We've been having trouble tuning the ALS DIFF matrix. Trying to see if the MC2 EXC can be cancelled in ALS DARM by adjusting the relative gains in ALSX and ALSY Phase Tracker outputs.
There's a bunch of intermittent behavior. Between different ALS locks, we get more or less cancellation. We were checking this by driving MC2 at ~100-400 Hz and checking the ALS response (with the ALS loops closed). We noticed that the X and Y readbacks were different by ~5-10 degrees and that we could not cancel this MC2 signal in DARM by more than a factor of 4-5 or so. In the middle of this, we had one lock loss and it came back up with 100x cancellation?
Attached is a PDF showing a swept sine measurement of the ALSX, ALSY, and DARM signals. You can see that there is some phase shift between the two repsonses leading to imperfect cancellation. Any ideas? Whitening filters? HOM resonance? Alignment? |