  9329   Fri Nov 1 19:09:01 2013 rana, evanConfigurationPSLPMC reflected beam nonsense

 While looking at the PMC REFL beam for the AOM diffracted beam, we noticed that although only one beam exists between the PMC and the first steering mirror, there are two afterwards and they both go to the PMC REFL  RFPD!!! This is madness. We only want one beam on our PDH diode.

The reason that we have two beams is that that first steering mirrors is actually a (W1-PW-1025-UV-1064-45P) non-wedged window with an AR coating on only one side. So two beams come out of it. There is a terrible and floppy and illegal anodized aluminum dump close to this beam which *someone* probably intended to use as a "scraper" to get rid of one of the beams.

Black anodized aluminum is a horrible beam dump material at 1064 - its about as grey as Steve's chair. And its so soft that it scatters light back into the PMC and makes more acoustic noise. And it is mounted so poorly (only one screw) that it can easily be bumped and twist and miss the beam. Punchline: only use anodized aluminum dumps for stray light around cameras or for HeNe for OL. Its NOT allowed anywhere where we care about interferometry of NIR beams.

It was also set to dump the dimmer beam. On Monday, we should order ~5 W1 and get them with a wedge of 1-2 deg. Then we use a black glass dump for the dim beam and orient the bright one to hit the REFL camera and the PMC REFL PD.

For the weekend, I have adjusted the crappy grey aluminum flapper to catch the bright beam so that the PMC REFL image no longer shows the interference fringe of two beams. Lets see how the PMC drifts over the next 3 days.

  10418   Thu Aug 21 02:42:17 2014 rana, ericqConfigurationGreen LockingGain changes on Green Y PDH

[rana, ericq]

We spent time trying to relieve the Yend green PDH of it troubles. 

We realized that the mixer in the PDH setup (mini circuits ZAD-8+), wants 7dBm of LO to properly function. However, we use one function generators output, through a splitter, to give signals to the laser PZT and the mixer LO. 

We don't want 7dBm of power hitting the laser PZT, though. The summing node that adds the servo output to the sideband signal was supposedly designed to do some of this attenuation. Rana measured that 10Vpp out of the function generator resulted in 20mVpp on the fast input to the NPRO, after the summing node. Hence, the 0.09V setting was only resulting in something like 0.2mV hitting the PZT. The PZT has something like 30 rad/V PM response, meaning we only had ~0.006 rad of modulation. 

Now, the function generator is set to 2 Vpp, meaning 4 mVpp hitting the PZT, meaning ~0.12 radians of modulation. The mixer is now getting +7dBm on its LO, and the PDH traces look much cleaner. However, the PDH error signal is now something like 100mVpp, which is much bigger than the PDH board is designed for, so there is now a 10dB attenuator between the reflection PD DC block and the RF input to the mixer. 

Here are screenshots of the Inmon channel (which has a gain of ~20) showing a sweep through some PDH signal, and the error signal while in green lock. Huge 60Hz harmonics are still observed. 

TEK00002.PNGTEK00003.PNG

 


Regarding these 60Hz issues, we need to make sure that we remove all situations where long BNCs are chained together with barrel connectors, or Ts are touching other ones. We also should glue or affix the pomona summing box to the shelf, so that its not just laying on the floor.

The concrete next step is to go fiddle with things, and see if we can get the 60Hz noise to go away, then measure the PDH loop and noises again. Hopefully, this should make the ALS much more reliable. 

  11303   Mon May 18 17:42:14 2015 rana, ericQUpdateGeneralsome status

Today at 5 PM we replaced the east N2 cylinder. The east pressure was 500 and the west cylinder pressure was 1000. Since Steve's elogs say that the consumption can be as high as 800 per day we wanted to be safe.

  1. We closed the black valve before the regulator and closed the valve on the cylinder.
  2. We unscrewed the brass fill line to the cylinder.
  3. We unchained the cylinder and put in the dolly (and attached the chains on there).
  4. We rolled in a fresh cylinder from outside using the red dolly (it should have chains).
  5. We put it in place, hooked up the chains, and screwed on the brass nozzle with the large adjustable wrench (need to put a non-adjustable here).
  6. Opened up the cylinder valve.
  7. Opened up the black valve.
  8. New east pressure reading is 2500 PSI. Regulated N2 pressure is 68 PSI.
Quote:

1) Checked the N2 pressures: the unregulated cylinder pressures are both around 1500 PSI. How long until they get to 1000?

 

  11315   Tue May 19 18:55:12 2015 rana, ericQUpdateGeneralsome status

After one day the pressures are east/west = 2200/450 PSI

Quote:

 

  1. New east pressure reading is 2500 PSI. Regulated N2 pressure is 68 PSI.

 

  11905   Mon Jan 4 14:45:41 2016 rana, eq, kojiConfigurationComputer Scripts / Programsnodus pwd change

We changed the password for controls on nodus this afternoon. We also zeroed out the authorized_keys file and then added back in the couple that we want in there for automatic backups / detchar.

Also did the recommended Ubuntu updates on there. Everything seems to be going OK so far. We think nothing on the interferometer side cares about the nodus password.

We also decided to dis-allow personal laptops on the new Martian router (to be installed soon).

  10723   Mon Nov 17 20:40:29 2014 rana, diegoUpdateIOOInvestigating the IMC WFS situation

We've known for years that the IMC WFS sensing chain is pointlessly bad, but until recently, we haven't thought it was worth it to fix.

There are problems in all parts of the chain:

  1. The WFS Photodetectors oscillate ~200 MHz when turned up to full gain. Diego and I confirmed this today by measuring the RF spectrum of the signals going into the WFS demod boards and seeing the oscillation change (not much) with RF gain. I recommend we switch the heads into the full gain mode (turn all of the attenuators OFF). At the moment we are operating with the 2dB and 8dB attenuators ON.
  2. The demod board has some bad gain allocation and noisy opamps.
  3. The whitening board has too much up/down of gain with noise injection along the way. And the range cannot fill up the ADC.
  10803   Tue Dec 16 01:50:27 2014 rana, diegoFrogsLSCMICH filter is nuts

 This is ridiculous.

