when working on rossa, these are the scripts that are called. Eventually, after a lot more testing, we can depracate the old scripts. Currently, where applicable, the MEDM screens allow for either the python2 or python3 version of the script to be called.

Please, for the time being, do not try and install any new packages on rossa unless you are prepared to debug any problems caused and return the machine to a workable state. If you find some issue with a missing package on rossa, (i) make a note of it on the elog, and (ii) if possible, set up your own conda environment for testing and install dependencies to that environment only.

  15471   Sun Jul 12 02:42:01 2020 gautamUpdateLSCLocking (on rossa)

Main goals tonight were:

  1. Check if I can lock the interferometer by working on rossa - indeed, I could! It is much snappier than the ageing pianosa. The viewing angle of the CRT monitors from this corner isn't so good though.
  2. Measure step responses for the arm ASC loops to see if any insight can be gained into these loops. Analysis forthcoming...
Attachment 1: ASCsteps.png
ASCsteps.png
  15472   Sun Jul 12 22:40:35 2020 gautamUpdateElectronicsWFS characterization - old SURF report

After some hunting, I found this old SURF report with the WFS head measurements. The y-axes don't make much sense to me, and I can't find the actual data anywhere (her wiki page doesn't actually exist). So I think it's still unknown if these heads ever had the advertised transimpedance gain, or if the measured transimpedance of ~1kohm was what it always was.

  15475   Mon Jul 13 12:37:05 2020 gautamUpdateComputersrossa: more developmental work

In fact, all these utilities are now available in python3. There may be some bugs (e.g. this), but I've checked basic functionality and things look usable enough for development to proceed. While we can have a python2 env on rossa, I think it's unnecessary.

Quote:

I too, would prefer py3 for everything, but aren't all the cdsutils / gaurdian things still python2?

  15476   Tue Jul 14 00:06:09 2020 gautamUpdateLSCLocking with POX for CARM

I tried using the POX_I error signal for the DC CARM_B path today a couple of times. Got to a point where the AO path could be engaged and the arm powers stabilized somewhat, but I couldn't turn the CARM_A path off without blowing the lock. Now the IMC has entered a temperemental state, so I'm abandoning efforts for tonight, but things to try tomorrow are:

  1. Check that the demod phase is set correctly
  2. With the CARM_B path engaged, measure some CARM OLTFs. Tonight, I was a bit over-optimistic I think, by expecting the scripted transition to take me all the way, but I think I'll have to fiddle around with the gains a bit.
  3. Check for offsets. The AO path should be AC coupled, but maybe the POX signal has some offset?

I have some data from a couple of days ago when the PRFPMI was locked as usual (CARM_B on REFL for both DC and AO paths), and the sensing lines were on, so I can measure the relative strength of the sensing lines in POX/REFL and get an estimate of what the correct digital gain should be.

The motivation here is to see if the sensing matrix looks any different with a modified locking scheme.

  15479   Tue Jul 14 15:29:25 2020 gautamUpdateBHDIn-air BHD - DCPD amplifier noise

For the first pass, it's probably easiest to use the existing DCPD amplifier. Looking at the gain and noise performance in Attachment #1, seems totally fine, the electronics noise will not be limiting if we have ~10mW of LO power. I assumed a transimpedance resistor of 1 kohm, and all other numbers as on the schematic (though who knows if the schematic is accurate). The noise should be measured to confirm that the box is performing as expected...

Attachment 1: DCPDamp.pdf
DCPDamp.pdf
  15480   Tue Jul 14 16:52:47 2020 gautamUpdateElectronicsCoil drivers for the test masses

Summary:

Koji and I had a discussion last Friday about the suspension electronics. I think there are still a few open questions - see Attachment #1. We should probably make a decision on these soon.

Other useful links:

  1. High-voltage coil driver circuit - D1900163
    • This board is ready to be fabricated and tested on the bench.
    • The way the connectors J2 and J3 are designed currently is meant to interface with the existing coil driver electronics.
    • Depending on the eventual coil driver we choose for the fast path, it may be benificial to change the signals on the connectors J2 and J3, to avoid the need for a custom interface board.
  2. HAM-A coil driver noise analysis.
    • The linked attachment evaluates the noise for the design value of the fast path series resistor, which is 1.2 kohms.
    • Iff we still have ambitions of measuring ponderomotive squeezing, we will need the resistance to be much higher, ~10 kohms (in the linked noise budget, only the Johnson noise of the series resistor is considered, but in reality, the OpAmp voltage and current noises also matter). 
    • This corresponds to a maximum current of 10V/10kohms = 100uA
    • Looking at signals to the ETMs from the current lock acquisition sequence, the RMS current to a single coil is approximately _____ (to be filled in later).
    • So we may need a version of the fast coil driver that supports a low noise mode (with large series resistance) and a high-range mode (with lower series resistance for lock acquisition).
  3. You can follow the links to DCC entries for other parts from Attachment #1.
Attachment 1: coilDriverSchem.pdf
coilDriverSchem.pdf
  15481   Tue Jul 14 17:28:29 2020 gautamUpdateLSCLocking with POX for CARM

From Attachment #1, looks like the phasing and gain for CARM on POX11 is nearly the same as CARM of REFL11, which is probably why I was able to execute a partial transition last night. The response in POY11 is ~10 times greater than POX11, as expected - though the two photodiodes have similar RF transimpedance, there is a ZFL-500-HLN at the POY11 output. The actual numerical values are 2.5e10 cts/m for CARM-->REFL11_I, 2.6e10 cts/m for CARM-->POX11_I, and 3.2e11 cts/m for CARM-->POY11_I.

