  10549   Mon Sep 29 12:47:51 2014 ericqUpdateGeneralVent update

Quote:
  •  Lock the SRM cage down fully (right now, has 2 clamps on, and one laying unused)
  • Align SRM for SRC flashes
  • Adjust SRM OSEM positions as needed
  • Adjust SRM oplev beam path, measure lever arm for calibration
  • Confirm beam spots on output mirrors in ITMY and BS chambers are ok

 [Koji, ericq]

We have completed the above points; the ITMY table is still level.

Despite what the wiki says, the SRM LR OSEM open voltage is ~1.97V instead of ~1.64, so we shot for half of that. 

The in-air steering of the SRM oplev return beam needs adjustment. I'll estimate the beam path length when I'm taking pictures and closing up. 

Left to do:

  • Now that AS is back on diode, lock arms and align everything. Confirm everyone's happiness. 
  • Take numerous pictures of ITMY chamber.
  • Center oplevs
  • Put doors on
  • Close shutters
  • Pump down
  • Replace MC refl Y1 with the beamsplitter
  • Turn PSL power back up

Related In-Air work:

  • Fix POY steering
  • Fix SRM oplev return steering
  10550   Mon Sep 29 17:10:51 2014 ericqUpdateGeneralVent update

Everything is aligned, AS and POY make it out of vacuum unclipped, OSEM readings look good.

I set up the SRM oplev, centered all oplevs.

Tomorrow, we just have to take pictures of the ITMY chamber before we put the heavy doors on. 

  10551   Mon Sep 29 18:12:24 2014 ericqUpdateGeneralVent update

I closed the PSL shutter as we didn't want to burn the mirror surface when we are not working.

  10552   Tue Sep 30 11:53:29 2014 ericqUpdateGeneralVent update

 

Photos have been taken of the ITMY chamber, and uploaded to picasa. Here's a slideshow:

  7306   Wed Aug 29 11:47:21 2012 ericqUpdateVACVenting

 [Steve, Eric]

I've been helping Steve vent this morning. The following things were done (from Steve's logbook):

  • Particle counts: 0.5 micron particles, 4200 counts per cubic ft
  • Vertex crane drive checked to be ok
  • Optical Levers set for local damping only
  • Saved some screens
  • PSL shutter and green shutters closed
  • HV Off checked, JAM nuts checked
  • Vac: Close V1, VM1, ans - VA6, open VM3 - RGA, cond: chamber open mode
  • 8AM: VV1 open to N2, regulator set  to 14 psi
  • 8:23AM: 35psi Instrument grade Air

(At this point, I took over the air canisters, while Steve made preparations around the lab. 

  • 9:00AM: 2nd air cylinder, 14 psi 
  • 9:40AM: 3rd air cyl
  • 10:20AM: 4th air cyl
  • 11:00AM: 5th air cyl

With the 5th cylinder, we began approaching 1 atm, so we slowed the regulator down to 5psi. Around 750 torr, Steve opened VV1 to air.

According to Steve, we will be at atmospheric pressure at  ~12:30pm.

  38   Wed Oct 31 10:31:23 2007 Andrey RodionovRoutineVACVenting is in progress

We (Steve, David, Andrey) started venting the vacuum system at 9.50AM Wednesday morning.
  13620   Thu Feb 8 00:01:08 2018 gautamUpdateCDSVertex FEs all crashed

I was poking around at the LSC rack to try and set up a temporary arrangement whereby I take the signals from the DAC differentially and route them to the D990694 differentially. The situation is complicated by the fact that, afaik, we don't have any break out boards for the DIN96 connectors on the back of all our Eurocrate cards (or indeed for many of the other funky connecters we have like IDE/IDC 10,50 etc etc). I've asked Steve to look into ordering a few of these. So I tried to put together a hacky solution with an expansion card and an IDC64 connector. I must have accidentally shorted a pair of DAC pins or something, because all models on the c1lsc FE crashedindecision. On attempting to restart them (c1lsc was still ssh-able), the usual issue of all vertex FEs crashing happened. It required several iterations of me walking into the lab to hard-reboot FEs, but everything is back green now, and I see the AS beam on the camera so the input pointing of the TTs is roughly back where it was. Y arm TEM00 flashes are also seen. I'm not going to re-align the IFO tonight. Maybe I'll stick to using a function generator for the THD tests, probably routing non AI-ed signals directly is as bad as any timing asynchronicity between funcGen and DAQ system...

Attachment 1: CDSrecovery_20180207.png
CDSrecovery_20180207.png
  15383   Mon Jun 8 18:14:55 2020 gautamUpdateCDSVertex FEs crashed

Summary:

Around 5pm local time, the three vertex FEs crashed. AFAIK, no one was in the lab or working on anything CDS related, so this is worrying.

Details:

  • Reboot script was used to bring all FEs back - only soft reboots were required.
  • The IMC and arms can now be locked.
  • I think combination of burt + SDF would have reverted all the settings as they should be, but if something appears off, it could be that some EPICS value didn't get reset correctly.
Attachment 1: FEcrash_CDSoverview.png
FEcrash_CDSoverview.png
  7770   Fri Nov 30 23:10:36 2012 CharlesUpdateElectronicsVertex Illuminators

 3 of the 4 remote controlled illuminators at the vertex are installed and can now be turned on via sitemap. There are a total of 15 controls for "Illum", but only the 3 labeled with MC, BS-PRM and ITMY-SRM are functional.

  4828   Thu Jun 16 08:45:14 2011 steveUpdateSUSVertex SUS Binary Output Boxes removed

Quote:

- I was investigating the SUS whitening issue.

- I could not find any suspension which can handle the input whitening switch correctly.

- I went to 1X5 rack and found that both of the two binary output boxes were turned off.
As far as I know they are pulling up the lines which are switched by the open collector outputs.

- I tried to turn on the switch. Immediately I noticed the power lamps did not work. So I need an isolated setup to investigate the situation.

- The cables are labelled. I will ask steve to remove the boxes from the rack.

 I shut down damping to the Vertex optics and removed Binary IO  Adapter chassy BO0 and BO1

About a week ago I discussed the BO0's power indicator lights with Kiwamu. They were  not on or they were blinking on-off.

