Message ID: 15906     Entry time: Thu Mar 11 20:18:00 2021
Author: gautam 
Type: Update 
Category: LSC 
Subject: High bandwidth POY 

I repeated the high bandwidth POY locking experiment.

  1. The "Q" demod output (SMA) was routed to the common mode board (it appears in the past I used the LEMO "MON" output instead but that shouldn't be a meaningful change).
  2. As usual, slow actuation --> ETMY, fast actuation --> IMC error point.
  3. Loop UGF measurement suggests that bandwidth ~25kHz, with ~25 degrees phase margin. Anyway the lock was pretty stable.

One thing I am not sure is - when looking at the in-loop error point spectra, the Y-arm error point did not get suppressed to the CM board's sensing noise floor - I would have thought that with the huge amount of gain at ~16 Hz, the usual structure we see in the spectra between 10-30Hz would be completely squished. Need to think about if this is signalling something wrong, because the loop TF measurements seemed as expected to me.

1020pm: plots uploaded. As I made the plot of the spectrum, I realized that I don't have the calibration for the Y-arm error point into displacement noise units, so it's in unphysical units for now. But I think the comment about the hump around 16 Hz not being crushed to some sort of flat electronics noise floor. For the TF plots, when the loop gain is high, this IN1/IN2 technique isn't the best (due to saturation issues) but I don't think there's anything controversial about getting the UGF this way, and the fact that the phase evolves as expected when the various gains are cranked up / boosts enabled makes me think that the CM board is itself just fine.


10am 12 March: i realized that the "Y-arm error point" plotted below is not the true error point - that would be the input to the CM board (before boosts etc), which we don't monitor digitally. The spectra are plotted for the CM_SLOW input which already has some transfer function applied to it. In the past, I routed the LEMO "MON" connector on the demod board to the CM board input, and hence, had the usual SMA outputs from the demod board going to the digital system. I hypothesize that plotting the spectra for that signal would have showed this expected suppression to the electronics noise floor.

In summary, on the basis of this test, I don't see any red flags with the CM board.

Attachment 1: OLGevolution.pdf  160 kB  Uploaded Thu Mar 11 22:16:10 2021  | Hide | Hide all
OLGevolution.pdf
Attachment 2: inLoopSpec.pdf  166 kB  Uploaded Thu Mar 11 22:23:42 2021  | Hide | Hide all
inLoopSpec.pdf