Message ID: 12878     Entry time: Thu Mar 9 20:38:19 2017
Author: rana 
Type: Configuration 
Category: IOO 
Subject: MC lock acquisition settings changed; no more HOM locks 

The MC was sort of misaligned. It was locking on some vertical HOMs. So I locked it and aligned the suspensions to the input beam (not great; we should really align the input beam to the centered spots on the MC mirrors).

With the HOMs reduced I looked at the MC servo board gains which Guatam has been fiddling with. It seems that since the Mod Depth change we're getting a lot more HOM locks. You can recognize this by seeing the longish stretches on the strip tool where FSS-FAST is going rail-to-rail at 0.03 Hz for many minutes. This is where the MC is locked on a HOM, but the autolocker still thinks its unlocked and so is driving the MC2 position at 0.03 Hz to find the TEM00 mode.

I lowered the input gain and the VCO gain in the mcdown script and now it very rarely locks on a HOM. The UGF in this state is ~3-4 kHz (I estimate), so its just enough to lock, but no more. I tested it by intentionally unlocking ~15 times. It seems robust. It still ramps up to a UGF of ~150 kHz as always. 'mcdown' commited to SVN.