I put screws into ps connectors in the back, but it did not helped.

Attachment 1: P1070894.JPG
P1070894.JPG
  4829   Thu Jun 16 23:19:09 2011 KojiUpdateSUSVertex SUS Binary Output Boxes removed

[Jamie, Koji]

- We found the reason why some of the LEDs had no light. It was because the LEDs were blown as they were directly connected to the power supply.
The LEDs are presumably designed to be connected to a 5V supply (with internal current-limiting resistor of ~500Ohm). The too much current
with the 15V (~30mA) made the LED blown, or the life-time of them shorter.

- Jamie removed all of the BO modules and I put 800Ohm additional resister such that the resultant current is to be 12mA.
The LEDs were tested and are fine now.

- The four BO boxes for C1SUS were restored on the rack. I personally got confused what should be connected where
even though I had labeled for BO0 and BO1. I just have connected CH1-16 for BO0. The power supplies have been connected only to BO0 and BO1.

- I tested the whitening of PRM UL sensor by exciting PRM UL sensor. The transfer function told us that the pendulum response can be seen
up to 10-15Hz. When the whitening is on, I could see the change of the transfer function in that freq band. This is good.
So the main reason why I could not see theis was that the power supply for the BOs were not turned on.

- I suppose Jamie/Joe will restore all of the BO boxes on the racks tomorrow. I am going to make a test script for checking the PD whitenings.

  4827   Thu Jun 16 00:43:36 2011 KojiUpdateSUSVertex SUS Binary Output Boxes were turned off / need investigation

- I was investigating the SUS whitening issue.

- I could not find any suspension which can handle the input whitening switch correctly.

- I went to 1X5 rack and found that both of the two binary output boxes were turned off.
As far as I know they are pulling up the lines which are switched by the open collector outputs.

- I tried to turn on the switch. Immediately I noticed the power lamps did not work. So I need an isolated setup to investigate the situation.

- The cables are labelled. I will ask steve to remove the boxes from the rack.

  16502   Fri Dec 10 21:35:15 2021 KojiSummarySUSVertex SUS DAC adapter ready

4 units of Vertex SUS DAC adapter (https://dcc.ligo.org/LIGO-D2100035) ready.

https://dcc.ligo.org/LIGO-S2101689

https://dcc.ligo.org/LIGO-S2101690

https://dcc.ligo.org/LIGO-S2101691

https://dcc.ligo.org/LIGO-S2101692

The units are completely passive right now and has option to extend to have a dewhitening board added inside.
So the power switch does nothing.

Some of the components for the dewhitening enhancement are attached inside the units.

 

 

Attachment 1: PXL_20211211_053155009.jpg
PXL_20211211_053155009.jpg
Attachment 2: PXL_20211211_053209216.jpg
PXL_20211211_053209216.jpg
Attachment 3: PXL_20211211_050625141-1.jpg
PXL_20211211_050625141-1.jpg
  11577   Fri Sep 4 15:20:31 2015 ericqUpdateLSCVertex Sensing

I've now made a collection of sensing matrix measurements. 

In all of the plots below, the radial scale is logarithmic, each grid line is a factor of 10. The units of the radial direction are calibrated into demod board output Volts per meter. The same radial scale is used on all plots and subplots.

I did two PRMI measurements: with MICH locked and excited with either the ITMS or the BS + PRM compensation. This tells us if our PRM compensation is working; I think it is indeed ok. I though I remembered that we came up with a number for the SRM compensation, but I haven't been able to find it yet. 

The CARM sensing int he PRFPMI measurement has the loop gain at the excitation frequency undone. All excitations were simultaneously notched out of all control filters, via the NotchSensMat filters. 

The angular scale is set to the analog I and Q signals; the dotted lines show the digitial phase rotation angle used at the time of measurement. 

Attachment 1: PRFMI_ITM.pdf
PRFMI_ITM.pdf
Attachment 2: PRFMI_BS.pdf
PRFMI_BS.pdf
Attachment 3: DRMI.pdf
DRMI.pdf
Attachment 4: PRFPMI.pdf
PRFPMI.pdf
  4225   Sat Jan 29 00:31:05 2011 SureshUpdateGeneralVertex crane upgrade completed

The Vertex crane is smarter and safer now.  This upgrade ensures that the two sections of I-beam (8ft, 4ft) remain firmly latched to form a straight member till the latch is released.

In specific, it ensures that problems such as this one do not occur in the future.

 

The new safety features are:

When the I-beam sections are latched together, a pneumatic piston ensures that the latch is secure. 

If the latch is not engaged the trolley does not move outward beyond the end of the 8-foot section of the I beam.

If the trolley is out on the 4-foot section of the beam then we cannot disengage the latch.

 

How does it work?

 

 Vertex_Crane-2.png Vertex_Crane-4.png

 

The state of the Limit Switch 1 changes when the trolly goes past it.    The Limit Switch 2 gets pressed when the two sections are latched together.

The pneumatic piston raises or lowers the latch.  The Pneumatic Latch Switch operates a pneumatic valve controlling the state of the piston.

 

 

Vertex_Crane-3.png P1280545.JPG

The new controller now has Pneumatic Latch Switch in addition to the usual Start, Stop, Up, Down, In and Out buttons. 

Each of the Up, Down, In and Out buttons have two operational states:  Half pressed (low speed) and Full pressed (High Speed).  Their functions remain the same as before.

 

The new Pneumatic Switch:

When this switch is 'Engaged' and the 4 ft section is swung in-line with the 8 ft section, the two sections get latched together.

To unlatch them we have to throw the switch into the 'Disengage' state.  This makes the piston push the latch open and a spring rotates the 4 ft section about its pivot.

Limit Switch 2 is not pressed (I-beams not aligned straight) ==> Limit Switch 1 will prevent the trolley from out going beyond the 8 ft section.

While Limit Switch 2 is pressed we cannot disengage the latch.

 

Note: 

   The pneumatic piston requires 80psi of pressure to operate.  However we have only 40psi in the lab and the piston seems to operate quite well at this pressure as well.  I believe a request has been made to get an 80psi line laid just for this application.