So I think I'll just have to fiddle around with the transition settings a little more tonight. 

One possible concern is that the POX and POY signals are digitized without preamplificatio, maybe this explains the larger uncertainty ellipse for the POX and POY photodiodes relative to the REFL11 photodiode? Maybe the high frequency noise is worse and is injecting junk in the AO path? I think it's valid to directly compare the POX and REFL spectra in Attachment #2, without correcting for any loops, because this signal is digitized from the LSC demodulator board output (not the preamplified one, which is what goes to the CM board, and hence, is suppressed by the CARM loop). Hard to be sure though, because while the heads are supposed to have similar transimpedance, and the POX photodiode has +12dB more whitening gain than REFL11, and I don't know what the relative light levels on these photodiodes are in lock.

Quote:

I have some data from a couple of days ago when the PRFPMI was locked as usual (CARM_B on REFL for both DC and AO paths), and the sensing lines were on, so I can measure the relative strength of the sensing lines in POX/REFL and get an estimate of what the correct digital gain should be

Attachment 1: PRFPMI_2020712sensMat.pdf
PRFPMI_2020712sensMat.pdf
Attachment 2: LSCerrSigs.pdf
LSCerrSigs.pdf
  15483   Wed Jul 15 19:11:40 2020 gautamUpdateBHDIn-air BHD - alignment into OMC

I forgot about the pointing - probably we will need another actuator to control the pointing of the AS beam onto the DCPDs. I found a few old PI PZTs (model number is S-320, which is a retired part), one is labelled broken but the others don't indicate a-priori that they are broken. I'll post a more detailed hardware survey later.

  Draft   Wed Jul 15 19:17:09 2020 gautamUpdateBHDIn-air BHD - alignment into OMC

You can activate all 3axis

 

  15485   Wed Jul 15 19:23:44 2020 gautamUpdateGeneralEmergency light on in control room

The emergency lamps above the exit sign on the NW entrance to the control room are on. I tried opening and closing the door, but it remains on. Probably nothing to worry about, but noting here anyway.

  15487   Wed Jul 15 20:58:40 2020 gautamUpdateGeneralEmergency light on in control room

True - it is now not on anymore.

Quote:

It happened before too. Doesn't it say it has occasional self-testing or something?

  15488   Wed Jul 15 21:08:43 2020 gautamUpdateElectronicsETM coil outputs DQed

To facilitate this investigation, I've DQed the 4 face coil outputs for the two ETMs. EX is currently running with 5 times the series resistance of EY, so it'll be a nice consistency check. Compilation, installation etc went smooth. But when restarting the c1scx model, there was a weird issue - the foton file, C1SCX.txt, got completely wiped (all filter coefficients were empty, even though the filter module names themselves existed). I just copied the chiara backup version, restarted the model, and all was well again.

This corresponds to 8 additional channels, recorded at 16k as float 32 numbers, so in the worst case (neglecting any clever compression algorithms), we are using disk space at a rate of ~4 MB/s more. Seems okay, but anyway, I will remove these DQ channels in a few days, once we're happy we have enough info to inform the coil driver design.

spoke too soon - there was an RFM error for the TRX channel, and restarting that model on c1sus took down all the vertex FEs. Anyways, now, things are back to normal I think. The remaining red light in c1lsc is from the DNN model not running - I forgot to remove those channels, this would've been a good chance! Anyways, given that there is an MLTI in construction, I'm removing these channels from the c1lsc model, so the next time we restart, the changes will be propagated.

For whatever reason, my usual locking scripts aren't able to get me to the PRFPMI locked state - some EPICS channel value must not have been set correctly after the model reboot 😞. I'll debug in the coming days.

Fun times lie ahead for getting the new BHD FEs installed I guess 🤡 ....

Quote:
 

Looking at signals to the ETMs from the current lock acquisition sequence, the RMS current to a single coil is approximately _____ (to be filled in later).

So we may need a version of the fast coil driver that supports a low noise mode (with large series resistance) and a high-range mode (with lower series resistance for lock acquisition).

Attachment 1: CDS.png
CDS.png
Attachment 2: coilOutDQed.png
coilOutDQed.png
  15489   Thu Jul 16 01:12:22 2020 gautamUpdateBHDIn-air BHD - preparing the LO path

Attachment #1 - The 80mW pickoff was getting clipped on a BNC cable, and not making it to the doubling oven. 😢 .

  • Since the PSL doubled beam isn't used for locking these days, I just didn't notice.
  • I blame the ringdown team, this crazy tee arrangement wasn't the case before.
  • I fixed the situation by changing the cabling such that the beam clears the cables comfortably.

Attachment #2 - PSL green shutter removed. Alignment into the doubling oven is extremely tedious, and so I opted to preserve the capability of recovering the green beam by simply removing a single mirror.

Attachment #3 - The beam path for coupling the LO beam into a fiber.

