  10175   Thu Jul 10 15:27:09 2014 HarryUpdateGeneralCoupling telescope design

 I designed this telescope to couple the 1064 NPRO into the PM980 fiber, using lenses from the Thorlabs LSB04-C kit.

The collimator is a CFC-2X-C, which has a variable focus length (2.0, 4.6, 7.5, and 11.0 mm) which gives corresponding angles of divergence of 0.298, 0.130, 0.79, and 0.054 degrees by the formula theta = (180*MFD) / (pi*f).

Then, using these values I calculated the spot size of a beam collimated by the CFC-2X-C, using f = w / tan(theta) where w is the spot size. This gave a value of 10.4 um.

I used this value (10.4 um) as a target waist for the telescope system, with the NPRO waist as a seed, at the origin.

telescopeBeamProfile.png

It consists of two lenses, one located at Z = 77cm f = 50cm, and the second located at Z = 85.88 cm f = 2.54cm, which yields a waist of 13um at Z = 88.32cm, (which is where the collimator would go) for an overlap of .974.

Note that the telescope is so far "downrange" from the NPRO waist because it's a virtual waist, and the NPRO itself is located at about Z = 73cm.

Find attached the alm code used.

Attachment 2: telescope.zip
  10176   Thu Jul 10 15:57:00 2014 SteveUpdateSUSflow bench is turned off

Flow bench effect on oplev error signal     is here.

I turned off the south X-end flow bench.

  10179   Thu Jul 10 18:25:18 2014 KojiUpdateGeneralCoupling telescope design

CFC-2X-C has a FIXED focal length of 2mm, but the collimator lens position is adjustable.
I'm not yet sure this affects your calculation or not as what you need is an approximate mode calculation;
once you couple the any amount of the beam into the fiber, you can actually measure it at the output of the fiber with a collimator attached.

  10180   Thu Jul 10 19:37:54 2014 ManasaUpdateGeneralPM980 fiber tested OK

[Harry, Manasa]

This is the update from yesterday that Harry missed to elog.

We pulled out the first spool of the PM980 fiber yesterday and checked it using the illuminator at the SP table. Harry will be using this for all his tests and characterisation of the fiber.

  10181   Fri Jul 11 08:11:56 2014 SteveUpdateCDSETMX needs help
Attachment 1: ETMX.png
ETMX.png
  10182   Fri Jul 11 09:41:43 2014 not KojiUpdateGeneralCoupling telescope design

Quote:

CFC-2X-C has a FIXED focal length of 2mm, but the collimator lens position is adjustable.
I'm not yet sure this affects your calculation or not as what you need is an approximate mode calculation;
once you couple the any amount of the beam into the fiber, you can actually measure it at the output of the fiber with a collimator attached.

 I don't believe it had any effect, as all the calculations gave me the same target waist.

  10183   Fri Jul 11 11:51:03 2014 NichinUpdateElectronicsPDFR: List of DC transimpedances

The following values are going to be entered in the param_[PDname].yml file for each PD. I am elogging them for future reference.

I got the values from combing schematics and old Elog entries. Please let me know if you believe the values are different.

  • AS55: 66.2 ohms
  • REFL11 : 66.2 ohms 
  • REFL33 : 50.2 ohms
  • REFL55: 50 ohms (Elog 4605)
  • REFL165: 50.2 ohms
  • POY11: 66.2 ohms
  • POX11: 50.2 ohms
  • REF (NF1611): 700 ohms
  • POP22: ?? (This is currently a Thorlab BBPD )
  10185   Fri Jul 11 15:29:26 2014 HarryUpdateGeneralTelescope With Collimator

 I used a la mode to make a design for the coupling telescope with a 3.3um target waist, that included the collimator in the overall design. The plot is below, and code is attached.

The components are as follows:

 label         z (m)     type    parameters         

   -----         -----     ----    ----------         

    lens1         0.7681    lens    focalLength: 0.5000

    lens2         0.8588    lens    focalLength: 0.0350

    collimator    0.8832    lens    focalLength: 0.0020

 

The z coordinates are as measured from the beam waist of the NPRO (the figure on the far left  of the plot).

Moving forward, this setup will be used to couple the NPRO (more specifically, the AUX lasers) light into the SM 980 fibers, as well as to help characterize the fibers themselves.

Ultimately, this will be a key component in the Frequency Offset Locking project that Akhil and I are working on, as it will transport the AUX light to the PSL, where the two beams will be beaten with each other to generate the input signal to the PID control loop, which will actuate the temperature servos of the AUX lasers.

Attachment 1: telescope_2.zip
Attachment 2: telescopeWCollimator.png
telescopeWCollimator.png
  10186   Fri Jul 11 17:49:12 2014 NichinUpdateElectronicsNew Prologix GPIB-Ethernet controller

I have configured a NEW Prologix GPIB-Ethernet controller to use with HP8591E Spectrum analyzer that sits right next to the control room computers.

Static IP: 192.168.113.109

Mask: 255.255.255.0

Gateway: 192.168.113.2

I have no clue how to give it a name like "something.martian" and to update the martian host table (Somebody please help!!)

 

  10187   Fri Jul 11 20:05:34 2014 JenneUpdateLSCQPD dark noise checkout

The other day, I took spectra of the Yend transmission QPD dark noise for several different configurations of the transimpedance gain and the whitening gains. 

I also calculated with Optickle the expected sensitivity vs. CARM offset for the sqrtInv CARM sensor. 

I put these things together to get a plot of transmission QPD noise vs. CARM offset, for a chosen set of analog settings.


Measurement of Yend transmission QPD dark noise

Since EricQ had already checked out the whitening filters (see elog 9637 and elog 9642), I didn't check on them.  I just left them (the analog whitening, and the digital antiwhitening filters) on. 