 

Attachment 1: Vertex_Crane-2.png
Vertex_Crane-2.png
Attachment 2: Vertex_Crane-4.png
Vertex_Crane-4.png
  4233   Mon Jan 31 16:12:11 2011 steveUpdateVACVertex crane upgrade shorth coming

 The upgrade is almost finished. I found that the passive latch lock  is not closing down all the way. It has about a 3/8" gap.    See Atm. 1 & 2

The service man was here this morning and agreed to fix it.  They will be back next week. The latch needs an other spring to push it into full lock. 

We tested all possible sequences of operation of the new upgrade. It performed to specification.

 

 

Quote:

 

Attachment 1: P1070364.JPG
P1070364.JPG
Attachment 2: P1070358.JPG
P1070358.JPG
  15035   Tue Nov 19 15:08:48 2019 gautamUpdateCDSVertex models rebooted

Jon and I were surveying the CDS situation so that he can prepare a report for discussion with Rolf/Rich about our upcoming BHD upgrade. In our poking around, we must have bumped something somewhere because the c1ioo machine went offline, and consequently, took all the vertex models out. I rebooted everything with the reboot script, everything seems to have come back smoothly. I took this opportunity to install some saturation counters for the arm servos, as we have for the CARM/DARM loops, because I want to use these for a watch script that catches when the ALS loses lock and shuts stuff off before kicking optics around needlessly. See Attachment #1 for my changes.

Attachment 1: armSat.png
armSat.png
  93   Mon Nov 12 10:53:58 2007 pkpUpdateOMCVertical Transfer functions
[Norna Sam Pinkesh]

These plots were created by injected white noise into the OSEMs and reading out the response of the shadow sensors ( taking the power spectrum). We suspect that some of the additional structure is due to the wires.
Attachment 1: VerticalTrans.pdf
VerticalTrans.pdf VerticalTrans.pdf VerticalTrans.pdf VerticalTrans.pdf
  105   Thu Nov 15 17:09:37 2007 pkpUpdateOMCVertical Transfer functions with no cables attached.
[Norna Pinkesh]

The cables connecting all the electronics ( DCPDs, QPDs etc) have been removed to test for the vertical transfer function. Now the cables are sitting on the OMC bench and it was realigned.
Attachment 1: VerticaltransferfuncnocablesattachedNov152007.pdf
VerticaltransferfuncnocablesattachedNov152007.pdf VerticaltransferfuncnocablesattachedNov152007.pdf VerticaltransferfuncnocablesattachedNov152007.pdf VerticaltransferfuncnocablesattachedNov152007.pdf
  12181   Wed Jun 15 09:52:02 2016 jamieUpdateCDSVery encouraging results from overnight split daqd test

laughVery encouraging results from the test last night.  The new configuration did not crash once overnight, and seemed to write out full, second trend, and minute trend frames without issueyes.  However, full validity of all the written out frames has not been confirmed.

overview

The configuration under test involves two separate daqd binaries instead of one.  We usually run with what is referred to as a "framebuilder" (fb) configuration:

  • fb: a single daqd binary that:
    • collect the data from the front ends
    • coallate full data into frame file format
    • calculates trend data
    • writes frame files to disk.

The current configuration separates the tasks into multiple separate binaries: a "data concentrator" (dc) and a "frame writer" (fw):

  • dc:
    • collect data from front ends
    • coallate full data into frame file format
    • broadcasts frame files over local network
  • fw:
    • receives frame files from broadcast
    • calculates trend data
    • writes frame files to disk

This configuration is more like what is run at the sites, where all the various components are separate and run on separate hardware.  In our case, I tried just running the two binaries on the same machine, with the broadcast going over the loopback interface.  None of the systems that use separated daqd tasks see the failures that we've been seeing with the all-in-one fb configuration (and other sites like AEI have also seen).

My guess frown is that there's some busted semaphore somewhere in daqd that's being shared between the concentrator and writer components.  The writer component probably aquires the lock while it's writing out the frame, which prevents the concentrator for doing what it needs to be doing while the frame is being written out.  That causes the concentrator to lock up and die if the frame writing takes too long (which it seems to almost necessarily do, especially when trend frames are also being written out).

results

The current configuration hasn't been tweaked or optimized at all.  There is of course basically no documentation on the meaning of the various daqdrc directives.  Hopefully I can get Keith Thorne to help me figure out a well optimized configuration.

There is at least one problem whereby the fw component is issuing an excessively large number of re-transmission requests:

2016-06-15_09:46:22 [Wed Jun 15 09:46:22 2016] Ask for retransmission of 6 packets; port 7097
2016-06-15_09:46:22 [Wed Jun 15 09:46:22 2016] Ask for retransmission of 8 packets; port 7097
2016-06-15_09:46:22 [Wed Jun 15 09:46:22 2016] Ask for retransmission of 3 packets; port 7097
2016-06-15_09:46:22 [Wed Jun 15 09:46:22 2016] Ask for retransmission of 5 packets; port 7097
2016-06-15_09:46:22 [Wed Jun 15 09:46:22 2016] Ask for retransmission of 5 packets; port 7097
2016-06-15_09:46:22 [Wed Jun 15 09:46:22 2016] Ask for retransmission of 5 packets; port 7097
2016-06-15_09:46:22 [Wed Jun 15 09:46:22 2016] Ask for retransmission of 5 packets; port 7097
2016-06-15_09:46:22 [Wed Jun 15 09:46:22 2016] Ask for retransmission of 6 packets; port 7097
2016-06-15_09:46:23 [Wed Jun 15 09:46:23 2016] Ask for retransmission of 1 packets; port 7097

It's unclear why.  Presumably the retransmissions requests are being honored, and the fw eventually gets the data it needs.  Otherwise I would hope that there would be the appropriate errors.

