  10488   Wed Sep 10 14:58:58 2014 JenneUpdateSUSOpLev test: New channels

Steve and EricG are moving their oplev test for aLIGO over to the SP table, so that we can have the SRM optical lever back.  

I have pulled out an Ontrak PSM2-10 position sensor and accompanying driver for the sensor.  This, like the POP QPD, has BNC outputs that we can take straight to the ADC.

In the c1pem model I have created 3 new filter modules:  C1:PEM-OLTEST_X, C1:PEM-OLTEST_Y, and C1:PEM-OLTEST_SUM.  I built, installed and restarted the model, and also restarted the daqd process on the frame builder.  On the AA breakout board on the 1X7 rack, these correspond to:

BNC # 29 = OLTEST_X

BNC # 30 = OLTEST_Y

BNC # 31 = OLTEST_SUM

By putting 1Vpp, 0.1Hz into each of these channels one at a time, I see on StripTool that they correspond as I expect.

Everything should be plug-and-play at this point, as soon as Steve is ready with the hardware.

  10490   Wed Sep 10 20:24:00 2014 JenneUpdateElectronicsPOY RF cable loose

Sitting down to work on the IFO, I couldn't lock the Yarm.  I looked at the error signal as well as the transmission on Dataviewer, as usual, and saw that the POY error signal was almost non-existant. 

Since there was work on the POY table today (Steve removed the oplev test setup, elog 10489 and Q centered the SRM oplev after doing SRMI alignment, no elog yet), I went out to have a look at the table. 

There was nothing occluding the POY beam, which I traced back to the edge of the table.  The beam looked nice and round, so I decided that wasn't it.  I jiggled the PD cables, and lo and behold, the POY RF out cable almost came off in my hand it was so loose.  My suspicion is that whomever was the last to put the POY RF out back didn't tighten the cable and then the work today jiggled the cable loose.  I tightened the cable, and by the time I was back to the control room the arm was locked and Koji was already running the alignment scripts.

  10491   Wed Sep 10 21:05:43 2014 JenneUpdateLSCHoly sensitivity, Batman!

Koji and Manasa did some work on the PSL green situation today (Koji is still writing that log post up), but I just measured the Yarm out of loop sensitivity, and WOAH. 

The beat is -11.5dBm at 42.8 MHz.  Koji said the sweet spot is around 30 MHz.  The out of loop sensitivity is 400 Hz RMS!  Something to note is that the Y beatnote still has a 20dB amplifier before going to the beatbox, but the X does not.  We had been worried about saturation issues with the X, so we took out the amplifier.  However, I might put it back if we win big like this.

Recall from elog 10462 that I had saved a reference of the out of loop noise for both X and Y, but Y was much noisier than X.  The references below are from that elog, and the new Y is in dark blue. (Edit, 9:18pm, updated plot measuring down to 0.01Hz.  This is the new reference on the ALS_outOfLoop_Ref.xml template).

 ALS_Y_outOfLoop_10Sept2014.pdf

EDIT:  (Don't worry, I'm going to measure X too, but right now the beam overlap on the camera is not good, as if something drifted after Koji and Manasa closed up the PSL table)

Touched up the alignment for X on the PSL table.  Current beatnotes are:  [Y, -13.5 dBm, 74.1 MHz], [X, -22 dBm, 13.9 MHz].  Red is the current X out of loop, and I've saved it as the new X reference on the template.

ALS_XY_outOfLoop_10Sept2014.pdf

  10493   Thu Sep 11 00:27:39 2014 JenneUpdateSUSSRM oplev touch-up

The sum vs. pitch and yaw signals for the SRM QPD weren't making sense to me - centering on the PD lowered the sum, etc.  So, I had a look at the SRM oplev setup. 

The beam going in to the chamber looked fine, but the beam coming out was weird, like it was being clipped, or diffracted off of a sharp edge.  The beam was spread out in yaw over almost 1cm as seen by eye.  I looked into the vacuum window, and the beam was sitting on the edge of one of the in-vac steering optics.  So, I adjusted the yaw of the beam-launching optic on the out of vac table so that I was roughly centered on both of the in-vac SRM steering mirrors.  This required moving the first out of vac mirror for the SRM oplev path on the way to the QPD to move a small amount to one side, since the beam was near-ish the edge of the optic.  I then centered the beam on the oplev (I had the SRM roughly aligned already). 

Now the SRM oplev makes more sense to me.  I have turned on FMs 1, 2, 5, 9 to match ITMY's loop shape.  I have set the gains to -10 for pitch and +10 for yaw, to make the upper UGF about 6 Hz.

  10494   Thu Sep 11 02:08:32 2014 JenneUpdateLSCHigher transmission powers

No breakthroughs tonight. 

DRMI didn't want to lock with either the recipe that we used a year ago (elog 9116) or that was used in May (elog 9968).  Being lazy and sleepy, I chickened out and went back to PRFPMI locking. 

Many attempts, I'll highlight 2 here.

(1) I had done the CARM -> sqrtInvTrans transition, and reduced the CARM offset to arm powers of about 7, and lost lock.  I don't remember now if I was trying to transition DARM to AS55, or if I was just prepping (measuring error signal ratio and relative sign).

Zoom_TRXTRYat7_CARMsqrtInvTrans_DARMalsdiff.png

(2) I stopped the carm_cm_up script just before it wanted to do the CARM -> sqrtInvTrans transition, and stayed with CARM and DARM both on ALS.  I got to reasonably high powers, and was measuring the error signal ratios I needed for CARM -> REFL DC and DARM -> AS55.  Things were too noisy to get good coherence for the DARM coefficient, but I thought I was in good shape to transition CARM to REFL DC (which looks like REFL11I, since REFLDC goes to the CM board, and the OUT2 of that board is used to monitor the input to the board. )  Anyhow, I set the offset such that it matched my current CARM offset value, and started the transition, but lost lock about halfway through.  CARM started ringing up here, and I think that's what caused this lockloss.  Could have been the CARM peak, which I wasn't considering / remembering at the time.

ALSonly_TryingTransitionStraightToREFLDC.png

Daytime activity for Thurs:  Lock DRMI, maybe first on 1f signals, but then also on 3f signals.

  10498   Fri Sep 12 00:40:23 2014 JenneUpdateLSCDRMI locking

 Tonight I worked on DRMI locking.  

I think the reason the May2014 DRMI recipe wasn't working for me is because I wasn't including the REFL11 -> SRCL element.  I had left it out because (a) I didn't think we should need it and (b) REFL11 is going through the CM board.  

Tonight, I flipped the switch on the CM screen so that OUT2 was seeing REFL11I, not REFLDC, so I had REFL11 in the usual place.  I reset the demod phase, since we had left it at zero for CM stuff. 

Setting demod phases for PRMI:

I locked PRMI on sideband, REFL 33 I&Q and drove PRM.  REFL55 was at 55deg, and I changed it to 33deg to minimize the peak in the Q-phase.  REFL11 was a 0deg, and I set it to 17deg.  I also checked the AS55 phase in the MICH-only case, and changed it from 14.75deg to 24.75 deg.

The May 2014 recipe (elog 9968) calls for adding 25 degrees to the REFL55 phase, so I put REFL55 at 58deg for DRMI locking.

After that, using the parameters in the May2014 recipe, the DRMI just locked.  Awesome!

I checked the demod phases with DRMI lock.  REFL11 stays at 17 degrees.  If I actuate the SRM, I get the largest peak in the I-phase of REFL55 with a phase of -143deg, but the acquisition is best with phase around 55deg.  [Note, as Q points out, I wonder if SRCL is mostly locked with REFL11I for some magical reason, which is why it didn't matter so much that I put a sign flip into REFL55...I wonder if fixing our macroscopic length offset in SRCL will fix this].  I also changed the REFL165 phase from -155.5deg to +145deg.

By looking at transfer functions at an excitation frequency, I expected that I should be able to hold SRCL and MICH on REFL165, with matrix elements -0.085 for REFL165I->SRCL and -0.23 for REFL165Q->MICH.  I was not able to acquire with these values, nor was I able to ramp the matrix elements while keeping lock.

So, I tried moving PRCL to REFL33I, which did work.  I used 1.245*REFL33I->PRCL, but left SRCL and MICH on REFL55 I&Q, with the REFL11I->SRCL element also there. This is where I started trying to get rid of the REFL11I element, but couldn't maintain lock most times, and could never acquire lock without it.

Next up, checking the MICH->SRCL coupling due to the output matrix.  I did as Koji did in elog 8816 , but first I copied the notches in FM10 of MICH over to PRCL and SRCL (old notch freqs were SRCL=566.1Hz, PRCL=675.1Hz, now they're all 475.1Hz).  I drove BS, and checked that the PRM element minimized the peak in REFL33I, the PRCL error signal.  I also added an SRM element to reduce the peak in REFL55I, the SRCL error signal.  I ended up with 0.5*BS, -0.284*PRM, -1.5*SRM for MICH drive, and unity in the PRM and SRM elements for PRCL and SRCL, respectively.

I measured the SRCL open loop gain, and the UGF was pretty low, so I increased the SRCL gain from 0.2 to 0.5 to make the UGF be around 70Hz.  I measured PRCL and MICH also, and they matched their references.

I worked a little bit on trying to remove REFL11 from the SRCL error signal, but didn't get anywhere.  I'm leaving the IFO to Q for the rest of the night.

To sum up, here is the set of parameters that worked for DRMI locking.  (These are saved as the template on the IFO Config screen.):

DEMOD PHASES:

REFL11:  17 deg

REFL33: 140.5 deg (not changed tonight)

REFL55: 58 deg  (58deg for DRMI, 33deg for PRMI)

REFL165: 145 deg

AS55: 24.75 deg

INPUT MATRIX

MICH = 0.15 * REFL55Q

PRCL = 1.245 * REFL33I

SRCL = -0.09 * REFL11I + 1.0 * REFL55I

DOF Triggers

MICH, PRCL, SRCL all on POP22I, 50:10

GAINS

MICH = 1.0

PRCL = -0.02

SRCL = 0.5

FM triggers 

MICH:  35:2, 2 sec delay, FM 2, 3, 6, 9

PRCL:  35:2, 0.5 sec delay, FM 2, 3, 6, 9

SRCL:  35:2, 5 sec delay, FM 3, 6, 9  (always lose lock trying to engage FM2).