How many RGs can I fit into one button???

Attachment 1: badMICHrg.pdf
badMICHrg.pdf
  6049   Wed Nov 30 02:04:26 2011 rana, den, jenne, kiwamu, jzweizigUpdateCDSFiltering Noise issue tracked down ???

 You can read through all of our past tests to see what didn't work in tracking things down. As Den mentions, there was actually a lot of evidence that there was some double->single precision action in the filter calculation causing the noise we saw.

However, it turns out that this is NOT the case.

This afternoon I was so confused that I enlisted JZ to help us out. He came over and I tried to replicate the error. When looking at the time series, we noticed that it wasn't random noise; the signals seem to be getting clipped as they crossed zero. Sort of like a stiction problem. JZ left to go replicate the error on an offline system.

This turned out to be the important clue. As we examine the code we find this inside of fm10Gen.c:

if((new_hist < 1e-20) && (new_hist > -1e-20)) new_hist = new_hist<0 ? -1e-20: 1e-20;

this is line is basically trapping the filter history at 1e-20, to prevent some kind of numerical underflow problem (?). Seems reasonable, except that some filters which have higher order low passing in them actually have an overall scale factor which can be small (even as small as 1e-23, as Den pointed out).

So the reason we saw such weird behavior is that the first filter in SUSPOS is an AC coupling filter. This takes the OSEM signal and remove the large mean value. Then the next filter multiplies it by 1e-23 before doing the filtering and you end up with this noise in the filter history.

I looked and this line is commented out in the new BiQuad code, but as far as I can tell this issue has been around in aLIGO, eLIGO, iLIGO, etc. for a long time and could have been causing many cases of excess noise whenever we ended up a tiny gain factor in an IIR filter. At the 40m, there are easily a hundred such cases.

For now, I suppose we can just change this number to 1e-40 or so. I don't know how to calculate what the right number should be. Not sure why this underflow is not an issue for the BiQuad, however.

  1551   Wed May 6 16:56:35 2009 rana, alex, joeConfigurationComputersdaqd log, cront, etc.
While Alex came over, we investigated the log file problems with DAQD and NDS on FB0. There was a lot of
the standard puzzling and mumbling, but eventually we saw that it doesn't create its log file and so it
doesn't write to it. The log file is /usr/controls/main_daqd.log. The other files called daqd.log.DATE
in the logs/ directory are actually not written to. Its awesome.

We also have put in a fix for the overflowing jobs/ directory. It gets a file written to it every time
you make and NDS request and our seisBLRMS has been overloading it. There's now a cron for it in the fb0
crontab which cleans out week-old files at 6:30 AM every day.

We also changed the time of the daily backup from 3:30 AM (when people are still working) to 5:50 AM
(by which time the seismic has ramped up and interferometerists should be asleep). I didn't like the
idea of a bandwidth hog nailing the framebuilder during the peak of interferometer work.

#
# Script to backup via rsync the most recent 40m minute trends and
# any changes to the /cvs/cds filesystem.
#
50 05 * * * /cvs/cds/caltech/scripts/backup/rsync.backup < /dev/null > /cvs/cds/caltech/scripts/\
backup/rsync.backup.log 2>&1

30 06 * * * find /usr/controls/jobs -mtime +7 -exec /bin/rm -f {} \;

seisBLRMS.m restarted on mafalda.
  1973   Tue Sep 8 15:14:26 2009 rana, alexConfigurationDAQRAID update to Framebuilder: directories added + lookback increased

 Alex logged in around 10:30 this morning and, at our request, adjusted the configuration of fb40m to have 20 days of lookback.

I wasn't able to get him to elog, but he did email the procedure to us:


1) create a bunch of new "Data???" directories in /frames/full
2) change the setting in /usr/controls/daqdrc file
       set num_dirs=480;

my guess is that the next step is:

3) telnet fb0 8087

    daqd>  shutdown

I checked and we do, in fact, now have 480 directories in /frames/full and are so far using up 11% of our 13TB capacity. Lets try to remember to check up on this so that it doesn't get overfull and crash the framebuilder.

  4112   Wed Jan 5 16:00:11 2011 rana, alexSummaryDAQFrameBuilder fails in a new way

Email from Alex:

Turned out to be the lack of current year information in the IRIG-B signal
received by the Symmetricom GPS card in the frame builder machine caused
this. I have added a constant in daqdrc to bring the seconds forward:

controls@fb /opt/rtcds/caltech/c1/target/
fb $ grep symm daqdrc
#set symm_gps_offset=-1;
set symm_gps_offset=31536001;

Hopefully we will be upgrading to the newer timing system at the 40M this
year, so this will not happen again next year.


 

Doing an 'ls -lrt' in /frames/full/ now shows that the names are correct:

drwxr-xr-x 2 controls controls 249856 Dec 30 19:06 9777
drwxr-xr-x 2 controls controls 192512 Dec 31 16:00 9778
drwxr-xr-x 2 controls controls 151552 Jan  1 08:53 9463
drwxr-xr-x 2 controls controls 266240 Jan  2 12:39 9464
drwxr-xr-x 2 controls controls 258048 Jan  3 16:26 9465
drwxr-xr-x 2 controls controls 253952 Jan  4 20:13 9466
drwxr-xr-x 2 controls controls 151552 Jan  5 13:54 9467
drwxr-xr-x 2 controls controls  12288 Jan  5 15:57 9783

  1151   Fri Nov 21 16:11:26 2008 rana, alberto, robUpdatePSLMach Zender alignment
The Mach Zender's dark port DC voltage had gone up too high (~0.5 V)and was turning yellow
on the screen. I re-aligned by touching the knobs on the 166 MHz path. Doing alignment after the
166 EOM had very little effect. The main improvement came from doing yaw on the turning mirror
just ahead of the 166 MHz EOM; this is the one which as no adjustment knobs (duh).