First, I checked the noise vs. transimpedance gain. There are a few too many settings to put them all on one plot, so I have them sorted by the original transimpedance:  0.5 kOhms vs 5 kOhms.  It's a little tricky to see, but all of the spectra that begin with the 5k transimpedance have a little extra noise around 10 Hz, although I don't know why.  In the legend I have made note of what the settings were.  x1 x1 is my representation of the "inactive" setting.

TransQPDY_darkNoise_vs_TransimpedanceGain_9July2014.pdf

I then looked at the noise with different whitening gain slider settings.  All but one of the traces are the 20 kOhm setting.  

TransQPDY_darkNoise_vs_WhiteningGain_9July2014.pdf

These .xml files are in /users/jenne/Arms/TransQPDnoise_July2014/


Calculation of inverse sqrt transmission sensitivity

I used Optickle to give me the power transmitted through the ETMs.  I first find the transmission in the FPMI case.  I use that to normalize the full PRFPMI transmission, so that the output units are the same as our C1:LSC-TR[x,y]_OUT units. 

I take the square root of the transmitted power (sum of transmissions from each arm) at each CARM offset point, add 1e-3 as we do in the front end model to prevent divide-by-zero problems, and then take the inverse.

I find the slope by taking the difference in power between adjacent points, divided by the CARM offset difference between those points. 

In this plot, I have taken the absolute value of the sensitivity, just for kicks.  I also display an arbitrarily scaled version of the log of the transmitted power, so that we can see that the highest sensitivity is at half the maximum power.

sqrtTrans_sensitivity.png


 Calculate the QPD dark noise in terms of meters

Finally, I put it all together, and find the dark noise of the QPD in terms of meters.  Since the spectra were measured in units where the single-arm transmission is unity, the already match the units that I used to calculate the sqrtInv sensitivity. 

I take the spectra of the QPD dark noise for the 20 kOhm case, and multiply it by the sensitivity calibration number at several different CARM offsets.  As we expect, the noise is the best at half-max transmission, where the sensitivity is maximal. 

sqrtTrans_Noise_vs_CARMoffset.png

  10189   Fri Jul 11 22:28:34 2014 ChrisUpdateCDSc1auxex "Unknown Host"

Quote:

c1auxex has forgotten who it is.  Slow sliders for the QPD head were not responding, so I did a soft reboot from telnet. The machine didn't come back, so I plugged the RJ45-DB9 cable into the machine and looked at it through a minicom session.  When I key the crate, it gives me an error that it can't load a file, with the error code 0x320001.  Looking that up on a List of VxWorks error codes, I see that it is:    S_hostLib_UNKNOWN_HOST (3276801 or 0x320001)

I'm not sure how this happened.  I unplugged and replugged in the ethernet cable on the computer, but that didn't help.  Rana is going in to wiggle the other end of the ethernet cable, in case that's the problem.  EDIT:  Replacing the ethernet cable did not help.

Former elogs that are useful:  10025, 10015

EDIT:  The actual error message is:

boot device          : ei                                                      
processor number     : 0                                                       
host name            : chiara                                                  
file name            : /cvs/cds/vw/mv162-262-16M/vxWorks                       
inet on ethernet (e) : 192.168.113.59:ffffff00                                 
host inet (h)        : 192.168.113.104                                         
user (u)             : controls                                                
flags (f)            : 0x0                                                     
target name (tn)     : c1auxex                                                 
startup script (s)   : /cvs/cds/caltech/target/c1auxex/startup.cmd             
                                                                               
Attaching network interface ei0... done.                                       
Attaching network interface lo0... done.                                       
Loading...                                                                     
Error loading file: errno = 0x320001.                                          
Can't load boot file!! 

 We fixed this problem (at least for now) by adding c1auxex to the /etc/hosts file on chiara (following a hint from this page). The DNS setup might be the culprit here.

  10190   Sun Jul 13 11:37:36 2014 JamieUpdateElectronicsNew Prologix GPIB-Ethernet controller

Quote:

I have configured a NEW Prologix GPIB-Ethernet controller to use with HP8591E Spectrum analyzer that sits right next to the control room computers.

Static IP: 192.168.113.109

Mask: 255.255.255.0

Gateway: 192.168.113.2

I have no clue how to give it a name like "something.martian" and to update the martian host table (Somebody please help!!) 

The instructions for adding a name to the martian DNS table are in the wiki page that I pointed you to:

https://wiki-40m.ligo.caltech.edu/Martian_Host_Table

  10191   Sun Jul 13 17:06:35 2014 AndresUpdate40m Xend Table upgradeXarm Table Upgrade Calculation and Diagrams of possible new table layout

 Current Mode Matching and Gouy Phase Between Steering Mirrors

We found in 40m elog ID 3330 ( http://nodus.ligo.caltech.edu:8080/40m/3330a documentation done by Kiwamu, where he measured the waist of the green. The waist of the green is about 35µm. Using a la mode, I was able to calculate the current mode matching, and the Gouy phase between the steering mirrors. In a la mode, I used the optical distances,which is just the distance measured times its index of refraction. I contacted someone from ThorLabs (which is the company that bought Optics For Research), and that person told that the Faraday IO-5-532-LP has a Terbium Gallium Garnet crystal of a length of 7mm and its index of refraction is 1.95. The current mode matching is 0.9343, and the current Gouy phase between steering mirrors is 0.2023 degrees. On Monday, Nick and I are planning to measure the actual mode matching. The attached below is the current X-arm optical layout. 