OUTPUT MATRIX

MICH = 0.5 * BS +  (-0.284)*PRM + (-1.5)*SRM

PRCL = 1*PRM

SRCL = 1*SRM

  10503   Fri Sep 12 15:10:09 2014 JenneUpdateASCMICH ASS

During the Sim meeting today, I added parts to the ASS model so that we can also dither the BS and minimize the power at AS. 

ASS screen has been updated. 

Model changes required a new sender from LSC for ASDC, so both LSC and ASS were compiled, installed and restarted.  Also on LSC, I added AS110 I&Q to DQ channels, since we haven't been recording them in the past.

I have done a quick trial in MICH-only lock, and it seems to work.  Gain of 10 for both Pit and Yaw servos. 

  10512   Wed Sep 17 01:40:55 2014 JenneUpdateLSCCloser to REFL DC? Maybe?

I tried a bunch of times to reduce my CARM offset so I could jump to REFLDC digitally, but I think I'm maybe being a little ambitious with the arm power I'm trying to get to.

I have modified the carm_cm_up script so that it does my new procedure.  Everything is the same through locking the PRCL and MICH on 3f.  Then it reduces the CARM offset to 1.5 nm.  This is where we *used* to transition to sqrtInvTrans.  Now I have it going a bit farther to 0.5 nm, and arm powers of about 1 before doing that transition.  Also, before it transitions it lowers the CARM gain and engages the 1kHz lowpass in FM9.  A gain of about 4 is fine to keep the gain peaking in the CARM loop to only about 10dB, and sets a UGF of 100Hz which is the peak of the phase bubble with the lowpass engaged. 

Once I got to this point (several times tonight), I turned on CARM and DARM oscillations and looked at the transfer functions between (CARM and REFLDC) and (DARM and AS55Q). I have 2 DTT templates setup for this, in /users/Templates/PRFPMI.  These templates assume that you have your new DARM signal (AS55) going to SRCL_IN1 and your new CARM signal (REFLDC, which is actually REFL11I coming through the CM board) going to MC_IN1. 

I'm not sure why I'm losing lock. I don't see anything terribly telling on the time series plots, in particular none of the loops look like they are oscillating.  Here is one of the better examples from this evening:

Zoom_TRXTRYupto25_1094967592.png

Other notes:

* I realigned the Xgreen on the PSL table (again) to maximize the beatnote amplitude.  Y was fine, but X was very poorly overlapped on the camera.

* I put the SR785 back by the LSC rack and plugged it into the CM board for transfer functions.  Didn't take any tonight.

* We have a small wishlist for scripting things:  (1) DRMI restore script should reset REFL11 to "normal" REFL11.  (2) CARM/DARM acquisition restore script should reset REFL11 to REFLDC.  (3) CARM/DARM acquisition restore should also set PRMI parameters (as Q noted last week).

  10513   Wed Sep 17 13:41:00 2014 JenneUpdateSUSNew calibrated channels for test QPD

Steve asked about calibrating the QPD, so I set up some new epics records so that we can have calibrated versions of the QPD output.

The new channels are called C1:ASC-TESTQPD_Y_Calc and C1:ASC-TESTQPD_X_Calc for pitch and yaw, respectively.


Details:

 * I modified /cvs/cds/caltech/target/c1iscaux/QPD.db to add 2 new channels.  Since we are currently plugged into the IPPOS channels, I didn't want to modify the units of IPPOS, which is why I created new channels.  The new channels are just the IPPOS normalized X and Y channels, multiplied by a calibration factor.  Steve has already done a rough calibration for his setup, so I used those numbers (0.15 urad/ct for pitch and 0.25 urad/ct for yaw). 

* Rebooted c1iscaux.  This required adding it to chiara's /etc/hosts file.

* Added the channels to the /opt/rtcds/caltech/c1/chans/daq/C0EDCU.ini file so that the channels would appear in dataviewer.

* Restarted the framebuilder daqd process.


How to modify the calibration:

1) On a control room workstation, cd /cvs/cds/caltech/target/c1iscaux to get to the right folder.  (Note that this is still in the old cvs/cds place, *not* the new opt/rtcds place)

2)  open the epics database file by typing sudo emacs QPD.db.  Since this is a protected file, you need to use the "sudo" command, and will have to type in the usual controls password. 

3)  Find the "records" that have the channel names C1:ASC-TESTQPD_Y_Calc and C1:ASC-TESTQPD_X_Calc by scrolling down.  (Right now they are on lines #550 and #561 of the text file).

4) For each of these 2 records, modify the calibration in the line that says something like field(CALC,"(A*0.25)").  In this example, the current calibration is 0.25 urad/oldCount.  Change the number to the new value.

5) Save the file.  If you followed the procedure in step2 and used the emacs program and you can't use the mouse, do the following:  Hold down the "ctrl" key.  Keeping ctrl pushed down, push the "x" key.  Still keeping ctrl pushed down, push the "s" key. 

6) Close the file.  If you followed the procedure in step2 and used the emacs program and you can't use the mouse, do the following:  Hold down the "ctrl" key.  Keeping ctrl pushed down, push the "x" key.  Still keeping ctrl pushed down, push the "c" key. 

7) Reboot the slow computer called c1iscaux.  You should be able to do this remotely by typing telnet c1iscaux, and then typing reboot.  If that doesn't work, you may have to go into the IFO room and power cycle the crate by turning the key.  This computer is in 1Y3, near the bottom.

8) Check that you can see your channels - you should be finished now!

For steps 3 and 4, here is a screenshot of the lines in the text file:

NewQPDcalibratedChans.png

  10516   Thu Sep 18 02:42:28 2014 JenneUpdateLSCAO path partly engaged

Tonight was a night of trying to engage the AO path.  The idea was to sit at arm powers of a few on sqrtInvTrans for CARM and ALS for DARM, and try to increase the gain for REFLDC->AO path.

No exciting nit-picky details in locking procedure.  Mostly it was just a night of trying many times. 

The biggest thing that Q and I found tonight was that the 2-pin lemo cable connecting the CM board's SERVO OUT to the MC board's IN2 is shitty.  The symptom that led to this investigation was that I could increase the AO path gain arbitrarily, and have no change in the measured analog CM loop transfer function. We checked that the CM board servo out spit out signals that were roughly what we expected based on our ~2kHz excitation.  However, if we look at digitized signals from the MC board, the noise level was very high, with loads of 60Hz lines, and a teensy-tiny signal peak.  We put a small drive directly into the MC board and could see that, so we determined that the cable is bad.  We have unplugged the white 2-pin lemo, and ran a long BNC cable between the 2 boards.  Tomorrow we need to make a new 2-pin lemo cable so that we can have the lower noise differential drive signal.

After putting in the temporary cable, we do see an excitation sent to the CM board showing up after the MC board.  For this monitoring, the MC_L cable to the ADC has been borrowed, so instead of being the OUT1, the regular length signal, MC_L is currently the OUT2 monitor right after the board inputs. 

At some point in the evening, around 1:15am, ETMX started exhibiting the annoying behavior of wandering off sometimes.  I went in and pushed on the SUS cables to the satellite box, and I think it has helped, although I still saw the drift at least once after the cable-squishing.

Other than that, it has just been many trials.

The best was one where I was holding the arm powers around 4, and got the CM board's AO gain to -8 dB and the MC board's IN2 AO gain to -4 dB. I lost lock trying to increase the CM board gain to -7 dB. 

I took several transfer functions, and used Q's nifty "SRmeasure" script to gather data, and Q made a plot to see the progress.

TF progress plots:

0020.pdf

Time series of that lockloss:

Zoom_TRXTRY4_EngagingAO_1095059162.png

I don't know yet if the polarity of the CM board should be plus or minus.  This series was taken with "minus".  But,  since the phase looked opposite of Q's single arm CM board checkout from several months ago, we did a few trials with the polarity switched to "plus".  I thought we weren't getting as high of AO path gains, so I switched back to "minus", but the last few trials didn't get even as far as the plus trials did.  So, I still don't know which sign we want.

  10519   Thu Sep 18 17:44:55 2014 JenneUpdateLSCOld AO cable pulled

[Q, Jenne]

We pulled the old 2-pin lemo cable after I had a look at the connectors.  When I unscrewed the connector on the MC side, one of the wires came off.  I suspect that it was still hanging on a bit, but my torquing it finally killed it. 

We pulled the cable with the idea of resoldering the connectors, but there are at least 2 places where the cable has been squished enough that the shielding or the inner wires are exposed.  These places aren't near enough the ends to just cut the cable short.

Downs doesn't have a spool of shielded twisted single-pair cable, so Todd is going to get me the part number for the cable they use, and I've asked Steve to order it tomorrow. 

For now, we will continue using the BNC cable that we installed last night - I don't think it's worth resoldering and putting in a crappy 2-pin lemo cable that we'll just throw out in a week.

  10521   Fri Sep 19 13:12:07 2014 JenneUpdateLSCAO path glitches

Quote:

Discontinuities / glitches could be seen in the CM board fast output when MC board gains were changed, which isn't so nice. Incidentally, I notice now that each lock loss corresponded to a step of AO gain on the CM board.

Back in May I looked at all the glitches that happen when we change the AO gain slider on the CM board - see elog 9938.   I wonder if the MC IN2 gain slider has the same issues.  I think I'll look at this this afternoon. Maybe we can set the CM board gain someplace, and just use the MC IN2 slider (if it's not as glitchy) for the delicate part where we're just about to cross unity, and then later we can again use the CM board's AO gain.

EDIT:  Yes, the glitches on the CM board AO path are *much* bigger, and more frequent.  Interestingly, the biggest glitches were every 4 dB.  When I went from -29 to -28, again from -25 to -24, -21 to -20, etc.  I saw the largest glitches on the MC IN2 slider going -29 to -28 and -17 to -16, but if there were small glitches at other transitions, they didn't hit my trigger levels.  I think next time I try engaging the AO path I'll try to do the delicate stuff by upping the MC IN2 gain rather than the CM board AO gain.