During this procedure, I had the MC off, the ISS off, and the MC autolocker off. After finishing
the alignment, the power on the ISS PDs had railed and the dark port power was ~0.29 V. So we
put in a ND0.2 on the ISS path and now the voltages are OK (~2 V on each PD). We have to remove the
ND filters and change the first ISS turning mirror into a ~10-20% reflector.


So now the MZ alignment seems good; Alberto is on the MC periscope alignment like a cheap suit.
Attachment 1: PB210051-1.JPG
PB210051-1.JPG
  1155   Fri Nov 21 20:29:43 2008 rana, alberto, robUpdatePSLMach Zender alignment

Quote:
So now the MZ alignment seems good; Alberto is on the MC periscope alignment like a cheap suit.


And alignment is now mostly done - MC locks on the TEM00.
                 REFL_DC

Unlocked           4.50  V
Locked noWFS       1.30  V
Locked + WFS       0.42  V
and the 29.5 MHz modulation depth is really small.

We should be able to rerun the Wiener analysis on this weekends MC data.

I don't know what our nominal StochMon numbers now are, but after Alberto tweaks up the alignment he can tell us if the RFAM has gotten any better.
  1161   Mon Nov 24 19:15:16 2008 rana, alberto, johnConfigurationEnvironmenttemperature
The PSL Room Temperature was alarming because it had gone above 23 C. This set off an unfortunate chain of events:

We found that the PSL HEPA was set low (20%). This is a fine setting for when no one is working in there but it
does raise the temperature since there are heat sources inside the blue box.

We tried to change the office area temperature to compensate and also the westmost sensor inside the lab area by 2 deg F.

The office area one was problematic - there was so much dust in it that the gas valve nipple was clogged. So we've
now blown it all clean with a compressed air can. We're now tuning the calibration screw to make our new
digital sensor agree with the setpoint on the controller.

Expect wild temperature swings of the office area for a couple days while Alberto and I tune the servo.
  1163   Tue Nov 25 19:29:15 2008 rana, alberto, johnConfigurationEnvironmenttemperature

Quote:
The PSL Room Temperature was alarming because it had gone above 23 C. This set off an unfortunate chain of events:

We found that the PSL HEPA was set low (20%). This is a fine setting for when no one is working in there but it
does raise the temperature since there are heat sources inside the blue box.

We tried to change the office area temperature to compensate and also the westmost sensor inside the lab area by 2 deg F.

The office area one was problematic - there was so much dust in it that the gas valve nipple was clogged. So we've
now blown it all clean with a compressed air can. We're now tuning the calibration screw to make our new
digital sensor agree with the setpoint on the controller.

Expect wild temperature swings of the office area for a couple days while Alberto and I tune the servo.


This morning Bob found 92F in the office area and in the control room of the lab. He turned down the thermostat and when I got in at about 9 I found 65. After a few hours of adjustment of the thermostat's calibration, I could stabilize the room temperature to about 72F. I also turned down the thermostats inside the lab of a couple of degrees F.
  1359   Thu Mar 5 01:09:29 2009 rana, albertoUpdateDMFstill not working
We tried to run DMF on mafalda, but it didn't work. I tried to compile it using Rob's elog instructions.
On mafalda, I started matlab and ran the mdv_config.m to set up mDV. I tested that the seisBLRMS.m
script ran and correctly produced changes in the seisBLRMS strip tool. but when I tried to compile it I got:
>> mcc -v -m -R -nojvm seisBLRMS.m
Warning: Duplicate directory name: /cvs/cds/caltech/apps/linux/matlab/toolbox/local.
Compiler version: 4.6 (R2007a)
Processing /cvs/cds/caltech/apps/linux/matlab/toolbox/matlab/mcc.enc
Processing /cvs/cds/caltech/apps/linux/matlab/toolbox/signal/mcc.enc
Processing /cvs/cds/caltech/apps/linux/matlab/toolbox/control/mcc.enc
Processing /cvs/cds/caltech/apps/linux/matlab/toolbox/filterdesign/mcc.enc
Processing /cvs/cds/caltech/apps/linux/matlab/toolbox/shared/controllib/mcc.enc
Processing /cvs/cds/caltech/apps/linux/matlab/toolbox/ident/mcc.enc
Warning: an error occurred while parsing class FilterDesignDialog.AbstractEditor:
Undefined function or variable 'DAStudio.Object'.

> In /cvs/cds/caltech/apps/linux/matlab/toolbox/shared/filterdesignlib/@FilterDesignDialog/@CoeffEditor/schema.p>schema at 9
Warning: an error occurred while parsing class FilterDesignDialog.CoeffEditor:
Invalid superclass handle.

Processing /cvs/cds/caltech/apps/linux/matlab/toolbox/fixedpoint/mcc.enc
terminate called after throwing an instance of 'ApplicationRedefinedException*'
Abort (core dumped)
"/cvs/cds/caltech/apps/linux/matlab/bin/mcc" -E "/tmp/fileRnU5Qj_31324": Aborted
??? Error executing mcc, return status = 134.

In the meantime, I've started up a green terminal on allegra which is ssh'd into megatron.
On megatron there is a regular matlab session which is running seisBLRMS.m as a matlab script
and the seis DMF channels are getting updated.
  1811   Wed Jul 29 19:46:04 2009 rana, albertoUpdatePEMDents = Bad??
It looks like the MC2 chamber and/or stack has been jarred and shifted. Please be careful and use much less force and speed around the MC2 chamber.

My guess is that the work with the accelerometers around there had made the MC2 angle and
position change last night. The reason that we don't see it in the OSEMs so much is that its
a change in the actual stack position and tilt.

To recover, we changed the MC2 alignment bias to get the beam through the Faraday. This did NOT get
the beam back onto the right place on the MC TRANS QPD. For tonight we decided to not recenter that
since Rob might not like this position. We did, however, zero out the MC WFS and the PSL POS/ANG.

