Entry  Sat Feb 26 01:01:40 2022, Tega, Update, VAC, Ongoing work to get the FRG gauges readouts to EPICs channels 
    Reply  Sat Feb 26 15:45:14 2022, Tega, Update, VAC, Ongoing work to get the FRG gauges readouts to EPICs channels Vac-gauges-flow-diagram.png
       Reply  Mon Feb 28 19:15:10 2022, Tega, Update, VAC, Ongoing work to get the FRG gauges readouts to EPICs channels 
       Reply  Tue Mar 1 20:38:49 2022, Tega, Update, VAC, Ongoing work to get the FRG gauges readouts to EPICs channels Working_Serial_Port_List_1.pngWorking_Serial_Port_List_2.pngWorking_Config_Ports#1-5.pngWorking_Config_Ports#7-8.png
          Reply  Wed Mar 2 11:50:39 2022, Tega, Update, VAC, Ongoing work to get the FRG gauges readouts to EPICs channels XGS600_IOLAN_settings_1.pngXGS600_IOLAN_settings_2.pngXGS600_USBserial_settings.pngXGS600_comm_test.pyXGS600_comm_test_result.txt
             Reply  Wed Mar 2 12:40:08 2022, Tega, Update, VAC, Ongoing work to get the FRG gauges readouts to EPICs channels 6x
                Reply  Sun Mar 6 18:14:45 2022, Tega, Update, VAC, Ongoing work to get the FRG gauges readouts to EPICs channels iolan_xgs_comm_investigation.pdf
                   Reply  Mon Mar 7 13:53:40 2022, Tega, Update, VAC, Ongoing work to get the FRG gauges readouts to EPICs channels iolan_xgs_comm_live.pdf
                      Reply  Mon Mar 7 14:52:34 2022, Koji, Update, VAC, Ongoing work to get the FRG gauges readouts to EPICs channels 
                         Reply  Tue Mar 8 12:08:47 2022, Tega, Update, VAC, Ongoing work to get the FRG gauges readouts to EPICs channels XGS600_calibration.pdf
Message ID: 16692     Entry time: Wed Mar 2 11:50:39 2022     In reply to: 16691     Reply to this: 16693
Author: Tega 
Type: Update 
Category: VAC 
Subject: Ongoing work to get the FRG gauges readouts to EPICs channels 

Here is the IOLAN SDS TCP socket setting and the USBserial setting for comparison.

I have also included the python script and output from the USBserial test from earlier.

Attachment 1: XGS600_IOLAN_settings_1.png  241 kB  | Hide | Hide all
XGS600_IOLAN_settings_1.png
Attachment 2: XGS600_IOLAN_settings_2.png  242 kB  | Hide | Hide all
XGS600_IOLAN_settings_2.png
Attachment 3: XGS600_USBserial_settings.png  116 kB  | Hide | Hide all
XGS600_USBserial_settings.png
Attachment 4: XGS600_comm_test.py  2 kB  | Hide | Hide all
#!/usr/bin/env python

#Created 2/24/22 by Tega Edo
'''Script to read/write to the XGS-600 Gauge Controller'''

import serial
import sys,os,math,time

ser = serial.Serial('/dev/cu.usbserial-1410') # open serial port 

... 74 more lines ...
Attachment 5: XGS600_comm_test_result.txt  1 kB  | Hide | Hide all
----- Multiple Sensor Read Commands -----

Sent to XGS-600 -> #0001\r : Read XGS contents
response : >FE4CFE4CFE4C

Sent to XGS-600 -> #0003\r : Read Setpoint States
response : >0000

Sent to XGS-600 -> #0005\r : Read software revision
response : >0206,0200,0200,0200
... 69 more lines ...