[Masayuki, Manasa]
I. ALS servo loops
After fixing things with the phase tracking loop, we checked if things were good with the ALS servo loops.
We measured the OLTF of the X and Y arm ALS servo loops. In both cases the phase margin was ~20 degrees. There was no room to set enough phase margin. So we looked at the servo filters. We tried to modify the filters so that we could bring enough phase margin, but could not get at it. So we put back the old filters as they were.
attachment1: OLTF of the ALS XARM and YARM control loops
attachment2: Current phase budget. FM4 and FM10 are the boost filters.
II. ALS in-loop noise
Also, I found that the overall noise of the ALS servo has gone up by about two orders of magnitude (in Hz/rtHz) over the whole range of frequencies for both the arms from the last time the measurements were made. I suspect this could be from some change in the calibration factor. Did anybody touch things around that could have caused this? Or can somebody recollect any changes that I made in the past which might have affected the calibration? Anyways, I will do the calibration again.
|