The data is being written out as expected:

 full/11500: total 182G
drwxr-xr-x  2 controls controls 132K Jun 15 09:37 .
-rw-r--r--  1 controls controls  69M Jun 15 09:37 C-R-1150043856-16.gwf
-rw-r--r--  1 controls controls  68M Jun 15 09:37 C-R-1150043840-16.gwf
-rw-r--r--  1 controls controls  68M Jun 15 09:37 C-R-1150043824-16.gwf
-rw-r--r--  1 controls controls  69M Jun 15 09:36 C-R-1150043808-16.gwf
-rw-r--r--  1 controls controls  69M Jun 15 09:36 C-R-1150043792-16.gwf
-rw-r--r--  1 controls controls  68M Jun 15 09:36 C-R-1150043776-16.gwf
-rw-r--r--  1 controls controls  68M Jun 15 09:36 C-R-1150043760-16.gwf
-rw-r--r--  1 controls controls  69M Jun 15 09:35 C-R-1150043744-16.gwf

 trend/second/11500: total 11G
drwxr-xr-x  2 controls controls 4.0K Jun 15 09:29 .
-rw-r--r--  1 controls controls 148M Jun 15 09:29 C-T-1150042800-600.gwf
-rw-r--r--  1 controls controls 148M Jun 15 09:19 C-T-1150042200-600.gwf
-rw-r--r--  1 controls controls 148M Jun 15 09:09 C-T-1150041600-600.gwf
-rw-r--r--  1 controls controls 148M Jun 15 08:59 C-T-1150041000-600.gwf
-rw-r--r--  1 controls controls 148M Jun 15 08:49 C-T-1150040400-600.gwf
-rw-r--r--  1 controls controls 148M Jun 15 08:39 C-T-1150039800-600.gwf
-rw-r--r--  1 controls controls 148M Jun 15 08:29 C-T-1150039200-600.gwf
-rw-r--r--  1 controls controls 148M Jun 15 08:19 C-T-1150038600-600.gwf

 trend/minute/11500: total 152M
drwxr-xr-x 2 controls controls 4.0K Jun 15 07:27 .
-rw-r--r-- 1 controls controls  51M Jun 15 07:27 C-M-1150023600-7200.gwf
-rw-r--r-- 1 controls controls  51M Jun 15 04:31 C-M-1150012800-7200.gwf
-rw-r--r-- 1 controls controls  51M Jun 15 01:27 C-M-1150002000-7200.gwf

The frame sizes look more or less as expected, and they seem to be valid as determined with some quick checks with the framecpp command line utilities.

  4266   Wed Feb 9 23:48:12 2011 SureshConfigurationCamerasVideo Cable work: New Labels

[Larisa, Aidan,Steve,Suresh]

   Today was the first session for implementing the new video cabling plan laid out in the document " CCD_Cable_Upgrade_Plan_Jan11_2011.pdf"  by Joon Ho attached to his elog entry 4139.  We started to check and label all the existing cables according to the new naming scheme. 

So far we have labeled the following cables. Each has been checked by connecting it to a monitor near the Video Mux and a camera at the other end.

C1:IO VIDEO 8ETMYF

C1:IO-VIDEO 6 ITMYF

C1:IO-VIDEO 21 SRMF

C1:IO-VIDEO 25 OMCT

C1:IO-VIDEO 19 REFL

C1:IO-VIDEO 22 AS

C1:IO-VIDEO 18 IMCR

C1:IO-VIDEO 14 PMCT

C1:IO-VIDEO 12 RCT

C1:IO-VIDEO 9 ETMXF

C1:IO-VIDEO 1 MC2T

 

Next we need to continue and finish the labeling of existing cables.  We then choose a specific set of cables which need to be laid together and proceed to lay them after attaching suitable lables to them.

 

 

  2304   Fri Nov 20 00:18:45 2009 ranaSummaryCamerasVideo MUX Selection Wiki page

Steve is summarizing the Video Matrix choices into this Wiki page:

http://lhocds.ligo-wa.caltech.edu:8000/40m/Electronics/VideoMUX

Requirements:

Price: < 5k$

Control: RS-232 and Ethernet

Interface: BNC (Composite Video)

Please check into the page on Monday for a final list of choices and add comments to the wiki page.

  4519   Wed Apr 13 16:38:17 2011 Larisa ThorneUpdateElectronicsVideo MUX camera/monitor check

 [Kiwamu, Larisa]

 

The following Video MUX inputs(cameras) and outputs(monitors) have been checked:

MC2F, FI, AS Spare, ITMYF, ITMXF, ETMYF, ETMXF, PSL Spare, ETMXT, MC2T, POP, MC1F/MC3F, SRMF, ETMYT, PRM/BS, CRT1(MON1), ETMY Monitor, CRT2(MON2), CRT4(MON4), MC1 Monitor, CRT3(MON3), PSL1 Monitor, PSL2 Monitor, CRT6(MON6), CRT5(MON5), ETMX Monitor, MC2 Monitor, CRT9, CRT7(MON7), CRT10, and Projector.

 

Their respective statuses have been updated on the wiki:   (wiki is down at the moment, I will come back and add the link when it's back up)

  16661   Thu Feb 10 21:10:43 2022 KojiUpdateGeneralVideo Mux setting reset

Now the video matrix is responding correctly and the web interface shows up. (Attachment 1)

Also the video buttons respond as usual. I pushed Locking Template button to bring the setting back to nominal. (Attachment 2)

Attachment 1: Screenshot_2022-02-10_21-11-21.png
Screenshot_2022-02-10_21-11-21.png
Attachment 2: Screenshot_2022-02-10_21-11-54.png
Screenshot_2022-02-10_21-11-54.png
  12694   Fri Jan 6 17:00:26 2017 ranaFrogsTreasureVideo of Lab Tour

In this video: https://youtu.be/iphcyNWFD10, the comments focus on the orange crocs, my wrinkled shirt, and the first aid kit.

  7945   Mon Jan 28 17:01:19 2013 DenUpdateLockingVideo of PRM-flat test cavity

What mode will you get if lock the cavity PRM - ITMY/ITMX/TEST MIRROR without PR2, PR3 and BS?

Is it possible to skip MC1, MC3 and lock the laser to this test cavity to make sure that this is not actuator/electronics noise?

