John, Yoichi

This morning John found that MZ servo is not working.
We were able to bring the MZ back by changing the output offset a bit. But we were not sure what was actually wrong.
So we pulled out the MZ board and checked several TPs to understand the behavior.
Here is the summary of what we learned this morning.

The MZ control board has the following stages:

[Mixer] -(error signal)-> [Sum amp for input offset] -(error + offset)-> [Variable Gain Amp] -> [Filter (x100 DC gain)] -(FB signal)-> [High Voltage Amp] -> output
(The HV amp also works as the sum amp for the output offset)

(1) We noticed that the Sum amp for the input offset has an output of -0.14V even when the offset input is 0V. This can be canceled by the input offset slider.
So for the moment, it is fine. But we might want to change the op-amp because the weird offset implies there might be something wrong with the chip.
The procedure to null the -0.14V offset is the following:
a) Enable Test 1 input on the MZ MEDM screen.
b) Move the input offset slider until the mixer monitor becomes 0V. Currently the input offset slider is at -7.5V to cancel the -0.14V offset.

(2) Because the gain of the Variable Gain Amp and the Filter combined is large, the Filter can be easily saturated if the output offset is not right.
This was the cause of the MZ problem this morning. The output offset slider was at a wrong position making the error signal slightly off centered from zero.
This residual DC error signal was amplified by the large gain chain and saturated the filter amp.
Our experience is that the output offset cannot go below -3V. We set it at 0V for now.