  10533   Wed Sep 24 16:02:58 2014 JenneUpdateVACvent is completed

[Steve, EricQ, Jenne]

ITMY and BS heavy doors are off, light doors are on.  Q is aligning the IFO.

  10538   Thu Sep 25 11:33:41 2014 JenneUpdateLSCPOY alignment laser

Quote:

I looked at the CAD layout and it seems like we will clearly be clipping POY if we move SRM by 7.5cm. Since POY is not visible at low power, we cannot be sure about the clipping.

 I was bad and forgot to elog this yesterday (bad grad student!), but I setup a laser pointer to show us where the POY beam is. 

To do this, I removed the tiny mirror that sends the beam to the POY RF PD (so we do not have POY to lock the Yarm right now.  I think Q has successfully been using AS).  The laser pointer goes through 2 temporary steering mirrors, then passes through the place that the tiny mirror usually sits, and then travels along the POY path into the vacuum system.  The idea here is that we should be able to adjust the laser pointer and the temp steering mirrors, and not touch any of the actual POY mirrors, but still get the green beam to go all the way to ITMY.  Yesterday I confirmed that the laser pointer was hitting the in-vac POY pickoff mirror, and today Q and Manasa are doing final adjustment to get the beam all the way to the ITM. 

  10539   Thu Sep 25 11:38:47 2014 JenneUpdatePEMSeismometers in place

[Zach, Jenne, Steve]

This work happened on Tuesday.  Bad Jenne for forgetting to elog it!

Zach brought the 40m's seismometers back (one Guralp and one T-240).  We have set the seismometers on their slabs.  Also, we ran the T240 cable from 1X5 over to the vertex slab.  Also, also, Zach and Steve mounted the T-240 readout box in the 1X5 rack.  We have not yet hooked it up to power, although there are fused power blocks available on that rack. 

So, the T-240 box needs power, and then we need to connect the seismometers to their respective boxes.  Also, we need to run medium-short BNC cables from the T-240 readout box to the PEM AA board over in 1X7.

  10570   Mon Oct 6 11:09:52 2014 JenneUpdateGeneralUnexpected power shutdown: slow computers

As per other slow computers, which Chris figured out in elog 10189, I added all the rest of the slow computers to Chiara's /etc/hosts file, so that they would come up when Manasa went and keyed the crates. 

Computers that were already there:

  • c1auxex
  • c1psl
  • c1iscaux

Computers that I added today:

  • c1susaux
  • c1auxey
  • c1iscaux2
  • c1pem1
  • c1aux
  • c1iool0
  • c1vac1

Manasa keyed all of these crates *except* for the vac computer, since Steve said that the vacuum system is up and running fine.

  10574   Tue Oct 7 00:18:12 2014 JenneUpdateLSCYgreen PSL alignment, ETMX strain relief

No exciting progress today.  I did PSL green alignment for the Yarm, although I now think that the Xarm green needs realigning too.

Also, I was foiled for a while by ETMX jumping around.  I think it's because the adapter board on the Xend rack didn't have any strain relief.  So, I zip tied the heavy cable in a few places so that it's no longer pulling on the connector.  Hopefully we won't see ETMX misbehaving as often now, so we won't have to go squish cables as often.

  10578   Tue Oct 7 16:41:12 2014 JenneUpdateGeneralChiara not responding

I put a little script into ...../scripts/Admin that will check the fullness of Chiara's disk.  We only have the mailx program installed on Nodus, so for now it runs on Nodus and sends and email when the chiara disk that nodus mounts is more than 97% full.

  10581   Wed Oct 8 03:20:46 2014 JenneUpdateLSCDo we need AO for acquisition?

As part of trying to determine whether we require the AO path for lock acquisition, or if we can survive on just digital loops, I looked at the noise suppression that we can get with a digital loop.

I took a spectrum of POX, and calibrated it using a line driving ETMX to match the ALSX_FINE_PHASE_OUT_HZ channel, and then I converted green Hz to meters. 

I then undid the LSC loop that was engaged at the time (XARM FMs 1,2,3,4,5,8 and the pendulum plant), to infer the free running arm motion. 

I also applied the ALS filters (CARM FMs 1,2,3,5,6) and the pendulum plant to the free running noise to infer what we expect we could do with the current digital CARM filters assuming we were not sensor noise limited.

In the figure, we see that the free running arm displacement is inferred to be about 0.4 micrometers RMS.  The in-loop POX signal is 0.4 picometers RMS, which (although it's in-loop, so we're not really that quiet) is already better than 1/10th the coupled cavity linewidth.  Also, the CARM filters that we use for the ALS lock, and also the sqrtInvTrans lock are able to get us down to about 1 pm RMS, although that is not including sensor noise issues. 

EstimatedNoisePerformance.png

For reference, here are the open loop gains for the LSC filters+pendulum and ALS filters+pendulum that we're currently using.  The overall gain of these loops have been set so the UGF is 150Hz.

 BodeLSCvsALS.png

It seems to me that as long as our sensors are good enough, we should be able to keep the arm motion down to less than 1/10th or 1/20th the coupled cavity linewidth with only the digital system.  So, we should think about working on that rather than focusing on engaging the AO path for a while.

Attachment 3: CARMnoise_7Oct2014.zip
  10583   Wed Oct 8 03:49:42 2014 JenneUpdateLSCPRFPMI, other sign of CARM offset

Other thoughts from talking with Rana earlier:

  • Is it possible to suppress CARM motion enough that we can use just a digital loop?  Can we do without the AO path?  What would said digital loop have to look like?
  • Q points out that there is a zero in the relative transfer function between CARM to transmission, and CARM to REFLDC.  Is that zero invertible?
  • We should look at some limits, like saturation limits.  How much will we need to actuate?
  • Rana is looking at making a more detailed CARM loop model in simulink to see if we can stay stable throughout our CARM offset reduction journey.

Also, Q and I squished on the suspension connectors earlier tonight.  MC2 was going wonky, which we feared might be because we were in that area working on Chiara earlier.  Then, after squishing the MC connectors, the PRM started misbehaving, so we went and gave all the corner suspension connectors another squish.  No suspension glitching problems since then.

  10585   Wed Oct 8 15:31:31 2014 JenneUpdateCDSComputer status

After the Great Computer Meltdown of 2014, we forgot about poor c0rga, which is why the RGA hasn't been recording scans for the past several months (as Steve noted in elog 10548).

Q helped me remember how to fix it.  We added 3 lines to its /etc/fstab file, so that it knows to mount from Chiara and not Linux1.  We changed the resolv.conf file, and Q made some simlinks.

Steve and I ran ..../scripts/RGA/RGAset.py on c0rga to setup the RGA's settings after the power outage, and we're checking to make sure that the RGA will run right now, then we'll set it back to the usual daily 4am run via cron.

EDIT, JCD:  Ran ..../scripts/RGA/RGAlogger.py, saw that it works and logs data again.  Also, c0rga had a slightly off time, so I ran sudo ntpdate -b -s -u pool.ntp.org, and that fixed it.

Quote:

 

In all of the fstabs, we're using chiara's IP instead of name, so that if the nameserver part isn't working, we can still get the NFS mounts.

On control room computers, we mount the NFS through /etc/fstab having lines like:

192.168.113.104:/home/cds /cvs/cds nfs rw,bg 0 0
fb:/frames /frames nfs ro,bg 0 0

Then, things like /cvs/cds/foo are locally symlinked to /opt/foo

For the diskless machines, we edited the files in /diskless/root. On FB, /diskless/root/etc/fstab becomes

master:/diskless/root                   /         nfs     sync,hard,intr,rw,nolock,rsize=8192,wsize=8192    0 0
master:/usr                             /usr      nfs     sync,hard,intr,ro,nolock,rsize=8192,wsize=8192    0 0
master:/home                            /home     nfs     sync,hard,intr,rw,nolock,rsize=8192,wsize=8192    0 0
none                                    /proc     proc    defaults          0 0
none                                    /var/log        tmpfs   size=100m,rw    0 0
none                                    /var/lib/init.d tmpfs   size=100m,rw    0 0
none                                    /dev/pts        devpts  rw,nosuid,noexec,relatime,gid=5,mode=620        0 0
none                                    /sys            sysfs   defaults        0 0
master:/opt                             /opt      nfs    async,hard,intr,rw,nolock  0 0
192.168.113.104:/home/cds/rtcds         /opt/rtcds      nfs     nolock  0 0
192.168.113.104:/home/cds/rtapps        /opt/rtapps     nfs     nolock  0 0

("master" is defined in /diskless/root/etc/hosts to be 192.168.113.202, which is fb's IP)

and /diskless/root/etc/resolv.conf becomes:

search martian

nameserver 192.168.113.104 #Chiara

 

 

 

  10588   Thu Oct 9 13:29:14 2014 JenneUpdatePSLPower outage II & recovery

Quote:

 

 IFO vacuum, air condition and PMC HV are still down. PSL out put beam is blocked on the table.

 PMC is fine.  There are sliders in the Phase Shifter screen (accessible from the PMC screen) that also needed touching. 

PSL shutter is still closed until Steve is happy with the vacuum system - I guess we don't want to let high power in, in case we come all the way up to atmosphere and particulates somehow get in and get fried on the mirrors. 

  10591   Thu Oct 9 18:30:59 2014 JenneUpdateLSCCARM W/N TFs

Okay, here (finally) is the optickle version.

I have the antispring case, starting at 501pm and going roughly every 10pm down to 1pm.  I also have the spring case, starting at -501pm and going down every 10pm to roughly -113pm.  Rossa crashed partway through the calculation, which is why it's not all the way.

In the .zip is a .mat file called PDs_vs_CARMoffset_WattsPerNewton.mat, which has (a) a list of the 50 CARM offsets, (b) a frequency vector, and (c) several transfer function arrays.  The transfer function arrays are supposed to be intuitively named, eg. REFLDC_antispring. 

In the .zip file are also the original .mat files that are a result of the tickle calculations, as well as a .m file for loading them and making the plots, etc.  For anyone who is trying to re-create the transfer function variables, I by-hand saved the variable called PD_WperN to the names like REFLDC_antispring.  Just kidding.  Those original mat files are over 100Mb each, and that's just crazy.  Anyhow, I think the .zip has everything needed to use the data from these plots.

