40m QIL Cryo_Lab CTN SUS_Lab TCS_Lab OMC_Lab CRIME_Lab FEA ENG_Labs OptContFac Mariner WBEEShop
  40m Log  Not logged in ELOG logo
Entry  Wed Jul 7 00:15:29 2010, tara,Rana, Summary, PSL, power spectral density from RefCav transmitted beam  RIN_rftrans.pngtara.png
    Reply  Wed Jul 7 10:42:29 2010, Koji, Summary, PSL, power spectral density from RefCav transmitted beam  
Message ID: 3164     Entry time: Wed Jul 7 10:42:29 2010     In reply to: 3163
Author: Koji 
Type: Summary 
Category: PSL 
Subject: power spectral density from RefCav transmitted beam  

How do you calibrate this to Hz/rtHz?

Quote:

I measured the RC transmitted light signals here at the 40m. I made all connections through the PSL patch panel. No optics/PD were touched.

I measured the spectral density of the signal of the transmitted beam behind RefCav in both time and frequency domain.

This will be compared with the result from PSL lab later, so I can see how stable the signal should be.

We re-aligned the beam into the cavity (the DC level increased from 2 V to 3.83V)

and the reflected beam to the center of the RFPD.

 

 

ELOG V3.1.3-