  7951   Tue Jan 29 10:50:02 2013 JenneUpdateLockingVideo of PRM-flat test cavity

 

I think Den accidentally edited and overwrote my entry, rather than replying, so I'm going to recreate it from memory:

I aligned the PRM-flat test cavity (although not as well as Jamie and Koji did later in the evening) and took some videos. Note that these may not be as relevant any more, since Jamie and Koji improved things after I left.

 

Also, before doing anything with the cavity, I tuned up the PMC since the pitch input alignment wasn't perfect (we were getting ~0.7 transmission), and also tuned up the MC alignment and remeasured the MC spot positions, to maintain a record.

  2314   Mon Nov 23 16:28:12 2009 steveSummaryCamerasVideo swicher options

Quote:

Steve is summarizing the Video Matrix choices into this Wiki page:

http://lhocds.ligo-wa.caltech.edu:8000/40m/Electronics/VideoMUX

Requirements:

Price: < 5k$

Control: RS-232 and Ethernet

Interface: BNC (Composite Video)

Please check into the page on Monday for a final list of choices and add comments to the wiki page.

 Composite video matrix switchers with 32 BNC in and 32 BNC channels out are listed.

  4498   Thu Apr 7 13:12:23 2011 KojiHowToVIDEOVideo switching tip

Long time ago, I looked at the manual of the video switcher.
http://media.extron.com/download/files/userman/Plus_Ultra_MAV_C.pdf
Here is the summary. This will be the basic of the more sophisticated switching program which may have GUI.

In principle, you can manually control the matrix via telnet. At the console machines, you can connect to the matrix using telnet

telnet 192.168.113.92

This opens TCP/IP port 23 of the specified machine. You will receive some messages.
Then type some command like:
--------------------

  • 1*2!       (connect input#1 to output#2)
  • 1,           (save the current setting into preset1)
  • 1.           (restore the setting from preset1)

--------------------

Basicaly that's all. There are many other features but I don't think we need them.

We can create a simple program with any of the language as any of the language has the capability of the TCP/IP connection.
e.g. C, Perl, Python. Tcl/Tk
Any of them are fine.

Now what we have to think about is how to implement the interface in the epics screen (or whatever).
It needs some investigation how the people is thinking as the ideal interface.
But, first of all, you should make the above three operations available as a simple UNIX command like:

videoswitch -i 192.168.113.92 1 2
videoswitch -i 192.168.113.92 -store 1
videoswitch -i 192.168.113.92 -recall 1
(There is no such command yet. These are showing what it should be!)

This can be done by a single day work and our life will be much better.

  4529   Fri Apr 15 02:30:24 2011 KojiHowToVIDEOVideo switching tip

I have made a small python script to handle the video matrix.

It is too far from the perfection, but I release it as it is already useful in some extent.

The script is in the /cvs/cds/rtcds/caltech/c1/scripts/general directory.

usage:

videoswitch.py in_ch_name out_ch_name

in_ch_name is one of the followings

MC2F, IFOPO, OMCR, FI, AS_Spare, ITMYF, ITMXF, ETMYF, ETMXF,
PMCR, RCR, RCT, PSL_Spare, PMCT, ETMXT, MC2T, POP, IMCR, REFL,
MC1F, SRMF, AS, ETMYT, PRM, OMCT, Quad1, Quad2, Quad3

out_ch_name is one of the followings

Mon1, Mon2, Mon3, Mon4, Mon5, Mon6, Mon7,
ETMY, MC1, PSL1, PSL2, ETMX, MC2, CRT9,CRT10,Projector,
Quad1_1, Quad1_2, Quad1_3, Quad1_4,
Quad2_1, Quad2_2, Quad2_3, Quad2_4,

Quad3_1, Quad3_2, Quad3_3, Quad3_4

  7839   Mon Dec 17 14:45:01 2012 JenneUpdateAlignmentVideos with PRMI locked

[Jamie, Jenne]

Koji and Jamie locked the PRMI, and then Jamie and I took some videos. 

Video 1:   https://www.youtube.com/watch?v=jszTeyETyxU shows the face of PR2.

Video 2:   https://www.youtube.com/watch?v=Tfi4I4Q3Mqw shows the back of PR3, the face of PR2, as well as REFL and AS.

Video 3:   https://www.youtube.com/watch?v=bLHNWHAWZBA is the camera looking at the face of PRM and (through a viewing mirror) BS.

 

If you watch video 1, you'll see how large the beam gets on the face of PR2.  The main spot, where the straight-through, no-cavity beam is, is a little high of center.  The rest of the inflated beam swirls around that point.

Video 2 shows the same behavior, but you also see that we're much too high on PR3, and too close to the right (as seen on the video) side.

Video 3 is very disconcerting to me.  The main, stationary beam spot seems nicely centered, but the resonant beam, since it inflates and gets big, is very close to the right side of the PRM (as seen on the video). 

It wouldn't surprise me if, were we able to quantify the beam clipping loss on PR3 and PRM, the clipping were the reason we have a crappy PRC gain.  This doesn't explain why we have such a weird inflated beam though.

  10394   Thu Aug 14 22:16:02 2014 JenneUpdateSUSViolin Mode filters for ETMs

The instigator of this was that we were seeing ring-ups of ETMs during our ALS locks this evening.  We measured the ETMY violin resonance to be 624.10 Hz, and Rana found an elog saying that the ETMX was around 631 Hz, so we made a 2 notch filter and added it to FM4 of the LSC-SUS filter banks for both ETMs. 

For the ETMY resonance, we measured the frequency in the DARM spectrum, and when we looked at the FINE_PHASE_OUT channels, the resonance was only in the Yarm sensor.  So, we conclude that it is coming from ETMY.

Also in the realm of filter modules, the FM3 boost for CARM, DARM, XARM and YARM was changed from zero crossing to ramp with a 1sec ramp time.

  10393   Thu Aug 14 20:52:36 2014 ranaUpdateWikiViolin Mode table added to Wiki

Mech Resonance Wiki

I've updated the wiki by trawling the elog for violin entries. Please keep it up to date so that we can make violin notches.