  • Primary goal was to have easy access to some steering mirrors so that I can optimize alignment into the fiber collimator.
  • I opted to use the NW corner of the PSL table - that's where most of our existing fiber hardware is anyways, and there was sufficient space and easy access over there.
  • 3 Y1 mirrors were installed, using the preferred Polaris mounts and 3/4" post + baseplate hardware. They were labelled Y1-1037-45P so that future workers need not be un-necessarily tortured. The third mirror is not visible in this photograph.
  • Once the collimator arrives, I will mode match this beam into the fiber. Plan is to use the fiber originally used for the mode spectroscopy project. It needs to be moved to the NW corner of the PSL table, and the other end needs to be routed to the AP table (it was brought back to the PSL table to facilitate Anjali's fiber MZ experiment). 
  • There is plenty of space in the beam path for mode-matching lens(es) and polarization control optics.

Attachment #4 shows the BHD photodiodes taken from QIL. 

  • Unfortunately, we could not find the readout electronics. 
  • In the worst case, we can just interface these PDs with the existing Satellite box (associated with the copper OMC).
  • It might be that the OMC cavity can simply be placed on this breadboard, making the whole setup nice and portable.
  • We may want to consider having an OFI between the OMC and the IFO AS beam at some point...
Attachment 1: IMG_8626.JPG
IMG_8626.JPG
Attachment 2: IMG_8627.JPG
IMG_8627.JPG
Attachment 3: IMG_8628.JPG
IMG_8628.JPG
Attachment 4: IMG_8629.JPG
IMG_8629.JPG
  15490   Thu Jul 16 14:41:22 2020 gautamUpdateGeneralFire extinguisher inspection

The (masked) tech accessed all areas in the lab (office area, control room, VEA) between ~230pm-3pm. The laser safety goggles he used have been kept aside for appropriate sanitaiton.

  15491   Fri Jul 17 00:18:13 2020 gautamUpdateGeneralLocking updat
  1. I found that an EPICS channel wasn't reset to the correct value by burtrestore after the FE bootfest yesterday.
    • This cost me the whole of last night, found it finally tonight. 
    • I'll try and modify the locking scripts to better capture such errors, but ideally, we should just use Guardian or something since it's made for this purpose already.
    • Anyways, tonight I was able to re-acquire the PRFPMI lock in a completely scripted way.
  2. Locking CARM on POX remains out of reach.
    • I think this has to do with the fact that the zero-crossing of the CARM and REFL error signals are dependent on the 3f PRCL/MICH error point offsets.
    • So even if the DC gain is right, the fact that we use POX for the digital AO path and REFL for the analog AO path is leading to some conflict I think.
    • Ran out of energy tonight, I'll try again tomorrow.

The DQ channels of the ETM coils were active tonight, so I'll make the coil driver actuation budget over the next couple of days.

  15493   Sun Jul 19 15:40:15 2020 gautamUpdateBHDIn-air BHD - CDS and wiring summary

Attachment #1 shows the proposed wiring and CDS topology for the in air BHD setup. The PDF document has hyperlinks you can follow to the DCC entries. Main points:

  1. I think we should run the realtime model on c1lsc. This will negate the need for any IPC between c1ioo and c1lsc machines.
  2. I think we have most of the electronics we need already, though I am still in the process of testing the various boards, especially the HV ones.
  3. We may choose to use the switchable whitening feature for the M2 ISS board
    • This would require some BIO channels
    • There are plenty spare in c1lsc, so it's not going to be a show stopper
    • This is why I've not explicitly included a whitening board for now...
  4. The main job seems to be to make a whole bunch of custom cables. For the most part, I think we have the long (~20m) long D9 cables, so I propose just snipping off the connector at one end, and soldering on the appropriate connectors to the correct conductors.
  5. For the homodyne phase control - the proposal is to use a PI PZT with 3 piezoelectric elements. We would drive the 3 elements with the same voltage, by shorting the conductors together (at least that's how I understood Koji's comment), so we'd only need a single DAC channel for this purpose.
    • Need to confirm that the parallel PZT capacitances (each element is ~300 nF so 3 in parallel would be ~900 nF) still allows sufficient actuation bandwidth.
    • If the relative actuation strength of the 3 elements needs to be individually tuned, we may have to use three DAC channels. The D980323 board will allow the driving of 3 independent channels. I have one of these boards in hand, but need to check if it works, and also implement the changes outlined here.
  6. The alignment control has not yet been accounted for
    • We could consider using the in-vacuum PZTs, these were verified to be working ~2018.
    • If we use only 1 steering PZT mirror, we have sufficient free DAC channels available in c1lsc. But if we need both (to avoid clipping for example), then we need more DAC channels - we can either free up one DAFI channel, or install a DAC in the c1lsc expansion chassis
  7. We may want to expand to have a second OMC at some point. In which case we'd need, at the very least
    • 1 more DAC card
    • A HV driver for the second OMC length (could use the Trek driver if we use D980323 for the homodyne phase control).

Please comment if I've overlooked something.