 

 

Calculation For the New Optical Layout

 

Since the current Gouy phase between the steering mirror is essentially zero, we need to find a way how to increase the Gouy Phase. We tried to add two more lenses after the second steering mirror, and we found that increasing the Gouy phase result in a dramatically decrease in mode matching. For instance, a Gouy phase of about 50 degrees results in a mode matching of about .2, which is awful. We removed the first lens after the faraday, and we added two more mirrors and two more lenses after the second steering mirror. I modified the photo that I took and I place where the new lenses and new mirrors should go as shown in the second pictures attached below. Using a la mode, we found the following solution:

 label                         z (m)            type                       parameters         

 -----                          -----              ----                        ----------         

 lens 1                       0.0800          lens                      focalLength: 0.1000

 First mirror              0.1550          flat mirror            none:            

 Second mirror         0.2800          flat mirror            none:            

 lens 2                      0.4275           lens                      focalLength: Inf   

 lens 3                     0.6549            lens                      focalLength: 0.3000

lens 4                      0.8968            lens                      focalLength: -0.250

Third mirror           1.0675            flat mirror            none:            

Fourth mirror         1.4183            flat mirror            none:            

lens 5                      1.6384            lens                     focalLength: -0.100

Fifth mirror            1.7351            flat mirror           none:            

Sixth mirror           2.0859            flat mirror           none:            

lens 6                     2.1621            lens                     focalLength: 0.6000

ETM                      2.7407            lens                    focalLength: -129.7

ITM                       40.5307          flat mirror          none:             

The mode matching is 0.9786. The different Gouy phase different between Third Mirror and Fourth Mirror is 69.59 degrees, Gouy Phase between Fourth and Fifth 18.80 degrees, Gouy phase between Fifth and Sixth mirrors is 1.28 degrees, Gouy phase between Third and Fifth 88.38 degrees, and the Gouy phase between Fourth and Sixth is 20.08 degrees. Bellow attached the a la Mode code and the Plots.

 

 

Plan for this week

I don't  think we have the lenses that we need for this new setup. Mostly, we will need to order the lenses on Monday. As I mention, Nick and I are going to measure the actual mode matching on Monday. If everything look good, then we will move on and do the Upgrade.

 

Attachment 1: CurrentOpticalLayout.png
CurrentOpticalLayout.png
Attachment 2: NewSetUp.PNG
NewSetUp.PNG
Attachment 3: AlaModeSolutionplots.png
AlaModeSolutionplots.png
Attachment 4: EntireScaleRangeAlaModeSolution.png
EntireScaleRangeAlaModeSolution.png
Attachment 5: NewXarmOptimizationFromFaraday.m
close all
clear all
% In this code we are using a la mode to optimatize the mode matching and
% to optimatize the Gouy phase between mirror 1 and mirror 2. All the units
% are in meter

w0=(50*1e-6)/sqrt(2); % The Waist of the laser measured after SHG
z0_laser=-0.0083; % position measured where the waist is located 
lamb= 532*10^-9; % wavelength of green light in mm
lFaraday=.0638; % Length of the faraday
... 209 more lines ...
  10192   Mon Jul 14 12:49:07 2014 NichinUpdateElectronicsNew Prologix GPIB-Ethernet controller

Quote:

Quote:

I have configured a NEW Prologix GPIB-Ethernet controller to use with HP8591E Spectrum analyzer that sits right next to the control room computers.

Static IP: 192.168.113.109

Mask: 255.255.255.0

Gateway: 192.168.113.2

I have no clue how to give it a name like "something.martian" and to update the martian host table (Somebody please help!!) 

The instructions for adding a name to the martian DNS table are in the wiki page that I pointed you to:

https://wiki-40m.ligo.caltech.edu/Martian_Host_Table

The instructions at https://wiki-40m.ligo.caltech.edu/Martian_Host_Table   are outdated!

The name server configuration is currently at /etc/bind/zones/martian.db [ source: elog:10067 ]

 

Anyway, I need superuser access to edit the files, which I don't have. Even if I did know the password, I don't think it's a good idea for me to be messing around. So any of the 40m folks please update the martian table to include:

santuzza.martian  192.168.113.109

 

  10195   Mon Jul 14 16:19:41 2014 AndresUpdate40m Xend Table upgradeTook the measurement for the Mode Matching

 Nick and I measured the reflected power of the green light in locked and unlocked. I'm working on the calculation of the mode matching. Tonight, I'll be posted my calculation I'm still working on it.

JCD:  Andres forgot to mention that they closed the PSL shutter, so that they could look at the green light that is reflected off the harmonic separator toward the IR trans path.  Also, the Xarm (and the Yarm) were aligned to IR using the ASS, and then ASX was used to align the green beam to the cavity.

  10196   Mon Jul 14 16:51:07 2014 NichinUpdateElectronicsMartian table updated, Named server restarted

 [Nichin, Jenne]

The martian lookup tables are located at /etc/bind/zones/martian.db  and etc/bind/zones/rev.113.168.192.in-addr.arpa

Jenne updated these to include santuzza.martian  192.168.113.109

 

 

The method to restart named server given at  https://wiki-40m.ligo.caltech.edu/Martian_Host_Table  also does not work.

I restarted it using  >sudo /etc/init.d/bind9 restart

The named server is now updated and works fine. :)  I will update the 40m wiki now.

  10197   Mon Jul 14 17:51:34 2014 NichinUpdateComputer Scripts / ProgramsMEDM for PDFR system

14.pngA

 Successfully completed the rudimentary GUI for PDFR system. (users/nichin/PDFR)

Pressing any of the buttons above runs the script that does the following:

1) Change RF mux channel to the required one.

2) Frequency sweep on the network analyzer. The common sweep parameters are in a file named param_NWAG4395A.yml . PD specific parameters are in param_[PD name].yml in their respective folders

3) The transimpedance is calculated and the plot is saved as PDF in the respective folder for the PD. Each set of measurement data and plot is in a timestamped subfolder.