 

  9057   Fri Aug 23 01:52:34 2013 JenneUpdateSUSViolin filters moved to LSC, from SUS

[Rana, Jenne]

We were meditating a little bit on what may be the story behind the PRM violin filter situation.  We locked the PRMI, and turned on and off the violin filters.  We noticed, very bizarrely, that when the violin filters were ON, the servos would oscillate.  Weird.  Also, probably because the oscillation was causing us to hit the limit we have in the MICH servo, we rung up a 3rd harmonic of one of the violin modes, which was at 1955 Hz. 

We took a transfer function of the PRCL servo, saw that the UGF was 300 Hz, and lowered it to ~180 Hz.  After later investigations, that high-ish UGF probably wasn't a problem.  Anyhow, we then took MICH servo transfer functions, and saw some very weird stuff. 

At frequencies where we had violin filter notches, we were seeing peaks in the transfer function, which came close to touching, or crossed the 0dB line!  We suspect that this may have something to do with the balancing of the drives to the optics, since we have PRCL driving PRM, but MICH driving BS and PRM.  What we did was move the violin filter notches into the LSC model.  There were already SUS filter banks in the LSC model (right side of the LSC screen).  In preparation for the DRMI, I have put the BS violin notches into the BS, PRM and SRM filter banks, as well as the PRM and SRM filters into all 3 banks.  Right now for PRMI, I have the BS and PRM notches (as well as the Vio3 notch) turned on in both BS and PRM.  All of the violin-related filters are turned off in the LSC filter bank inside the SUS models.  When we did this, the servo oscillations no longer are excited when we turn on the notches, and when we take a new transfer function, there are no longer weird peaks at the notch frequencies.  More meditation needs to be done.

Also, for the violin3 filter, Rana noted that at 1955 Hz, after we confirmed that the REFL 55 phase was set correctly (and we're using REFL 55 I&Q for PRMI locking), the I-phase had a response of 0.36, while the Q-phase had a response of 0.20.  I should be able to think about these numbers, and decide if the vio3 is for the BS or the PRM violin mode.

 MICHloopNotches_23Aug2013.pdf

the above series of Bode plots shows the MICH Open Loop Gain as the REFL55 phase is adjusted (PURPLE, ORANGE) with the notches in the SUS and then (RED) as the notches are moved to the LSC and made the same for all optics.

In other news, I have the parts that Jamie ordered to make a new 110 demod board, so that's one of my daytime activities now, so we can have both POP110 and AS110.

  9058   Fri Aug 23 14:19:28 2013 ranaUpdateSUSViolin filters moved to LSC, from SUS

  Just to rephrase somewhat:

  • We balance the BS & PRM actuators in the LSC Output matrix so that there is no MICH signal going into REFL_I @ 100 Hz.
  • REFL Phase is adjusted by driving PRM and minimizing Q/I to within ~1 deg (Q/I ratio of ~2%)
  • The REFL_I sensing matrix element is ~50x bigger than REFL_Q (in W/m)
  • When we turn ON the violin mode notch in the BS SUS, it changes the MICH drive into a purely PRCL drive at that frequency !!!
  • So, putting notches into the SUS screens is bad since it imbalances the drives.
  8620   Wed May 22 18:24:19 2013 JenneUpdateSUSViolin mode survey

Quote:

It was too embarassing to see that the actuation frequency was set at the violin mode frequency in order to avoid designing a new filter!?

 Ooops, definitely my bad.  I think I was the one who put in the PRM violin filter, so I should have recognized that frequency.  However, I couldn't think of a reason why violin mode filters should be in the LSC filter banks, since we usually put them in C1:SUS-optic_LSC filter banks. 

Anyhow, so that I don't make a mistake like that again, I was looking through all of the violin mode filters for all the optics, so I could write down the frequencies.  The result: confusion.

Violin filters in C1:SUS-optic_LSC filter banks:

The PRM's violin mode filter is set correctly to 627.75Hz:  elog 8533.

One of BS or SRM has probably been measured (presumably BS), since they have the same filters centered around 645Hz. 

Neither ITM has a violin filter.

The ETMs have violin filters in the 440's, which I assume was correct back in the MOS days, before 2010.

Vio2 filters in C1:SUS-optic_LSC filter banks:

PRM, SRM, BS, ITMX, ITMY:  Centered around 1285 Hz, which matches the violin notch frequencies in the BS and SRM.

ETMY:  Centered around 883.5Hz, which matches the old 440Hz frequency

ETMX:  Centered around 631Hz .  So, this could have been measured, but it was put into the wrong filter module.

 

Koji tells me that we don't really need to worry about all these violin filters unless they are required (as with the PRM and the obnoxious hum a few weeks ago), so I 'm not going to do any measuring / adjusting of these filters for now.

  4991   Tue Jul 19 20:36:08 2011 ranaUpdateComputersVirtualBox + Windows 7 on rossa

I installed Virtual Box on rossa. Then I put Windows 7 in there and am now installing Altium.

You can run Windows on rossa by just clicking Applications -> System Tools -> Virtual Box.

  7239   Tue Aug 21 00:35:25 2012 ranaSummaryIOOVisibilities and Chrome

MC and PMC vis:

MC REFL Unlocked    = 4.4

MC REFL Locked      = 0.67

 1 - Locked/Unlocked = 85%

 

PMC REFL Unlocked   = 0.270

PMC REFL Locked     = 0.013

 1 - Locked/Unlocked = 95%

I checked (by looking through recent trends) that the zero level is zero on both channels. I tried to do a proper mode scan, but we have lost the PSL fast channels during the upgrade sadly. Also, the DC signal for the PMC REFL needs some gain. Unlocked level should be more like 3-5 V.

Also used the instructions from this page to add Google's sources to rosalba's apt-get list and then installed Chrome.

Attachment 1: Untitled.pdf
Untitled.pdf
  13450   Wed Nov 22 17:52:25 2017 gautamUpdateOptimal ControlVisualizing cost functions

I've attempted to visualize the various components of the cost function in the way I've defined it for the current iteration of the Oplev optimal control loop design code. For each term in the cost function, the way the cost is computed depends on the ratio of the abscissa value to some threshold value (set by hand for now) - if this ratio is >1, the cost is the logarithm of the ratio, whereas if the ratio is <1, the cost is the square of the ratio. Continuity is enforced at the point at which this transition happens. I've plotted the cost function for some of the terms entering the code right now - indicated in dashed red lines are the approximate value of each of these costs for our current Oplev loop - the weights were chosen so that each of the costs were O(10) for the current controller, and the idea was that the optimizer could drive these down to hopefully O(1), but I've not yet gotten that to happen.