Anyhow.  Here are plots of what are in the various transfer function arrays:

 TRX_antispring.pngTRX_spring.png

REFLDC_antispring.pngREFLDC_spring.png

REFL11I_antispring.pngREFL11I_spring.png

Attachment 6: ForElog.zip
  10595   Fri Oct 10 03:25:11 2014 JenneUpdateLSCWhich side of optical spring are we on (simulation)

I have a simulated version of the differences that we expect to see between the 2 different sides of the CARM resonance.  The point is that we can try to compare these results with Q's measured results (elog 10594) to see if we know if we are on the spring or antispring side.


I calculated the same transfer functions vs CARM offset again, although tonight I do it in steps of 20pm because I was getting bored of waiting forever.  Anyhow, this is important because my previous post (elog 10591) didn't have spring side calculations all the way down to 1pm.

This is similarly true for that elog 10591, but here are some notes on how I am currently getting the W/N units out of Optickle.  First of all, I am still using old Optickle1.  I don't know if there are significant units ramifications for that, but just in case I'll write it down.  Nic tells me that to get [W/N] out of Optickle1, I need to multiply sigAC (units of [W/m]) by my simple pendulum (units of [m/N]).  Both of these "meters" in the last sentence are "mevans meters", which are the meters you would get per actuation if radiation pressure didn't exist.  So, I guess they're supposed to cancel out?  I need to camp out in Nic's office until I figure this out and get it untangled in my head.

Plots of transfer functions for both sides of CARM resonance (same as prev. elog), as well as the ratio between the spring and antispring transfer functions at each CARM offset:

TFs_TRX_vsCARMoffset_PRFPMI_antispring.pngTFs_TRX_vsCARMoffset_PRFPMI_spring.pngTFs_TRX_vsCARMoffset_PRFPMI_differentials.png

 

TFs_REFLDC_vsCARMoffset_PRFPMI_antispring.pngTFs_REFLDC_vsCARMoffset_PRFPMI_spring.pngTFs_REFLDC_vsCARMoffset_PRFPMI_differentials.png

TFs_REFL11I_vsCARMoffset_PRFPMI_antispring.pngTFs_REFL11I_vsCARMoffset_PRFPMI_spring.pngTFs_REFL11_vsCARMoffset_PRFPMI_differentials.png

The take-away message from the 3rd column is that other than a sign flip, we don't expect to see very much difference between the 2 sides of the CARM resonance, particularly above a few hundred Hz.  (Note that we do not see the sign flip in Q's measurements because he is looking at CARM_IN1, which is after the input matrix, and the input matrix elements have opposite signs between the signs of the CARM offsets.  So, the sign flip between spring and antispring around the UGF is implied in the measurements, just not explicit).

Also, something that Rana pointed out to me, and I still don't know why it's true:  The antispring transfer functions (at least for the transmission) don't have all the phase features that we expect to see based on their magnitudes.  If you look at the TRX antispring plot, blue trace (which is about 500pm from resonance), you'll see that the magnitude starts flat at DC, has some slope in an intermediate region, and then at high frequencies has 1/f^2.  However, the phase seems to not know about this intermediate region, and magically waits until the 1kHz resonance to flip the full 180 degrees. 

Attachment 10: ForElog_9Oct2014.zip
  10600   Mon Oct 13 16:08:49 2014 JenneUpdateCDSFrame builder is mad

I think the daqd process isn't running on the frame builder. 

Daqd_problem_maybe.png

I tried telnetting' to fb's port 8087 (telnet fb 8087) and typing "shutdown", but so far that is hanging and hasn't returned a prompt to me in the last few minutes.  Also, if I do a "ps -ef | grep daqd" in another terminal, it hangs. 

I wasn't sure if this was an ntp problem (although that has been indicated in the past by 1 red block, not 2 red blocks and a white one), so I did "sudo /etc/init.d/ntp-client restart", but that didn't make any change.  I also did an mxstream restart just in case, but that didn't help either. 

I can ssh to the frame builder, but I can't do another telnet (the first one is still hung).  I get an error "telnet: Unable to connect to remote host: Invalid argument"

Thoughts and suggestions are welcome!

  10603   Mon Oct 13 21:20:56 2014 JenneUpdateLSCWhich side of optical spring are we on? (No progress)

[Jenne, Diego]

In order to distinguish between the spring and antispring sides of the CARM resonance, we need to have transfer function measurements down to at least 100 Hz (although lower is better). 

We tried to get some transfer functions the same way Q did, but noticed that (a) we couldn't get any low frequency coherence, and (b) that when we increased the amplitude of the white (well, lowpass at 5kHz) noise, the coherence between the AO injection and REFL DC went down.  Not clear why this is.

Anyhow, we tried taking good ol' fashioned swept sine transfer functions, although eventually the lightbulb came on that the AO path has a highpass in it.  Duh, Jenne.  So, we started trying to actuate on MC2 position rather than the AO path laser frequency.  We didn't get too far though before El Salvadore decided to have a few 7.4 earthquakes.  We're bored of aftershocks knocking us out of lock, so we're going to come back to this tomorrow.

 

  10607   Wed Oct 15 02:58:03 2014 JenneUpdateLSCWhich side of optical spring are we on?

Some measurements.  Unclear meaning.  

We tried both positive and negative numbers in the CARM offset, and then looked at transfer functions at various arm powers. The hope is to be able to compare these with some simulation to figure out which side of the CARM resonance we are on.

The biggest empirical take-away is that we repeatedly (3 times in a row) lost lock when holding at arm powers of about 5 with negative CARM offsets.  However, we were repeatedly (2+ times tonight) able to sit and hold at arm powers of 10+ with positive CARM offsets.

 

 

 

I am not sure that we get enough information out of these plots to tell us which side of the CARM resonance we are really on.  Q is working on taking some open loop CARM measurements (actuating and measuring at SUS-MC2_LSC) to see if we can compare those more directly to Rana's plots.

Positive number in the digital CARM offset:

PRFPMI_MC2injection_PosCARMoffset_REFLDC_TRX_response_14Oct2014_TRX.pdf

PRFPMI_MC2injection_PosCARMoffset_REFLDC_TRX_response_14Oct2014_REFLDC.pdf

Negative numbers in digital CARM offset:

PRFPMI_MC2injection_NegCARMoffset_REFLDC_TRX_response_14Oct2014_TRX.pdf

PRFPMI_MC2injection_NegCARMoffset_REFLDC_TRX_response_14Oct2014_REFLDC.pdf

  10609   Wed Oct 15 13:38:33 2014 JenneUpdateLSCCARM W/N TFs

Here are the same plots, but the legend also includes the arm power that we expect at that CARM offset.  


Here is what the arm powers look like as a function of CARM offset according to Optickle.  Note that the cyan trace's maximum matches what Q has simulated in Mist with the same high losses.  For illustration I've plotted the single arm power, so that you can see it's normalized to 1.  Then, the other traces are the full PRFPMI buildup, with various amounts of arm loss.  The "no loss" case is with 0ppm loss per ETM.  The "150 ppm loss" case is with 150 ppm of loss per ETM.  The "high loss" case is representative of what Q has measured, so I have put 500 ppm loss for ETMX and 150 ppm loss for ETMY.

ArmPower_vs_Loss.png


And, the transfer functions (all these, as with all TFs in the last week, use the "high loss" situation with 500ppm for ETMX and 150ppm for ETMY).

TFs_TRX_vsCARMoffset_PRFPMI_antispring.pngTFs_TRX_vsCARMoffset_PRFPMI_spring.png

TFs_REFLDC_vsCARMoffset_PRFPMI_antispring.pngTFs_REFLDC_vsCARMoffset_PRFPMI_spring.png

TFs_REFL11I_vsCARMoffset_PRFPMI_antispring.pngTFs_REFL11I_vsCARMoffset_PRFPMI_spring.png

  10611   Wed Oct 15 17:18:10 2014 JenneUpdateComputer Scripts / ProgramsDataviewer fix with Ubuntu 12.04

 

 I have modified the Dataviewer launcher (which runs when you either click the icon or type "dataviewer" in the terminal).  

A semi-old problem was that it would open in the file /users/Templates, but our dataviewer templates start in /users/Templates/Dataviewer_Templates.  Now this is the folder that dataviewer opens into.  This was not related to the upgrade to Ubuntu 12, but will be overwritten any time someone does a checkout of the /ligo/apps/launchers folder.

A problem that is related to the Ubuntu 12 situation, which we had been seeing on Ottavia and Pianosa for a few weeks, was that the variable NDSSERVER was set to fb:8088, which is required for cdsutils to work.  However, dataviewer wants this variable to be set to just fb.  So, locally in the dataviewer launcher script, I set NDSSERVER=fb.  NB: I do not export this variable, because I don't want to screw up the cdsutils.  This may need to be undone if we ever upgrade our Dataviewer.

  10612   Wed Oct 15 19:56:38 2014 JenneUpdateLSCWhich side of optical spring are we on? Meas vs Model

 I have plotted measured data from last night (elog 10607) with a version of the result from Rana's simulink CARM loop model (elog 10593).

The measured data that was taken last night (open circles in plots) is with an injection into MC2 position, and I'm reading out TRX.  This is for the negative side of the digital CARM offset, which is the one that we can only get to arm powers of 5ish.

The modeled data (solid lines in plots) is derived from what Rana has been plotting the last few days, but it's not quite identical.  I added another excitation point to the simulink model at the same place as the "CARM OUT" measurement point.  This is to match the fact that the measured transfer functions were taken by driving MC2.  I then asked matlab to give me the transfer function between this new excitation point (CARM CTRL point) and the IN1 point of the loop, which should be equivalent to our TRX_OUT.  So, I believe that what I'm plotting is equivalent to TRX/MC2.  The difference between the 2 plots is just that one uses the modeled spring-side optical response, and the other uses the modeled antispring-side response.