Attachment 1: wiringDiagram.pdf
wiringDiagram.pdf
  15494   Mon Jul 20 17:23:46 2020 gautamUpdateElectronicsCoil drivers for the test masses

Summary:

Looking at the signals to the test mass coils, it seems borderline to me that we will be able to acquire lock and run in a low noise configuration with the same series resistor in the coil driver circuit. The way I see it, options are:

  1. Use a moderately high series resistance (e.g. 5 kohms) for the time being, and go ahead with the HAM-A coil driver.
    • This will mean a current noise of ~3pA/rtHz, which translates to ~3e-18 m/rtHz @ 100 Hz in DARM displacement noise (assuming the ITMs have much higher series resistance than the ETMs).
    • If the lock acquisiton looks smooth, double the resistance to 10 kohms.
    • With 5 kohm series resistance, there is negligible possibility of measuring ponderomotive squeezing for any of the input powers we consider feasible, but this is under the assumption that we will expose coil driver noise, which is very optimistic imho.
  2. Re-design a new coil driver that allows switchable impedance, so we can have a higher noise acquisition mode for acquiring and holding the ALS lock, then transition to a lower noise, lower range config once the RF / BHD lock has been acquired.
    • On paper, this solves all the problems, but the design of such a circuit is probably pretty non-trivial and time consuming.

Details:

I only looked at the ETMs for this study. The assumption is that we will have no length actuation on the ITMs, only local damping and Oplev loops (and maybe some ASC actuation?), which can be sufficiently low-pass filtered such that even with coil de-whitening, we won't have any range issues.

Attachment #1 shows the time-domain traces of the coil driver signals as we transition from POX/POY lock to the ALS lock. There are some transients, but I think we will be able to hold the lock even with a 5 kohm resistor (~twice what is on ETMX right now). From just these numbers, it would seem we can even go up to 10 kohms right away and still be able to acquire lock, especially if we re-design the digital feedback loop to have better low-pass filtering of the high-frequency ALS noise, see the next attachment.

Attachment #2 shows the f-domain picture, once the arm lengths are fully under ALS control (~25 seconds onwards in Attachment #1). The RMS is dominated by high frequency ALS length loop noise, which we can possibly improve with better design of the digital control loop.

Finally, Attachment #3 shows the situation once DARM control has been transitioned over to AS55_Q. Note that the vertex DoFs are still under 3f control, so there is the possibility that we can make this even lower noise. However, one thing that is not factored in here is that we will have to de-whiten these signals to low-pass filter the DAC noise (unless there is some demonstrated clever technique with noise-mons or something to subtract the DAC noise digitally). Nevertheless, it seems like we can run safely with 5 kohms on each ETM coil and still only use ~2000 cts RMS, which is ~1/10th the DAC range (to allow for dealing with spurious transients etc). 

Quote:

Looking at signals to the ETMs from the current lock acquisition sequence, the RMS current to a single coil is approximately _____ (to be filled in later).

Attachment 1: ALSlock_timeDomain.pdf
ALSlock_timeDomain.pdf
Attachment 2: ALSlock.pdf
ALSlock.pdf
Attachment 3: RFlock.pdf
RFlock.pdf
  15495   Mon Jul 20 17:55:15 2020 gautamUpdateBHDIn-air BHD - preparing the LO path

The LO pickoff has been coupled into a fiber with ~90% MM (8 mW / 9 mW input). While I wait for the DCPD electronics to be found in the Cryo lab, I want to monitor the stability of the pointing, polarization etc, so I'd like to clear some space on the AP table that was occupied for the mode spectroscopy project. If there are no objections before 2pm tomorrow July 21 2020, I will commence this work.

  15497   Tue Jul 21 00:30:24 2020 gautamUpdateBHDIn-air BHD - LO RIN

Attachment #1 shows the RIN of the local oscillator beam delivered to the AP table via fiber. I used a PDA520 to make this measurement, while the electronics for the DCPDs are pending. I don't really have an explanation for the difference between the locked IFO trace vs the not locked trace - we don't have an ISS running (but this first test suggests we should) and the beam is picked off before any cavities etc, so this is a reflection of the state of the FSS servo at the times of measurement?


Tried locking CARM using the hybrid REFL (for AO path) and POX 11 (for MCL path) scheme a bunch of times today, but I had no luck. When the CARM offset is zeroed, the PRMI lock is lost almost immediately. Maybe this is indicative of some excess noise in the POX data stream relative to the REFL signal? The one thing I haven't tried is to take the IFO all the way to the locked state, and then transition the MCL actuation from CM_SLOW to POX11_I.


An SR785 is sitting on the North side of the AP table in the walkway - I will clear it tomorrow.

Attachment 1: LO_RIN.pdf
LO_RIN.pdf
  15498   Tue Jul 21 16:41:46 2020 gautamUpdateBHDPMC assembly space

I decided to use the old EY auxiliary optics table, which is now stored along the east arm about 10 m from the end, as a workspace for assembling the little PMCs. I wiped everything down with isopropanol for general cleanliness, removed the metal plate on the south edge of the table enclosure to allow access, covered the table with some clean Aluminium foil, and then moved the plastic box with PMC parts to the table - see Attachment #1. I haven't actually done any assembly just yet, waiting for more info (if available) on the procedure and implements available...

Attachment 1: IMG_8635.JPG
IMG_8635.JPG
  15506   Thu Jul 30 16:16:43 2020 gautamUpdateSUSSuspension recovery

This earthquake and friends had tripped all watchdogs. I used the scripted watchdog re-enabler, and released the stuck ITMX (this operation is still requires a human and hasn't been scripted yet). IMC is locked again and all Oplevs report healthy optic alignment.