If the interferometer locking is OK tonight, then we (Steve and whoever else is here at 7 AM)
should recenter IP POS and IP ANG and also fix up the PSL POS and PSL ANG QPDs. You can see
in the attached picture that there are two problems to fix:
1) The knobs (circled in red and blue) are wrapped in foil. Why???
2) The handedness of the mirror mount with the orange arrow is wrong. This should be unmounted and clocked
by 90 deg. Right now the beam is nearly clipping on the mount. Also, we need to change the channel names
on the PSL POS (or maybe its ANG). It has the horizontal/vertical channels misnamed.
  2469   Wed Dec 30 20:33:36 2009 rana, albertoConfigurationCamerasITMY & MC2 Camera work

We restored the good state of the ITMY camera and neatened both the MC2 and ITMY camera.

The MC2 camera was driving a triple T jungle into some random cables and spoiling the image. We removed all T's and the MC2 camera now drives only The Matrix.

The ITMY camera was completely unmounted and T'd. So it was misaligned just by the force of gravity acting on its BNC cable. We swapped the lens for a reasonable sized one and remounted it in its can. We then used orange cable ties to secure the power and BNC cable for the MC2 and ITMY cameras so that tugging on the cables doesn't misalign the cameras. This is part of the camera's SOP.

No more driving 50 Ohm cables and T's with video cables, Steve! If you need a portable video, just use a spigot of the Matrix and then you can control it with a web browser.

DSC_1064.JPGDSC_1065.JPGDSC_1066.JPG

I also wiped out the D40's memory card after uploading all of the semi-useful files to the Picasa page.

  9898   Fri May 2 02:22:56 2014 rana, QUpdateIOOMC alignments

We were having unstable MC locking so we did some physical alignment touch up. Use MC suspension bias to have good MC alignment. Unlock MC and align DC beams to center on WFS. Re-lock and things are now stable.

Someone had been feeding bad info to Eric about MC alignments; we do not center the MC WFS beams with the MC locked.

However, in any case, today the MC2-TRANS servo was not being good so I turned it off. We need the real matrix measurement to bring it back.

  9932   Thu May 8 17:00:56 2014 rana, QSummaryLSCREFL_DC handoff didn't work last night

Last night after checking cabling and turning on ISS, we tried several times to handoff to REFL_DC but it didn't work at all.

Some issues:

  1. The ISS was injecting a lot of very low frequency power fluctuations because of bad AC coupling.
  2. The SR560 @ LSC rack was saturating a lot with the x10 gain that Jenne and Rana put in; we turned it back to G = 1.
  3. The ISS was also saturating a lot. We turned it off around 4 AM, but still no success.
  4. The ALS sequence for finding the Red Resonance takes too long (~2 minutes), so we're trying a faster scheme tonight.
  8555   Thu May 9 00:05:12 2013 rana, Koji, JenneSummaryLSCAA and AI change

We would like to increase the UGF of the PRC loop so as to allow more suppression of the PRC signal and less pollution of the MICH signal (remember that the PRC/MICH optical gain ratio is huge).

We were already losing phase because of delay in the LSC - SUS digital link. In addition to that, a major source of delay is the analog anti-aliasing (on the LSC error signals before they enter the ADC) and the analog anti-imaging (between the SUS DAC and the coil driver).

 IN addition to these, the other major sources of phase lag in the system are the FM5 filter in the LSC-PRC filter bank, the digital upsampling and downsampling filters, and the DAC sample and hold.

In the near term, we want to modify these analog filters to be more appropriate for our 64 kHz ADC/DAC sample rate. Otherwise, we are getting the double phase lag whammy.

 

Staring at the schematics for the AA (D000076-01) and the AI (D000186-A), we determined a plan of action.

For the AA, we want to remove the multi-pin AA chip filter from Frequency Devices, Inc. and replace it with a passive LC low pass. Hopefully, these chips are socketed. Rana will design an appropriate LC combo and elog; we should make the change on a Wednesday afternoon so that we have enough soldering help.

For the AI, the filter is a dual bi-quad using discrete components and LT1125 opamps. Not so clear what to do with these. The resistors are all the noisy thick film kind and maybe should be replaced. Koji will find some online design tool for these or do it in LISO. Changing the TF is easy; we can just scale the capacitors. But we also want to make sure that the noise of the AI does not destroy the noise reduction action of the dewhitening board which precedes it.

Jenne should figure out how low the noise needs to be at the input to the coil driver.

 

P.S. the matlab code which defines these filters

>> [z,p,k] = ellip(4,4,60,2*pi*7570,'s');
>> misc.ai = zpk(z,p,k*10^(4/20)) * zpk([],-2*pi*13e3,2*pi*13e3);
>>
>> % Fudged Anti-Imaging Filter
>> [z,p,k] = ellip(8,0.001,80,2*pi*7570,'s');
>> misc.aa = zpk(z,p,k*10^(0.001/20)) * zpk([],-2*pi*32768,2*pi*32768);

Attachment 1: AAAI.pdf
AAAI.pdf
  15341   Wed May 20 20:10:34 2020 rana, John ZUpdateComputer Scripts / ProgramsNDS2 server / conf updated - seems OK now

We noticed about a week ago that the NDS2 channel lists were not getting updated on megatron. JZ and I investigated; he was able to fix it all up this afternoon by logging in and snooping around Megatron.

Please try it out and tell me about any problems in getting fresh data.


  1. The NDS2 server is what we connect to through our python NDS2 client software to download some data.
  2. It has been working for years, but it looks like there was a file corruption of the channel lists that it makes back in 2017.
  3. Since the NDS2 server code tries to make incremental changes, it was failing to make a new channel list. Was failing to parse the corrupted file.
  4. there was a controls crontab entry to restart the server every morning, but the file name in that tab had a typo, so that wasn't working. I commented it out, since it shouldn't be necessary (lets see how it goes...)
  5. the nds2mgr account also has a crontab, but that was failing since it didn't have sudo permission. JZ added nds2mgr to the sudoers list so that should work now.
  6. I was able to get new channels as of 4 PM today, so it seems to be working.

