Author: Gabriele 
Type: Summary 
Category: LSC 
Subject: Effect of PRC length mismatch on error signals 

 [Gabriele, EricQ]


Actually it is difficult to see any laser frequency line in the dark fringe signal, since the Schnupp asymmetry is small. It is much better to use a differential MICH excitation which gives a better signal at the dark port.

We repeated the simulation explained before. We can use both the AS55 or the AS11 signals, bout the first one has a limited linear range and the expected 4cm value is very close to saturation.

as11.pngas55.png