Message ID: 2336     Entry time: Wed Nov 25 16:44:52 2009     In reply to: 2335
Author: Koji 
Type: Update 
Category: PSL 
Subject: Measured MC length--FSS trend 

I checked C1:PSL-FSS_VCODETPWR. The attached is the 4 months trend of the FSS RCTRANS / RFPDDC(=FSS REFL) / VCODETPWR / VCOMODLEVEL.

Although VCO modulation level setting was mostly constnt, VCODETPWR, which presumably represents the RF level, changes time by time.
It coincides with the recent reduction of the RCTRANS/RFPDDC. Actually, my touch restored the VCO to the previous more stable state.
One can see that this is not only a single occation, but it happened before too. (In the middle of Aug.)

This could be explained by the bad contact of some cable or connector.

Nevertheless we need more careful investigation:

1. Understand what VCODETPWR is exactly.
2. Investigate relationship between VCOMODLEVEL / VCODETPWR / AOM deflection efficiency / RCTRANSPD
3. Confirm the frequency matching between the VCO and AOM.

Quote:

but the increase in both the RCtrans and the RCrefl is consistent with my theory that the power going to the RC has increased ; its not just an increase in the visibility.

We should scan the AOM/VCO to make sure the frequency is matched to the resonance to within 0.5 dB.

 