[Anchal, Paco, Rana]
We locked the XARM using POX11 and made a noise budget for the single arm displacement; see Attachment #1. The noise budget is rough in that we use simple calibrations to get it going; for example we calibrate the measured error point C1:LSC-XARM_IN1_DQ using the single cavity pole and some dc gain to match the UGF point. The control point C1:LSC-XARM_OUT_DQ is calibrated using the actuator gain measured recently by Yuta. We also overlay an estimate of the seismic motion using C1:PEM-SEIS_BS_X_OUT_DQ (calibrated using a few poles to account for stack and pendulum), and finally the laser frequency noise as proxied by the mode cleaner C1:IOO-MC_F_DQ.
A couple of points are taken with this noise budget, apart from it needing a better calibration;
- Overall the inferred residual displacement noise is high, even for our single arm cavity.
- By looking at the sim OLTF in foton, it seemed that the single arm cavity loop TF could easily become unstable due to some near-UGF-funkiness likely from FM3 (higher freq boost), so we disabled the automatic triggering on it; the arm stayed locked and we changed the error signal (light blue vs gold (REF1) trace)
- The arm cavity is potentially seeing too much noise from the IMC in the 1 to 30 Hz band in the form of laser frequency noise.
- Need IMC noise budget to properly debug.
- At high frequency (>UGF), there seem to be a bunch of "wiggles" which remain unidentified.
- We actually tried to investigate a bit into these features, thinking they might have something to do with misalignment, but we couldn't really find significant correlation.
RXA edit:
- we also noticed some weirdness in the calibration of MC_F v. Arm. We think MC_F should be in units of Hz, and Paco calculated the resulting motion as seen by the arm, but there was a factor of several between these two. Need to calibrate MC_F and check. In principle, MC_F will show up directly in ALS_BEATX (with the green PDH lock off), and I assume that one is accurately calibrated. Somehow we should get MC_F, XARM, and ALS_BEAT to all agree. JC is working on calibrating the Mini-Circuits frequency counter, so once that is done we will be in good shape.
- we may need to turn on some MC_L feedback for the IMC, so that the MC length follows the NPRO frequency below ~20 Hz.
- Need to estimate where the IMC WFS noise is in all of this. Does it limit the MC length stability in any frequency band? How do we determine this?
- Also, we want to redo this noise budget today, whilst using AS55 instead of POX. Please measure the Schnupp asymmetry by checking the optimum demod phase in AS55 for locking Xarm v Yarm.
|