Further work:

To take transimpedance readings for each PD and create a canonical set of data that can be used to compare with data obtained for every measurement run.

  10198   Tue Jul 15 00:16:52 2014 JenneUpdateSUSPRM significantly pitched

I'm starting to lock for the night, and I noticed that PRM is very, very pitched.  Why?  The PRM pitch slider is 5 full integer units higher than the backup (and the backup value is about where I like it, around -0.2).

I am not aware of any scripts that touch the PRM slider values.  The PRM ASS (which I haven't used in ages) offloads the biases to the SUS screen fast channels, so even if someone turned that on and then saved the values, it wouldn't leave the PRM so very, very misaligned.

I have restored it, and relocked the PRMI, so all is well, but it's very weird to have found it so misaligned.

  10200   Tue Jul 15 01:41:43 2014 JenneUpdateLSCData for DARM on sqrtInv investigation

I took some data tonight for a quick look at what combinations of DC signals might be good to use for DARM, as an alternative to ALS before we're ready for RF.

I had the arms locked with ALS, PRMI with REFL33, and tried to move the CARM offset between plus and minus 1.  The PRMI wasn't holding lock closer than about -0.3 or +0.6, so that is also a problem.  Also, I realized just now that I have left the beam dumps in front of the transmission QPDs, so I had prevented any switching of the trans PD source.  This means that all of my data for C1:LSC-TR[x,y]_OUT_DQ is taken with the Thorlabs PDs, which is fine, although they saturate around arm powers of 4 ever since my analog gain increase on the whitening board.  Anyhow, the IFO didn't hold lock for much beyond then anyway, so I didn't miss out on much.  I need to remember to remove the dumps though!!

Self:  Good stuff should be between 12:50am - 1:09am.  One set of data was ./getdata -s 1089445700 -d 30 -c C1:LSC-TRX_OUT_DQ C1:LSC-TRY_OUT_DQ C1:LSC-CARM_IN1_DQ C1:LSC-PRCL_IN1_DQ

  10201   Tue Jul 15 11:04:05 2014 SteveUpdatesafetysafety glasses must be worn all times!

This is the third safety glasses that I found laying around in the IFO lab lately. Safety glasses must be worn ALL times in the IFO room!

This rule is essential for your protection! Please do not enter if you can not put up with this regulation!

 

 

Attachment 1: safetyglasses.jpg
safetyglasses.jpg
  10202   Tue Jul 15 12:36:17 2014 NichinUpdateElectronicsRF cables rerouted

Quote:

 

I have not connected them to the RF switch yet. ( until I figure out how to get both the switches working properly)

 I went into the lab and connected the RF cables to the Mux. Will take measurements for each PD henceforth.

  10203   Tue Jul 15 17:34:01 2014 manasaUpdatePSLproposing AOM re-alignment

I am going to tweak the alignment of the beam into the AOM (before the PMC) tomorrow morning. If anybody has any objections to this, please raise a red flag.

Proposed alignment procedure:

1. Reduce PSL power to say 10% 

2.  Since the AOM is not on any sort of a mechanical stage, I will have to just play around carefully until I see a maximum power rejection into first order.

I am assuming that moving the AOM is not going to affect the input pointing because all these activities are happening before the PMC. So as long as I have the output beam from the AOM aligned to the PMC at the end, everyone should be happy.

  10204   Tue Jul 15 18:26:40 2014 HarryUpdateGeneralBeam Waist, Telescope, and Fiber Coupling

 Goal

To design an optical setup (telescope / lens) to couple 1064nm NPRO light into PANDA PM980 fibers in order to characterize the fibers for further use in the frequency offset locking setup.

Design

 fiberTestingSetup.png

Calculations

 The beam waist of the NPRO was determined as 233um 6cm in front of the NPRO. This was used as the seed waist in ALM.

The numerical aperture of the fiber was given as 0.12, which allowed me to calculate the maximum angle of light it would accept, with respect to the optical axis, as NA = sin(theta) where theta is that angle.

Given that the coupler has a focal length of 2mm, I used the formula r = f * tan(theta), to yield a "target waist" for efficient coupling into the fiber. This ended up being 241.7um.

Since there was not a huge difference between the natural beam width of the NPRO and our target waist, I had no need for multiple lenses.

I used 230um as a target waist for a la mode, to leave myself some room for error while coupling. This process gave me a beam profile with a lens (f=0.25m), and a target waist of 231um, located 38.60cm from the coupling lens

I have attached ALM code, as well as the beam profile image. Note that the profile takes zero to be the location of the NPRO waist.

fiberTestingCouplingDesign.png

Next Steps

 

After this setup is assembled, and light is coupled into the fibers, we will use it to run various tests to the fiber, for further use in FOL. First of all, we wish to measure the coupling efficiency, which is the purpose of the powermeter in the above schematic. We will measure optical power before and after the fibers, hoping for at least ~%60 coupling. Next is the polarization extinction ratio measurement, for which we will control the input polarization to the fibers, and then measure what proportion of that polarization remains at the output of the fiber. 

Attachment 3: fiberTesting.zip
  10205   Tue Jul 15 18:39:04 2014 HarryUpdateGeneralWeekly Plan (7.16.14)

 The Past Week

 

Attempted to design coupling telescope, turned out waist measurement was still off. Took another waist measurement, this time more reasonable.

Used recent waist measurement to actually design a coupling system to couple NPRO light into Panda PM980 fibers (see recent elog)

The Next Week

Assemble fiber coupling system

Measure coupling efficiency, ensure it's at least 60%

Begin measuring Polarization Extinction ratio

Materials

 

PLCX lens with f = 0.25m ------> status: here

Fiber Coupled Powermeter//PD ------> status: unknown (have any laying around?)

