40m QIL Cryo_Lab CTN SUS_Lab TCS_Lab OMC_Lab CRIME_Lab FEA ENG_Labs OptContFac Mariner WBEEShop
  40m Log  Not logged in ELOG logo
Entry  Wed Sep 15 01:58:28 2010, valera, Update, PSL, FSS locking DSC_2510.JPG
    Reply  Wed Sep 15 03:08:26 2010, Koji, Update, PSL, FSS locking 
Message ID: 3575     Entry time: Wed Sep 15 03:08:26 2010     In reply to: 3574
Author: Koji 
Type: Update 
Category: PSL 
Subject: FSS locking 

Brilliant! This is the VERY way how the things are to be conquered!

Quote:

The RefCav is locked and aligned. I changed the fast gain sign by changing the jumper setting on the TTFSS board. The RefCav visibility is 70%. The FSS loop ugf is about 80 kHz (plot attached)  with FSS common gain max out at 30 dB. There is about 50 mW coming out of the laser and a few mW going to RefCav out of the back of the PMC. So the ugf can be made higher at full power. I have not made any changes to account for the PMC pole (the FSS is after the PMC now). The FSS fast gain was also maxed out at 30 dB to account for the factor of 5 smaller PZT actuation coefficient - it used to be 16 dB according to the (previous) snap shot. The RefCav TRANS PD and camera are aligned. I tuned up the phase of the error signal by putting cables in the LO and PD paths. The maximum response of the mixer output to the fast actuator sweep of the fringe was with about 2 feet of extra cable in the PD leg.

I am leaving the FSS unlocked for the night in case it will start oscillating as the phase margin is not good at this ugf.

 