  15508   Thu Aug 6 22:57:20 2020 gautamUpdatesafetyNew live HV Supplies

Be aware that there is now a KEPCO HV supply that is energized, sitting on the floor immediately adjacent to the OMC rack, east of the AP table. It is currently set to 100 V DC, and a PI PZT installed on the AP table has its 3 PZTs energized by said supply (via an OMC piezo driver). I will post pictures etc of the work from the last 10 days over the weekend.

  15513   Mon Aug 10 16:52:04 2020 gautamUpdateBHDWorkable setup prepared

All the details are in E2000436, and documents linked from there, I think an elog would be much too verbose. In summary, a workable setup consisting of

  • 2 DCPDs interfaced with the realtime CDS system. Note that because this circuit is single-ended, while the AA and ADC are differential receiving, there is an overall gain of 0.5. Explicitly, for the 300 ohm DC transimpedance, the conversion is ~350 cts/mW.
  • A local oscillator beam delivered via fiber that is mode-matched (roughly) with the IFO AS beam.
  • A PZT mounted mirror to control the homodyne phase. The PZT (S320) is an obsolete part and it's hard to find a datasheet for it, but if its specs are comparable to the more modern S330, the full stroke is 10 um, for a max applied voltage of 100 V DC, so 100nm/V. c.f. 200V for 3um full stroke of the Noliac.

was prepared.

Last night, I locked the PRMI with the carrier resonant, and convinced myself that the DCPD null stream was sensing the MICH degree of freedom (while it was locked on AS55_Q) with good SNR below ~60 Hz. Above ~60 Hz, in this configuration, the ADC noise was dominating, but by next week, I'll have a whitening board installed that will solve this particular issue. With the optical gain of MICH in this configuration, the ADC noise level was equivalent to ~500 nrad/rtHz of phase noise above ~60 Hz (plots later).

Now, I can think about how to commission this setup interferometrically.

  15514   Tue Aug 11 23:20:29 2020 gautamUpdateBHDSome first tests with air BHD setup

Some tests done today:

All of these tests were done with the PRMI locked with carrier resonant in the recycling cavity (i.e. sidebands rejected to REFL port). I then actuated the BS length DOF with a sine wave at 311.1 Hz, 40 cts amplitude (corresponding to ~8 pm of peak-to-peak displacement).

  1. Attempt to balance the DCPDs
    • I tried to tune the digital gains of the two DCPDs so as to minimize the appearance of this line in the SUM channel
    • but no matter how I tuned the gains, I couldn't make the line in the SUM channel disappear entirely - in fact, the best I could do was to make the line height in SUM and NULL channels (yes I recognize the poor channel name choice, I'll change "NULL" to "DIFF" at the next model recompile) the same. See Attachment #1.
    • The lobes around the main peak are indicative of some scattering?
    • Attachment #2 shows a wider frequency range. The homodyne phase isn't controlled, so the "NULL" channel is not necessarily measuring the correct quadrature to be sensing MICH motion.
    • I think I can back out something about the contrast defect from this fact, but I need to go back to some modeling.
  2. A simple test of the homodyne phase actuator
    • I wanted to check that this PI S320 piezo actually allows me to actuate the optical path length of the local oscillator.
    • I'm using the OMC HV driver to drive said PZT - so there are two DAC channels available, one to dither the optic and one to apply a control signal. I think mainly this is to avoid using up DAC range for the dither signal, the overall dynamic range is still limited by the HV supply.
    • I can't find the maximum voltage that can be applied on the datasheet - so conservatively, I limited the HV output to saturate at 100 V DC, as this is the maximum for the S330 piezos used for green steering, for which there is a manual.
    • The S320 manual does say the full stroke of each PZT element is 10 um - so the actuation coefficient is ~100 nm/V. I then drove this actuator with a sine wave of 500 cts amplitude, at 314.1 Hz (corresponding to 15 nm of motion). With only the LO beam incident on the PDs, I saw no signal in either DCPD - as expected, so this was good.
    • Then, with the PRMI locked, I repeated the test. If there is no DC light field (as expected for the PRMI in this configuration), I wouldn't expect this drive signal to show up in the DCPDs. But in fact, I do. Again, this supports the presence of some (for now unquantified) contrast defect.

While it would seem from these graphs that the RIN of the LO beam at these frequencies is rather high, it is because of the ADC noise. More whitening (to be installed in the coming days) will allow us to get a better estimate, should be ~1e-6 I think.

I was just playing today, still need to setup some more screens, DTT templates etc to do more tests in a convenient way.

Now, I can think about how to commission this setup interferometrically.

Attachment 1: PRMI_RFlock.pdf
PRMI_RFlock.pdf
Attachment 2: PRMI_RFlock_fullscale.pdf
PRMI_RFlock_fullscale.pdf
  15515   Wed Aug 12 17:36:42 2020 gautamUpdateCDSTiming distribution slot availability

See Attachment #1. J8 was connected to a "LASTI timing slave" sitting in the rack that Chiara lives in - we don't use this for anything and I confirmed that there was no effect on the RTCDS when I pulled that fiber out. The LASTI timing slave also had a blinky that was blinking when the fiber was plugged in - which I take to believe that the slot works. 