Quarter Wave Plate, Polarizing Beamsplitter, Photodiodes ------> status: here

other components from original razorblade measurement  setup

  10206   Tue Jul 15 21:43:28 2014 JenneUpdateLSCQPDs back

I removed the dumps in front of the trans QPDs.  The Yend QPD needed re-normalization, so I did that.

  10207   Tue Jul 15 22:23:51 2014 AndresUpdate40m Xend Table upgradeScan the Xarm for the mode matching

 Nick and I with the help of Jenne scan the green light when the cavity is unlocked. Nick placed a Beam dump on the IR so that we can just scan the green, but it was removed as soon as we finished with the measurement. I'm working on the calculation, and i'll be posted solution tonight.

  10208   Wed Jul 16 01:04:09 2014 JenneUpdateLSCData for DARM on sqrtInv investigation

I realized while I was looking at last night's data that I had been doing CARM sweeps, when really I wanted to be doing DARM sweeps.  I took a few sets of data of DARM sweeps while locked on ALSdiff.  However, Rana pointed out that comparing ALSdiff to TRX-TRY isn't exactly a fair comparison while I'm locked on ALSdiff, since it's an in-loop signal, so it looks artificially quiet. 

Anyhow, I may consider transitioning DARM over to AS55 temporarily so that I can look at both as out-of-loop sensors. 

Also, so that I can try locking DARM on DC transmission, I have added 2 more columns to the LSC input matrix (now we're at 32!), for TRX and TRY.  We already had sqrt inverse versions of these signals, but the plain TRX and TRY were only available as normalization signals before.  Since Koji put in the facility to sqrt or not the normalization signals, I can now try:

Option 1:  ( TRX - TRY ) / (TRX + TRY)

Option 2:  ( TRX - TRY ) / sqrt( TRX + TRY )

DARM does not yet have the facility to normalize one signal (DC transmission) and not another (ALS diff), so I may need to include that soon.  For tonight, I'm going to try just changing matrix elements with ezcastep.

Since I changed the c1lsc.mdl model, I compiled it, restarted the model, and checked the model in.  I have also added these 2 columns to the AUX_ERR sub-screen for the LSC input matrix.  I have not changed the LSC overview screen.

  10212   Wed Jul 16 01:46:41 2014 NichinUpdateElectronicsTest run of PDFR system

A test run was conducted on the PDFR system last afternoon and transimpedance plots were generated for 6 of the PDs. The laser was shut down after the test run.

I have not verified (yet) if the transimpedance values indicated by the plots are correct or not. The values mostly look INCORRECT. But the peaks are exactly where they need to be. *phew!*

Reasons: Incorrect calibration, Light other than from the PDFR system fibers on the PDs

Will have to work on debugging all this.

Attachment 1: PDFR_testRun_15-07-2014.pdf
PDFR_testRun_15-07-2014.pdf PDFR_testRun_15-07-2014.pdf PDFR_testRun_15-07-2014.pdf PDFR_testRun_15-07-2014.pdf PDFR_testRun_15-07-2014.pdf PDFR_testRun_15-07-2014.pdf
  10213   Wed Jul 16 01:54:25 2014 NichinUpdateGeneralWork plan for next week

1) Debugging transimpedance calculations in the PDFR

Requires presence of an expert whenever I get inside the lab to take DC measurements or check the illuminating fibers.

2) Creating and incorporating canonical data plots with every measurement of PDFR.

3) Transfer function fitting for transimpedance

4) Improve the Spectrum analyzer scan scripts as mentioned in my elog.

  10214   Wed Jul 16 02:22:10 2014 KojiUpdateElectronicsTest run of PDFR system

Log-log ... 

  10215   Wed Jul 16 07:51:52 2014 AkhilUpdateGeneralWeekly Update

Work Done:

  • Solved all the timing issues pertaining to the R Pi and the FC.
  • Took all the measurements for complete characterization of the frequency counter(Phase Plots to follow shortly).
  • Finished  installation of the FC on the martian and created a channel  for the FC frequencies(will be tested in this week).

Plans for this Week:

  • Testing of the EPICS soft IOC created for the FC as a channel access server and hence completing the installation of the FC.
  • Placing the FC inside the lab( plan discussed in this elog: http://nodus.ligo.caltech.edu:8080/40m/10163) with proper supervision.
  • Characterization of the temperature actuator.

Inside the 40m Lab:        

I will need to be inside the lab to place the FC . This will be done in the morning session (on thursday) with supervision of Manasa and Steve(if required).

 

 

 

 

 

 

 

 

  10216   Wed Jul 16 15:26:48 2014 SteveUpdatePSLPSL Innolight controller fan is dying

Quote:

Quote:

Also, while I was working on the PSL table, I heard noise that sounded like a bearing rolling around.  I suspected the HEPAs, since the one on the north east corner of the table has a problem when it's turned up high (we've known about this for a long time), however turning off the HEPAs didn't affect the noise.  The noise is strongest near the back of the PSL controller on the shelf above the table, and the PSL controller box is vibrating.  So, I suspect that the fan on the PSL controller box is about to give out.

EDIT:  To clarify, I mean the Innolight's controller.

 The bearing is chirping in the back of the 2W Innolight laser controller. It is loud enough to hear it. I placed 4 soft  rubber feet under the controller to avoid shaking other things on self.

The HEPA filter bearing becomes noisy at 50V

 Keep it at 20V for low noise

 The fan is dying. It is changing speed erratically and stops for short time periods. It is very likely to stop rotating soon. It will halt all operations in the lab. We can not see the PMC-T power because Manasa is working on

 AOM alignment.

 

  10217   Wed Jul 16 17:06:41 2014 NichinUpdateComputer Scripts / ProgramsHP8591E spectrum analyzer remote scan

Updated script does the following:

1) Gets the highest 2 peaks

2) Puts a marker on the peaks. Now it looks very similar to the spectrum analyzer display.