AntispringModel_NegOffsetMeas_comparison.pdf

SpringModel_NegOffsetMeas_comparison.pdf

I have zoomed the X-axis of these plots to be between 30 Hz - 3 kHz, which is the range that we had coherence of better than 0.8ish last night in the measurements.  The modeled data is all given the same scale factor (even between plots), and is set so that the lowest arm power traces (pink) line up around 150 Hz. 

I conclude from these plots that we still don't know what side of the CARM resonance we are on. 

 I have not plotted the measurements from the positive side of the digital CARM offset, because those transfer functions were to sqrtInvTRX, not plain TRX, whereas the model only is for plain TRX. There should only be an overall gain difference between them though, no phase difference.  If you look at last night's data, you'll see that the positive side of the CARM offset measured phase has similar characteristics to the negative offset, i.e. the phase is not flat, but it is roughly flat in both modeled cases, so even with that data, I still say that we don't know what side of the CARM resonance we are on.

 

 

  10614   Wed Oct 15 22:39:17 2014 JenneUpdateLSCThe Plan

 [Rana, Jenne]

We're summarizing the discussions of the last few days as to the game plan for locking.  

  1. PRMI on REFL165.  The factor of 5 in frequency will give us more MICH signal.    We want this.
    1. Drive CARM, measure coupling to PRCL, MICH while locked on REFL33.
    2. Switch to REFL165, re-measure CARM coupling.
    3. Hopefully this will reduce the AS port fluctuations, and reduce the POP22 power decrease as CARM offset decreases. 
  2. DARM transition from ALSdiff to an intermediate signal.  Simulate, and try empirically.
    1. Maybe try ASDC normalized by sum of transmissions?
    2. Maybe try difference of transmissions divided by sum of transmissions?  
  3. Look at data on disk.
    1. Do we have anything specific causing our locklosses (lately there haven't been obvious loop instabilities causing the locklosses)?
    2. How much do we think our lengths are actually changing right now (particularly DARM on ALSdiff)?
    3. Are there better ways of combining error signals that could be useful?
    4. Do we need to work on angular loops?
      1. Oplevs
      2. POP ASC for sidebands
      3. POP QPD or Trans QPDs for arms
  4.  Think about what could be causing ETMX to be annoying.  The connection that is most suspect has been ziptied, but we're still seeing ETMX move either at locklosses or sometimes just spontaneously.
  5.  RAM.  What kind of RAM levels do we have right now, and how do they affect our locking offsets?  Do we have big offsets, or negligible offsets?
  10615   Thu Oct 16 03:13:23 2014 JenneUpdateLSCPRMI on REFL165, and more

 The first thing I looked at tonight was locking the PRMI on REFL 165.

I locked the PRMI (no arms), and checked the REFL 165 demod phase. I also found the input matrix configuration that allowed me to acquire PRMI lock directly on REFL165.  After locking the arms on ALS, I tried to lock the PRMI with REFL 165 and failed.  So, I rechecked the demod phase and the relative transfer functions between REFL 165 and REFL 33.  The end of the story is that, even with the re-tuned demod phase for CARM offset of a few nanometers, I cannot acquire PRMI lock on REFL 165, nor can I transition from REFL 33 to REFL 165.  We need to revisit this tomorrow.

IFO configuration CARM offset [cts] REFL 165 demod phase [deg]
Found as-is N/A +145
PRMI, no arms N/A -135
PRFPMI +3 +110
PRFPMI +2 +110
PRFPMI +1 +110
PRFPMI +0.5 +120

 

IFO configuration REFL 33 I / REFL 165 I (PRCL) REFL 33 Q / REFL 165 Q (MICH)
PRMI, no arms +0.1 +0.22, although easier to acquire lock with +0.1
PRFPMI, CARM offset = +3 -0.09  (TF measured, no lock) +0.033  (TF measured, no lock)

For the PRMI-only case, I ended up using 0.1's in the input matrix, and I added an FM 1 to the MICH filter bank that is a flat gain of 2.2, and then I had it trigger along with FM2.

I turned this FM1 off (and no triggering) while trying to transition from REFL33 to REFL165 in the PRFPMI case, but that didn't help.  I think that maybe I need to remeasure my transfer functions or something, because I could put values into the REFL165 columns of the input matrix while REFL33 was still 1's, but I couldn't remove (even if done slowly) the REFL33 matrix elements without losing lock of the PRMI.  So, we need to get the input matrix elements correct.


I also recorded some time series for a quick RAM investigation that I will work on tomorrow.  

I left the PRM aligned, but significantly misaligned both ITMs to get data at the REFL port of the RAM that we see.  I also aligned the PRMI (no arms) and let it flash so that I can see the pk-pk size of our PDH signals.  I need to remember to calibrate these from counts to meters.  

Raw data is in /users/jenne/RAM/ .


I have not tried any new DARM signals, since PRMI wasn't working with 3f2.  

We should get to that as soon as we fix the PRMI-3f2 situation.

  10616   Thu Oct 16 03:18:48 2014 JenneUpdateCDSDaqd segfaulting again

 The daqd process on the frame builder looks like it is segfaulting again.  It restarts itself every few minutes.  

The symptoms remind me of elog 9530, but /frames is only 93% full, so the cause must be different.  

Did anyone do anything to the fb today?  If you did, please post an elog to help point us in a direction for diagnostics.

Q!!!!  Can you please help?  I looked at the log files, but they are kind of mysterious to me - I can't really tell the difference between a current (bad) log file and an old (presumably fine) log file.  (I looked at 3 or 4 random, old log files, and they're all different in some ways, so I don't know which errors and warnings are real, and which are to be ignored).

  10622   Fri Oct 17 13:19:48 2014 JenneUpdateLSCPOP22 ?!?!

We've seen this before, but we need to figure out why POP22 decreases with decreased CARM offset.  If it's just a demod phase issue, we can perhaps track this by changing the demod phase as we go, but if we are actually losing control of the PRMI, that is something that we need to look into.

In other news, nice work Q!

Quote:

TRdifflock.png

 

  10625   Fri Oct 17 17:52:55 2014 JenneUpdateLSCRAM offsets

Last night I measured our RAM offsets and looked at how those affect the PRMI situation.  It seems like the RAM is not creating significant offsets that we need to worry about.


Words here about data gathering, calibration and calculations.

Step 1:  Lock PRMI on sideband, drive PRM at 675.13Hz with 100 counts (675Hz notches on in both MICH and PRCL).  Find peak heights for I-phases in DTT to get calibration number.

Step 2:  Same lock, drive ITMs differentially at 675.13Hz with 2,000 counts.  find peak heights for Q-phases in DTT to get calibration number.

Step 3:  Look up actuator calibrations.  PRM = 19.6e-9/f^2 meters/count and ITMs = 4.68e-9/f^2 meters/count.  So, I was driving PRM about 4pm, and the ITMs about 20pm.

Step 4:  Unlock PRMI, allow flashes, collect time series data of REFL RF siganls.

Step 5: Significantly misalign ITMs, collect RAM offset time series data.

Step 6: Close PSL shutter, collect dark offset time series data.

Step 7: Apply calibration to each PD time series.  For each I-phase of PDs, calibration is (PRM actuator / peak height from step 1).  For each Q-phase of PDs, calibration is (ITM actuator / peak height from step 2).

Step 8:  Look at DC difference between RAM offset and dark offset of each PD.  This is the first 4 rows of data in the summary table below.

Step 9:  Look at what peak-to-peak values of signals mean.  For PRCL, I used the largest pk-pk values in the plots below.  For MICH I used a calculation of what a half of a fringe is - bright to dark.  (Whole fringe distance) = (lambda/2), so I estimate that a half fringe is (lambda/4), which is 266nm for IR.  This is the next 4 rows of data in the table.

Step 10: Divide.  This ratio (RAM offset / pk-pk value) is my estimate of how important the RAM offset is to each length degree of freedom. 


Summary table:

  PRCL (I-phase) MICH (Q-phase)
RAM offsets    
11 4e-11 m 2.1e-9 m
33 1.5e-11 m ~2e-9 m
55 2.2e-11 m ~1e-9 m
165 ~1e-11 m ~1e-9 m
Pk-pk (PDH or fringes) PDH pk-pk from flashes MICH fringes from calculation
11 5.5e-9 m 266e-9 m
33 6.9e-9 m 266e-9 m
55 2.5e-9 m 266e-9 m
165 5.8e-9 m 266e-9 m
Ratio: (RAM offset) / (pk-pk)    
11 7e-3 8e-4
33 2e-3 7e-3
55 9e-3 4e-3
165 2e-3 4e-3

 


Plots (Left side is several PRMI flashes, right side is a zoom to see the RAM offset more clearly):

RAM_PRMI_PRMIflashing_REFL11I.pdfRAM_PRMI_Zoom_REFL11I.pdf

RAM_PRMI_PRMIflashing_REFL11Q.pdfRAM_PRMI_Zoom_REFL11Q.pdf

RAM_PRMI_PRMIflashing_REFL33I.pdfRAM_PRMI_Zoom_REFL33I.pdf

RAM_PRMI_PRMIflashing_REFL33Q.pdfRAM_PRMI_Zoom_REFL33Q.pdf

RAM_PRMI_PRMIflashing_REFL55I.pdfRAM_PRMI_Zoom_REFL55I.pdf

RAM_PRMI_PRMIflashing_REFL55Q.pdfRAM_PRMI_Zoom_REFL55Q.pdf

RAM_PRMI_PRMIflashing_REFL165I.pdfRAM_PRMI_Zoom_REFL165I.pdf

RAM_PRMI_PRMIflashing_REFL165Q.pdfRAM_PRMI_Zoom_REFL165Q.pdf

  10626   Mon Oct 20 17:50:30 2014 JenneUpdateLSCCARM W/N TFs (Others were all wrong!)

I realized today that I had been plotting the wrong thing for all of my transfer functions for the last few weeks! 

The "CARM offsets" were correct, in that I was moving both ETMs, so all of the calculations were correct (which is good, since those took forever). But, in the plots I was just plotting the transfer function between driving ETMX and the given photodiode.  But, since just driving a single ETM is an admixture of CARM and DARM, the plots don't make any sense.  Ooops.