Can we get away with just using these two available slots, J8 and J13? Do we really need three new expansion chassis?

Attachment 1: IMG_8706.JPG
IMG_8706.JPG
  15516   Wed Aug 12 17:42:58 2020 gautamUpdateElectronicsPhotodiode inventory

See Attachments #1 and #2. We don't have any Q3000 QPDs in hand, at least not in the photodiode box stored in the clean optics cabinet at the south end. I also checked a cabinet along the east arm where we store some photodiodes - but didn't find any there either. The only QPDs we have in hand are the YAG-444-4AH, which I believe is what is used in the iLIGO WFS heads.

So how many do we want to get?

Attachment 1: IMG_8709.JPG
IMG_8709.JPG
Attachment 2: IMG_8708.JPG
IMG_8708.JPG
  15517   Wed Aug 12 18:08:54 2020 gautamUpdateElectronicsNumber of the beast

The "source" output of the SR785 has a DC offset of -6.66 V. I couldn't make this up.

Upshot is, this SR785 is basically not usable for TF measurements. I was using the unit to characterize the newly stuffed ISC whitening board. The initial set of measurements were sensible, and at some point, I started getting garbage data. Unclear what the cause of this is. AFAIK, we don't have any knob to tune the offset - adjusting the "offset" in the source menu, I can change the level of the offset, but only by ~1 V even if I apply an offset of 10 V. I also tried connecting the ground connection on the rear of the SR785 to the bench power supply ground, no change.

Do we have to send this in for repair?

Attachment 1: IMG_8710.JPG
IMG_8710.JPG
  15521   Thu Aug 13 11:30:19 2020 gautamUpdateCDSTiming distribution slot availability

That's great. I wonder if we can also get away with not adding new Dolphin infrastructure. I'd really like to avoid changing any IPC drivers.

Quote:

I believe we will use two new chassis at most. We'll replace c1ioo from Sun to Supermicro, but we recycle the existing timing system.

  15523   Thu Aug 13 18:10:22 2020 gautamUpdateGeneralPower outage

There was a power outage ~30 mins ago that knocked out CDS, PSL etc. The lights in the office area also flickered briefly. Working on recovery now. The elog was also down (since nodus presumably rebooted), I restarted the service just now. Vacuum status seems okay, even though the status string reads "Unrecognized".

The recovery was complete at 1830 local time. Curiously, the EX NPRO and the doubling oven temp controllers stayed on, usually they are taken out as well. Also, all the slow machines and associated Acromag crates survived. I guess the interruption was so fleeting that some devices survived.

The control room workstation, zita, which is responsible for the IFO status StripTool display on the large TV screen, has some display driver issues I think - it crashed twice when I tried to change the default display arrangement (large TV + small monitor). It also wants to update to Ubuntu 18.04 LTS, but I decided not to for the time being (it is running Ubuntu 16.04 LTS). Anyways, after a couple of power cycles, the wall StripTools are up once again.

  15524   Fri Aug 14 00:01:55 2020 gautamUpdateCDSBHD / OMC model channels now added to autoburt

I added the EPCIS channels for the c1omc model (gains, matrix elements etc) to the autoburt such that we have a record of these, since we expect these models to be running somewhat regularly now, and I also expect many CDS crashes.

  15529   Mon Aug 17 15:18:26 2020 gautamUpdateEquipment loanBeam Profiler + peripherals --> 40m

Gabriele left the DataRay beam profiler + peripherals (see Attachment #1) in his office. I picked them up just now and brought them over to the 40m.

Attachment 1: IMG_8719.JPG
IMG_8719.JPG
  15530   Mon Aug 17 21:24:43 2020 gautamUpdateGeneralFire extinguisher inspection

A technician came to the lab today at ~4pm. He entered the VEA (with booties and googles), and also the clean and bake lab. The whole procedure lasted ~10 minutes. I did not follow him around, but was available in the control room throughout the process. I think the whole episode went without incident.

BTW, this guy didn't ring the doorbell, I just happened to be here when he came by. I don't know if this is usual practise - are we happy with the technicians entering the VEA and/or clean and bake labs without supervision? AFAIK, this wasn't scheduled.

  15531   Mon Aug 17 23:36:10 2020 gautamUpdateALSWhitening and ALS noise

finally managed to install a differential-receiving whitening board in 1Y2 - 4 channels are available at the moment. As I claimed, one stage of 15:150 Hz z:p whitening does improve the ALS noise a little, see Attachment #1. While the RMS (from 1kHz-0.5 Hz) does go down by ~10 Hz, this isn't really going to make any dramatic improvement to the 40m lock acquisiton. Now we're really sitting on the unsuppressed EX laser noise above ~30 Hz. This measurement was taken with the arm cavities locked with POX/POY, and end lasers locked to the arm cavities with uPDH boxes as usual. This was just a test to confirm my suspicion, the whitening board is to be used for the air BHD channels, but when we get a few more stuffed, we can install it for the ALS channels too.

