Message ID: 4369     Entry time: Wed Mar 2 18:08:43 2011
Author: Aidan 
Type: Update 
Category: Green Locking 
Subject: Ghost beams on green 

Kiwamu and I noticed that there is a ghost beam on the green beam going into the ETM. What we see is some interference fringes on the edge of the transmission of the green beam through the dichroic beam splitter (DCBS). If we look at the reflection from the dichroic beam splitter these are much more pronounced.

The spacing of the fringes (about 2 per 10mm) indicates an angle between the fields of around 0.1 mrad.

We were able to cause significant motion of the fringes by pushing on the knobs of the steering mirrors that steer the beam into the DCBS. A rough calculation of the derivative of optical path difference between the ghost and the primary beam as a function of input angle gives about 15 microns per mrad. What filtering the effect the arm cavity will have on the ghost beam is not immediately clear, but the numbers shouldn't be too difficult to determine.

 

Attachment 1: Ghost_Beam_at_ETM_DCBS.pdf  115 kB  Uploaded Wed Mar 2 19:11:06 2011  | Hide | Hide all