[Anchal, Paco]
We started the day by taking a spectrum of C1:HPC-LO_PHASE_IN1, the BHD error point, and confirming the absence of 268 Hz peaks believed to be violin modes on LO1. We then locked the LO phase by actuating on LO2, and AS1. We couldn't get a stable loop with AS4 this morning. In all of these trials, we looked to see if the noise increased at 268 Hz or its harmonics but luckily it didn't. We then decided to add the necessary output filters to avoid exciting these violin modes. The added filters are in the C1:SUS-LO1_LSC bank, slots FM1-3 and comprise bandstop filters at first, second and third harmonics observed previously (268, 536, and 1072 Hz); bode plots for the foton transfer functions are shown in Attachment #1. We made sure we weren't adding too much phase lag near the UGF (~ 1 degree @ 30 Hz).
We repeated the LO phase noise measurement by actuating on LO1, LO2 and AS1, and observe no noise peaks related to 268 Hz this time. The calibrated spectra are in Attachment #2. Now the spectra look very similar to one another, which is nice. The rms is still better when actuating with AS1.
[Paco]
After the above work ended, I tried enabling FM1-3 on the C1:HPC_LO_PHASE control filters. These filters boost the gain to suppress noise at low frequencies. I carefully enabled them when actuating on LO1, and managed to suppress the noise by another factor of 20 below the UGF of ~ 30 Hz. Attachment #3 shows the screenshot of the uncalibrated noise spectra for (1) unsupressed (black, dashed), (2) suppressed with FM4-5 (blue, solid), and (3) boosted FM1-5 suppression (red).
Next steps:
- Compare LO-ITMY and LO-ITMX single bounce noise spectra and MICH.
- Compare DC locking scheme versus BH55 once it's working.
|