Attachment 1: ALSimprovement.pdf
ALSimprovement.pdf
  15532   Mon Aug 17 23:41:50 2020 gautamUpdateBHDWhitening and air BHD dark noise

Summary:

With the chosen transimpedance of 300 ohms, in order to be able to see the shot noise of 10 mW of light in the digitized data streams, we'd need all 3 stages of whitening. If we want to be shot noise limited with 1 mW of LO light, we'd need to increase said transimpedance I think.

Details:

The measurements were taken with

  1. No light incident on the DCPDs.
  2. The flat whitening gain was set to 0 dB.
  3. Whitening engaged sequentially, stage by stage, shown as (Blue, Red, Orange and Green) curves corresponding to (0, 1, 2, 3) stages of whitening.

Of course, it's unlikely we're going to be shot noise limited for any configuration in the short run. But this was also a test of 

  1. My soldering.
  2. Change of whitening corner frequencies.
  3. Test of the overall whitening board assembly.

All 3 tests passed.

Attachment 1: BHD_whitening.pdf
BHD_whitening.pdf
  15534   Thu Aug 20 00:21:51 2020 gautamUpdateElectronicsFirst look at HV coil driver

Summary:

A single channel of this board was stuffed (and other channels partially populated). The basic tests passed, and nothing exploded! Even though this is a laughably simple circuit, it's nice that it works.

HV power supplies:

A pair of unused KEPCO BHK300-130 switching power supplies that I found in the lab were used for this test. I pulled the programmable cards out at the rear, and shorted the positive output of one unit to the negative of the other (with both shorted to the supply grounds as well), thereby creating a bipolar supply from these unipolar models. For the purposes of this test, I set the voltage and current limits to 100V DC, 10mA respectively. I didn't ramp up the supply voltage to the rated 300 V maximum. The setup is shown in Attachment #1.

Tests:

  1. With the input to the channel shorted to ground, I confirmed with a DMM that the output was (nearly) zero (there was an offset of ~40mV but I think this is okay).
  2. Used the calibrated voltage source, and applied +/- 3 V in steps of ~0.5 V, while monitoring the output with a DMM. Confirmed the output swing of ~ +/-90 V, which is what is expected, since the design voltage gain of this circuit is 31.
  3. Drove a 0.1 Hz, 500mVpp sine wave at the input while monitoring the output and the Vmon testpoints, see Attachment #2. Note the phasing between input and output, and also the fact that the gain is slightly lower than the expected gain of 31, because there are three poles at ~0.7 Hz, which already start showing some influence on the transfer function at 0.1 Hz.
  4. Noise measurement 
    • The whole point of this circuit is to realize sub 1pA/rtHz current noise to the coil, when it is connected.
    • For this test, no load was connected (i.e. voltage noise was measured at the output of the 25 kohm resistor), and the input was shorted to ground so that the DC value of the output was close to 0 (the idea was to not overload the SR560/SR785 with high voltage).
    • An SR560 preamp with gain x50 (DC coupled) was used to preamplify the signal. This was the maximum gain that could be used with the unit DC coupled, due to the small DC offset. I opted to keep the DC coupling to get a look at the low frequency noise as well, but in hindsight, maybe I should have used AC coupling as we only care about the current noise at ~100 Hz.
    • See Attachment #3 for results. The measurement is close to the model above ~100 Hz

Need to think more about how to better characterize this noise. An estimate of the required actuation can be found here.

Attachment 1: IMG_8724.JPG
IMG_8724.JPG
Attachment 2: timeDomain.pdf
timeDomain.pdf
Attachment 3: HVampNoise.pdf
HVampNoise.pdf
  15535   Fri Aug 21 15:27:00 2020 gautamUpdateBHDBetter BHD mode-matching

Summary:

The mode-matching between the LO and AS beams is now ~50%. This isn't probably my most average mode-matching in the lab, but I think it's sufficient to start doing some other characterization and we can try squeezing out hopefully another 20-30% by putting the lenses on translation stages, tweaking alignment etc.

Details:

The main change was to increase the optical path length of the IFO AS path, see Attachment #1. This gave me some more room to put a lens and translate it.

  • The LO path uses two lenses, f=200mm and f=100mm to focus the collimator output beam, which is supposedly ~1200um diameter, to something like 400um diameter (measured using beam profiler but not very precisely).
  • This beam is  fairly well collimated, and the beam size is close to what the PMC cavity will want, I opted not to tweak this too much more.
  • For the AS beam, the single bounce reflection from ITMY was used for alignment work.
  • There is a 2" f=600mm lens upstream (not seen in Attachment #1). This supposedly makes a beam with waist ~80um, but I couldn't numerically find a good solution numerically if this assumption is true, so I decided to do the mode-matching empirically.
  • A single f=150mm lens got me a beam that seemed pretty well collimated, and roughly the same size as the LO beam, so I opted to push ahead with that. Later, I measured with the beam profiler that the beam is ~600um in diameter, so the beam isn't very well matched to the LO spot size, but I decided to push ahead nevertheless.
  • Patient alignment work enabled me to see interference fringes.
    • Note that the ITM reflection registers 30 cts (~80 uW). Assuming 800mW transmission through the IMC, I would have expected more like 800mW * 5.637% * 50% * 98.6% * 50% * 10% * 30% * 50% * 50% = 80uW, so this is reasonable I guess. The chain of numbers corresponds to T_PRM * T_BS * R_ITM * R_BS * T_SRM * T_vac_OMC_pickoff * R_in_air_BS * R_homodyneBS.
    • The IFO AS beam appears rather elliptical to the eye (and also on the beam profiler). It already looks like this coming out of the vacuum so not much we can do about it right now I guess. By slightly rotating the f=150mm focusing lens so that the beam going through it at ~10 degrees instead of normal incidence, I was able to get a more circular beam as measured using the beam profiler.
    • With the AS beam blocked, the LO beam registers 240 cts on each DCPD (~0.7 mW). 
    • The expected fringe should then be (sqrt(240) + sqrt(30))^2 - (sqrt(240) - sqrt(30))^2 ~ 440 cts pp.
    • The best alignment I could get is ~200 cts pp, see Attachment #2.