* we should remember to rebuild the NDS2 server code for Ubuntu. The thing running on there is for CentOS / SL7, but we moved to Ubuntu recently since the SL7 support is going away.

** the nds2 code & conf files are not backed up anywhere since its not on /cvs/cds. It has 52 GB(!!) of txt channel lists & archives which we don't need to backup

  3609   Sun Sep 26 18:29:23 2010 rana, JohnUpdateCDSModified front end medm screens

Issues I notice on first glance:

  1. The OSEM Sensor Input matrix and the DOF2COIL Output matrix screens should be their own screens and linked from the overview. Right now they are not. Where is the input matrix?
  2. The SIDE GAIN looks like zero on the main screen, but the side OSEM signal seems to be getting through to the SIDE filter bank. . I think the wiring of the SIDE signal through the input matrix is bogus.
  3. The OUTPUT matrix seems to be the transpose of the previous OUTPUT matrix and we have lost the wires that connect the inputs and outputs to the matrix. We ought to think about how best to represent things on the OVERVIEW screen; probably only need to have a minimal representation and allow power users to open up the detailed screen.
  4. The TIME string is whited out. How will this be done? Does each FE display its local time on its EPICS screens?
  5. So far unable to get any channels on DV. How do we look at channels / test points?
  6. As far as we can tell, there is no connection between the output of the SUSPOS, etc. filter banks and the OUTPUT MATRIX. So....nothing actually goes to the coil driver. Its hard to imagine that this new SUS could have ever worked. Is there any evidence that the damping actually worked in the past, or was it something like "well, the watchdog values came down to small numbers eventually..." ???
  7. We are trying to debug the simulink file, but....the wiki entry on how to do this is out of date (yet updated as recently as August!) some path stuff just probably needs to be edited.

 megamind-poster3.jpg

Basically the suspensions are not functioning yet and we can't attempt locking of the MC.

  1926   Tue Aug 18 19:57:47 2009 rana, JenneUpdatePSLMZ

- we finished the MZ alignment; the contrast is good.

- we did the RFAM tuning using a new technique: a bubble balanced analyzer cube and the StochMon RFPD. This techniques worked well and there's basically no 33 or 166 RFAM. The 133 and 199 are as expected.

- the MC locked right up and then we used the periscope to align to it; the transmission was ~75% of max before periscope tuning. So the beam pointing after the MC should be fine now.

- the Xarm locked up with TRX = 0.97 (no xarm alignment).

 

If Rob/Yoichi say the alignment is now good, the we absolutely must center the IOO QPDs and IP POS and IP ANG and MC TRANS  today so that we have good references.

 

-----------------------------------

The first photo is of our nifty new setup to get the beam to the StochMon PD.  The MZ transmitted beam enters the photo from the bottom right corner, and hits the PBS (which we leveled using a bubble level).  The P-polarization light is transmitted through the cube, and the S-polarization is reflected to the left.  The pure S-polarized light hits a Beam Splitter, which we are using as a pickoff to reduce the amount of light which gets to the PD.  Most of the light is dumped on an aluminum dump.  The remaining light hits a steering mirror (Y1 45-S), goes through a lens, and then hits the StochMon PD.  While aligning the MZ to maximize visibility, we look at the small amount of P-polarized light which passes through the PBS on an IR card, and minimize it (since we want to be sending purely S-polarized light through the EOMs and into the MC).

The second photo is of a spectrum analyzer which is directly connected to the RF out of the StochMon PD.  To minimize the 33MHz and 166MHz peaks, we adjust the waveplates before each of the EOMs, and also adjusted the tilt of the EOM holders.

The final photo is of the EOMs themselves with the Olympus camera.

Once we finished all of our MZ aligning, we noticed that the beam input to the MC wasn't perfect, so Rana adjusted the lower periscope mirror to get the pointing a little better.  

The MZ refl is now at 0.300 when locked.  When Rana reduced the modulation depth, the MZ refl was about 0.050 .  Awesome!

 

Attachment 1: MZ_RFAMmon_setup_small.jpg
MZ_RFAMmon_setup_small.jpg
Attachment 2: MZ_RFAMmon_SpecAnalyzer_small.jpg
MZ_RFAMmon_SpecAnalyzer_small.jpg
Attachment 3: MZ_EOM_IRrefl2_small.jpg
MZ_EOM_IRrefl2_small.jpg
  2665   Tue Mar 9 12:06:53 2010 rana, JenneUpdatePEMStyrofoam Cooler on the Seismos

Looks like the GUR2_X signal is bad. Jenne says that we need to center it mechanically before the signals will become useful again. Maybe Steve will do this - instructions are in the manuel ?

  8531   Mon May 6 21:05:06 2013 rana, Jamie, KOjiUpdateIOOmode cleaner not locking

As it turned out, the setting of +26dB for the FSS Fast gain was making the NPRO PZT / Pockels cell crossover too unstable. This was the cause of the large ~30 kHz peak that Jamie was seeing in his spectra (more to follow in the morning).

After recovering the locking, etc. by fiddling with the gains, we went about systematically setting all of the gains/offsets. Jamie's elog will show all of the various spectra with different FSS gains.

For offset setting, this was the procedure:

  1. We want the MC servo board to have zero voltage on its MCL and MCF outputs (aka MC_SLOW_MON and MC_FAST_MON) with the Boosts ON, so we switched ON the 40:4000 and the 2 Super Boosts and put the VCO gain into its nominal +25 dB setting. This saturates the outputs and makes them impossible to use as readbacks so you have to be careful. Get the outputs close to zero as you turn on each boost. Finally, you will see that +/- 1mV of input offset (aka MC_REFL_OFFSET) will flip the FAST output between +/- 10V. This is the right setting.
  2. With the MC board adjusted to send 0 V into the FSS error point, adjust the FSS input offset (with the Common and Fast gains at the nominal values) so that the FAST output voltage goes to +5.5 V. This is the middle of the range for the high voltage driver which drives the NPRO.
  3. Let the MC lock and go through the UP script. When the MCL comes on with the integrator, the FAST voltage will shift from +5.5 V to something else. Use the following line: ezcaservo -r C1:PSL-FSS_FAST -g -1 -s 5.5 C1:SUS-MC2_MCL_OFFSET, to automatically tune the MCL offset.

