40m QIL Cryo_Lab CTN SUS_Lab TCS_Lab OMC_Lab CRIME_Lab FEA ENG_Labs OptContFac Mariner WBEEShop
  40m Log  Not logged in ELOG logo
Entry  Tue Sep 15 17:19:09 2020, gautam, Update, IOO, MC F spectrum  MCF_MCL.pdf
    Reply  Tue Sep 15 19:17:30 2020, rana, Update, IOO, MC F spectrum  
       Reply  Wed Sep 16 09:43:41 2020, gautam, Update, IOO, MC F spectrum  MCF_MCL.pdf
Message ID: 15574     Entry time: Tue Sep 15 19:17:30 2020     In reply to: 15573     Reply to this: 15576
Author: rana 
Type: Update 
Category: IOO 
Subject: MC F spectrum  

that's a very curious disconnection

the "Pentek" whitening board that carries the MC channels has jumpers to enable either 1 or 2 stages of 15:150 whitening. Looks lik MC_F has 2 and MC_L has 1.

I guess the MC_F signal is so low because of the high gain on the FSS board.  We could lower the FSS common gain and increase the IMC board's VCO gain to make up for this. Maybe 6 dB would be enough. IF that is risky, we could also up the analog gain on the whitening board.