3) The refresh rate is still 3 seconds. It might become better if the analyzer was hooked up to a wired martian LAN port rather than the wireless module I am using now.

PFA a sample pdf

Attachment 1: HP8591E_View.pdf
HP8591E_View.pdf
  10218   Wed Jul 16 17:34:11 2014 HarryUpdateGeneralFiber Coupled

 Purpose

To couple the spare NPRO into our Panda PM980 fibers, in order to carry out tests to characterize the fibers, in order to use them in FOL.

Design

 Manasa and I spent this morning building the setup to couple NPRO light into the fibers. We used two steering mirrors to precisely guide the beam into the coupler (collimator).

We also attached the lens to a moveable stage (in the z axis), so the setup could be fine tuned to put the beam waist precisely at the photodiode.

The fiber was attached to a fiber-coupled powermeter, so I would be able to tell the coupling efficiency.

fiberTestCouplingSchematic.png

Methods 

During alignment, the NPRO was operating at 1.0 amps, roughly half of nominal current (2.1A).

I first placed the coupler at the distance that I believed the target waist of 231um to be.

Using the steering mirrors and the stage that holds the couple, I aligned the axes of the coupler and the beam.

Finally, I used the variable stage that the lens is attached to to fine tune the location of the target waist.

Results

Once I was getting readings on the powermeter (~0.5nW), the laser was turned up to nominal current of 2.1A.

At this point, I and getting 120nW through the fiber.

While far from "good" coupling, it is enough to start measuring some fiber characteristics.

Moving Forward

Tomorrow, I hope to borrow the beam profiler once again so as to measure the fiber mode.

Beyond this, I will be taking further measurements of the Polarization Extinction Ratio, the Frequency Noise within the fiber, and the effects of a temperature gradient upon the fiber.

Once these measurements are completed, the fiber will have been characterized, and will be ready for implementation in FOL.

  10220   Wed Jul 16 21:23:35 2014 ManasaUpdatePSLPSL Innolight controller fan is dying

[Koji, Manasa]

The air flow from the dying fan was kinda weak and we decided to give a help with an external fan.

Koji brought a fan taken from a junk found at EE shop in W.Bridge.
The fan has been tied to the cage of the existing fan using cable ties to provide air circulation.
So even if the existing one dies anytime, we still don't super-heat anything.
The power supply for the fan rests next to the controller. 

The air from the fan ventilation was hot, and now with the additional fan this hot air is actually sucked
out with stronger flow. So this is relieving for now.

Attachment 1: IMG_1659.JPG
IMG_1659.JPG
Attachment 2: IMG_1658.JPG
IMG_1658.JPG
  10221   Wed Jul 16 21:24:41 2014 ReetikaUpdateElectronicsVCO Driver inside 40m

  

I found the VCO driver, that Rana asked me to locate, inside the 40m. I already have one VCO from PSL lab. Now, I have kept both of them inside the 40m lab(one on the cart in the side of the Y-arm and the other near the X-arm electronics table).

  10224   Thu Jul 17 00:38:30 2014 JenneUpdateLSCRIN in arm transmission

[Rana, Jenne]

We had a look at the RIN of the transmission signals TRX and TRY, when the arms were individually locked on IR.  If the intensity noise is very bad, then the transmission signals aren't really a good option to use for locking.  So, if the RIN is bad, we need to work on our intensity stabilization. 

We need to understand what the situation is with the AOM, and why it isn't working as expected, so that we can reinstall it.  We also need to decide if we're going to use the SR560 setup, or if the Chas ISS is sufficiently characterized for us to use. 

The RIN is certainly bad.  Also, I don't know why the Xarm's RIN is worse between 10 Hz and a few hundred Hz than the Yarm.

TransRIN.pdf

  10225   Thu Jul 17 01:24:35 2014 ranaUpdateIOOMC / EOM Stability Mystery Solved!

MC loop is near unstable. Common gain reduced. Needs more loop tuning.

We've often seen that the MC gets into a state where it keeps losing lock and the EOM drive shows a large RMS. We've usually been looking at the noise spectrum to diagnose this.

Tonight we finally just measured the OLG. The attached plot shows the loop gain measured with the 4395 on the MC servo board

Although the phase margin is a healthy 45 degrees, its close to instability at 1 MHz. For this plot, I reduced the gain by 3 dB and now the margin is ~7 dB. So usually its pretty close to unstable and at least its always making a noise peak.

That whole TF above a few hundred kHz is weird. We should tune out whatever makes it so flat and also remove the resonance that makes the 1 MHz peak; maybe its from some post mixer low pass?

Anyone interested in helping in the investigation ought to measure the TF of the MC demod board, the MC servo board, and the FSS box.

Silver lining: if we fix this loop shape, we might be able to have a much more stable IMC and IFO.

Attachment 1: MC_OLG.pdf
MC_OLG.pdf
  10226   Thu Jul 17 02:57:32 2014 AndresUpdate40m Xend Table upgradeFInish Calculation on Current X-arm mode Matching

Data and Calculation for the Xarm Current Mode Matching

Two days ago, Nick, Jenne, and I took a measurement for the Green Transmission for the X-arm. I took the data and I analyzed it. The first figure attached below is the raw data plotted. I used the function findpeaks in Matlab, and I found all the peaks. Then, by taking close look at the plot, I chose two peaks as shown in the second figure attached below. I took the ratio of the TEM00 and the High order mode, and I average them. This gave me a Mode Matching of 0.9215, which this value is pretty close to the value that I predicted by using a la Mode in http://nodus.ligo.caltech.edu:8080/40m/10191, which is 0.9343. Nick and I measured the reflected power when the cavity is unlocked and when the cavity is locked, so we measured the PreflUnLocked=52+1µW and PreflOnLocked=16+2µW and the backgroundNoise=0.761µW. Using this information we calculated  Prefl/Pin=0.297. Now, since Prefl/Pin=|Eref/Ein|2, we looked at the electric fields component by using the reflectivity of the mirror we calculated 0.67. The number doesn't agree, but this is because we didn't take into account the losses when making this calculation. I'm working in the calculation that will include the losses.