     

    That's all. I have left the FSS common gain at +10.5 and the Fast gain at +23.5. These seem to be close to the optimum. Jamie will have more tuning tomorrow.
     
    I have also changed the 'mcup' script to set the MCL offset and set the FSS Slow setpoint to shoot for +5.5 V of FSS_FAST.
     
    MC_REFL_OFFSET = +1.176 V
    MC2_MCL_OFFSET = +47.8 counts
    FSS_INOFFSET   = -0.85 V
  6081   Wed Dec 7 09:56:14 2011 rana imitation of steveOmnistructureEnvironmentTelephone Map

 This is the fone storree 

we should turn off 3977,3979 and 2351 has no function

 

Edit, JD:  x8925 is at the Visitor desk, and isn't on the map.

  1   Wed Oct 17 18:46:33 2007 ranaConfigurationGeneraleLog Change
This is the first entry in the new 40m eLog.

Its GWs or bust now! Big grin



[Hnull][/Hnull]
  2   Thu Oct 18 14:52:35 2007 ranaRoutineASCtest
test

X-(:P;(:))
  9   Tue Oct 23 09:01:00 2007 ranaOtherOMCPZT calibration/ transfer function.
Are you sure that the error signal sweep is not saturated on the top ends? This is usually the downfall
of this calibration method.
  13   Thu Oct 25 00:01:21 2007 ranaSoftware InstallationCDSGEO DV => LIGO DV
Martin Hewitson of GEO600 fame has modified the cool GEO DV
to work with the LIGO NDS system with some NDS advice from Rolf (who's over in Germany this week).

I've moved it onto the 40m CDS system and installed it on the AdhikariLab computer named 'django'. It worked immediately.

I modified the main .m file to include the 40m's NDS server. When you run it you have to include the path to the NDS
client written by Ben Johnson.

The attached is a screenshot of it working on a Mac; it looks as cool on Linux.

Its installed in /cvs/cds/caltech/apps/ligoDV/. In matlab you navigate to that directory and then
type addpath('/cvs/cds/caltech/apps/linux/UNIX_NDS_Client_beta2/') to add the NDS client.
On the Solaris machines, type type addpath('/cvs/cds/caltech/apps/solaris9/UNIX_NDS_Client_beta2/') instead.

Then type ligoDV to start it up. Then click away and have fun.

In the example I've selected
C1:PEM-BS_ACC_EAST_Z
and plotted its specgram.

Big grin
Attachment 1: Picture_1.png
Picture_1.png
  22   Sun Oct 28 03:03:42 2007 ranaConfigurationIOOThree Way Excitement
We've been trying to measure the MC mirror internal mode frequencies so that we can measure
their absorption before and after drag wiping.


It looked nearly impossible to see these modes as driven by their thermal excitation level;
we're looking at the "MC_F" or 'servo' output directly on the MC servo board.

Today, I set up a band limited noise drive into the 'Fast POS' inputs of the 3 MC coil
driver boards (turns out you can do this with either the old HP or the SR785).

Frequencies:
MC1     28.21625 kHz
MC2     28.036   kHz
MC3     28.21637 kHz

I don't really have this kind of absolute accuracy. These are just numbers read off of the SR785.

The other side of the setup is that the same "MC_F" signal is going into the SR830 Lock-In which
is set to 'lock-in' at 27.8 kHz. The resulting demodulated 'R" signal (magnitude) is going into
our MC_AO channel (110B ADC).

As you can see from the above table, MC1 and MC3 are astonishingly and annoyingly very close in
frequency. I identified mirrors with peaks by driving one at a time and measuring on the spectrum
analyzer. I repeated it several times to make sure I wasn't fooling myself; it seems like they
are really very close
but distinct peaks. I really wish we had chipped one of these mirrors
before installing them.



Because of the closeness of these drumhead modes, we will have to measure the absorption by making long
measurements of this channel.
  29   Tue Oct 30 00:47:29 2007 ranaOtherIOOMC Ringdowns
I did a bunch of MC ringdown measurements using the PD that Rob set up. The idea is to put a fast PD (PDA255)
looking at the transmission through MC2 after focusing by a fast lens. The input to the MC is turned off fast
by flipping the sign of the FSS (Andri Gretarsson's technique).

With the laptop sitting on the MC can, its easy to repeat many ringdowns fast:
- Turn off the MC autolocker. Relock the MC with only the acquisition settings; no boosts
  and no RGs. This makes it re-acquire fast. Turn the MC-WFS gain down to 0.001 so that
  it keeps it slowly aligned but does not drift off when you lose lock.

- Use low-ish gain on the FSS. 10 dB lower than nominal is fine.

- Setup the o'scope (100 MHz BW or greater) to do single shot trigger on the MC2 trans.

- Flip FSS sign.

- Quickly flip sign back and waggle common gain to get FSS to stop oscillating. MC
  should relock in seconds.

Clearly one can scriptify this all just by hooking up the scope to the ethernet port.


Attached are a bunch of PNG of the ringdowns as well as a tarball with the actual data. A sugar
napoleon to whomever can explain the 7 us period of the wiggle before the vent!
Attachment 1: tek00000.png
tek00000.png
Attachment 2: tek00001.png
tek00001.png
Attachment 3: tek00004.png
tek00004.png
Attachment 4: MC2ringdown.tar.gz
  34   Wed Oct 31 08:33:54 2007 ranaProblem FixedSUSVent measurements
There was a power outage during the day yesterday; whoever was around should post something here about the
exact times. Andrey and David and Tobin got the computers back up - there were some hiccups which you can
read about in David's forthcoming elog entry.