In these revised plots (and the .mat file attached to this elog), for each PD I extract from sigAC the transfer function between driving ETMX and the photodiode.  I also extract the TF between driving ETMY and the PD.  I then  sum those two transfer functions and divide by 2.  I multiply by the simple pendulum, which is my actuator transfer function to get to W/N, and plot.

The antispring plots don't change in shape, but the spring side plots do.  I think that this means that Rana's plots from last week are still true, so we can use the antispring side of TRX to get down to about 100 pm.

Here are the revised plots:

TFs_TRX_vsCARMoffset_PRFPMI_antispring.pngTFs_TRX_vsCARMoffset_PRFPMI_spring.png

TFs_REFLDC_vsCARMoffset_PRFPMI_antispring.pngTFs_REFLDC_vsCARMoffset_PRFPMI_spring.png

TFs_REFL11I_vsCARMoffset_PRFPMI_antispring.pngTFs_REFL11I_vsCARMoffset_PRFPMI_spring.png

Attachment 1: PDs_vsCARMoffset_20Oct2014.mat.zip
  10630   Wed Oct 22 02:35:45 2014 JenneUpdateLSCEfforts at hopping PRMI to REFL165

[EricQ, Jenne]

The first half of our evening was spent working on CARM and DARM in PRFPMI, and then we moved on to the PRMI part.

I moved the DARM ALSdiff -> TransDiff transition to be after the CARM ALScomm -> SqrtInvTrans transition in the carm_cm_up script.  After I did that, I succeeded every time (at least  10?  We did it many times) to get both CARM and DARM off of the ALS signals. 

We tried for a little while looking at transitioning to REFL11 normalized by the sum of the transmissions, but we kept losing lock.  We also several times lost lock at arm powers of a few, when we thought we weren't touching the IFO for any transitions.  Looking at the lockloss time series did not show any obvious oscillations in any of the _IN1 or _OUT channels for the length degrees of freedom, so we don't know why we lost lock, but it doesn't seem to be loop oscillations caused by changing optical gain.  Also, one time, I tried engaging Rana's "Lead 350" filter in FM7 of the CARM filter bank when we were on sqrtInvTrans for CARM, and the arm powers were around a few, but that caused the transmission signals to start to oscillate, and after one or two seconds we lost lock.  We haven't tried the phase lead filter again, nor have we tried the Boost2 that is in FM8. 

We increased the REFL11 analog gain from 0dB to 12dB, and then reset the dark offsets, but still weren't able to move CARM to normalized REFL11. Also, I changed the POP22 demod phase from 159 degrees to 139 degrees. This seems to be where the signal is maximized in the I-phase, while the arms are held off resonance, and also partway up the resonance peak. 

We then decided that we should go back to the PRMI situation before trying to reduce the CARM offset further.  We can robustly and quickly lock the PRMI on REFL33 while the arms are held off resonance with ALS.  So, we have been trying to acquire on  REFL33 I&Q, and then look at switching to REFL 165 I&Q.  It seems pretty easy to get PRCL over to REFL165 I (while leaving MICH on REFL33 I).  For REFL33, both matrix elements are +1.  For PRCL on REFL165, the matrix element is -0.08.  We have not successfully gotten MICH over to REFL 165 ever this evening. 

We went back and set the REFL165 I&Q offsets so that the outputs after the demod phase were both fluctuating around 0.  I don't know if they were around +/-100 because our dark offsets were bad or what, but we thought this would help.  We were still able to get PRCL transitioned no problem, but even after remeasuring the MICH REFL33 vs. REFL165 relative gains, we still can't transition MICH.  It seems like it's failing when the REFL33Q matrix element finally gets zeroed out, so we're not really getting enough signal in REFL165Q, or something like that, and throughout the rest of the transition we were depending entirely on REFL33Q. 

So. Plan:

  • Get PRMI on REFL165 while arms are held off resonance. 
    • May require PRCL-MICH FF decoupling, by combining error signals?
    • May require looking back at simulations to see what we expect the relative gains and signs to be.
  • Look at CARM loop stability in simulation for REFLDC, REFL11, and normalized REFL11.  Is there a stable loop path from about 100pm down to 0pm on normalized REFL11?
  10633   Thu Oct 23 01:39:34 2014 JenneUpdateCDSDaqd "fixed"?

Merging of threads. 

ChrisW figured out that it looks like the problem with the frame builder is that it's having to wait for disk access.  He has tweaked some things, and life has been soooo much better for Q and I this evening!  See Chris' elog at elog 10632.

In the last few hours we've had 2 or maybe 3 times that I've had to reconnect Dataviewer to the framebuilder, which is a significant improvement over having to do it every few minutes.

Also, Rossa is having trouble with DTT today, starting sometime around dinnertime.  Ottavia and Pianosa can do DTT things, but Rossa keeps getting "test timed out". 

  10634   Thu Oct 23 02:08:40 2014 JenneUpdateLSCIncreased DARM gain

I changed the carm_cm_up.sh script so that it requires fewer human interventions.  Rather than stopping and asking for things like "Press enter to confirm PRMI is locked", it checks for itself.  The sequence that we have in the up script works very reliably, so we don't need to babysit the first several steps anymore. 

Another innovation tonight that Q helped put in was servoing the CARM offset to get a certain arm power.  A failing of the script had been that depending on what the arm power was during transition over to sqrtInvTrans, the arm power was always different even if the digital offset value was the same.  So, now the script will servo (slowly!!) the offset such that the arm power goes to a preset value.

The biggest real IFO progress tonight was that I was able to actually measure the CARM and DARM loops (thanks ChrisW!), and so I discovered that even though we are using (TRX-TRY)/(TRX+TRY) for our IR DARM error signal, we needed to increase the digital gain for DARM as the CARM offset was reduced.  For ALS lock and DC trans diff up to arm powers of 3, we use the same ol' gain of 6.  However, between 3 - 6, we need a gain of 7.  Then, when we go to arm powers above 6 we need a gain of 7.5.  I was also measuring the CARM loop at each of these arm powers (4, 6, 7, 8, 9), but the gain of 4 that we use for sqrtInvTrans was still fine. 

So, the carm_cm_up script will do everything that it used to without any help (unless it fails to find IR resonance for ALS, or can't lock the PRMI, in which case it will ask for help), and then once it gets to these servo lines to slowly increase the arm power and increase the DARM gain, it will ask you to confirm before each step is taken.  The script should get you all the way to arm powers of 9, which is pretty much exactly 100pm according to Q's Mist plot that is posted.

The CARM and DARM loops (around the UGFs) don't seem to be appreciably changing shape as I increase the arm powers up to 9 (as long as I increase the DARM loop gain appropriately).  So, we may be able to go a little bit farther, but since we're at about 100pm, it might be time to look at whether we think REFL11 or REFLDC is going to be more promising in terms of loop stability for the rest of the way to resonance.


Here are some plots from this evening. 

First, the time I was able to get to and hold at arm powers of 9.  I have a striptool to show the long time trends, and then zooms of the lockloss.  I do not see any particular oscillations or anything that strikes me as the cause for the lockloss.  If anyone sees something, that would be helpful.

ArmPowers9_1am_22Oct2014.png

Lockloss_ArmsAt9_1am_22Oct2014.png

LocklossZoom_ArmsAt9_1am_22Oct2014.png

This next lockloss was interesting because the DARM started oscillating as soon as the normalization matrix elements were turned on for DARM on DC transmissions.  The script should be measuring values and putting in matrix elements that don't change the gain when they are turned on, but perhaps something didn't work as expected.  Anyhow, the arm powers were only 1ish at the time of lockloss.  There was some kind of glitch in the DARM_OUT (see 2nd plot below, and zoom in 3rd plot), but it doesn't seem to have caused the lockloss.

Lockloss_ArmsAt1ish_TransOscillation_110am_22Oct2014.png

LocklossZoom_ArmsAt1ish_TransOscillation_110am_22Oct2014.png

LocklossZoomGlitch_ArmsAt1ish_TransOscillation_110am_22Oct2014.png

  10636   Thu Oct 23 17:45:54 2014 JenneUpdateGeneralPianosa frozen

Not sure why, but Pianosa was frozen.  Also couldn't ssh or ping.  So, I hard power cycled it.

  10637   Fri Oct 24 02:14:05 2014 JenneUpdateLSCIncreased DARM gain even more

[Jenne, Diego]

We spent the afternoon working on the new scan for IR resonance script.  It is getting much closer, although we need to work on a plan for the fine scanning at the end - so far, the result from the wavelet thing mis-estimates the true peak phase, and so if we jump to where it recommends, we are only at about half of the arm resonance.  So, in progress, but moving forward.

Tonight we repeated the process of reducing the CARM offset and measuring the DARM loop gain as we went.  I'm not sure if I just had the wrong numbers yesterday, or if the gains are changing day-by-day.  The gains that it wanted at given arm buildups were constant throughout this evening, but they are about a factor of 2 higher than yesterday.  If they really do change, we may need to implement a UGF servo for DARM.  New gains are in the carm_cm_up script.

We also actually saved our DARM loop measurements as a function of CARM offset (as indicated by arm buildups).  The loop stays the same through arm powers of 4.  However, once we get to arm powers of 6, the magnitude around 100 Hz starts to flatten out, and we get some weird features in the phase.  It's almost like the phase bubble has a peak growing out of it.  I saw these yesterday, and they just keep getting more pronounced as we go up to arm powers of 7, 8 and 9 (where we lost lock during the measurement).  The very last point in the power=9 trace was just before/during the lockloss, so I don't know if we trust it, or if it is real and telling us something important.  But, I think that it's time to see about getting both CARM and DARM onto a different set of error signals now that we're at about 100pm.

 DARM_23Oct2014.pdf

  10642   Mon Oct 27 22:19:17 2014 JenneUpdateCDSc1iscex restarted

I'm not sure why, but c1iscex did not want to do an mxstream restart.  It would complain at me that  "* ERROR:  mx_stream is already stopping."
Koji suggested that I reboot the machine, so I did.  I turned off the ETMX watchdog, and then did a remote reboot.  Everything came back nicely, and the mx_stream process seems to be running.

