[Rana, Koji]
REFL55 was modified. The noise level confirmed. The PD is now ready to be installed.
Kevin's measurement report told us that something was wrong with REFL55 PD. The transimpedance looked OK, but the noise level was terrible (equivalent to the shotnoise of 14mA DC current).
Rana and I looked at the circuit, and cleaned up the circuit, by removing unnecessary 11MHz notch, 1k shunt resister, and so on.
I made a quick characterization of the PD.
First page:
The transimpedance ws measured as a function of the frequency. The resonance was tuned at 55MHz. The notch was tuned at 110MHz in order to reject the second harmonics. The transimpedance was ~540V/A at 55MHz. (For the calibration, I believed the DC transimpedance of 50V/A and 10000V/A for the DC paths of this PD and #1611, respectively, as well as the RF impedance (700V/A0 of #1611.
Second page:
Output noise levels were measured with various amount of photocurrent using white light from a light bulb. The measurement was perforemed well above the noise level of the measurement instruments.
Third page:
The measured output noise levels were converted into the equivalent current noise on the PD. The dark noise level agrees with the shot noise level of 1.5mA (i.e. 22pA/rtHz). In deed, the noise level went up x~1.5 when the photocurrent is ~1.4mA.
|