We restarted a few of the locking scripts on op340m: FSSSlowServo, MCautolocker. Along with the updates
to the cold restart procedures we have to put an entry in there for op340m and a list of what scripts
to restart.

David tuned up the FSS Slow PID parameters a little; he and Andrey will log some entry about the proper
PID recipe very soon. We tested the new settings and the step response looks good.

We got the MC locking with no fuss. The 5.6 EQ in San Francisco tripped all of the watchdogs and I upped
the trip levels to keep them OK. We should hound Rob relentlessly to put the watchdog rampdown.pl into
the crontab for op340m.
  35   Wed Oct 31 08:34:35 2007 ranaOtherIOOloss measurements
In the end, we were unable to get a good scatter measurement just because we ran out of steam. The idea was to get a frame
grab image of MC2 but that involves getting an unsaturated image.

In the end we settle for the ringdowns, Rob's (so far unlogged) cavity pole measurement, and the MC transmission numbers. They
all point to ~100-150 ppm scatter loss per mirror. We'll see what happens after wiping.
  36   Wed Oct 31 08:38:35 2007 ranaProblem FixedIOOMC autolocker
The MC was having some trouble staying locked yesterday. I tracked this down to some steps in the last
half of the mcup script; not sure exactly which ones.

It was doing something that made the FAST of the PSL go to a rail too fast for the SLOW to fix.
So, I broke the script in half so that the autolocker only runs the first part. We'll need to
fix this before any CM locking can occur.

We also need someone to take a look at the FSS Autolocker; its ill.
  61   Sun Nov 4 23:55:24 2007 ranaUpdateIOOFriday's In-Vac work
On Friday morning when closing up we noticed that we could not get the MC to flash any modes.
We tracked this down to a misalignment of MC3. Rob went in and noticed that the stops were
still touching. Even after backing those off the beam from MC3 was hitting the east edge of
the MC tube within 12" of MC3.

This implied a misalignment of MC of ~5 mrad which is quite
large. At the end our best guess is that either I didn't put the indicator blocks in the
right place or that the MC3 tower was not slid all the way back into place. Since there
is such a strong stickiness between the table and the base of the tower its easy to
imagine the tower was misplaced.

So we looked at the beam on MC2 and twisted the MC3 tower. This got the beam back onto the
MC2 cage and required ~1/3 if the MC3 bias range to get the beam onto the center. We used
a good technique of finding that accurately: put an IR card in front of MC2 and then look
in from the south viewport of the MC2 chamber to eyeball the spot relative to the OSEMs.

Hitting MC2 in the middle instantly got us multiple round trips of the beam so we decided
to close up. First thing Monday we will put on the MC1/MC3 access connector and then
pump down.


Its possible that the MC length has changed by ~1-2 mm. So we should remeasure the length
and see if we need to reset frequencies and rephase stuff.
  62   Mon Nov 5 07:29:35 2007 ranaUpdateIOOFriday's In-Vac work
Liyuan recently did some of his pencil beam scatterometer measurements measuring not the
BRDF but instead the total integrated power radiated from each surface point
of some of the spare small optics (e.g. MMT, MC1, etc.).

The results are here on the iLIGO Wiki.

So some of our loss might just be part of the coating.
  91   Sun Nov 11 21:05:55 2007 ranaHowToSUSMC Touching or not
I wrote a script: SUS/freeswing-mc.csh, which gives the MC mirrors the appropriate kicks
needed to make a measurement of the free swinging peaks in the way that Sonia did.
#!/bin/csh

set ifo = C1
set sus = ${ifo}:SUS-

foreach opt (MC1 MC2 MC3)

  set c = `ezcaread -n ${sus}${opt}_PD_MAX_VAR`
  ezcastep ${sus}${opt}_PD_MAX_VAR +300

  ezcaswitch ${sus}${opt}_ULCOIL OFFSET ON
  ezcawrite ${sus}${opt}_ULCOIL_OFFSET 30000
  sleep 1
  ezcawrite ${sus}${opt}_ULCOIL_OFFSET 0
  sleep 1
  ezcawrite ${sus}${opt}_ULCOIL_OFFSET 30000
  sleep 1
  ezcawrite ${sus}${opt}_LATCH_OFF 0

  ezcawrite ${sus}${opt}_ULCOIL_OFFSET 0
  ezcaswitch ${sus}${opt}_ULCOIL OFFSET OFF

  ezcawrite ${sus}${opt}_PD_MAX_VAR $c

end

echo
date
echo

It basically ups the watchdog threshold, wacks it around at the pendulum frequency, and then disables the
optic so that there are no electronic forces applied to it besides the bias. The date command at the end
is so that you know when to start your DTT or mDV or lalapps code or whatever.
  92   Sun Nov 11 21:21:04 2007 ranaHowToComputersNew DV
To use the new ligoDV (previously GEO DV) to look at 40m data, open up a matlab, set up for mDV as usual,
and then from the /cvs/cds/caltech/apps/ligoDV/ directory, type 'ligoDV'.

Then select which NDS server you want to look at and then start clicking to get some plots.
Attachment 1: Screenshot-1.png
Screenshot-1.png
  128   Wed Nov 28 04:21:46 2007 ranaUpdatePSLFSS

Quote:
Rana, Tobin

We looked at the RF PD signal to the FSS (siphoning off a signal via a minicircuits directional coupler) and also took an open loop transfer function of the FSS. In the transfer function we saw the step at 100 kHz (mentioned by Rob) as well as some peculiar behavior at high frequency. The high frequency behavior (with a coupling of ~ -20 dB) turns out to be bogus, as it is still present even with the beam blocked. Rearranging the cabling had no effect; the cause is apparently inside the FSS. The step at 100 kHz turns out to be a saturation effect, as it moved as we lowered the signal amplitude, disappearing as we approached -60 dBm. (Above the step, the measurement data is valid; below, bogus.)