Based on the meeting yesterday, some possible ideas:

  1. For minimizing the control noise injection - we know the transfer function from the Oplev control signal coupling to MICH from measurements, and we also have a model for the seismic noise. So one term could be a weighted integral of (coupling - seismic) - the weight can give more importance to f>30Hz, and even more importance to f>100Hz. Right now, I don't have any suc frequency dependant requirement on the control signal.
  2. Try a simpler problem first - pendulum local damping. The position damping controller for example has fewer roots in the complex plane. Although it too has some B/R notches, which account for 16 complex roots, and hence, 32 parameters, so maybe this isn't really a simpler problem?
  3. How do we pick the number of excess poles compared to zeros in the overall transfer function? The OL loop low-pass filters are elliptic filters, which achieve the fastest transition between the passband and stopband, but for the Oplev loop roll-off, perhaps its better to have a just have some poles to roll off the HF response?
Attachment 1: globalCosts.pdf
globalCosts.pdf
  3206   Tue Jul 13 13:56:29 2010 GopalConfigurationOptic StacksVitol Material Properties

Viton flouroelastomers have somewhat variable material properties. The following parameters are being used for eigenfrequency analysis.

Young's Modulus: 72,500-87,000 psi (cite: http://www.row-inc.com/pfa.html) *Accurate for PFAs

Poisson's Ratio: 0.48-0.50 (cite: http://www.engineeringtoolbox.com/poissons-ratio-d_1224.html) *Accurate for rubber

Density: 1800 kg/m^3 (cite: http://physics.nist.gov/cgi-bin/Star/compos.pl?matno=275)

  3222   Wed Jul 14 19:00:56 2010 GopalConfigurationOptic StacksVitol Material Properties, REVISED

Quote:

Viton flouroelastomers have somewhat variable material properties. The following parameters are being used for eigenfrequency analysis.

Young's Modulus: 72,500-87,000 psi (cite: http://www.row-inc.com/pfa.html) *Accurate for PFAs

Poisson's Ratio: 0.48-0.50 (cite: http://www.engineeringtoolbox.com/poissons-ratio-d_1224.html) *Accurate for rubber

Density: 1800 kg/m^3 (cite: http://physics.nist.gov/cgi-bin/Star/compos.pl?matno=275)

 The Young's Modulus for PFA turned out to be orders of magnitude greater than for Viton. The revised values produced much more accurate results:

Young's Modulus for Viton-75: 1950 psi or 6.6 MPa

(Courtesy Row, Inc. and Steve Vass)

This website contains other details: http://www.row-inc.com/viton.html

  15084   Sun Dec 8 20:27:11 2019 ranaUpdateComputersViviana upgrade to Ubuntu 16

The IBM laptop at EX was running Ubuntu 14, so I allowed it to start upgrading itself to Ubuntu16 as it desired. After it is done, I will upgrade it to 18.04 LTS. We should have them all run LTS.

  1713   Thu Jul 2 05:27:12 2009 ClaraUpdatePEMVoltage Divider Oops

I tested the voltage dividers and was getting up to about 3V. I retested the mic w/o the voltage divider in place, and, lo and behold, I was able to generate about 70-75V (previously, I maxed out at 45V). I'm not 100% sure why this was, but it occurs to me that, before, the sounds I was generating were short in duration (loud claps, short yelps). This time, I tried yelling continuously into the microphone. So, probably, I simply wasn't seeing the real peak before on the scope because it was too short to pick up. I have corrected the voltage dividers (by replacing the 25.5 kOhm resistors, which were in parallel with the ADC, with 10 kOhm resistors, taking the voltage ratio to ~60:1) and tested them. I haven't been able to generate more than 1500 mV, so I think they are safe. (It's possible we would have been fine with the old setup, since I think it would be hard to get any noises as loud as I was making, but better safe than sorry, right?)

I'm attaching a diagram of the new-and-improved voltage dividers.

voltage_divider_diagram.png

  2288   Wed Nov 18 00:38:33 2009 ranaSummaryElectronicsVoltage Noise of the SR560's OUTPUTs (the back panel)

I've measured the voltage noise of the SR560's lead acid battery outputs; they're not so bad.

Steve ordered us some replacement lead-acid batteries for our battery powered pre-amps (SR560). In the unit he replaced, I measured the noise using the following setup:

SR560                              Busby Box

(+12V/GND) -------------AC Input      Out  ----------------   SR785

The SR785 was DC coupled and auto-ranged. The input noise of the SR785 was measured via 50 Ohm term to be at least 10x less than the SR560's noise at all frequencies.

sr560.png

Its clear that this measurement was spoiled by the low frequency noise of the Busby box below 10 Hz. Needs a better pre-amp.

  8778   Thu Jun 27 23:18:46 2013 jamieUpdateComputer Scripts / ProgramsWARNING: Matlab upgraded

Quote:

I moved the old matlab directory from /cvs/cds/caltech/apps/linux64/matlab_o to /cvs/cds/caltech/apps/linux64/matlab_oo

and moved the previously current matlab dir from /cvs/cds/caltech/apps/linux64/matlab to /cvs/cds/caltech/apps/linux64/matlab_o.

And have installed the new Matlab 2013a into /cvs/cds/caltech/apps/linux64/matlab.

Since I'm not sure how well the new Matlab/Simulink plays with the CDS RCG, I've left the old one and we can easily revert by renaming directories.

Be careful with this.  If Matlab starts re-saving models in a new file format that is unreadable by the RCG, then we won't be able to rebuild models until we do an svn revert.  Or the bigger danger, that the RCG *thinks* it reads the file and generates code that does something unexpected.

Of course this all may be an attempt to drive home the point that we need an RCG test suite.