  10643   Tue Oct 28 01:12:57 2014 JenneUpdateSUSETMX bad :(

ETMX is misbehaving again.  I went to go squish his cable at the rack and at the satellite box, but it still happened at least once.

Anecdotally and without science, it seems to happen when ETMX is being asked to move a "big" amount.  If I move the sliders too quickly (steps of 1e-3, but holding down the arrow key for about 1 second) or if I offload the ASS outputs when they're too large (above 10ish?), ETMX jumps so that it's about 50 urad off in yaw according to the oplev (sometimes right, more often left), and either 0 or 50urad off in pitch (up if right in yaw, down if left in yaw). 

So far, by-hand slowly offloading the ASS outputs using the sliders seems to keep it happy.

I would ask if this is some DAC bit flipping or something, but it's happening for outputs through both the fast front ends (ASS offloading) and the slow computers (sliders moved too fast).  So.  I don't know what it could be, except the usual cable jiggling out issue.

Anyhow, annoying, but not a show stopper.

  10644   Tue Oct 28 02:44:08 2014 JenneUpdateSUSETMX bad :(

Okay, now ETMX's badness is a show-stopper.  I'm not sure why, but after this last lockloss, ETMX won't stay put.  Right now (as opposed to earlier tonight) it seems to only be happening when I enable LSC pushing on the SUS.  ETMX is happy to sit and stay locked on TEM00 green while I write this entry, but if I go and try to turn on the LSC it'll be wacky again.  Daytime work.

Anyhow, this is too bad, since I was feelin' pretty good about transitioning DARM over to AS55. 

I had a line on (50 counts at 503.1 Hz pushing differentially on the ETMs), and could clearly see the sign flip happen in normalized AS55Q between arm powers of 4 and 6.  The line also told me that I needed a matrix element of negative a few x10^-4 in the AS55Q -> DARM spot.  Unfortunately, I was missing a zero (so I was making my matrix element too big by a factor of 10) in my ezcastep line, so both times I tried to transition I lost lock. 

So.  I think that we should put values of 0.5 into the power normalization for our test case (I was using SRCL_IN1 as my tester) since that's the approximate value that the DCtrans uses, and see what size AS55Q matrix element DARM wants tomorrow (tonight was 1.6-3 x 10^-4, but with 1's in the normalization matrix).  I feel positive about us getting over to AS55. 

Also, Q is (I assume) going to work some more tomorrow on PRMI->REFL165, and Diego is going to re-test his new IR resonance finding script.  Manasa, if you're not swamped with other stuff, can you please see if you can have a look at ETMX?  Maybe don't change any settings, but see what things being turned on makes ETMX crazy (if it's still happening in the morning). 

  10652   Thu Oct 30 01:21:37 2014 JenneUpdateLSCNo MICH in REFL165

[Koji, Jenne, Diego]

Summary:  We really don't have any MICH signal in REFL 165.  Why is still a mystery.

We made several transfer function measurements while PRMI was locked on REFL33 with the arms held off resonance, and compared those to the case where the ETMs are misaligned.  We fine-tuned the REFL165 demod phase looking at the transfer function between 10-300 Hz (using bandpassed white noise injected in the MICH FF filter bank and looking at REFL165Q), rather than just a single line.  We did that at CARM offset of 3 counts (ALS locked), and then saw that as we reduced the CARM offset, the coherence between MICH injection and REFL165Q just goes down.  Any signal that is there seems to be dominated by PRCL. 

So, we're not sure why having the arms eats the MICH 165 signal, but it does.  Everyone should dream tonight about how this could happen. 

Koji suggested that if the signal is just lost in the noise, perhaps we could increase our modulation depth for 55MHz (currently at 0.26, a pretty beefy number already).  Alternatively, if instead the problem is that the MICH signal has rotated to be in line with the PRCL signal, there may be no hope (also, why would this happen?).

Anyhow, we'd like to understand why we don't have any MICH signal in REFL165 when the arm cavities are involved, but until we come up with a solution we'll stick with REFL33 and see how far that gets us. 

The only really worthwhile plot that I've got saved is the difference in these transfer functions when PRMI-only locked and PRMI+arms locked.  Green is PRMI-only, with the demod phase optimized by actuating on PRM and minimizing the peak in the Q signal.  Blue is PRMI with the arms held off resonance using the ALS signals, with the demod phase set again, in the same way.  We were expecting (at least, hoping) that the blue transfer function would have the same shape as the green, but clearly it doesn't.  The dip that is around 45 Hz can be moved by rotating the demod phase, which changes how much PRCL couples into the Q phase.  Weird.  At ~3nm we had somewhat reasonable coherence to RELF165Q, and were able to pick -102deg as the demod phase where the dip just disappears.  However, as the CARM offset is reduced, we lost coherence in the transfer functions.

MICH_to_REFL165_29Oct2014.pdf

  10686   Fri Nov 7 16:15:53 2014 JenneUpdateLSC3F RFPD RF spectra

I have found an SHP-150, but no SHP-175's (also, several 200's, and a couple of 500's).

Why do you say the SHP-150 isn't enough?  The blue peak at 10*fmod in your plot looks like it's at -12 dBm.  -13 dB on top of that will leave that peak at -25 dBm.  That should be enough to keep us from saturation, right?  It's not a lot of headroom, but we can give it a twirl until a 175-er comes in.  

Koji also suggests putting in a 110 MHz notch, combined with either an SHP-150 or SHP-175, although we'll have to measure the combined TF to make sure the notch doesn't spoil the high pass's response too much.

Quote:

165_HPatPD.png


Minicircuits' SHP-150 only has 13dB suppression at 110MHz, which would not be enough either. SHP-175 has 31dB suppression at 110MHz and 0.82dB at 160MHz, maybe this is what we want.

 

  10696   Tue Nov 11 03:48:46 2014 JenneUpdateLSC3f DRMI

I was able to lock the DRMI on 3f signals this evening, although the loops are not stable, and I can hear oscillations in the speakers.  I think the big key to making this work was the placement of the SHP-150 high pass filter at the REFL165 PD, and also the installation of Koji's 110 MHz notch filter just before the amplifier, which is before the demod board for REFL165.  These were done to prevent RF signal distortion.

DRMI 3f:   With DRMI locked on 1f (MICH gain = 1, PRCL gain = -0.05, SRCL gain = 2, MICH = 1*REFL55Q, PRCL = 0.1*REFL11I, SRCL = 1*REFL165I), I excited lines, and found the signs and values for 3f matrix elements.  I was using the same gains, but MICH = 0.5*REFL165Q, PRCL = 0.8*REFL33I and SRCL = -0.2*REFL165I.  Part of the problem is likely that I need to include off-diagonal elements in the input matrix to remove PRCL from the SRCL error signal. 

With the DRMI locked on 1f, I took a sensing matrix measurement.  I don't think we believe the W/ct of the photodiode calibration (we need to redo this), but otherwise the sensing matrix measurement should be accurate.  Since the calibration of the PDs isn't for sure, the relative magnitude for signals between PDs shouldn't be taken as gospel, but within a single PD they should be fine for comparison. 

As a side note, we weren't sure about the MICH -> ITMs balancing, so we checked during a MICH-only, and with the locking apparatus we are unable to measure a difference between 1's for both ITMs in the output matrix, and 1 for ITMX and 0.99 for ITMY.  So, we can't measure 1% misbalances in the actuator, but we think we're at least pretty close to driving pure MICH. 

We kind of expect that SRCL should only be present in the 55 and 165 PDs, although we see it strongly in all of the REFL PDs.  Also, PRCL and SRCL are not both lined up in the I-phase.  So, I invite other people to check what they think the sensing matrix looks like. 

  • The excitation lines (and matching notches) were on from 12:14am (
  • Nov 11 2014 08:14:00 UTC / GPS 1099728856) to 12:20am (
  •  
  • Nov 11 2014 08:20:00 UTC / GPS 
  • 1099729216) for 360sec. 
  • MICH was driven with 800 counts at 675.13 Hz, with +1*ITMY, -1*ITMX. 
  • PRCL was driven with 1000 counts at 621.13 Hz with the PRM. 
  • SRCL was driven with 800 counts at 585.13 Hz using the SRM. 

All 3 degrees of freedom have notches at all 3 of those frequencies in the FM10 of the filter banks (and they were all turned on).  During this time, DRMI was locked with 1f signals. 

DRMI sensing matrix:

 SensMatMeas_10Nov2014_DRMI.png

Earlier in the evening, I also took a PRMI sensing matrix, with the PRMI locked on REFL33 I&Q.  Watch out for the different placement of the plots - I couldn't measure AS55 in the DRMI case, since cdsutils.avg freaked out if I asked for more than 14 numbers (#PDs * #dofs) at a time.

SensMatMeas_10Nov2014_PRMI.png

Rana, Koji and I were staring at the DRMI sensing matrix for a little while, and we aren't sure why PRCL and SRCL aren't co-aligned, and why they aren't orthogonal to MICH.  Do we think it's possible to do something to digitally realign them?  Will the solution that we choose be valid for many CARM offsets, or do we have to change things every few steps (which we don't want to do)? 

Things to work on:

* Reanalyze DRMI sensing matrix data from 12:14-12:20am. 

* Make a simulated scan of higher order mode resonances in the arm cavities.  Is it possible that on one or both sides of the CARM resonance we are getting HOM resonances of the sidebands? 

* Figure out how to make DRMI 3f loops stable.

* Try CARM offset reduction with DRMI, and / or PRMI on REFL 165.

  10701   Wed Nov 12 03:22:23 2014 JenneUpdateLSC3f DRMI sensing mat

Koji pointed out something to me that I think he had told me ages ago, and Rana alluded to last night:  Since I'm not tuning my "demod phase" for the sensing matrix lockins, unless I happened to get very lucky, I was throwing away most of the signal.  Lame.  

So, now the magnitude is sqrt(real^2 + imag^2), where real and imag here are the I and Q outputs of the lockin demodulator, after the 0.1Hz lowpass.  (I put in the low pass into all of the Q filter banks).  To keep the signs consistent, I did do a rough tuning of those angles, so that I can use the sign of the real part as the sign of my signal.  When I was PRMI locked, I set the phase for all things acutated by MICH to be 79deg, all things actuated by PRCL to be 20 deg, and when DRMI locked set all things SRCL to be 50deg. 