Next steps:

Try the PRMI experiments again, now that I have some confidence that the beams are actually interfering.

See Attachment #3 for the updated spectra - the configuration is PRMI locked with carrier resonant and the homodyne phase is uncontrolled. There is now much better clearance between the electronics noise and the MICH signal as measured in the DCPDs. The "LO only" trace is measured with the PSL shutter closed, so the laser frequency isn't slaved to the IMC length. I wonder why the RIN (seen in the SUM channel) is different whether the laser is locked to the IMC or not? The LO pickoff is before the IMC.

Attachment 1: IMG_7548.JPG
IMG_7548.JPG
Attachment 2: BHD_MM.png
BHD_MM.png
Attachment 3: PRMI_DCPDs.pdf
PRMI_DCPDs.pdf
  15536   Sun Aug 23 23:36:58 2020 gautamUpdateElectronicsFirst look at HV coil driver

Summary:

A more careful analysis has revealed some stability problems. I see oscillations at frequencies ranging from ~600kHz to ~1.5 MHz, depending on the voltage output requested, of ~2 V pp at the high-voltage output in a variety of different conditions (see details). My best guess for why this is happening is insufficient phase margin in the open-loop gain of the PA95 high voltage amplification stage, which causes oscillations to show up in the closed loop. I think we can fix the problem by using a larger compensation capacitor, but if anyone has a better suggestion, I'm happy to consider it

Details:

The changes I wanted to make to the measurement posted earlier in this thread were: (i) to measure the noise with a load resistor of 20 ohms (~OSEM coil resistance) connected, instead of the unloaded config previously used, and (ii) measure the voltage noise on the circuit side (= TP5 on the schematic) with some high voltage output being requested. The point was to simulate conditions closer to what this board will eventually be used in, when it has to meet the requirement of <1pA/rtHz current noise at 100 Hz. The voltage divider formed by the 25 kohm series resistor and the 20 ohm OSEM coil simulated resistance makes it hopeless to measure this level of voltage noise using the SR785. On the other hand, the high voltage would destroy the SR785 (rated for 30 V max input). So I made a little Pomona box to alllow me to do this measurement, see Attachment #1. Its transfer function was measured, and I confirmed that the DC high voltage was indeed blocked (using a Fluke DMM) and that the output of this box never exceeded ~1V, as dictated by the pair of diodes - all seemed okay .

Next, I wanted to measure the voltage noise with ~10mA current flowing through the output path - I don't expect to require more than this amount of current for our test masses. However, I noticed some strange features in the spectrum (viewed continuously on the SR785 using exponential averaging setting). Closer investigation using an oscilloscope revealed:

  1. 600kHz to 1 MHz oscillations visible, depending on output voltage.
  2. The oscillations vanish if I drive output above +30 V DC (so input voltage > 1 V).
  3. The oscillations seem to be always present when the output voltage is negative.
  4. No evidence of this offset if circuit is unloaded and voltage across 25k resistor is monitored. But they do show up on scope if connected to circuit side even in this unloaded config.

Some literature review suggested that the capacitor in the feedback path, C4 on the schematic, could be causing problems. Specifically, I think that having that capacitor in the feeddback path necessitates the use of a larger compensation capacitor than the nominal 33pF value (which itself is higher than the 4.7pF recommended on the datasheet, based on experience of the ESD driver circuit which this is based on, oscillations were seen there too but the topology is a bit different). As a first test of this idea, I removed the feedback capacitor, C4 - this seemed to do the trick, the oscillations vanished and I was able to drive the output between the high voltage supply rails. However, we cannot operate in this configuration because we need to roll off the noise gain for the input voltage noise of the PA95 (~6 nV/rtHz at 100 Hz will become ~200 nV/rtHz, which I confirmed using the SR785). Using a passive RC filter at the output of the PA95 (a.k.a. a "snubber" network) is not an option because we need to sum in the fast actuation path voltage at the output of the 25 kohm resistor.

Some modeling confirms this hypothesis, see Attachment #2.  The quantity plotted is the open-loop gain of the PA95 portion of the circuit. If the phase is 0 degrees, then the system goes unstable.

So my plan is to get some 470pF capacitors and test this idea out, unless anyone has better suggestions? I guess usually the OpAmps are compensated to be unconditionally stable, but in this case maybe the power op-amp is more volatile?

Quote:

Need to think more about how to better characterize this noise. An estimate of the required actuation can be found here.

Attachment 1: IMG_5379.JPG
IMG_5379.JPG
Attachment 2: stabilityCriterion.pdf
stabilityCriterion.pdf