Quote:

It looks like something wrong happened around the PSL front end.  One of the PSL channel, C1:PSL-PMC_LOCALC, got crazy. 

We found it by the donkey alarm 10 minutes ago.

The attached picture is a screen shot of the PMC medm screen.

The value of C1:PSL-PMC_LOCALC ( middle left on the picture ) shows wired characters. It returns "nan" when we do ezcaread.

Joe went to the rack and powered off / on the crate, but it still remains the same. It might be an analog issue (?)

The problem seems to be a software one.

In any case, Kiwamu and I looked at the at the PMC crystal board and demod board, in search of a possible bad connection. We found a weak connection of the RG cable going into the PD input of the demod board. The cable was bent and almost broken.

I replaced the SMA connector of the cable with a new one that I soldered in situ. Then I made sure that the connection was good and didn't have any short due to the soldering.