Transfer functions will be attached to this entry.

Some things to check tomorrow: the RF signal to the PC, RF AM generation by the PC, LO drive level into the FSS, RF reflection from the PC, efficiency of FSS optical path, quality of RF cabling.


I would also add to Tobin's entry that we believe what Rob was seeing was saturation.

With the bi-directional coupler in there, the RF signal into the FSS board clearly went UP if moved the offset slider away from zero.
With a scope looking at the IN2 testpoint, we can see that there's less than 2 mV offset at zero slider offset.

One tangential thing we noticed with the coupler is that, in lock, the amount of reflected RF is around the same as that going in to the mixer.
I have always wanted to look at this but have only had uni-directional couplers in the past. I think that the double balanced mixer is inherently
not a 50 Ohm device during the times where the diodes are being switched. IF that's the case we might do better in the future by having an RF
buffer on board just before the mixer to isolate the PD head from these reflections.
  132   Wed Nov 28 16:46:28 2007 ranaConfigurationComputersscientific linux 5.0
I tried installing Scientific Linux on Tiramisu. The installation process was so bad (really)
that I quit after 15 minutes. Its back to booting Ubuntu as if nothing had ever happened. Let
us never speak of Scientific Linux again.
  133   Wed Nov 28 17:15:26 2007 ranaConfigurationSUSETMY damping / watchdogs
Steve has noted that ETMY was often tripping its watchdog. I saw this again today.

So I checked the damping settings. Someone had set the SIDE gain to +1. The gain which gives
it a Q of ~10 is +10. I set the SIDE gain to +20. I checked and the ETMX gain is -16 so now
they're at least similar. I have updated the snapshot to reflect the new value.

Hopefully now it will be more well behaved.
  143   Thu Nov 29 19:35:14 2007 ranaHowToComputer Scripts / ProgramsGPIB Scripts

Quote:
I've spent a lot of time trying to configurate the GPIB-USB interface for the HP4195. After installing 1) the Agilent libraries, 2) the drivers, 3) the matlab Instrument Toolbox, 4) Jamie script, 5) Alice's script the computer can see the HP but still they can't 'talk' to each other.
I give up. I asked Alice Wang how she managed to get data. I'm not sure she used the GPIB interace. Rob said she might have used the old fashion floppy disks that we can't read anymore here.
I would really appreciate any suggestion by anyone who happened to have the same problems.


Alice and Jamie used the USB-GPIB interface. You should just try using the black laptop which already has this capability or ask Jamie Rollins
who actually knows something.
  147   Fri Nov 30 19:11:05 2007 ranaConfigurationElectronicsETMX oplev dead again
I removed the ETMX HeNe and put in on a test table and it fired up fine. In its
previous location the light on the HeNe power supply was not lighting up. If
that's still on over the weekend we'l blame the power strip; the HeNe is a JDS
2.7 mW laser from 2002.
  148   Fri Nov 30 19:29:14 2007 ranaConfigurationSUSnew screen
Andrey is working on a new screen to show us the drift of the optics by alarming on
their osem values. You can find it under SUS as 'Drift Mon' from the site map.

To aid in this I ran the following csh commands which effect all optics:
foreach opt (ETMX ETMY ITMX ITMY MC1 MC2 MC3 BS PRM SRM)
  foreach dof (POS PIT YAW)
     ezcawrite C1:SUS-${opt}_SUS${dof}_INMON.PREC 0
  end
end

This should make the DOF readouts more readable.
  149   Fri Nov 30 19:46:58 2007 ranaConfigurationComputersEPICS Time Bad again
The time on the EPICS screens is off by 10 minutes again. Por Que?

Its because the ntpd on scipe25 wasn't restarted after the last boot. If someone
knows how to put the ntpd startup into that machine, please do so.

This time I started it up by just going sshing in as controls and then entering:

sudo /usr/sbin/ntpd -c /etc/ntp.conf

which runs it as root and points to the right file.

It takes a few minutes to get going because all of the martian machines have to first fail to
connect to the worldwide pool servers (e.g. 0.pool.ntp.org) before they move on and try linux1
which has a connection to the world. Once it gets it you'll see the time on the EPICS screens
freeze. It then waits until the ntp time catches up with its old, wrong time before updating
again.

According to the Wikipedia, this time is then good to 128 ms or less.
  152   Fri Nov 30 21:27:24 2007 ranaDAQPEMweather / stacis / c1pem1
I was trying to add some Seis BLRMS channels to the c1pem1 processor so that we could have DMT trends.

Then I found that none of the Weather channels have been working for a year or so. I could also not
telnet into it. I tried resetting it but no luck. There was no entry in the Wiki for it so I added
a place holder.

Have the weather channels ever worked? Do we have those sensors? I think I've never actually looked
for this. Seems like a fine ugrad job.
  153   Sun Dec 2 17:37:33 2007 ranaOmnistructureComputersNetwork Cabling in the Office
We all know that we've spent many integrated man hours trying to figure out why our network connections
in the office area don't work. Usually its because of the bad hub around the Tobin/Osamu desk.

I pried open some of the wall conduit today and it looks pretty easy to fish cables through. I think
its time we finally did that. It may be a little disruptive, but I propose we get Larry to come over
and figure out what needs to happen for us to get regular 100 Mbit ports on the walls. These can
then all go over and get connected to a switch in the rack that holds linux1.

Opinions / comments ?
  154   Sun Dec 2 21:02:12 2007 ranaConfigurationIOOMC SUS re-alignment
The spot on MC2 was not centered, so I put it back in the center:

  • Made sure MC trans was high with the WFS off.
  • Moved the Sliders on the MC Align screen until spot was centered (by eye)
  • Moved some more until power was maximized.
  • Unlock MC
  • Center spots on McWFS
  • Re-enable autolocker and McWFS loops.