Subject: MC/FSS Frequency Noise 
The FSS frequency noise is not very bad.

I compared the MC_F spectra between Hanford and the 40m using DTT and its 'User NDS' option.
After Sam, Jenne, and DavidM installed the new MC Servo some time ago, the MC_F spectrum here
has had some whitening before it goes into the DAQ (on board; same as LLO & LHO). The tuning
coefficient of the VCO is also basically the same between all PSLs since everyone has the same
chip in the VCO driver.

Therefore, at the frequencies where the MC gain is more than ~4, the MC_F signal calibration is
the same here as anywhere. Since its the servo control signal, its basically a measure of the
frequency noise incident on the MC -- its just what comes out of the FSS with the table noise on
top. At low frequencies (< 100 Hz) its a measure of the motion of the MC mirrors.

Above 200 Hz ours is the same as theirs; except for the enormous power line spikes. I think that's
either all on the light. But our acoustics are better and the noise above 1 kHz levels off at the
same flat floor (the phase noise of the VCO) as H1. The huge lump around 100 Hz is the MC2 DAC noise and
it goes down to the H1 levels when we flip on the dewhites. The giant excess from 5-50 Hz is just the fact
that our stacks don't do much until 20-30 Hz.

So we can stop blaming the FSS and move on with life as soon as Tobin gets the ISS back in shape.