Today, Nick and I ordered the lenses and the mirrors. I'm working in putting together a representation of how much improvement the new design will give us in comparison to the current setup.

Attachment 1: RawDataForTheModeGreenScan.png
RawDataForTheModeGreenScan.png
Attachment 2: ResultForModeMatching.png
ResultForModeMatching.png
Attachment 3: DataAndCalculationOfModeMismatch.zip
  10227   Thu Jul 17 16:07:34 2014 Emily UpdateElectronicsVCO Driver

I took back he VCO driver that Reetika brought over to the 40m from the PSL lab.  

  10229   Thu Jul 17 16:39:34 2014 NichinUpdateElectronicsPDFR debugging attempt : REFL11

In a attempt to debug the values of transimpedance generated by the PDFR system, I did a manual measurement for REFL11 PD.

  • Took the tops off AS and POY tables. (REFL11 and REF PD) Under the supervising eye of Manasa
  • Verify that no extra light is falling on REFL11.
  • Retake DC voltage readings, power readings.
  • Manually set the sweep parameters and record readings from network analyzer.
  • Put the tops back on the tables
  • Calculate transimpedance 

Results:

REF PD(1611):

Pinc = 1.12 mW                 T_dc = 10000 V/A (datasheet)

Vdc = 7.68 V                      T_rf = 700 V/A (datasheet)

Calculated Responsivity = 0.68 A/W (Which matches perfectly with the datasheet value of 0.68 A/W) 

REFL11:

Pinc = 0.87 mV             T_dc = 66.2 V/A (schematic)

Vdc = 32.5 mV      

Calculated Responsivity = 0.56 A/W

 

 

Network analyzer reading at 11 MHz : 0.42

Calculated RF Transimpedance = 460 V/A

40m Wiki : RF Transimpedance = 4 kV/A

I ran the same measurement using PDFR system and got the same results.

Attached: the automatic data and plot obtained.

Conclusion:  The PDFR system and manual measurements agree with each other. However the values do not match with 40m Wiki. I have no clue about which measurement is correct or any mistakes I might be making in the calculations. 

 

Attachment 1: REFL11_17-07-2014_154534.pdf
REFL11_17-07-2014_154534.pdf
Attachment 2: REFL11_17-07-2014_154534.zip
  10230   Thu Jul 17 17:08:58 2014 HarryUpdateGeneral1X2 Rack Changes

 Purpose

 

Steve and I moved some things around in the 1X2 rack in order to make room (roughly 6") for the electronics box that will contain rf frequency counters, ADC, and Raspberry Pi's for use in the Frequency Offset Locking apparatus

Picture

1X2Changes.png

Occurrences

First, we killed power by removing the fuse that the boxes we were moving were running through.

Then, we moved the boxes. I dropped//lost a washer. It didn't seem to cause any problems, so no further attempts to locate it were made.

The fuse was reinstalled, and everything was reconnected.

Moving Forward

We are now working on putting together the electronics box, which will contain ADC, and raspberry pi's. The frequency counters will be mounted on the front of the box.

Once complete, it will be installed for use in FOL.

  10231   Thu Jul 17 17:19:25 2014 SteveUpdatePSLPSL Innolight controller fan is dying with extra fan

Quote:

[Koji, Manasa]

The air flow from the dying fan was kinda weak and we decided to give a help with an external fan.

Koji brought a fan taken from a junk found at EE shop in W.Bridge.
The fan has been tied to the cage of the existing fan using cable ties to provide air circulation.
So even if the existing one dies anytime, we still don't super-heat anything.
The power supply for the fan rests next to the controller. 

The air from the fan ventilation was hot, and now with the additional fan this hot air is actually sucked
out with stronger flow. So this is relieving for now.

     PMC transmission as an indicator of laser controller with extra fan solution: 8 and 1day plot

Attachment 1: fanadded.png
fanadded.png
  10232   Thu Jul 17 17:39:57 2014 KojiUpdateElectronicsPDFR debugging attempt : REFL11

What is the coupling factor between the RF in and the RF mon of the demodulator?
I don't assume you have the same amount RF power at those two points unless you have an RF amplifier in the mon path.

  10233   Thu Jul 17 21:01:28 2014 ManasaUpdateGeneral1X2 Rack Changes

Quote:

 Purpose

 

Steve and I moved some things around in the 1X2 rack in order to make room (roughly 6") for the electronics box that will contain rf frequency counters, ADC, and Raspberry Pi's for use in the Frequency Offset Locking apparatus

Picture

1X2Changes.png

Occurrences

First, we killed power by removing the fuse that the boxes we were moving were running through.

Then, we moved the boxes. I dropped//lost a washer. It didn't seem to cause any problems, so no further attempts to locate it were made.

The fuse was reinstalled, and everything was reconnected.

Moving Forward

We are now working on putting together the electronics box, which will contain ADC, and raspberry pi's. The frequency counters will be mounted on the front of the box.

Once complete, it will be installed for use in FOL.

Additional comments:

This was done based on the earlier proposed setup plan for the frequency counters that will be used to measure the beat note frequencies [Akhil's elog]

I switched off the power supply to the green PDs so that we don't cause any damage while moving the amplifier panel for the beat signals and beatbox. 