  4816   Tue Jun 14 12:23:44 2011 Jamie, JoeUpdateCDSWE ARE ALL GREEN! LSC back up and running in new configuration.

After moving the c1lsc computer to 1X4, then connecting c1lsc to it's IO chassis in 1Y3 by a fiber PCIe extension cable, everything is back up and running and the status screen is all green.  c1lsc is now directly connected to c1sus via a short copper Dolphin cable.

After lunch we will do some more extensive testing of the system to make sure everything is working as expected.

  1313   Mon Feb 16 21:49:06 2009 Kakeru, RanaUpdateIOOWFS

We centerd the input of WFS QPD.

  6066   Sun Dec 4 13:56:54 2011 DenUpdateIOOWFS

Yesterday I locked the MC and left at 8 pm. Analyzing the data I saw that MC was locked all time from 8 pm to 12.30 am when it lost lock. Moreover there was no light on transmition and reflected screens at all. I went to the PSL and saw that no red light comes to the MC from PSL, only green. I took infrared sensos to track the laser light. Then I came back to control room to study the medm diagram of the PSL. Then I came back and saw that the laser beam goes to the MC! I returned to control room and saw light on the MC screens. Does someone do something parallel with me through ssh?

I enabled the auto locker and saw the MC locked for a couple of seconds. After that the WFS were turned on automatically and I saw that the signal of the OSEM local sensors of the MC mirrors began to increase. So the WFS master provides not good feedback signal. I thought that it is due to my recompilation of c1mcs with a fixed if-statement line. And may be if c1mcs workes without digital noise and c1ioo with it then there might occur some mismatches and the signal is corrupted. For this assumption I've recompiled c1mcs back to 1e-20 in the if-statement and so added the digital noise back that I saw in the dtt tools.

However, the problem was still present - WFS feedback signal crashed the MC lock. I open the WFS master window and disabled the output to MC. I can see that the C1:IOO-WFS1_PIT_INMON and other input channels have reasonable values 8 - 20 but the output continues to increase up to 1000000. The output was off so the MC stayed at lock. As for now I turned off WFS so no feedback is applied to MC mirros.

  6067   Sun Dec 4 23:49:38 2011 DenUpdateIOOWFS

Quote:

Yesterday I locked the MC and left at 8 pm. Analyzing the data I saw that MC was locked all time from 8 pm to 12.30 am when it lost lock. Moreover there was no light on transmition and reflected screens at all. I went to the PSL and saw that no red light comes to the MC from PSL, only green. I took infrared sensos to track the laser light. Then I came back to control room to study the medm diagram of the PSL. Then I came back and saw that the laser beam goes to the MC! I returned to control room and saw light on the MC screens. Does someone do something parallel with me through ssh?

I enabled the auto locker and saw the MC locked for a couple of seconds. After that the WFS were turned on automatically and I saw that the signal of the OSEM local sensors of the MC mirrors began to increase. So the WFS master provides not good feedback signal. I thought that it is due to my recompilation of c1mcs with a fixed if-statement line. And may be if c1mcs workes without digital noise and c1ioo with it then there might occur some mismatches and the signal is corrupted. For this assumption I've recompiled c1mcs back to 1e-20 in the if-statement and so added the digital noise back that I saw in the dtt tools.

However, the problem was still present - WFS feedback signal crashed the MC lock. I open the WFS master window and disabled the output to MC. I can see that the C1:IOO-WFS1_PIT_INMON and other input channels have reasonable values 8 - 20 but the output continues to increase up to 1000000. The output was off so the MC stayed at lock. As for now I turned off WFS so no feedback is applied to MC mirros.

With the help of Suresh we have adjusted optics near PMC and input to the MC on the PSL and in the black box where WFS are. Surprisingly, some optics near WFS was not attached to the table. But these mirrors are not used. One screw was near the hole but not screwed in. This mirror is used. Suresh could also rotate other screws. I thought that they must be attached to the table more rigidly.

Then we found that WFS output matrix is wrong and Suresh recalculated it. After that we've locked the MC using WFS. C1:IOO-MC_RFPD_DCMON is 0.7-0.8. 

We also recompiled and reinstalled C1MCS and C1IOO with fixed if-statement and again saw how MC_F curve moves down. WFS error signals are also improved. But still some more work on output matrix is needed.

  15266   Wed Mar 11 18:12:53 2020 gautamSummaryPSLWFS Demod board modifications

[koji, gautam]

Attachment #1 shows the relevant parts of the schematic of the WFS demod board (not whitening board). 

  • The basic problem was that the switchable gain channels were not accounted for in the Acromag channel list 😒.
  • What this meant was that the DC gain was set to the default x100 (since the two DG211s that provide the switchable x10 and x1 gain options had their control logic pins pulled up to +5V because these pins weren't connected to any sinking BIO channel).
  • Rather than set up new connections to Acromags inside the chassis (though we have plenty of spares), Koji and I decided to make these fixed to x1 gain.
  • The actual fix was implemented as shown in the annotated schematic. There are some pictures 📷 of the PCB in the DCC entry linked above.
  • Amusingly, this board will now require a sourcing BIO unit if we want to still have the capability of switching gains.

Before removing the boards from the eurocrate: 

  • I dialled down the Kepco HV supplies
  • disconnected all the cabling to these boards after noting down cable numbers etc.

After Koji effected the fix, the boards were re-installed, HV supplies were dialled back up to nominal voltage/currents, and the PMC/IMC were re-locked. The WFS DC channels now no longer saturate even when the IMC is unlocked 👏 👏 . I leave it to Yehonathan / Jon to calibrate these EPICS channels into physical units of mW of power. We should also fix the MEDM screen and remove the un-necessary EPICS channels.

Later in the evening, I took advantage of the non-saturated readbacks to center the beams better on the WFS heads. Then, with the WFS servos disabled, I manually aligned the IMC mirrors till REFLDC was minimized. Then I centered the beam on the MC2 transmission QPD (looking at individual quadrants), and set the WFS1/2 RF offsets and MC2 Trans QPD offsets in this condition.

Quote:

WFS DC channels are saturating when the IMC is unlocked.