After doing this, the phases of my sensing matrix output matches Koji's careful analyses.  I don't know where the W/ct numbers I was using came from (I don't think I made them up out of the blue, but I didn't document where they're from, so I need to remeasure them).  Anyhow, for now I have 1's in the calibration screen for the W/ct calibration for all PDs, so my sensing matrices are coming out in cts/m, which is the same unit that Koji's analysis is in. (Plot for comparing to Koji is at end of entry).

While reducing the CARM offset, I left the sensing matrix lines on, and watched how they evolved.  The phases don't seem to change all that much, but the magnitudes start to decrease as I increase the arm power.

For this screenshot, the left plot is the phases of the sensing matrix elements (all the REFL signals, MICH and PRCL), and the right plot is the magnitudes of those same elements.  Also plotted is the TRX power, as a proxy for CARM offset.  The y-scale for the TRX trace is 0-15.  The y-scale for all the phases is -360 to +360.  The y-scale of the magnitude traces are each one decade, on a log scale.

SensMatVsPower_UpToArms10.png

Bonus plot, same situation, but the next lock held for 20 minutes at arm powers of 8.  We don't know why we lost lock (none of the loops were oscillating, that I could see in the lockloss plot).

PRMI_arms8_20minutes.png


Here are some individual sensing matrix plots, for a single lock stretch, at various times.  One thing that you can see in the striptool screenshots that I don't know yet how to deal with for the radar plots is the error bars when the phase flips around by 360 degrees.  Anyhow, the errors in the phases certainly aren't as big as the error boxes make them look.

PRMI just locked, CARM offset about 3nm, CARM and DARM on ALS comm and diff, arm powers below 1:

SensMatMeas_11Nov2014_PRMIarms_ArmPowSmall.png

PRMI still on REFL33 I&Q, CARM and DARM both on DC transmissions, arm powers about 4:

SensMatMeas_11Nov2014_PRMIarms_ArmPow3pt8.png

CARM offset reduced further, arm powers about 6:

SensMatMeas_11Nov2014_PRMIarms_ArmPow6.png

CARM offset reduced even more, arm powers about 7:

SensMatMeas_11Nov2014_PRMIarms_ArmPow7.png


For this plot for comparing with Koji's analysis, I had not yet put 1's in the calibration screen, so this is still in "W"/m, where "W" is meant to indicate that I don't really know the calibration at all.  What is good to see though is that the angles agree very well with Koji's analysis, even though he was analyzing data from yesterday, and this data was taken today.  This sensing matrix is DRMI-only (no arms), 1f locking.

SensMatMeas_11Nov2014_DRMI_fixedMags.png

Bonus plot, PRMI-only sensing matrix, with PRMI held using REFL 33 I&Q:

SensMatMeas_11Nov2014_PRMI_fixedMags.png

 

  10703   Wed Nov 12 18:08:35 2014 JenneUpdateLSCRIN in transmission a problem?

In my previous meditations about RIN, particularly elog 10258, I was only thinking about the RIN contribution at the offset that I was currently sitting at.  Also, In elog 10258 I was comparing to the ALS signals and just said that the trans signals are better which is true, although isn't super helpful when thinking of reduced CARM offsets. 

My summary today is that I think we want to reduce the RIN in arm transmissions by a factor of 3.


Rather than dig around, I just remeasured the RIN, for both the single arm transmissions and the MC transmission.  (Data attached as .xml file)

The RMS RIN for the Xarm is 1.3e-2.  The RMS RIN for the Yarm is 8.9e-3.  The RMS RIN for MCtrans is 4.0e-3.  For the simulations below, I will use 1e-2 as an average RIN for the arms.

RIN_TRX_TRY_MCtrans_12Nov2014.pdf


As an estimate of the RIN's contribution to cavity fluctuations, I divide the RIN by the slope of the CARM transmission peak.  The slope (from optickle) gives me [ delta-W / delta-m ], and the inverse of that gives me [ delta-m / delta-W ].  I multiply this by RIN, which is [ delta-W / W ] to get [delta-m / W]. 

Then, since I'm using the DC transmission signals as my error signals, I use just TRX (normalized to be 1 for single arm resonance) as my Watts.

So, in total, the traces plotted are { TRX * RIN / slope }. 

The 2 plots are the same data, one with linear-x and the other with log-x.  They both include my estimate of the cavity length fluctuations due to RIN at the arm transmission, as well as an estimate of the cavity length fluctuations if the arm RIN was as good as the MC RIN.  I also show the DRFPMI CARM linewidth (23 pm for HWHM), and 1% of that linewidth.  The last trace is 1% of the half-width of the transmission peak, at the current CARM offset.  For example, 1000 pm away from full resonance the half-width is 1000 pm and 1% of that is 10 pm. 

 RINcontribution_12Nov2014_linearXscale.pngRINcontribution_12NOv2014_logXscale.png

What we want to see here is that the solid blue line is below one of the dotted lines.  I think that using the overall linewidth (purple dotted line) isn't really the right thing to look at.  Our goal is to prevent excursions that will get too close to the resonance peak, and cause a lockloss.  A one picometer excursion is a much bigger problem (relatively) below say 100 pm, as opposed to above 100 pm.  So, I think that we should be looking at the half-width of the resonance peak at whatever the current CARM offset is (orange dotted line).  Above 25 pm, the blue line is below the orange line for all offsets plotted.  If we made the arm RIN as good as the MC RIN, that would be true down to 12-ish pm. 

We should be able to safely transition to non-normalized RF signals at 10pm or below.  This implies that (since any RF signals normalized by this RIN-y trans signal will have the RIN), we want to improve the RIN of the transmission PDs by about a factor of 3. (This will lower the blue line such that it crosses the orange dotted line at 10 pm).

 

Attachment 1: RIN_TRX_TRY_MCtrans_12Nov2014_zip.xml.gz
  10709   Thu Nov 13 04:28:28 2014 JenneUpdateLSCRIN in transmission a problem?

[Jenne, Rana, Koji]

We did some thinking on what could be causing the excess RIN that we see in the arm transmissions but not in the MC transmission.  Unfortunately, I don't think we have anything conclusive yet. 

We thought about:

  • Test mass oplevs
  • Input tip tilt jitter
  • MC motion

Oplevs

As Rana reported in elog 10708, we tuned the oplev servos for ITMX, ETMX, ITMY and ETMY.  They all now look like the new SRM oplevs that Rana described in elog 10694.  However, when we re-looked at the RIN after the oplev tuning, we did not see a noticeable change.  So, fixing up the oplevs didn't fix up the RIN.

Side notes:

  • Several optics had low gains for suspos, which were increased to give Qs of ~5ish.
  • When we gave ITMX a 500 count step in pitch (the same size used for all other optics in both pit and yaw), it didn't come back afterward.  This is a little disconcerting.  Rana had to move the alignment slider to get it back so that we had MICH fringing at the AS port again.

Input Tip Tilts

Koji did some work, reported in elog 10706, on how much the jitter of the input pointing tip tilts should affect us.  We don't think that they are moving enough to be the cause of the excess RIN that we see.


Mode Cleaner Motion

We see some coherence between MC2 suspit and TRX/TRY, so we suspect that the MC's motion could be causing problems. 

I looked at the WFS vs. TRX & TRY, and saw significant coherence at the 3 Hz stack resonance.  I think it's clear that the WFS can help suppress this motion more.  The WFS noise level was too bad to see any other coherence at other frequencies. (Side note:  We should consider increasing the analog WFS signal.  As Rana mentioned back in 2008 in elog 454, the signal is super small.  Increasing the analog gain could allow us to suppress motion at more frequencies, although it would be a bit of a pain.)

To try and get some more signal, I routed the IPPOS beam over to the PRM oplev temporarily.  The idea was to be able to look at the IPPOS port, but with a fast channel.  I turned off the BS/PRM HeNe, and removed the last steering mirror before the QPD.  I put in 2 Y1 steering mirrors to get the IPPOS beam across the table and pointing at IPPOS.  I took my measurements 3 times, with different beam sizes on the QPD, to try to image different gouy phases.  I used absorptive ND filters (0.6 + 0.1) to get the light level on the PD such that I had about 10,000 counts per quadrant, where 16,000 counts seemed to be the saturation point. I also reset the dark offsets of the QPD quadrants, although they were so small that I don't think it did much.  I also took out the optical lever calibration from counts to microradians, since that number isn't meaningful for what I was doing.  So, the IPPOS signals (using the PRM oplev channels) are in raw ADC counts.  The first measurement had no lens, and the beam was probably at least half the size of the QPD.  The second measurement had a lens, and the beam on the QPD was about half the original size.  The third measurement had the lens closer to the QPD, so that the beam was about 1mm on the diode.  In none of these cases do I see any significant coherence with the TRX/TRY RIN signals, except at 3 Hz. After my measurements I put the oplev back including all of the digital settings, although for now I just left the IPPOS beam dumped on a razor dump, since it wasn't being used anyway.  I need to realign IPPOS when it's not the middle of the night.

Some thoughts that we have:

  • Maybe it's time to resurrect seismic feedforward on MC length, to suppress some of this 3 Hz motion?
  • Maybe we should be using the MC_L path to offload some of the frequency feedback, so that we're not pushing on MC2 so hard (because if we have bad F2P coupling, this creates beam motion)

I have plots for each of my IPPOS vs. TRX/TRY measurements.  The data is attached.  For each, I looked at the transfer function between IPPOS (using the SUS-PRM_OPLEV channels) and TRX/TRY to get the 'calibration' between input beam motion and transmission RIN.  In all cases, at 3 Hz the coefficient was about 1, so in the power spectra on the right side, there is no calibration applied to the IPPOS signals. 

IPPOS vs. Transmission RIN, no lens, big beam on QPD:

RIN_TRX_TRY_IPPOS_NoLens_12Nov2014.pdf

(Just kidding about the other 2 plots - the elog can't handle it.  They're in the zippyzip file though, and I don't think they look qualitatively different from the no-lens case).

 

Attachment 1: zippyzip.zip