  10234   Thu Jul 17 22:08:14 2014 KojiUpdateGeneral1X2 Rack Changes

It sounds like the work was done carefully. Even so, Jenne or Manasa have to run the ALS (X and Y) to check if they are still functional.

  10235   Fri Jul 18 14:59:07 2014 EvanUpdateIOOMC servo TFs

[Rana, Evan]

This morning we took several TFs of the MC servo board using the HP4395A.

The 4395 source was teed, with one output of the tee going to 4395 R and the other output going to the board's IN1. We then took TFs of (4395 A) / (4395 R), where 4395 A was one of the following four points on the servo board:

  • OUT2
  • A TEST1
  • B TEST1
  • SERVO

For each of these points, we took a TF at two gain settings: IN1 and VCO gains both at 0 dB, and then IN1 and VCO gains both at 20 dB.

Before doing these measurements, we calibrated out the cable delay. Additionally, SERVO was always loaded with 50 Ω—either from the 4395 or from a terminator.

The attached png shows the servo board settings when these TFs were taken with the 0 dB gain settings. The settings for the 20 dB measurements are identical, except for the higher IN1 and VCO gains.

Attachment 1: mcServoTFSettings.png
mcServoTFSettings.png
Attachment 2: MCtfs.pdf
MCtfs.pdf
  10236   Fri Jul 18 15:21:12 2014 ericqUpdateComputer Scripts / ProgramsLocal Chiara backups

Quote:

I've also written a backup script at scripts/backup/rsync_chiara.backup which keeps its books in scripts/backup/rsync_chiara.backup.log 

I'm adding a entry to the root crontab on chiara to execute the script every day at 7am. 

 I had some syntax errors in the script that prevented the script from doing the right thing. The backup is now up to date, and the cronjob should work. 

  10237   Fri Jul 18 16:52:56 2014 AndresUpdate40m Xend Table upgradeFInish Calculation on Current X-arm mode Matching

Quote:

Data and Calculation for the Xarm Current Mode Matching

Two days ago, Nick, Jenne, and I took a measurement for the Green Transmission for the X-arm. I took the data and I analyzed it. The first figure attached below is the raw data plotted. I used the function findpeaks in Matlab, and I found all the peaks. Then, by taking close look at the plot, I chose two peaks as shown in the second figure attached below. I took the ratio of the TEM00 and the High order mode, and I average them. This gave me a Mode Matching of 0.9215, which this value is pretty close to the value that I predicted by using a la Mode in http://nodus.ligo.caltech.edu:8080/40m/10191, which is 0.9343. Nick and I measured the reflected power when the cavity is unlocked and when the cavity is locked, so we measured the PreflUnLocked=52+1µW and PreflOnLocked=16+2µW and the backgroundNoise=0.761µW. Using this information we calculated  Prefl/Pin=0.297. Now, since Prefl/Pin=|Eref/Ein|2, we looked at the electric fields component by using the reflectivity of the mirror we calculated 0.67. The number doesn't agree, but this is because we didn't take into account the losses when making this calculation. I'm working in the calculation that will include the losses.

Today, Nick and I ordered the lenses and the mirrors. I'm working in putting together a representation of how much improvement the new design will give us in comparison to the current setup.

We want to be able to graphically see how much better it is the new optical table setup in comparison to the current optical table setup. In other words, we want to be able to see how displacement of the beam and how much angle change can be obtained at the ETM from changing the mirrors angles independently. Depending on the spread of the mirrors' vectors we can observe whether the Gouy phase is good. In the plot below, the dotted lines correspond to the current set up, and we can see that the lines are not spread from each other, which essentially mean that changing the angles of the two mirrors just contribute to small change in angle and in the displacement of the beam at the ETM, and therefore the Gouy phase is not good. Now on the other hand. The other solid lines correspond to the new setup mirrors. We can observe that the spread of the line of mirror 1 and mirror 4 is almost 90 degrees, which just implies that there is a good Gouy phase different between these two mirrors. For the angles chosen in the plot, I looked at how much the PZT yaw the mirrors from the elog http://nodus.ligo.caltech.edu:8080/40m/8912. In this elog, they give a plot in mrad/v for the pitch and yaw, so I took the range that the PZT can yaw the mirrors, and I converted into mdegrees/v and then I plotted as shown below. I plot for the current setup and for the new setup in the same plot. The matlab code is also attached below.

Attachment 1: OldAndNewSetupPlotsOfDisplacementAndAngleAtTheETM.png
OldAndNewSetupPlotsOfDisplacementAndAngleAtTheETM.png
Attachment 2: OldSetUpDisplacementAndNewSetup.m.zip
  10240   Sat Jul 19 01:59:34 2014 HarryUpdateGeneralFiber Mode Measurement

Purpose 

We wanted to measure the mode coming out of the fibers, so we can later couple it to experimental setups for measuring different noise sources within the fiber. i.e. Polarization Extinction Ratio, Frequency Noise, Temperature Effects.

Methods

I used the beamscan mounted on a micrometer stage in order to measure the spot sizes of the fiber coupled light at different points along the optical axis, in much the same way as in the razorblade setup I used earlier in the summer.

fiberModeMeasurement.png

Analysis

I entered my data (z coordinates, spot size in x, spot size in y) into a la mode to obtain the beam  profile (waist size, location)

 fiberModeMeasurement1.png 

Code is attached in .zip file.

Moving Forward

After I took these measurements, Manasa pointed out that I need points over a longer distance. (These were taken over the range of the micrometer stage, which is 0.5 inches.)

I will be coming in to the 40m early on Monday to make these measurements, since precious beamscan time is so elusive.

Eventually, we will use this measurement to design optical setups to characterize Polarization Extinction Ratio, Frequency Noise, and temperature effects of the fibers, for further use in FOL.

Attachment 3: fiberModeMeasurement1.zip