Entry  Tue Mar 13 07:04:55 2012, kiwamu, Update, LSC, evolution of the sensing matrix in PRMI as a function of time PRCL.pngMICH.png
    Reply  Tue Mar 13 16:40:06 2012, kiwamu, Update, LSC, evolution of the sensing matrix in PRMI as a function of time: details lockins_MICH.png
    Reply  Tue Mar 13 16:56:19 2012, kiwamu, Update, LSC, evolution of the sensing matrix in PRMI as a function of time 
    Reply  Wed Mar 14 21:01:36 2012, keiko, Update, LSC, evolution of the sensing matrix in PRMI as a function of time 
Message ID: 6419     Entry time: Wed Mar 14 21:01:36 2012     In reply to: 6403
Author: keiko 
Type: Update 
Category: LSC 
Subject: evolution of the sensing matrix in PRMI as a function of time 

This is the simulated signals to compare with the original post #6403

 

 

PRMI configuration, PRCL signal

[W/m] Simulation Measured
REFL11 575440

 

~10000

REFL33 4571 ~50
REFL55 288400 ~5000
REFL165 891 NA
AS55 71 70

 

PRMI configuration, MICH signal

[W/m] Simulation Measured
REFL11 2290

 

~600

REFL33 36 ~4
REFL55 5623 ~200
REFL165 17 NA
AS55 6456 ~200
 

Simulated DC REFL power is 9mW (before the attenuator). AS DC is 0.3mW.

They don't agree. I suspect the PR gain for the SBs are somehow different. It is about 40 (or a bit less) in the simulation for 11MHz.

 

 

 

 